An here is the fix: in asyn/vxi11/drvVxi11.c, procedure vxiConnectPort, line
906, vxiCreateDeviceLink gets called. Before this is done, the member
abortPort must be cleared (set to 0). Thus I added
pvxiPort->abortPort = 0;
at line 905 and now re-connects work as expected.
Cheers
Ben
On Thursday, December 16, 2010, Benjamin Franksen wrote:
> The reconnect feature in asyn is still buggy. I am using it with a
> gpib/lan gateway (Agilent E5810A). Any time I reboot the gpib box I get
> these messages:
> epics> 2010/12/16 13:07:44.150 L3 PAHRP:NRVD:cmdTrg: asynError in read:
> L3 RPC failed
> 2010/12/16 13:07:44.150 L3 PAHRP:NRVD:cmdTrg: I/O error after reading 0
> bytes: ""
> 2010/12/16 13:07:44.150 L3 PAHRP:NRVD:cmdTrg: Protocol aborted
> 2010/12/16 13:07:45.747 L3 vxiDestroyDevLink RPC error : RPC: Unable to
> send; errno = Broken pipe
> 2010/12/16 13:07:45.747 L3 vxiDisconnectPort : RPC: Unable to send; errno
> = Broken pipe
> 2010/12/16 13:07:47.750 L3 vxiDestroyDevLink RPC error : RPC: Unable to
> L3 port not connected
> 2010/12/16 13:07:47.750 L3 asynGpib:pollAddr srqEnable
> 2010/12/16 13:07:47.750 L3 PAHRP:NRVD:rdPow1: asynError in write: L3 RPC
> failed
> 2010/12/16 13:07:47.750 cbLow PAHRP:NRVD:rdPow2 lockRequest:
> pasynManager-
> >queueRequest() failed: port L3 not connected
> 2010/12/16 13:07:47.766 gpib0 vxiCreateDeviceLink abort channel TCP port
> changed from 1004 to 1005
> 2010/12/16 13:07:47.772 gpib0,20 vxiCreateDeviceLink abort channel TCP
> port changed from 1004 to 1005
> Afterwards there are no further communication attempts; rebooting the IOC
> correctly establishes a new connection.
> The IOC runs on a GNU/linux system. Version info:
> kernel 2.6.18-6-686
> EPICS base-3.14.8.2
> asyn-4.15
> streamDevice-2-snapshot20100615
> Cheers
> Ben