1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  <2013>  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  <2013>  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026
Subject: RE: Problem with mca R7.2 and Rontec XFlash
To: "'Zachary Brown'" <[email protected]>, "[email protected]" <[email protected]>
Date: Mon, 11 Feb 2013 20:11:22 +0000
You can diagnose problems like this by turning on asynTrace on the underlying serial port.  In order to have the asynTrace output not impact the performance significantly you can send the output to a file:
# Send the asynTrace output to a file.  Substitute the name of your asyn port for the serial device below.
asynSetTraceFile “serial1”,0,”serial_output.txt”
# ASYN_TRACEIO_HEX
asynSetTraceIOMask “serial1”,0,4
# ASYN_TRACE_ERROR and ASYN_TRACEIO_DRIVER
asynSetTraceMask “serial1”,0,9
To turn off the debugging output
# ASYN_TRACE_ERROR
asynSetTraceMask “serial1”,0,1
To set the file back to stderr
asynSetTraceFile “serial1”,0,””
The output file will give you a series of time-stamped messages when messages were sent to the device and when replies were received.  That should help to determine if the serial I/O is the bottleneck.
Are you running at the fastest possible baud rate?
From: [email protected] [mailto:[email protected]] On Behalf Of Zachary Brown
Sent: Friday, February 08, 2013 10:25 AM
To: [email protected]
Cc: [email protected]
Subject: Re: Problem with mca R7.2 and Rontec XFlash
We are using RS232.  Using the XFlash unit as configured through spec does not show the same delay as the IOC.  Disabling the Temperature monitoring PV (:ReadTemperature) which is on a 1s scan-rate, seems to alleviate the problem a bit.
-ZB
On 2/6/13 11:00 AM, Phillip Sorensen wrote:
> Hello,
> We are tring to setup the Rontec XFlash MCA with epics using the R7.2
> version of mca.
> The problem we are having is there seems to be a delay of some number of
> seconds after sending a command until we see an update in the MEDM
> screen.  For example clicking the start and stop buttons for the
> acquisition show this.  I we look at the XFlash unit it seems to update
> the front panel immediately, but the MEDM screen shows a delay.  The
> spectrum update (which is set to 1 second) also seems to take 5 to 10
> seconds per update.
> We are not getting any timeout errors.  Any suggestions?
Hi, Phil.
Over what physical link are you communicating with the XFlash?  If it's
RS-232, it could just take that long to transmit the data over the slow
link.
Lewis
Re: Problem with mca R7.2 and Rontec XFlash Zachary Brown
Prev: RE: header files in snc code Mark Rivers
Next: RE: streamDevice: how to read a value as fast as possible? Mark Rivers
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  <2013>  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026
Prev: Re: Problem with mca R7.2 and Rontec XFlash Zachary Brown
Next: Does anyone have an Ethernet based scaler ? Kate Feng