On 20/02/20 15:40, "Shen, Guobao" <gshen at anl.gov> wrote:
> the target delay must always be set so that it is greater than the physical delay, including internal delays. Otherwise some events will be missed
That's interesting and good to know.
It seems a firmware issue. Did you get any response from Jukka?
This is not a firmware issue in the sense that it could be "corrected" by some change in the firmware. It is by nature of the delay compensation system; the system simply cannot go back in time.
I got some information related to occasional unlocking of the delay compensation feedback loop that we have seen a few times in the lab.
In essence, Jukka is able to reproduce the issue although it seems to be very rare. The current assumption, as far as I understood, is that this is probably due to some threshold settings in the Xilinx DCM (Digital Clock Manager) being set too low. Also, as well as I understand, this does not cause any jumps in the delays. There will be updates to this but I think this is a very minor issue.
(sorry that this is missing some context for tech-talk readers; it would take a lot of text to replicate the whole discussion here. We can post a summary later if there is interest.)
Best regards,
Timo
> We are starting the commissioning phase soon so we will have better results.
Would love to know your experience.
Guobao
On 2/20/20, 2:40 AM, "Javier Cereijo García" <javier.cereijogarcia at ess.eu> wrote:
Hi all,
Yes, here at ESS we have done some tests in the lab in the past,
although they were quite simple. I can share what we have learnt. These
tests were done using mTCA-EVR-300 and mTCA-EVM-300, not VME.
The first feeling is that everything works as expected, and changing the
target delay works smoothly, as well as changing the length of the
optical fiber. There is one thing though that seems strange to me, and
it is that the target delay must always be set so that it is greater
events will be missed. Even when using a 230-series EVG with the delay
compensation disabled on the EVR side the target delay needs to be set
correctly.
We didn't perform any complex jitter test, although monitoring the "DC
lock" shows that it gets unlocked from time to time, usually recovering
quickly. We didn't worry too much at that moment, the measured delay
wasn't changing that quickly, I think that the lock monitor is quite
sensitive. I can't recall from the top of my head if we were using the
EVM's fractional synthesizer or an external RF source, so this may
easily disappear in a more stable environment.
Best regards,
Javier
On 19/02/2020 18:17, Michael Davidsaver wrote:
> On 2/18/20 10:43 AM, Shen, Guobao via Tech-talk wrote:
>> Hi all,
>> Is there anyone who has been using MRF EVM-300 delay compensation function for any purpose, in the lab, test stand, real machine, etc?
> ESS did some testing as well.
>> If yes, would you mind to share us your experience?
> As I recall, it worked well on a test stand. My personal knowledge ends there.
>> Thanks,
>> Guobao
>> Guobao Shen
>> gshen at anl.gov <mailto:gshen at anl.gov>
>> Controls Group Leader
>> Advanced Photon Source
>> Argonne National Laboratory
>> 9700 S. Cass Avenue, Lemont, IL 60439
Embedded System Engineer
HW&I group, ICS division
European Spallation Source ERIC
Mobile: +46 (0)72 179 23 34
E-mail: javier.cereijogarcia at ess.eu
- Re: MRF Delay Compensation Function Jukka Pietarinen via Tech-talk
- MRF Delay Compensation Function Shen, Guobao via Tech-talk
- Re: MRF Delay Compensation Function Michael Davidsaver via Tech-talk
- Re: MRF Delay Compensation Function Javier Cereijo García via Tech-talk
- Re: MRF Delay Compensation Function Shen, Guobao via Tech-talk
Re: [EXTERNAL] matrix multiplication and inversions Pearson, Matthew R. via Tech-talk
Buffer full issue with ADAravis Katie Matusik via Tech-talk
<2020>
<2020>