Subject: RE: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway
To: "'Hu, Yong'" <[email protected]>
Date: Thu, 9 Jul 2020 14:33:10 +0000
Hi Yong,
Please send the complete output when your IOC starts.  In your original message was that the complete stream protocol file?
Sent: Wednesday, July 8, 2020 6:24 PM
Cc: [email protected]
Subject: Re: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway
Hi Mark R.,
Thanks for your suggestion. Yes, I found the version in /usr/lib/epics/include/asynDriver.h: 4.23.0.
Yes, you are right. Both Asyn and StreamDevice are very old. I am going to upgrade them to the latest releases.
I have set the trace mask to 0xFF. Attached please find the detailed output messages for a few cycles. Below is the simplified version with important timestamps highlighted.
1) The record is scanned at every 5-second so that we get 2020/07/08 15:16:31.567 à 2020/07/08 15:16:36.567 à 2020/07/08 15:16:41.567, which makes perfect sense. Then it jumps to 2020/07/08 15:16:51.567, which is 10-second gap meaning one scan is skipped. This kind of skipping is confirmed by Michael’s suggestion of setting TPRO. I guess “scan5: Active SR:C03-BI{DMM:1}V-I_” means the record is scanned at 5-second and its PACT is true so that the record processing is simply skipped and it waits for the next scan.
2) Could you briefly explain what these mean: queueRequest timeout, vxiSetEos 1, vxiSetEos 0? I am surprised to see “\n 0a” after vxiSetEos 1. The time difference between “vxiRead” and “vxiSetEos 1” is 3.4 seconds, which means it takes 3.4 seconds to read a floating number “-4.008121752E+00”.
scan5: Process SR:C03-BI{DMM:1}V-I_
2020/07/08 15:16:31.567 L0 addr 6 queueRequest priority 0 not lockHolder
2020/07/08 15:16:31.567 L0 schedule queueRequest timeout
2020/07/08 15:16:31.567 asynManager::portThread port=L0 callback
2020/07/08 15:16:31.567 L0 6 vxiWrite numchars 14
2020/07/08 15:16:31.569 L0 6 vxiWrite
DCV 10, 1E-6
DCV 10, 1E-6\r\n
44 43 56 20 31 30 2c 20 31 45 2d 36 0d 0a
2020/07/08 15:16:31.569 L0 addr 6 queueRequest priority 0 from lockHolder
2020/07/08 15:16:31.569 L0 schedule queueRequest timeout
2020/07/08 15:16:31.569 asynManager::portThread port=L0 callback
2020/07/08 15:16:31.569 L0 vxiSetEos 1
\n
0a
2020/07/08 15:16:34.961 L0 6 vxiRead
-4.008121752E+00
-4.008121752E+00\r\n
2d 34 2e 30 30 38 31 32 31 37 35 32 45 2b 30 30 0d 0a
2020/07/08 15:16:34.961 L0 vxiSetEos 0
2020/07/08 15:16:36.567 L0 addr 6 queueRequest priority 0 not lockHolder
2020/07/08 15:16:40.407 L0 vxiSetEos 0
2020/07/08 15:16:41.567 L0 addr 6 queueRequest priority 0 not lockHolder
2020/07/08 15:16:46.957 L0 vxiSetEos 0
2020/07/08 15:16:51.567 L0 addr 6 queueRequest priority 0 not lockHolder ##note: one scan is skipped
2020/07/08 15:16:55.007 L0 vxiSetEos 0
2020/07/08 15:16:56.567 L0 addr 6 queueRequest priority 0 not lockHolder
Date: Wednesday, July 8, 2020 at 10:12 AM
To: "Hu, Yong" <[email protected]>
Hi Yong,
Ø  PS: I had difficulty finding the exact version number of asyn/streamdevice since I built my IOC against Michael Davidsaver’s Debian packages (https://epics.nsls2.bnl.gov/debian/ ) and the packages provide everything under /usr/lib/epics/
The Debian packages must also install the asyn header files, probably in /usr/include/epics.  Look at asynDriver.h, it should contain lines like the following:
#define ASYN_VERSION       4
#define ASYN_REVISION     39
#define ASYN_MODIFICATION  0
Let me know what version it is.  Since your streamDevice is more than 7 years old I suspect that asyn is very old as well.  If so, there have been quite a few fixes to the asyn VXI-11 support since 2013.  I suggest you build a test IOC with the latest version of asyn and streamDevice and see if he problem still occurs.
You should probably enable all of the asyn trace flags on that driver, i.e. set the trace mask to 0xFF.  There may be some hints as to the timing issues if you do that.
From: Tech-talk <[email protected]> On Behalf Of Hu, Yong via Tech-talk
Sent: Wednesday, July 8, 2020 8:38 AM
To: [email protected]
Subject: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway
Hello everyone,
I have a simple StreamDevice protocol for an old GPIB-based device HP3458A which is connected to Agilent E5810 LAN/GPIB Gateway (100 Mbps) talking to a softIOC. Here is the record to read the voltage data at 0.2Hz (5 second).
#only showing the must-have fields
record(ai, "SR:C03-BI{DMM:1}V-I_")
    field(DTYP, "stream")
    field(INP,  "@hp3458a.proto ReadData L0 6")
    field(SCAN, "5 second")
#StreamDevice protocol
ReadData {
         out "DCV  10, 1E-6";
         in "%f";
And this is the EPICS shell output after I turn on the Asyn trace.
2020/07/07 17:55:46.280 L0 6 vxiWrite
DCV  10, 1E-6
2020/07/07 17:55:49.630 L0 6 vxiRead
-4.016502336E+00
2020/07/07 17:55:52.396 L0 6 vxiWrite
DCV  10, 1E-6
2020/07/07 17:55:55.756 L0 6 vxiRead
-4.015675668E+00
2020/07/07 17:56:00.680 L0 6 vxiWrite
DCV  10, 1E-6
2020/07/07 17:56:04.030 L0 6 vxiRead
-4.014579704E+00
I get the expected response (i.e. -4.016502336E+00) after the command (DCV  10, 1E-6) is sent. However, as you see from the timestamps for each vxiWrite, the command is not sent at every 5-second. It takes more than 5 seconds and it has a pattern: 6.116, 8.284, 6.116, 8.284, …. ; the response time (vxiRead – vxiWrite) is consistent at 3.35 seconds although it seems ridiculously slow.
So, the system is slower than expected. However, I am surprised that there is no any alarm for the record and no any warning message from the EPICS shell output when that record can not keep up its scanning rate. Is this kind of silence supposed to be? Just to be clear, the softIOC itself and the physical server are perfectly fine: I added the classical “dbExample1.db” and its calc record can be scanned at exactly 5-second.
This system used to work as expected at 0.2Hz (5 second) despite it is ridiculously slow. Now it becomes even slower after our ITD upgraded the network switch. If anyone has experience on these old devices (HP3458A, E5810), please advise what I should do next.
If the version of Asyn/StreamDevice matters for this case, here we go:
epics> dbior
Driver: drvAsyn
L0 multiDevice:Yes canBlock:Yes autoConnect:Yes
    vxi11, host name: 10.0.132.67
Driver: stream
  StreamDevice 2.6.0 built May 29 2013 18:45:57
  registered bus interfaces:
    DebugInterface
    DummyInterface
AsynDriverInterface
PS: I had difficulty finding the exact version number of asyn/streamdevice since I built my IOC against Michael Davidsaver’s Debian packages (https://epics.nsls2.bnl.gov/debian/ ) and the packages provide everything under /usr/lib/epics/. I am sure there are many ways (ldd, strings, configure/RELEASE, etc.) to find out the version number of EPICS support module. I just found “dbior” seems a handy tool for reporting EPICS driver’s version. I am not sure if Mark Rivers has added this kind of report in his latest Asyn release.
Best regards,
Yong Hu
NSLS-II Controls Group
Re: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway Hu, Yong via Tech-talk
RE: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway Mark Rivers via Tech-talk
Prev: RE: Get/Set Local Variable in motorAxis From sub Record Pearson, Matthew R. via Tech-talk
Next: Announcing aioca: Python asyncio Channel Access client using libca via ctypes Cobb, Tom (DLSLtd,RAL,LSCI) via Tech-talk
Prev: Re: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway Hu, Yong via Tech-talk
Next: Re: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway Hu, Yong via Tech-talk
