1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  <20222023  2024  2025  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  <20222023  2024  2025
Subject: RE: Eurotherm2604 modbus support
From: Mark Rivers via Tech-talk <tech-talk at aps.anl.gov>
To: "Sintschuk, Michael" <michael.sintschuk at bam.de>, tech-talk <tech-talk at aps.anl.gov>
Date: Mon, 5 Dec 2022 16:47:37 +0000
Hi Michael,
Ø  drvAsynSerialPortConfigure("EMODBUS","COM1",0,0,1)
Ø  modbusInterposeConfig(EMODBUS, 0, 2000, 0)
The above 2 lines do not look internally consistent to me.  The first line says you are using a serial port for communication.  The second line says that the link type (second argument) is 0, which is Modbus TCP.  Normally when you communicate over a serial link type needs to be 1=RTU or 2=ASCII.
From: Sintschuk, Michael <michael.sintschuk at bam.de>
Sent: Monday, December 5, 2022 10:30 AM
To: tech-talk <tech-talk at aps.anl.gov>
Cc: kirstaedter at fhi-berlin.mpg.de; jwlodek at bnl.gov; Mark Rivers <rivers at cars.uchicago.edu>
Subject: Eurotherm2604 modbus support
Hello EPICS-community,
I have trouble connecting a Eurotherm2604 temperature controller via modbus with this EPICS-module:
https://github.com/jwlodek/eurotherm2k
A couple of months ago this package was refurbished as you can see from this tech-talk:
https://epics.anl.gov/tech-talk/2022/msg01319.php
I can compile the module and can build the IOC, which works fine via the EI-bisynch protocol. But I can’t establish any connection when it comes to modbus (I need the modbus protocol to read some additional info that is not implemented in EI-bisynch). FYI: the modbus communication on the very same COM-port (same cable) just works fine with the vendor software iTools. This is how my IOC-modbus startup looks like (I’m using the latest modbus-master R3-2):
C:\epics\IOC\EurothermIOC\iocBoot\iocEurotherm>C:\epics\IOC\EurothermIOC\bin\windows-x64-static\Eurotherm.exe st.cmd
#!../../bin/windows-x64-static/Eurotherm
< envPaths
epicsEnvSet("IOC","iocEurotherm")
epicsEnvSet("TOP","C:/epics/IOC/EurothermIOC")
epicsEnvSet("SUPPORT","C:/epics/inst-7.0.3.1")
epicsEnvSet("ASYN","C:/epics/inst-7.0.3.1/asyn-R4-39")
epicsEnvSet("EUROTHERM2K","C:/epics/inst-7.0.3.1/eurotherm2k")
epicsEnvSet("MODBUS","C:/epics/inst-7.0.3.1/modbus")
epicsEnvSet("EPICS_BASE","C:/epics/base-7.0.3.1")
cd "C:/epics/IOC/EurothermIOC"
dbLoadDatabase "dbd/Eurotherm.dbd"
Eurotherm_registerRecordDeviceDriver pdbbase
asynSetOption("EMODBUS", -1, "baud",    "9600")
asynSetOption("EMODBUS", -1, "bits",    "8")
asynSetOption("EMODBUS", -1, "stop",    "1")
asynSetOption("EMODBUS", -1, "parity",  "none")
#asynSetOption("EMODBUS", -1, "crtscts", "Y")
#asynSetOption("EMODBUS", -1, "clocal",  "N")
#asynSetOption("EMODBUS", -1, "ixon", "Y")
#asynOctetSetInputEos( "EMODBUS", -1, "\r\n")
#asynOctetSetOutputEos("EMODBUS", -1, "\r")
eurothermModbusCtrlConfigure(EMODBUS, 1)
2022/12/05 16:51:56.791 drvModbusAsyn::doModbusIO port Eurotherm_EMODBUS_1_0_Rd_
ATUNE error calling writeRead, error=asynManager::queueLockPort queueRequest tim
ed out, nwrite=0/6, nread=0
2022/12/05 16:51:57.797 drvModbusAsyn::doModbusIO port Eurotherm_EMODBUS_1_0_Wr_
ATUNE error calling writeRead, error=, nwrite=6/6, nread=0
eurothermModbusLoopConfigure(EMODBUS, 1, 1)
2022/12/05 16:52:00.802 drvModbusAsyn::doModbusIO port Eurotherm_EMODBUS_1_1_Rd_
PV error calling writeRead, error=asynManager::queueLockPort queueRequest timed
2022/12/05 16:52:01.808 drvModbusAsyn::doModbusIO port Eurotherm_EMODBUS_1_1_Wr_
PV error calling writeRead, error=, nwrite=6/6, nread=0
2022/12/05 16:52:03.814 drvModbusAsyn::doModbusIO port Eurotherm_EMODBUS_1_1_Wr_
RR error calling writeRead, error=, nwrite=6/6, nread=0
2022/12/05 16:52:05.816 drvModbusAsyn::doModbusIO port Eurotherm_EMODBUS_1_1_Wr_
MAN error calling writeRead, error=asynManager::queueLockPort queueRequest timed
2022/12/05 16:52:06.809 drvModbusAsyn::doModbusIO port Eurotherm_EMODBUS_1_1_Rd_
2022/12/05 16:52:07.813 drvModbusAsyn::doModbusIO port Eurotherm_EMODBUS_1_1_Rd_
2022/12/05 16:52:07.814 drvModbusAsyn::doModbusIO port Eurotherm_EMODBUS_1_1_Wr_
PID error calling writeRead, error=asynManager::queueLockPort queueRequest timed
## Load record instances
dbLoadRecords("db/example_modbus_expanded.db","user=bl")
cd "C:/epics/IOC/EurothermIOC/iocBoot/iocEurotherm"
iocInit
Starting iocInit
## EPICS R7.0.3.1
## EPICS Base built Mar  9 2020
2022/12/05 16:52:09.814 drvModbusAsyn::doModbusIO port Eurotherm_EMODBUS_1_1_Rd_
PID error calling writeRead, error=, nwrite=6/6, nread=0
## Start any sequence programs
#seq sncxxx,"user=bl"
epics> 2022/12/05 16:52:12.321 drvModbusAsyn::doModbusIO port Eurotherm_EMODBUS_
1_1_Wr_PV error calling writeRead, error=asynManager::queueLockPort queueRequest
timed out, nwrite=0/6, nread=0
2022/12/05 16:52:12.323 EUROTHERM2K:SP devAsynInt32::processCallbackOutput proce
ss write error
2022/12/05 16:52:12.325 EUROTHERM2K:P devAsynInt32::processCallbackOutput proces
s write error
2022/12/05 16:52:12.330 EUROTHERM2K:RR devAsynInt32::processCallbackOutput proce
ss write error
2022/12/05 16:52:13.414 drvModbusAsyn::doModbusIO port Eurotherm_EMODBUS_1_1_Rd_
RR error calling writeRead, error=asynManager::queueLockPort queueRequest timed
2022/12/05 16:52:13.416 drvModbusAsyn::doModbusIO port Eurotherm_EMODBUS_1_1_Rd_
2022/12/05 16:52:14.332 drvModbusAsyn::doModbusIO port Eurotherm_EMODBUS_1_0_Rd_
ed out, nwrite=6/6, nread=0
2022/12/05 16:52:16.415 drvModbusAsyn::doModbusIO port Eurotherm_EMODBUS_1_1_Rd_
2022/12/05 16:52:22.410 drvModbusAsyn::doModbusIO port Eurotherm_EMODBUS_1_0_Rd_
2022/12/05 16:52:22.419 drvModbusAsyn::doModbusIO port Eurotherm_EMODBUS_1_1_Rd_
2022/12/05 16:52:24.413 drvModbusAsyn::doModbusIO port Eurotherm_EMODBUS_1_1_Rd_
2022/12/05 16:52:24.420 drvModbusAsyn::doModbusIO port Eurotherm_EMODBUS_1_1_Rd_
2022/12/05 16:52:25.423 drvModbusAsyn::doModbusIO port Eurotherm_EMODBUS_1_1_Rd_
2022/12/05 16:52:26.419 drvModbusAsyn::doModbusIO port Eurotherm_EMODBUS_1_1_Rd_
2022/12/05 16:52:27.419 drvModbusAsyn::doModbusIO port Eurotherm_EMODBUS_1_1_Rd_
2022/12/05 16:52:28.418 drvModbusAsyn::doModbusIO port Eurotherm_EMODBUS_1_1_Rd_
Can someone help?
Michael
Michael Sintschuk
Bundesanstalt für Materialforschung und –prüfung (BAM)
8.5 Mikro-ZfP
Unter den Eichen 87
12205 Berlin
GERMANY
P: +49 30 8062-15063 (BESSY, Adlershof)
P: +49 30 8104-4065 (BAM, Steglitz)
michael.sintschuk at helmholtz-berlin.de
AW: Eurotherm2604 modbus support Sintschuk, Michael via Tech-talk
Prev: Eurotherm2604 modbus support Sintschuk, Michael via Tech-talk
Next: AW: Eurotherm2604 modbus support Sintschuk, Michael via Tech-talk
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  <20222023  2024  2025
ANJ, 05 Dec 2022 Valid HTML 4.01! · Home · News · About · Base · Modules · Extensions · Distributions ·
