1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  <2023>  2024  2025  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  <2023>  2024  2025  2026
Subject: RE: Ways to know if instrument connected via TCP/IP through StreamDevice has lost connection
From: "Wang, Andrew via Tech-talk" <[email protected]>
To: Mark Rivers <[email protected]>, EPICS tech-talk <[email protected]>
Date: Tue, 7 Mar 2023 00:09:00 +0000
HI Mark,
Thank you for the reply. I have additional questions, but I will post it on EPICS tech-talk so that everybody can see it.
Andy
Sent: Monday, March 6, 2023 3:41 PM
To: Wang, Andrew <[email protected]>; EPICS tech-talk <[email protected]>
Hi Andy,
With generic asyn device support the timeout is specified in the link:  "@asyn(port, addr, timeout)drvUserString").
With streamDevice it is specified in the protocol file, and there are 3 values that affect the timeout on reads:
LockTimeout (default=5000 ms)
ReplyTimeout (default=1000 ms)
ReadTimeout (default=100 ms)
These are explained here: https://paulscherrerinstitute.github.io/StreamDevice/protocol.html#sysvar
Sent: Monday, March 6, 2023 1:59 PM
To: Mark Rivers <[email protected]>; EPICS tech-talk <[email protected]>
Hi Mark,
As a follow up question, how is the read timeout set? Is this separate from the read timeout that is specified on the protocol file listed under System variables.
Andy
Sent: Monday, December 12, 2022 2:11 PM
Subject: Re: Ways to know if instrument connected via TCP/IP through StreamDevice has lost connection
Mark,
This is extremely helpful. I will let you know if I have follow up questions.
Andy
Sent: Thursday, December 8, 2022 6:51 PM
To: EPICS tech-talk <[email protected]>; Wang, Andrew <[email protected]>
Hi Andy,
> I feel like there is a better way than me trying to use a calcout record and check if any record has its STAT field set to COMM. Like, is it possible to monitor for connection status without needing to process a record whose device type is StreamDevice and check its STAT field?
Yes, there is a better way.
First, you probably want to use this line in your startup script:
asynSetOption("$(PORT)",0, "disconnectOnReadTimeout", "Y")
That will cause the port to disconnect if there is a timeout.  If you don't do that then the asyn won't know that the port is disconnected for a very long time on Linux, it will just keep giving "temporarily unavailable" messages.
In your startup script load an asyn record that is connected to that asyn port.  Here is an example:
dbLoadRecords("$(ASYN)/db/asynRecord.db", "P=IOC:, R=asyn1, PORT=ARIES1, ADDR=0, OMAX=256, IMAX=256")
You can monitor the  IOC:asyn1.CNCT field,  which will either be "Connect" or "Disconnect".
From: Tech-talk <[email protected]> on behalf of Wang, Andrew via Tech-talk <[email protected]>
Sent: Thursday, December 8, 2022 8:21 PM
Subject: Ways to know if instrument connected via TCP/IP through StreamDevice has lost connection
Hi all,
Basically, I was wondering if there is a better way to know if communication to an instrument has been lost. I understand that this can be done either by simplying looking at the IOC shell for disconnection errors or checking if the STAT field is set to COMM.
I was wondering if there is another method of determining if the instrument is still connected or not besides the two methods described above. Additionally, I would like to have a record called "is_instrument_connected", that holds a value like 1 to indicate the instrument is connected and a 0 if it is not. I feel like there is a better way than me trying to use a calcout record and check if any record has its STAT field set to COMM. Like, is it possible to monitor for connection status without needing to process a record whose device type is StreamDevice and check its STAT field?
Thank you,
Andy
RE: Ways to know if instrument connected via TCP/IP through StreamDevice has lost connection Wang, Andrew via Tech-talk
RE: Ways to know if instrument connected via TCP/IP through StreamDevice has lost connection Mark Rivers via Tech-talk
Prev: RE: Ways to know if instrument connected via TCP/IP through StreamDevice has lost connection Mark Rivers via Tech-talk
Next: Re:Re: Segmentation fault (core dumped) using SimDetector ioc lynn via Tech-talk
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  <2023>  2024  2025  2026
Next: Thorlab and Standa Motion stages Ernesto Paiser via Tech-talk
