1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  <2024>  2025  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  <2024>  2025  2026
Subject: Re: StreamDevice+regex for escaping and unescaping HEX messages
From: Sebastian Eckert via Tech-talk <[email protected]>
To: Zimoch Dirk <[email protected]>
Date: Mon, 8 Jan 2024 16:33:10 +0100
Hi Dirk,
It seems that the main issue is the use of bytes outside the ascii range. Using the octal representation for 0x80 and 0x9E seems to work:
"%#/\x5E(?=J)J/\x0A/"
"%#/\x5E(?=Q)Q/\x11/"
"%#/\x5E(?=M)M/\x0D/"
"%#/\x5E(?=S)S/\x13/"
"%#/\x5E(?=\o{200})\o{200}/\x40/"
"%#/\x5E(?=\o{236})\o{236}/\x5E/"
I'll test this on the device.
Best,
Sebastian
On 12/24/23 11:35, Zimoch Dirk wrote:
Does [^] work?
Dirk
Am 23.12.2023 um 15:20 schrieb Sebastian Eckert <[email protected]>:
Hi Dirk,
Thanks for replying while you are on vacation! \^ yields the same behavior.
I would be happy if you could test this when you are back at work.
Happy Holidays!
Sebastian
Am 23.12.2023 um 09:44 schrieb Zimoch Dirk <[email protected]>:
Hi Sebastian,
I am currently on vacation and thus cannot test anything. But you may try to escape the problematic byte as \^
Dirk
Am 22.12.2023 um 20:16 schrieb Sebastian Eckert via Tech-talk <[email protected]>:
Dear StreamDevice+regex experts,
I am trying to communicate with a device with a communication protocol, where messages contain:
- 0x0D (Start of a message)
- a source and a destination byte (to assign replies to requests)
- 0x04 or 0x05 or 0x08 (read, write, data)
- address of the register to access (can be multiple bytes)
- data (can be multiple bytes, depending on data type, little endian)
- CRC16 xmodem checksum (without leading 0x0D)
- 0x0A
Before dressing the message with 0x0D and 0x0A, special characters need to be escaped by adding a 0x5E byte before the corresponding byte and increasing the decimal value of the byte by 64 (0x0A to 0x5E 0x4A, 0x0D to 0x5E 0x4D, 0x11 to 0x5E 0x51, 0x13 to 0x5E 0x53, 0x40 to 0x5E 0x80, 0x5E to 0x5E 0x9E). When a message is received, the message (without the leading 0x0D and 0x0A in the end) needs to be un-escaped in the same fashion.
An example for reading a gain value (0x88 0x13 0x00 0x00, signed 32bit integer, escaped: 0x88 0x5E 0x53 0x00 0x00, dec: 5000) is:
out: 0x0D 0xFF 0x03 0x04 0x01 0x03 0x07 0x02 0x74 0xA6 0x0A
in: 0x0D 0x03 0x42 0x08 0x01 0x88 0x5E 0x53 0x00 0x00 0x90 0xA1 0x0A
Here 0x01 0x03 0x07 0x02 is the address of the register we want to read. In the reply only the first byte of the register (0x01), which was read, is mentioned.
Below is the protocol file, in which I escape and unescape the messages with regular expressions. It works correct for sending commands to the device. The received messages are not un-escaped correctly. In the example above, it just replaces the 0x5E by 0x11. My guess is that the 0x5E (^) creates the biggest problems due to its distinct role in regex. The outcome of the byte transformation also drastically depends on the order in which the individual un-escape regex-expressions are applied. I would appreciate any idea of what could by the problem with the expressions.
Best regards!
Sebastian
---------proto file---------------------
Terminator = LF;
preo = 0x0D,0xFF,0x03;
prei = 0x0D,0x03,0x42;
get = 0x04;
write = 0x05;
gain = 0x01,0x03,0x07,0x02;
esco =    "%#/\x5E/\x5E\x9E/"
                   "%#/\x11/\x5E\x51/"
                   "%#/\x13/\x5E\x53/"
                   "%#/\x40/\x5E\x80/"
                   "%#/(?<!^)\x0D/\x5E\x4D/"
                   "%#/\x0A/\x5E\x4A/";
unesci =    "%#/\x5E\x4A/\x0A/"
                   "%#/\x5E\x51/\x11/"
                   "%#/\x5E\x80/\x40/"
                   "%#/\x5E\x4D/\x0D/"
                   "%#/\x5E\x53/\x13/"
                   "%#/\x5E\x9E/\x5E/";
getGain{
out $preo $get $gain "%1.0<xmodem>" $esco;
in $unesci $prei BS 0x01 "%#4r" "%1.0<xmodem>";
setGain{
out $preo $write $gain "%#4.4r%1.0<xmodem>" $esco;
in  $unesci $prei ETX $write 0x01 "%1.0<xmodem>";
@init{
      in $unesci $prei BS 0x01 "%#4r%1.0<xmodem>";
----------db file---------------------
record (longin,"$(P):gain")
   field (DTYP, "stream")
   field (INP,  "@syncroStream.proto getGain $(PORT)")
   field (PINI,"YES")
record (longout,"$(P):gain_set")
   field (OUT,  "@syncroStream.proto setGain $(PORT)")
Sebastian Eckert
Helmholtz-Zentrum Berlin für Materialien und Energie
BESSY II - PS-ISRR
Albert-Einstein-Str. 15
12489 Berlin
Germany
E-mail: [email protected]
Tel.: +49 30 8062 14465
Prev: RE: Behavior of the soft motor's DINP field Pearson, Matthew via Tech-talk
Next: Help with device support for aao (waveform) Gabriel Fedel via Tech-talk
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  <2024>  2025  2026
Prev: Re: StreamDevice+regex for escaping and unescaping HEX messages Sebastian Eckert via Tech-talk
