Dear Mark,
On 05/11/10 23:24, Mark Rivers wrote:
>>> When you do intermediate cagets while your camonitor is not getting
>>> updates, what timestamps are these showing?
>> They seem to be showing some old timestamp - e.g. few hours back,
> while
>> it should show current date&time stamp.
> That really suggests that your record is not processing, i.e. that the
> problem is not with Channel Access callbacks to clients, but with
> callbacks from your driver to device support.
Right.
Since the I/O Intr scans are used could it be that they do not get sent
out due to high CPU load? For example by changing the scanning to 1
second seems to remedy the situation - scanning periodically is more
deterministic and possibly with higher priority?
>>> Are you sure asyn gets the value you expect it to into the record?
>>> (Check by enabling asyn debug.)
>> I've checked the status of the low level stuff (kernel driver and
>> hardware) and the new value is there..
> But you didn't answer the question: does the RECORD have the correct
> value? Do a "dbpr" on the record at the IOC prompt. If the record does
> not have the correct value then the problem is not with Channel Access
> callbacks, it is with asyn callbacks from your driver to the record.
No, the monitored record DOES NOT have the correct value - hence the old
time stamp with the old value, I reckon.
> You should probably put some printf calls in your driver where it does
> the interrupt callbacks, and/or in devAsynInt32.c where it gets
> interrupt callbacks from the driver.
Will do.
FYI, I followed Jeff's advice and checked the CPU load - one of the
units has constant load over 90% and this is the offending one, that
does not send out CA_PROTO_EVENT_ADD responses - probably because the
monitor did not see the interrupt callback yet!?.
By relaxing the EPICS IOC work, CPU load dropped below 50% and the
CA_PROTO_EVENT_ADD responses start arriving smoothly!
I'll need to investigate further to be sure what the cause/remedy will be.
Best regards,
Hinko
> Mark
> -----Original Message-----
> From: [email protected]
> [mailto:[email protected]] On Behalf Of Hinko Kocevar
> Sent: Tuesday, May 11, 2010 11:01 AM
> To: Ralph Lange
> Cc: [email protected]
> Subject: Re: No monitor refresh
> Hello Ralph,
> On 05/11/10 16:47, Ralph Lange wrote:
>> Hello Hinko,
>> hm ..... it's an I/O scanned mbbi connected to asyn, and I have no
> asyn
>> experience. (So I'm sort of out-of-the-game at this point.)
> Yes it is a record defined like this:
> record(mbbi, "$(DEVICE)ENV_AGC_MONITOR") {
> field(DTYP, "asynInt32")
> field(INP, "@asyn($(PORT), 725, 1.0)")
> field(SCAN, "I/O Intr")
> field(DESC, "AGC")
> field(ZRVL, "0")
> field(ONVL, "1")
> field(ZRST, "MANUAL")
> field(ONST, "AUTO")
> field(PINI, "YES")
> record(mbbo, "$(DEVICE)ENV_AGC_SP") {
> field(OUT, "@asyn($(PORT), 725)")
> field(DESC, "AGC")
> field(ZRVL, "0")
> field(ONVL, "1")
> field(ONST, "AUTO")
> field(PINI, "YES")
> ENV_AGC_MONITOR is used with camonitor, ENV_AGC_SP is used with
> caget/caput. When I set the value with ENV_AGC_SP, ENV_AGC_MONITOR
> should refresh the value.
>> Nevertheless - some thoughts:
>> Your NOBT is zero, but your mask is correct, probably something that
>> asyn does for you.
>> Looked at the record code: An mbbi should definitely send VAL updates
>> when it processes and the value changed.
> Yes it should, but this does not happen always.
>> Are you sure the record gets processed by asyn when you expect it to?
>> (Check by setting the trace field.)
> I want it to react as soon as the value in hardware changes - hence the
> I/O scanning. Hmm, maybe I could try some other Periodic scans to see if
> it helps..
> Will try the asyn debug and traces too..
>> When you set up multiple camonitor clients - do they all behave the
> same?
> Yes, they do.
> I've also tried running camonitor on several clients, even locally on
> the IOC server itself (ARM cross compiled), but without any success. It
> seems to me that EPICS TCP transport layer is not reacting to the change
> They seem to be showing some old timestamp - e.g. few hours back, while
>> (BTW: You only have two states with raw values 0 and 1 defined - any
>> particular reason for using mbbi as opposed to bi?)
> No particular reason at all.. :)
>> Someone with more asyn background could probably say more....
> Thank you for the valuable pointers!
> Best regards,
> Hinko
>> Good luck!
>> Ralph
>> On Tue 11 May 2010 10:07:12 Hinko Kocevar wrote:
>>> Hello Ralph,
>>> On 05/11/10 15:21, Ralph Lange wrote:
>>>> Hello Hinko,
>>>> what EPICS versions are you using for client and IOC?
>>> IOC and client use 3.14.10 EPICS base version.
>>>> Can you show a dump (dbpr <name>, 4) of the record that you are
>>>> connecting to?
>>> Sure, here it is:
>>> epics> dbpr LIBERA01:ENV:ENV_AGC_MONITOR 4
>>> ACKS: NO_ALARM ACKT: YES ASG: ASP:
>>> 0x00000000
>>> BKPT: 0x00 COSV: NO_ALARM DESC: AGC DISA: 0
>>> DISP: 0 DISS: NO_ALARM DISV: 1 DPVT:
>>> 0x50c72c00
>>> DSET: 0x7c620840 DTYP: asynInt32 EIST: EISV:
>>> NO_ALARM
>>> EIVL: 0x0 ELST: ELSV: NO_ALARM ELVL: 0x0
>>> EVNT: 0 FFST: FFSV: NO_ALARM FFVL: 0x0
>>> FLNK:CONSTANT 0 FRST: FRSV: NO_ALARM FRVL: 0x0
>>> FTST: FTSV: NO_ALARM FTVL: 0x0 FVST:
>>> FVSV: NO_ALARM FVVL: 0x0 INP:INST_IO @asyn(ENV, 725,
> 1.0)
>>> LALM: 1 LCNT: 0 LSET: 0xe8752e00 MASK:
>>> 4294967295
>>> MLIS: 0x38d0320060d0320003000000 MLOK: 0x28700b00 MLST: 1
>>> NAME: LIBERA01:ENV:ENV_AGC_MONITOR NIST: NISV:
>>> NO_ALARM
>>> NIVL: 0x0 NOBT: 0 NSEV: NO_ALARM NSTA:
>>> NO_ALARM
>>> ONST: AUTO ONSV: NO_ALARM ONVL: 0x1 ORAW: 1
>>> PACT: 0 PHAS: 0 PINI: YES PPN:
>>> 0x00000000
>>> PPNR: 0x00000000 PRIO: LOW PROC: 0 PUTF: 0
>>> RDES: 0xa8a10300 RPRO: 0 RSET: 0x34b70a40 RVAL: 1
>>> SCAN: I/O Intr SDEF: 1 SDIS:CONSTANT SEVR:
>>> NO_ALARM
>>> SHFT: 0 SIML:CONSTANT SIMM: NO SIMS:
>>> NO_ALARM
>>> SIOL:CONSTANT SPVT: 0x388a2f00 STAT: NO_ALARM SVAL: 0
>>> SVST: SVSV: NO_ALARM SVVL: 0x0 SXST:
>>> SXSV: NO_ALARM SXVL: 0x0 TEST: TESV:
>>> NO_ALARM
>>> TEVL: 0x0 THST: THSV: NO_ALARM THVL: 0x0
>>> TIME: 2010-05-11 11:18:40.100440000 TPRO: 0 TSE: 0
>>> TSEL:CONSTANT TTST: TTSV: NO_ALARM TTVL: 0x0
>>> TVST: TVSV: NO_ALARM TVVL: 0x0 TWST:
>>> TWSV: NO_ALARM TWVL: 0x0 UDF: 0 UNSV:
>>> NO_ALARM
>>> VAL: 1 ZRST: MANUAL ZRSV: NO_ALARM ZRVL: 0x0
>>>> What field of that record are you connecting to?
>>> I did no specify any field - so it is VAL, I guess by default.
>>> Thank you!
>>> Regards,
>>> Hinko
>>>> Regards,
>>>> Ralph
>>>> On Tue 11 May 2010 6:26:48 Hinko Kocevar wrote:
>>>>> Hi all,
>>>>> I'm experiencing some strange behavior when using camonitor against
> an
>>>>> EPICS IOC.
>>>>> For some reason a CA_PROTO_EVENT_ADD response from IOC to CA client
>>>>> does
>>>>> not arrive, or arrives several minutes later after the PV value has
>>>>> changed. Note that first CA_PROTO_EVENT_ADD response is seen
> always,
>>>>> only latter CA_PROTO_EVENT_ADD responses do not arrive (or arrive
>>>>> late)?!
>>>>> This behavior is not seen always - sometimes the IOC would promptly
>>>>> respond with CA_PROTO_EVENT_ADD response when needed.
>>>>> I can see the CA_PROTO_ECHO requests/responses fine on the same
>>>>> camonitor channel link, only CA_PROTO_EVENT_ADD response is not
> seen.
>>>>> Inspecting the IOC I can see that the value did change (on lower
>>>>> levels)
>>>>> and caput response is also valid - updated with new value.
>>>>> I've ruled out the network problems, since the issue is raised even
>>>>> when
>>>>> using cross-over cable to communicate with the IOC. Actually even I
>>>>> update the value in the IOC shell, the camonitor is not refreshed!!
>>>>> Has someone experienced something similar?
>>>>> Are there some conditions that need to fulfilled in order for IOC
> to
>>>>> send a subsequent CA_PROTO_EVENT_ADD response(s)?
>>>>> TIA & HAND,
>>>>> Hinko
Hinko Kocevar
Technical support software engineer
Instrumentation Technologies
Velika pot 22, SI-5250 Solkan - Slovenia
T:+386 5 3352600, F:+386 5 3352601
mailto: [email protected]
http://www.i-tech.si - When your users demand stability
The information transmitted is intended solely for the addressee and may
contain confidential and/or privileged information. Any review, retention,
disclosure or other use by persons other than the intended recipient is
prohibited. If you received this in error, please notify the sender and
delete all copies.
- RE: No monitor refresh Mark Rivers
- Re: No monitor refresh Ralph Lange
- Re: No monitor refresh Hinko Kocevar
RE: MAXv vs XPS Phillip Brown
<2010>
2025
2026
<2010>
2025
2026