1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  <2010>  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  <2010>  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026
Subject: RE: problem with MCA module (Struck 3820 board), ACQUIRING command
To: "Jay Steele" <[email protected]>, <[email protected]>
Date: Fri, 29 Oct 2010 23:16:53 -0500
Hi Jay,
Yes, that is the correct solution.  Thanks for catching that, I'm surprised no one has found it before.  The device will actually work OK without that record, it just increases the latency for reading the MCAs one last time when acquisition completes.
I am not sure about the problem with not seeing counts on input #9.  Can you try a +5V pulse train, maybe 3.3V is not quite enough?
Sent: Fri 10/29/2010 10:26 PM
To: [email protected]
Subject: problem with MCA module (Struck 3820 board), ACQUIRING command
Hi Tech-talk,
    I'm using epics-3-14-11, mca-s55 to try to control a Struck (SIS) 3820 VME board. I'm using a PPC mother board running vxWorks.
    I run the following in my startup script using the examples in mca-s55/iocBoot/iocVxWorks for SIS3820.
mcaSIS3820Setup(0xA0000000, 224, 6)
mcaSIS3820Config("mcaSIS3820/1", 0, 2048, 32, 2, 0, 1000000)
dbLoadTemplate("SIS3820_32.substitutions")
I get the following error after pmacVmeInit:
drvMcaSIS3820Asyn::drvUserCreate, unknown command=ACQUIRING
21:D3:BNP:scalar1:ReadCallback devAsynInt32::initCommon drvUserCreate
    This error seems to come from the following in SIS3820_32.substitutions.
file "$(MCA)/SIS3820.db"
pattern
{P,         INP}
{21:D3:BNP:scalar1:,  "@asyn(mcaSIS3820/1 0)"}
and SIS3820.db:
record(ai,"$(P)ReadCallback") {
       field(SCAN,"I/O Intr")
       field(DTYP, "asynInt32")
       field(INP, "$(INP)ACQUIRING")
       field(FLNK, "$(P)ReadAll.PROC PP MS")
When I change "ACQUIRING" to "MCA_ACQUIRING" (based on a quick review of the source code), the script seems to work correctly. Is this the correct solution?
Also assuming that the correct command is "MCA_ACQUIRING" - when I run scaler32_full.adl to see the data from the SIS3820 board and have a signal generator generating a 3.3V pulse train at 25 Hz connected to input #9, I only see the counts from the internal clock at input #1. Anybody have any hints on how to activate input #9? Attached is the screen shot from the MEDM screen.
Jay Steele
XRadia Corporation
The information in this email, including any attachments, is confidential and intended only for the recipient(s) listed. Any use of this email for any other purpose is prohibited. If you have received this email in error, please notify me immediately by reply email, delete this email, and do not disclose its contents to anyone.
problem with MCA module (Struck 3820 board), ACQUIRING command Jay Steele
Prev: problem with MCA module (Struck 3820 board), ACQUIRING command Jay Steele
Next: RE: cygin-x86-debug behavior Mark Rivers
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  <2010>  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026
Next: readline for iocSh Rod Nussbaumer
ANJ, 31 Oct 2010 · Home · News · About · Talk · Base · Modules · Extensions ·