1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  <2014>  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  <2014>  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026
Subject: RE: Areadetector IOC for Pilatus
To: "'Specht, Eliot D.'" <[email protected]>, "[email protected]" <[email protected]>
Date: Fri, 15 Aug 2014 21:59:00 +0000
Hi Eliot,
The only error messages I see are this one from camserver:
*Illegal exposure period
and this one from the IOC:
2014/08/15 22:21:55.924 pilatusDetector:readCamserver unexpected response from camserver, no OK, response=15 ERR Illegal exposure period
They are really the same error.  It just means that the EPICS AcquirePeriod PV is not at least 6 ms or so larger than the AcquireTime (I am not sure of the exact value required on the 200K).  If you increase the AcquirePeriod to AcquireTime + 0.01 I think those messages will go away.
But those errors are not "serious" in that camserver will still behave sensibly.  They are not causing this issue:
> but PVs such as 13PIL1:cam1:ArraySizeY_RBV are not found.
What do you mean by that?  Do you have a channel access client like medm that is not able to see those PVs?  If so, that could be because you have a firewall running on your machine.
You can try:
sudo /etc/init.d/iptables stop
To disable the iptables firewall permanently:
sudo chkconfig iptables off
-----Original Message-----
From: [email protected] [mailto:[email protected]] On Behalf Of Specht, Eliot D.
Sent: Friday, August 15, 2014 3:59 PM
To: [email protected]
Subject: Areadetector IOC for Pilatus
I'm setting up a new Pilatus 3R 200K-A detector. The Dectris runtvx utility communicates properly with the detector, and camserver responds to telnet commands. I'm trying to use epics, which is installed on the computer controlling the detector. I'm running areadetector 1.9.1 straight out of the box; I changed only the hostname and the detector XSIZE.
There is some communication between camserver and areadetector, but PVs such as 13PIL1:cam1:ArraySizeY_RBV​ are not found. I'm a novice with epics, and I don't know what to make of the "unexpected response" and "illegal exposure period" diagnostics.
The lengthy output from camserver and the ioc follow.
Eliot Specht
Oak Ridge National Lab
camserver output
camserver
Read detector hardware definitions from:
 /var/local/lib/dectris/config/cam_data/pidet.def
Building standard xor table
Configured from /etc/dectris/camrc
Camserver configured for PILATUS3 200K, S/N 2-0191
Setting default LUT directory for rate correction
Reading LUT from
 /var/local/lib/dectris/config/cam_data/ratecorrection/ContinuousStandard_v1.1
After rate correction, cutoff = 1165942 counts
Code release: 7.5.14
DCBe scan:
DCBe #1 firmware version: 0x0104
DCBe #1 board assembly version: 1
DCBe #1 readout clock frequency: 150.0 MHz
Temperature/Humidity sensor channel #3 enabled on DCBe #1
Camera image format: 487(w) x 407(h) pixels
Camserver listening on port 41234
Reading camserver startup file:
 /var/local/lib/dectris/config/cam_data/pidet.set
# set of startup commands for camserver
set B15_M7_VTRM 1.8
Setting B**_M**_VTRM voltage to integer: 0x2e6 = ~1.801V
set B15_M7_VRFS 0.7
Setting B**_M**_VRFS voltage to integer: 0x11f = ~0.699V
set B15_M7_VDEL 0.8
Setting B**_M**_VDEL voltage to integer: 0x149 = ~0.801V
set B15_M7_VRF -0.2
Setting B**_M**_VRF voltage to integer: 0xd8 = ~-0.200V
set B15_M7_VCAL 0.3
Setting B**_M**_VCAL voltage to integer: 0x7a = ~0.299V
set B15_M7_VCCA 0.7
Setting B**_M**_VCCA voltage to integer: 0x11f = ~0.699V
set B15_M7_VCMP0 0.7
Setting B**_M**_VCMP0 voltage to integer: 0x11f = ~0.699V
set B15_M7_VCMP1 0.7
Setting B**_M**_VCMP1 voltage to integer: 0x11f = ~0.699V
set B15_M7_VCMP2 0.7
Setting B**_M**_VCMP2 voltage to integer: 0x11f = ~0.699V
set B15_M7_VCMP3 0.7
Setting B**_M**_VCMP3 voltage to integer: 0x11f = ~0.699V
set B15_M7_VCMP4 0.7
Setting B**_M**_VCMP4 voltage to integer: 0x11f = ~0.699V
set B15_M7_VCMP5 0.7
Setting B**_M**_VCMP5 voltage to integer: 0x11f = ~0.699V
set B15_M7_VCMP6 0.7
Setting B**_M**_VCMP6 voltage to integer: 0x11f = ~0.699V
set B15_M7_VCMP7 0.7
Setting B**_M**_VCMP7 voltage to integer: 0x11f = ~0.699V
set B15_M7_VCMP8 0.7
Setting B**_M**_VCMP8 voltage to integer: 0x11f = ~0.699V
set B15_M7_VCMP9 0.7
Setting B**_M**_VCMP9 voltage to integer: 0x11f = ~0.699V
set B15_M7_VCMP10 0.7
Setting B**_M**_VCMP10 voltage to integer: 0x11f = ~0.699V
set B15_M7_VCMP11 0.7
Setting B**_M**_VCMP11 voltage to integer: 0x11f = ~0.699V
set B15_M7_VCMP12 0.7
Setting B**_M**_VCMP12 voltage to integer: 0x11f = ~0.699V
set B15_M7_VCMP13 0.7
Setting B**_M**_VCMP13 voltage to integer: 0x11f = ~0.699V
set B15_M7_VCMP14 0.7
Setting B**_M**_VCMP14 voltage to integer: 0x11f = ~0.699V
set B15_M7_VCMP15 0.7
Setting B**_M**_VCMP15 voltage to integer: 0x11f = ~0.699V
prog CHSEL 0xffff
Setting B**_M**_CHSEL PATTERN to 0xffff (hardware pattern: 0xffff)
THread
Channel 3: Temperature = 30.1C, Rel. Humidity = 2.7%
dbglvl 1
SetRetriggerMode 1
Retrigger mode is set
Reading LUT from
After rate correction, cutoff = 1163856 counts
SetCu
Requested setting: autog; threshold: 4024 eV
Trim directories:
 /var/local/lib/dectris/config/calibration/calib_v/
 /var/local/lib/dectris/config/calibration/calib_fe/
100% [========================================================================]
2 excluded pixels were read
Loading commands from:
 /tmp/setthreshold.cmd
Imgonly /tmp/trim_values_img.tif
Flat field correction loaded from
  /var/local/lib/dectris/config/calibration/FF_p2-0191_E8048_T4024_vrf_m0p100.tif
Bad pixels loaded from
  /var/local/lib/dectris/config/calibration/Mask/badpix_mask.tif
Energy setting: 8048 eV
 Settings: autog (vrf_avg = -0.093); threshold: 4024 eV; vcmp: 0.714 V
 Trim file:
 /var/local/lib/dectris/config/calibration/p2-0191_E8048_T4024.bin
"/tmp/setthreshold.cmd" was successful
Reading LUT from
After rate correction, cutoff = 1409688 counts
Gapfill -1
Set detector gap-fill to: -1
Connection request from 160.91.158.17
Setting B01_M01_VTRM voltage to integer: 0x238 = ~1.380V
Setting B01_M01_VRF voltage to integer: 0x112 = ~-0.093V
Setting B01_M01_VRFS voltage to integer: 0x172 = ~0.900V
Setting B01_M01_VCAL voltage to integer: 0x10b = ~0.651V
Setting B01_M01_VDEL voltage to integer: 0x162 = ~0.861V
Setting B01_M01_VADJ voltage to integer: 0x343 = ~1.197V
Setting B01_M01_VCCA voltage to integer: 0x149 = ~0.801V
Setting B01_M01_VCMP0 voltage to integer: 0x125 = ~0.714V
Setting B01_M01_VCMP1 voltage to integer: 0x11f = ~0.699V
Setting B01_M01_VCMP2 voltage to integer: 0x125 = ~0.714V
Setting B01_M01_VCMP3 voltage to integer: 0x120 = ~0.701V
Setting B01_M01_VCMP4 voltage to integer: 0x11d = ~0.694V
Setting B01_M01_VCMP5 voltage to integer: 0x121 = ~0.704V
Setting B01_M01_VCMP6 voltage to integer: 0x124 = ~0.711V
Setting B01_M01_VCMP7 voltage to integer: 0x11f = ~0.699V
Setting B01_M01_VCMP8 voltage to integer: 0x123 = ~0.709V
Setting B01_M01_VCMP9 voltage to integer: 0x123 = ~0.709V
Setting B01_M01_VCMP10 voltage to integer: 0x120 = ~0.701V
Setting B01_M01_VCMP11 voltage to integer: 0x11e = ~0.697V
Setting B01_M01_VCMP12 voltage to integer: 0x121 = ~0.704V
Setting B01_M01_VCMP13 voltage to integer: 0x122 = ~0.706V
Setting B01_M01_VCMP14 voltage to integer: 0x11d = ~0.694V
Setting B01_M01_VCMP15 voltage to integer: 0x121 = ~0.704V
Setting B02_M01_VTRM voltage to integer: 0x238 = ~1.380V
Setting B02_M01_VRF voltage to integer: 0x112 = ~-0.093V
Setting B02_M01_VRFS voltage to integer: 0x172 = ~0.900V
Setting B02_M01_VCAL voltage to integer: 0x10b = ~0.651V
Setting B02_M01_VDEL voltage to integer: 0x162 = ~0.861V
Setting B02_M01_VADJ voltage to integer: 0x33a = ~1.185V
Setting B02_M01_VCCA voltage to integer: 0x149 = ~0.801V
Setting B02_M01_VCMP0 voltage to integer: 0x11d = ~0.694V
Setting B02_M01_VCMP1 voltage to integer: 0x128 = ~0.721V
Setting B02_M01_VCMP2 voltage to integer: 0x12b = ~0.728V
Setting B02_M01_VCMP3 voltage to integer: 0x12f = ~0.738V
Setting B02_M01_VCMP4 voltage to integer: 0x12b = ~0.728V
Setting B02_M01_VCMP5 voltage to integer: 0x127 = ~0.718V
Setting B02_M01_VCMP6 voltage to integer: 0x12f = ~0.738V
Setting B02_M01_VCMP7 voltage to integer: 0x123 = ~0.709V
Setting B02_M01_VCMP8 voltage to integer: 0x128 = ~0.721V
Setting B02_M01_VCMP9 voltage to integer: 0x12c = ~0.731V
Setting B02_M01_VCMP10 voltage to integer: 0x12f = ~0.738V
Setting B02_M01_VCMP11 voltage to integer: 0x128 = ~0.721V
Setting B02_M01_VCMP12 voltage to integer: 0x121 = ~0.704V
Setting B02_M01_VCMP13 voltage to integer: 0x11e = ~0.697V
Setting B02_M01_VCMP14 voltage to integer: 0x127 = ~0.718V
Setting B02_M01_VCMP15 voltage to integer: 0x126 = ~0.716V
Setting B**_M**_CHSEL0 to 1
Setting B**_M**_CHSEL PATTERN to 0x0 (hardware pattern: 0x0)
Setting B**_M**_CHSEL1 to 1
Setting B**_M**_CHSEL2 to 1
Setting B**_M**_CHSEL3 to 1
Setting B**_M**_CHSEL4 to 1
Setting B**_M**_CHSEL5 to 1
Setting B**_M**_CHSEL6 to 1
Setting B**_M**_CHSEL7 to 1
Setting B**_M**_CHSEL8 to 1
Setting B**_M**_CHSEL9 to 1
Setting B**_M**_CHSEL10 to 1
Setting B**_M**_CHSEL11 to 1
Setting B**_M**_CHSEL12 to 1
Setting B**_M**_CHSEL13 to 1
Setting B**_M**_CHSEL14 to 1
Setting B**_M**_CHSEL15 to 1
Connection has full control; pid = 11582
*mxsettings: Oscillation_axis X, CW
*mxsettings: cbf_template_file 0
*Image path:
 /root/images/
*N images set to: 1
*Exposures per frame set to: 1
*Exposure time set to: 1.0000000 sec.
*Delay time set to: 0.0000000 sec.
*tau = 0.0 s
*N images set to: 1
*tau = 0.0 s
*mxsettings: Alpha 0.000000
*mxsettings: Angle_increment 0.100000
*mxsettings: Beam_xy 0.000000,0.000000
*mxsettings: Chi 0.000000
*N images set to: 1
*tau = 0.0 s
*mxsettings: Detector_2theta 0.000000
*mxsettings: Detector_distance 1.000000
*mxsettings: Detector_Voffset 0.000000
*mxsettings: Energy_range 0.000000,0.000000
*mxsettings: Filter_transmission 1.000000
*mxsettings: Flux 0.000000
*mxsettings: Kappa 0.000000
*mxsettings: Phi 0.000000
*mxsettings: Polarization 0.990000
*mxsettings: Start_angle 0.000000
*mxsettings: Wavelength 1.540000
*N images set to: 1
*tau = 0.0 s
*N images set to: 1
*tau = 0.0 s
*mxsettings: N_oscillations 1
*N images set to: 1
*tau = 0.0 s
*N images set to: 1
*tau = 0.0 s
*Code release: 7.5.14
*Channel 3: Temperature = 30.1C, Rel. Humidity = 2.7%
Closing the socket & exiting child process
start_epics output
./start_epics
< envPaths
epicsEnvSet("ARCH","linux-x86_64")
epicsEnvSet("IOC","iocPilatus")
epicsEnvSet("TOP","/opt/epics/synApps_5_7/support/areaDetector-1-9-1")
epicsEnvSet("SUPPORT","/opt/epics/synApps/support")
epicsEnvSet("ASYN","/opt/epics/synApps/support/asyn-4-21")
epicsEnvSet("CALC","/opt/epics/synApps/support/calc-3-2")
epicsEnvSet("BUSY","/opt/epics/synApps/support/busy-1-6")
epicsEnvSet("SSCAN","/opt/epics/synApps/support/sscan-2-9")
epicsEnvSet("AUTOSAVE","/opt/epics/synApps/support/autosave-5-1")
epicsEnvSet("AREA_DETECTOR","/opt/epics/synApps/support/areaDetector-1-9-1")
epicsEnvSet("EPICS_BASE","/opt/epics/base")
errlogInit(20000)
dbLoadDatabase("/opt/epics/synApps/support/areaDetector-1-9-1/dbd/pilatusDetectorApp.dbd")
medmInit: Unable to load font widgetDM_4
  Trying default (fixed) instead
medmInit: Unable to load font widgetDM_6
medmInit: Unable to load font widgetDM_8
medmInit: Unable to load font widgetDM_10
medmInit: Unable to load font widgetDM_12
medmInit: Unable to load font widgetDM_14
medmInit: Unable to load font widgetDM_16
medmInit: Unable to load font widgetDM_18
medmInit: Unable to load font widgetDM_20
medmInit: Unable to load font widgetDM_22
medmInit: Unable to load font widgetDM_24
medmInit: Unable to load font widgetDM_30
medmInit: Unable to load font widgetDM_36
medmInit: Unable to load font widgetDM_40
medmInit: Unable to load font widgetDM_48
medmInit: Unable to load font widgetDM_60
MEDM Version 3.1.9: Loading aliased fonts.................
pilatusDetectorApp_registerRecordDeviceDriver(pdbbase)
epicsEnvSet("PREFIX", "13PIL1:")
epicsEnvSet("PORT",   "PIL")
epicsEnvSet("QSIZE",  "20")
epicsEnvSet("XSIZE",  "487")
epicsEnvSet("YSIZE",  "407")
epicsEnvSet("NCHANS", "2048")
# Create the asyn port to talk to the Pilatus on port 41234.
drvAsynIPPortConfigure("camserver","pilatus:41234")
# Set the input and output terminators.
asynOctetSetInputEos("camserver", 0, "\030")
asynOctetSetOutputEos("camserver", 0, "\n")
pilatusDetectorConfig("PIL", "camserver", 487, 407, 50, 200000000)
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/ADBase.template", "P=13PIL1:,R=cam1:,PORT=PIL,ADDR=0,TIMEOUT=1")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDFile.template", "P=13PIL1:,R=cam1:,PORT=PIL,ADDR=0,TIMEOUT=1")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/pilatus.template","P=13PIL1:,R=cam1:,PORT=PIL,ADDR=0,TIMEOUT=1,CAMSERVER_PORT=camserver")
# Create a standard arrays plugin
NDStdArraysConfigure("Image1", 5, 0, "PIL", 0, 0)
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDPluginBase.template","P=13PIL1:,R=image1:,PORT=Image1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PIL,NDARRAY_ADDR=0")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDStdArrays.template", "P=13PIL1:,R=image1:,PORT=Image1,ADDR=0,TIMEOUT=1,TYPE=Int32,FTVL=LONG,NELEMENTS=94965")
# Load all other plugins using commonPlugins.cmd
< ../commonPlugins.cmd
# Create a netCDF file saving plugin.
NDFileNetCDFConfigure("FileNetCDF1", 20, 0, "PIL", 0)
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDPluginBase.template","P=13PIL1:,R=netCDF1:,PORT=FileNetCDF1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PIL,NDARRAY_ADDR=0")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDFile.template",      "P=13PIL1:,R=netCDF1:,PORT=FileNetCDF1,ADDR=0,TIMEOUT=1")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDFileNetCDF.template","P=13PIL1:,R=netCDF1:,PORT=FileNetCDF1,ADDR=0,TIMEOUT=1")
# Create a TIFF file saving plugin
NDFileTIFFConfigure("FileTIFF1", 20, 0, "PIL", 0)
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDPluginBase.template","P=13PIL1:,R=TIFF1:,PORT=FileTIFF1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PIL,NDARRAY_ADDR=0")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDFile.template",      "P=13PIL1:,R=TIFF1:,PORT=FileTIFF1,ADDR=0,TIMEOUT=1")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDFileTIFF.template",  "P=13PIL1:,R=TIFF1:,PORT=FileTIFF1,ADDR=0,TIMEOUT=1")
# Create a JPEG file saving plugin
NDFileJPEGConfigure("FileJPEG1", 20, 0, "PIL", 0)
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDPluginBase.template","P=13PIL1:,R=JPEG1:,PORT=FileJPEG1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PIL,NDARRAY_ADDR=0")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDFile.template",      "P=13PIL1:,R=JPEG1:,PORT=FileJPEG1,ADDR=0,TIMEOUT=1")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDFileJPEG.template",  "P=13PIL1:,R=JPEG1:,PORT=FileJPEG1,ADDR=0,TIMEOUT=1")
# Create a NeXus file saving plugin
NDFileNexusConfigure("FileNexus1", 20, 0, "PIL", 0)
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDPluginBase.template","P=13PIL1:,R=Nexus1:,PORT=FileNexus1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PIL,NDARRAY_ADDR=0")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDFile.template",      "P=13PIL1:,R=Nexus1:,PORT=FileNexus1,ADDR=0,TIMEOUT=1")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDFileNexus.template", "P=13PIL1:,R=Nexus1:,PORT=FileNexus1,ADDR=0,TIMEOUT=1")
# Create an HDF5 file saving plugin
NDFileHDF5Configure("FileHDF1", 20, 0, "PIL", 0)
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDPluginBase.template","P=13PIL1:,R=HDF1:,PORT=FileHDF1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PIL,NDARRAY_ADDR=0")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDFile.template",      "P=13PIL1:,R=HDF1:,PORT=FileHDF1,ADDR=0,TIMEOUT=1")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDFileHDF5.template",  "P=13PIL1:,R=HDF1:,PORT=FileHDF1,ADDR=0,TIMEOUT=1")
# Create a Magick file saving plugin
NDFileMagickConfigure("FileMagick1", 20, 0, "PIL", 0)
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDPluginBase.template","P=13PIL1:,R=Magick1:,PORT=FileMagick1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PIL,NDARRAY_ADDR=0")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDFile.template",      "P=13PIL1:,R=Magick1:,PORT=FileMagick1,ADDR=0,TIMEOUT=1")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDFileMagick.template","P=13PIL1:,R=Magick1:,PORT=FileMagick1,ADDR=0,TIMEOUT=1")
# Create 4 ROI plugins
NDROIConfigure("ROI1", 20, 0, "PIL", 0, 0, 0)
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDPluginBase.template","P=13PIL1:,R=ROI1:,  PORT=ROI1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PIL,NDARRAY_ADDR=0")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDROI.template",       "P=13PIL1:,R=ROI1:,  PORT=ROI1,ADDR=0,TIMEOUT=1")
NDROIConfigure("ROI2", 20, 0, "PIL", 0, 0, 0)
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDPluginBase.template","P=13PIL1:,R=ROI2:,  PORT=ROI2,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PIL,NDARRAY_ADDR=0")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDROI.template",       "P=13PIL1:,R=ROI2:,  PORT=ROI2,ADDR=0,TIMEOUT=1")
NDROIConfigure("ROI3", 20, 0, "PIL", 0, 0, 0)
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDPluginBase.template","P=13PIL1:,R=ROI3:,  PORT=ROI3,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PIL,NDARRAY_ADDR=0")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDROI.template",       "P=13PIL1:,R=ROI3:,  PORT=ROI3,ADDR=0,TIMEOUT=1")
NDROIConfigure("ROI4", 20, 0, "PIL", 0, 0, 0)
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDPluginBase.template","P=13PIL1:,R=ROI4:,  PORT=ROI4,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PIL,NDARRAY_ADDR=0")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDROI.template",       "P=13PIL1:,R=ROI4:,  PORT=ROI4,ADDR=0,TIMEOUT=1")
# Create a processing plugin
NDProcessConfigure("PROC1", 20, 0, "PIL", 0, 0, 0)
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDPluginBase.template","P=13PIL1:,R=Proc1:,  PORT=PROC1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PIL,NDARRAY_ADDR=0")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDProcess.template",   "P=13PIL1:,R=Proc1:,  PORT=PROC1,ADDR=0,TIMEOUT=1")
# Create 5 statistics plugins
NDStatsConfigure("STATS1", 20, 0, "PIL", 0, 0, 0)
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDPluginBase.template","P=13PIL1:,R=Stats1:,  PORT=STATS1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ROI1,NDARRAY_ADDR=0")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDStats.template",     "P=13PIL1:,R=Stats1:,  PORT=STATS1,ADDR=0,TIMEOUT=1,HIST_SIZE=256,XSIZE=487,YSIZE=407,NCHANS=2048")
NDStatsConfigure("STATS2", 20, 0, "ROI1", 0, 0, 0)
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDPluginBase.template","P=13PIL1:,R=Stats2:,  PORT=STATS2,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ROI2,NDARRAY_ADDR=0")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDStats.template",     "P=13PIL1:,R=Stats2:,  PORT=STATS2,ADDR=0,TIMEOUT=1,HIST_SIZE=256,XSIZE=487,YSIZE=407,NCHANS=2048")
NDStatsConfigure("STATS3", 20, 0, "ROI2", 0, 0, 0)
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDPluginBase.template","P=13PIL1:,R=Stats3:,  PORT=STATS3,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ROI3,NDARRAY_ADDR=0")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDStats.template",     "P=13PIL1:,R=Stats3:,  PORT=STATS3,ADDR=0,TIMEOUT=1,HIST_SIZE=256,XSIZE=487,YSIZE=407,NCHANS=2048")
NDStatsConfigure("STATS4", 20, 0, "ROI3", 0, 0, 0)
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDPluginBase.template","P=13PIL1:,R=Stats4:,  PORT=STATS4,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ROI4,NDARRAY_ADDR=0")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDStats.template",     "P=13PIL1:,R=Stats4:,  PORT=STATS4,ADDR=0,TIMEOUT=1,HIST_SIZE=256,XSIZE=487,YSIZE=407,NCHANS=2048")
NDStatsConfigure("STATS5", 20, 0, "ROI4", 0, 0, 0)
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDPluginBase.template","P=13PIL1:,R=Stats5:,  PORT=STATS5,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PIL,NDARRAY_ADDR=0")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDStats.template",     "P=13PIL1:,R=Stats5:,  PORT=STATS5,ADDR=0,TIMEOUT=1,HIST_SIZE=256,XSIZE=487,YSIZE=407,NCHANS=2048")
# Create a transform plugin
NDTransformConfigure("TRANS1", 20, 0, "PIL", 0, 0, 0)
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDPluginBase.template","P=13PIL1:,R=Trans1:,  PORT=TRANS1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PIL,NDARRAY_ADDR=0")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDTransform.template", "P=13PIL1:,R=Trans1:,  PORT=TRANS1,ADDR=0,TIMEOUT=1")
# Create an overlay plugin with 8 overlays
NDOverlayConfigure("OVER1", 20, 0, "PIL", 0, 8, 0, 0)
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDPluginBase.template","P=13PIL1:,R=Over1:, PORT=OVER1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PIL,NDARRAY_ADDR=0")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDOverlay.template",   "P=13PIL1:,R=Over1:, PORT=OVER1,ADDR=0,TIMEOUT=1")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDOverlayN.template","P=13PIL1:,R=Over1:1:,NAME=ROI1,   SHAPE=1,O=Over1:,XPOS=13PIL1:ROI1:MinX_RBV,YPOS=13PIL1:ROI1:MinY_RBV,XSIZE=13PIL1:ROI1:SizeX_RBV,YSIZE=13PIL1:ROI1:SizeY_RBV,PORT=OVER1,ADDR=0,TIMEOUT=1")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDOverlayN.template","P=13PIL1:,R=Over1:2:,NAME=ROI2,   SHAPE=1,O=Over1:,XPOS=13PIL1:ROI2:MinX_RBV,YPOS=13PIL1:ROI2:MinY_RBV,XSIZE=13PIL1:ROI2:SizeX_RBV,YSIZE=13PIL1:ROI2:SizeY_RBV,PORT=OVER1,ADDR=1,TIMEOUT=1")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDOverlayN.template","P=13PIL1:,R=Over1:3:,NAME=ROI3,   SHAPE=1,O=Over1:,XPOS=13PIL1:ROI3:MinX_RBV,YPOS=13PIL1:ROI3:MinY_RBV,XSIZE=13PIL1:ROI3:SizeX_RBV,YSIZE=13PIL1:ROI3:SizeY_RBV,PORT=OVER1,ADDR=2,TIMEOUT=1")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDOverlayN.template","P=13PIL1:,R=Over1:4:,NAME=ROI4,   SHAPE=1,O=Over1:,XPOS=13PIL1:ROI4:MinX_RBV,YPOS=13PIL1:ROI4:MinY_RBV,XSIZE=13PIL1:ROI4:SizeX_RBV,YSIZE=13PIL1:ROI4:SizeY_RBV,PORT=OVER1,ADDR=3,TIMEOUT=1")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDOverlayN.template","P=13PIL1:,R=Over1:5:,NAME=Cursor1,SHAPE=1,O=Over1:,XPOS=junk,                  YPOS=junk,                  XSIZE=junk,                   YSIZE=junk,                   PORT=OVER1,ADDR=4,TIMEOUT=1")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDOverlayN.template","P=13PIL1:,R=Over1:6:,NAME=Cursor2,SHAPE=1,O=Over1:,XPOS=junk,                  YPOS=junk,                  XSIZE=junk,                   YSIZE=junk,                   PORT=OVER1,ADDR=5,TIMEOUT=1")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDOverlayN.template","P=13PIL1:,R=Over1:7:,NAME=Box1,   SHAPE=1,O=Over1:,XPOS=junk,                  YPOS=junk,                  XSIZE=junk,                   YSIZE=junk,                   PORT=OVER1,ADDR=6,TIMEOUT=1")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDOverlayN.template","P=13PIL1:,R=Over1:8:,NAME=Box2,   SHAPE=1,O=Over1:,XPOS=junk,                  YPOS=junk,                  XSIZE=junk,                   YSIZE=junk,                   PORT=OVER1,ADDR=7,TIMEOUT=1")
# Create 2 color conversion plugins
NDColorConvertConfigure("CC1", 20, 0, "PIL", 0, 0, 0)
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDPluginBase.template",   "P=13PIL1:,R=CC1:,  PORT=CC1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PIL,NDARRAY_ADDR=0")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDColorConvert.template", "P=13PIL1:,R=CC1:,  PORT=CC1,ADDR=0,TIMEOUT=1")
NDColorConvertConfigure("CC2", 20, 0, "PIL", 0, 0, 0)
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDPluginBase.template",   "P=13PIL1:,R=CC2:,  PORT=CC2,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PIL,NDARRAY_ADDR=0")
dbLoadRecords("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db/NDColorConvert.template", "P=13PIL1:,R=CC2:,  PORT=CC2,ADDR=0,TIMEOUT=1")
# Load scan records
dbLoadRecords("/opt/epics/synApps/support/sscan-2-9/sscanApp/Db/scan.db", "P=13PIL1:,MAXPTS1=2000,MAXPTS2=200,MAXPTS3=20,MAXPTS4=10,MAXPTSH=10")
# Load sseq record for acquisition sequence
dbLoadRecords("/opt/epics/synApps/support/calc-3-2/calcApp/Db/yySseq.db", "P=13PIL1:, S=AcquireSequence")
set_requestfile_path("./")
set_requestfile_path("/opt/epics/synApps/support/areaDetector-1-9-1/ADApp/Db")
set_requestfile_path("/opt/epics/synApps/support/calc-3-2/calcApp/Db")
set_requestfile_path("/opt/epics/synApps/support/sscan-2-9/sscanApp/Db")
set_savefile_path("./autosave")
set_pass0_restoreFile("auto_settings.sav")
set_pass1_restoreFile("auto_settings.sav")
save_restoreSet_status_prefix("13PIL1:")
dbLoadRecords("/opt/epics/synApps/support/autosave-5-1/asApp/Db/save_restoreStatus.db", "P=13PIL1:")
#asynSetTraceMask("$(PORT)",0,255)
asynSetTraceMask("PIL",0,3)
iocInit()
Starting iocInit
## EPICS R3.14.12.4 $Date: Mon 2013-12-16 15:51:45 -0600$
## EPICS Base built Aug 12 2014
reboot_restore: entry for file 'auto_settings.sav'
reboot_restore (v5.1): entry for file 'auto_settings.sav'
reboot_restore: Found filename 'auto_settings.sav' in restoreFileList.
*** restoring from './autosave/auto_settings.sav' at initHookState 6 (before record/device init) ***
reboot_restore: done with file 'auto_settings.sav'
2014/08/15 22:21:55.403 asynInt32SyncIO read: 0
2014/08/15 22:21:55.403 asynInt32SyncIO read: 487
2014/08/15 22:21:55.403 asynInt32SyncIO read: 407
2014/08/15 22:21:55.405 asynInt32SyncIO read: 0
2014/08/15 22:21:55.405 asynInt32SyncIO read: 5
2014/08/15 22:21:55.405 asynInt32SyncIO read: 2
2014/08/15 22:21:55.406 asynInt32SyncIO read: 0
*** restoring from './autosave/auto_settings.sav' at initHookState 7 (after record/device init) ***
The CA server's beacon address list was empty after initialization?
# save things every thirty seconds
create_monitor_set("auto_settings.req", 30,"P=13PIL1:,D=cam1:")
2014/08/15 22:21:55.916 13PIL1:cam1:ArraySizeX_RBV devAsynInt32::interruptCallbackInput new value=487
2014/08/15 22:21:55.916 13PIL1:cam1:ArraySizeY_RBV devAsynInt32::interruptCallbackInput new value=407
2014/08/15 22:21:55.916 13PIL1:cam1:ArraySizeZ_RBV devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.916 13PIL1:cam1:ArraySize_RBV devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.916 13PIL1:cam1:DataType_RBV devAsynInt32::interruptCallbackInput new value=5
2014/08/15 22:21:55.916 13PIL1:cam1:ColorMode_RBV devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.916 13PIL1:cam1:ArrayCounter_RBV devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.916 13PIL1:cam1:WriteFile devAsynBusy::interruptCallback new value=0
2014/08/15 22:21:55.916 13PIL1:cam1:WriteFile_RBV devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.916 13PIL1:cam1:ReadFile devAsynBusy::interruptCallback new value=0
2014/08/15 22:21:55.916 13PIL1:cam1:ReadFile_RBV devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.916 13PIL1:cam1:Capture devAsynBusy::interruptCallback new value=0
2014/08/15 22:21:55.916 13PIL1:cam1:Capture_RBV devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.916 13PIL1:cam1:WriteStatus devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.916 13PIL1:cam1:NumCaptured_RBV devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.916 13PIL1:cam1:MaxSizeX_RBV devAsynInt32::interruptCallbackInput new value=487
2014/08/15 22:21:55.916 13PIL1:cam1:MaxSizeY_RBV devAsynInt32::interruptCallbackInput new value=407
2014/08/15 22:21:55.916 13PIL1:cam1:DetectorState_RBV devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.916 13PIL1:cam1:NumImagesCounter_RBV devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.916 13PIL1:cam1:NumExposuresCounter_RBV devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.916 13PIL1:cam1:TimeRemaining_RBV devAsynFloat64::interruptCallbackInput new value=0.000000
2014/08/15 22:21:55.916 13PIL1:cam1:ShutterStatus_RBV devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.916 13PIL1:cam1:Acquire devAsynBusy::interruptCallback new value=0
2014/08/15 22:21:55.916 13PIL1:cam1:Acquire_RBV devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.916 13PIL1:cam1:SizeX_RBV devAsynInt32::interruptCallbackInput new value=487
2014/08/15 22:21:55.916 13PIL1:cam1:SizeY_RBV devAsynInt32::interruptCallbackInput new value=407
2014/08/15 22:21:55.916 13PIL1:cam1:ImageMode_RBV devAsynInt32::interruptCallbackInput new value=2
2014/08/15 22:21:55.916 13PIL1:cam1:TriggerMode_RBV devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.916 13PIL1:cam1:Armed devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.916 13PIL1:cam1:NumBadPixels devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.916 13PIL1:cam1:FlatFieldValid devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.916 13PIL1:cam1:Temp0_RBV devAsynFloat64::interruptCallbackInput new value=0.000000
2014/08/15 22:21:55.916 13PIL1:cam1:Temp1_RBV devAsynFloat64::interruptCallbackInput new value=0.000000
2014/08/15 22:21:55.916 13PIL1:cam1:Temp2_RBV devAsynFloat64::interruptCallbackInput new value=0.000000
2014/08/15 22:21:55.916 13PIL1:cam1:Humid0_RBV devAsynFloat64::interruptCallbackInput new value=0.000000
2014/08/15 22:21:55.916 13PIL1:cam1:Humid1_RBV devAsynFloat64::interruptCallbackInput new value=0.000000
2014/08/15 22:21:55.916 13PIL1:cam1:Humid2_RBV devAsynFloat64::interruptCallbackInput new value=0.000000
2014/08/15 22:21:55.916 13PIL1:cam1:OscillAxis devAsynOctet: writeIt
2014/08/15 22:21:55.916 13PIL1:cam1:BadPixelFile devAsynOctet: writeIt
2014/08/15 22:21:55.916 13PIL1:cam1:ArraySizeX_RBV devAsynInt32::getCallbackValue from ringBuffer value=487
2014/08/15 22:21:55.916 13PIL1:cam1:ArraySizeY_RBV devAsynInt32::getCallbackValue from ringBuffer value=407
2014/08/15 22:21:55.916 13PIL1:cam1:ArraySizeZ_RBV devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.916 13PIL1:cam1:ArraySize_RBV devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.916 13PIL1:cam1:ArrayCounter_RBV devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.916 13PIL1:cam1:WriteFile_RBV devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.916 13PIL1:cam1:ReadFile_RBV devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.916 13PIL1:cam1:Capture_RBV devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.916 13PIL1:cam1:WriteStatus devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.916 13PIL1:cam1:NumCaptured_RBV devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.916 13PIL1:cam1:MaxSizeX_RBV devAsynInt32::getCallbackValue from ringBuffer value=487
2014/08/15 22:21:55.916 13PIL1:cam1:MaxSizeY_RBV devAsynInt32::getCallbackValue from ringBuffer value=407
2014/08/15 22:21:55.916 13PIL1:cam1:DetectorState_RBV devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.916 13PIL1:cam1:NumImagesCounter_RBV devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.916 13PIL1:cam1:NumExposuresCounter_RBV devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.916 13PIL1:cam1:TimeRemaining_RBV devAsynFloat64::getCallbackValue from ringBuffer value=0.000000
2014/08/15 22:21:55.916 13PIL1:cam1:ShutterStatus_RBV devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.916 13PIL1:cam1:Acquire_RBV devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.916 13PIL1:cam1:ImageMode_RBV devAsynInt32::getCallbackValue from ringBuffer value=2
2014/08/15 22:21:55.916 13PIL1:cam1:TriggerMode_RBV devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.916 13PIL1:cam1:Armed devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.916 13PIL1:cam1:NumBadPixels devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.916 13PIL1:cam1:FlatFieldValid devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.916 13PIL1:cam1:Temp0_RBV devAsynFloat64::getCallbackValue from ringBuffer value=0.000000
2014/08/15 22:21:55.916 13PIL1:cam1:Temp1_RBV devAsynFloat64::getCallbackValue from ringBuffer value=0.000000
2014/08/15 22:21:55.916 13PIL1:cam1:Temp2_RBV devAsynFloat64::getCallbackValue from ringBuffer value=0.000000
2014/08/15 22:21:55.916 13PIL1:cam1:Humid0_RBV devAsynFloat64::getCallbackValue from ringBuffer value=0.000000
2014/08/15 22:21:55.916 13PIL1:cam1:Humid1_RBV devAsynFloat64::getCallbackValue from ringBuffer value=0.000000
2014/08/15 22:21:55.916 13PIL1:cam1:Humid2_RBV devAsynFloat64::getCallbackValue from ringBuffer value=0.000000
2014/08/15 22:21:55.918 13PIL1:cam1:CbfTemplateFile devAsynOctet: writeIt
2014/08/15 22:21:55.918 13PIL1:cam1:FileName devAsynOctet: writeIt
2014/08/15 22:21:55.919 13PIL1:cam1:FilePath devAsynOctet: writeIt
2014/08/15 22:21:55.919 13PIL1:cam1:FileTemplate devAsynOctet: writeIt
2014/08/15 22:21:55.919 13PIL1:cam1:FlatFieldFile devAsynOctet: writeIt
2014/08/15 22:21:55.919 13PIL1:cam1:HeaderString devAsynOctet: writeIt
2014/08/15 22:21:55.919 13PIL1:cam1:NDAttributesFile devAsynOctet: writeIt
2014/08/15 22:21:55.919 13PIL1:cam1:PoolMaxMem devAsynFloat64 process value=190.734863
2014/08/15 22:21:55.928 13PIL1:cam1:AcquirePeriod_RBV devAsynFloat64::interruptCallbackInput new value=0.000000
2014/08/15 22:21:55.928 13PIL1:cam1:NumImages_RBV devAsynInt32::interruptCallbackInput new value=1
2014/08/15 22:21:55.928 13PIL1:cam1:NumExposures_RBV devAsynInt32::interruptCallbackInput new value=1
2014/08/15 22:21:55.928 13PIL1:cam1:AcquireTime_RBV devAsynFloat64::interruptCallbackInput new value=1.000000
2014/08/15 22:21:55.928 13PIL1:cam1:DelayTime_RBV devAsynFloat64::interruptCallbackInput new value=0.000000
2014/08/15 22:21:55.928 13PIL1:cam1:GapFill_RBV devAsynInt32::interruptCallbackInput new value=-2
2014/08/15 22:21:55.928 13PIL1:cam1:AcquirePeriod devAsynFloat64 process val 0.000000
2014/08/15 22:21:55.928 13PIL1:cam1:AcquirePeriod_RBV devAsynFloat64::getCallbackValue from ringBuffer value=0.000000
2014/08/15 22:21:55.928 13PIL1:cam1:NumImages_RBV devAsynInt32::getCallbackValue from ringBuffer value=1
2014/08/15 22:21:55.928 13PIL1:cam1:NumExposures_RBV devAsynInt32::getCallbackValue from ringBuffer value=1
2014/08/15 22:21:55.928 13PIL1:cam1:AcquireTime_RBV devAsynFloat64::getCallbackValue from ringBuffer value=1.000000
2014/08/15 22:21:55.928 13PIL1:cam1:DelayTime_RBV devAsynFloat64::getCallbackValue from ringBuffer value=0.000000
2014/08/15 22:21:55.928 13PIL1:cam1:GapFill_RBV devAsynInt32::getCallbackValue from ringBuffer value=-2
2014/08/15 22:21:55.933 pilatusDetector:readCamserver unexpected response from camserver, no OK, response=15 ERR Illegal exposure period
2014/08/15 22:21:55.937 13PIL1:cam1:AcquireTime devAsynFloat64 process val 1.000000
2014/08/15 22:21:55.938 13PIL1:cam1:Alpha devAsynFloat64 process val 0.000000
2014/08/15 22:21:55.939 13PIL1:cam1:AngleIncr devAsynFloat64 process val 0.100000
2014/08/15 22:21:55.940 13PIL1:cam1:BeamX devAsynFloat64 process val 0.000000
2014/08/15 22:21:55.941 13PIL1:cam1:BeamY devAsynFloat64 process val 0.000000
2014/08/15 22:21:55.942 13PIL1:cam1:Chi devAsynFloat64 process val 0.000000
2014/08/15 22:21:55.947 pilatusDetector:readCamserver unexpected response from camserver, no OK, response=15 ERR Illegal exposure period
2014/08/15 22:21:55.949 13PIL1:cam1:DelayTime devAsynFloat64 process val 0.000000
2014/08/15 22:21:55.951 13PIL1:cam1:Det2theta devAsynFloat64 process val 0.000000
2014/08/15 22:21:55.952 13PIL1:cam1:DetDist devAsynFloat64 process val 1000.000000
2014/08/15 22:21:55.953 13PIL1:cam1:DetVOffset devAsynFloat64 process val 0.000000
2014/08/15 22:21:55.954 13PIL1:cam1:EnergyHigh devAsynFloat64 process val 0.000000
2014/08/15 22:21:55.955 13PIL1:cam1:EnergyLow devAsynFloat64 process val 0.000000
2014/08/15 22:21:55.956 13PIL1:cam1:FilterTransm devAsynFloat64 process val 1.000000
2014/08/15 22:21:55.958 13PIL1:cam1:Flux devAsynFloat64 process val 0.000000
2014/08/15 22:21:55.958 13PIL1:cam1:Gain_RBV devAsynFloat64::interruptCallbackInput new value=1.000000
2014/08/15 22:21:55.958 13PIL1:cam1:Gain devAsynFloat64 process val 1.000000
2014/08/15 22:21:55.958 13PIL1:cam1:ImageFileTmot devAsynFloat64 process val 20.000000
2014/08/15 22:21:55.958 13PIL1:cam1:Gain_RBV devAsynFloat64::getCallbackValue from ringBuffer value=1.000000
2014/08/15 22:21:55.959 13PIL1:cam1:Kappa devAsynFloat64 process val 0.000000
2014/08/15 22:21:55.960 13PIL1:cam1:Phi devAsynFloat64 process val 0.000000
2014/08/15 22:21:55.961 13PIL1:cam1:Polarization devAsynFloat64 process val 0.990000
2014/08/15 22:21:55.961 13PIL1:cam1:ShutterCloseDelay_RBV devAsynFloat64::interruptCallbackInput new value=0.000000
2014/08/15 22:21:55.961 13PIL1:cam1:ShutterCloseDelay devAsynFloat64 process val 0.000000
2014/08/15 22:21:55.961 13PIL1:cam1:ShutterOpenDelay_RBV devAsynFloat64::interruptCallbackInput new value=0.000000
2014/08/15 22:21:55.961 13PIL1:cam1:ShutterOpenDelay devAsynFloat64 process val 0.000000
2014/08/15 22:21:55.961 13PIL1:cam1:ShutterCloseDelay_RBV devAsynFloat64::getCallbackValue from ringBuffer value=0.000000
2014/08/15 22:21:55.961 13PIL1:cam1:ShutterOpenDelay_RBV devAsynFloat64::getCallbackValue from ringBuffer value=0.000000
2014/08/15 22:21:55.962 13PIL1:cam1:StartAngle devAsynFloat64 process val 0.000000
2014/08/15 22:21:55.962 13PIL1:cam1:Temperature_RBV devAsynFloat64::interruptCallbackInput new value=25.000000
2014/08/15 22:21:55.962 13PIL1:cam1:Temperature devAsynFloat64 process val 25.000000
2014/08/15 22:21:55.962 13PIL1:cam1:ThresholdEnergy_RBV devAsynFloat64::interruptCallbackInput new value=25.000000
2014/08/15 22:21:55.962 13PIL1:cam1:ThresholdEnergy devAsynFloat64 process val 10.000000
2014/08/15 22:21:55.962 13PIL1:cam1:Temperature_RBV devAsynFloat64::getCallbackValue from ringBuffer value=25.000000
2014/08/15 22:21:55.962 13PIL1:cam1:ThresholdEnergy_RBV devAsynFloat64::getCallbackValue from ringBuffer value=25.000000
2014/08/15 22:21:55.963 13PIL1:cam1:Wavelength devAsynFloat64 process val 1.540000
2014/08/15 22:21:55.963 13PIL1:cam1:ArrayCallbacks_RBV devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.963 13PIL1:cam1:ArrayCallbacks devAsynInt32 process value 0
2014/08/15 22:21:55.963 13PIL1:cam1:AutoIncrement_RBV devAsynInt32::interruptCallbackInput new value=1
2014/08/15 22:21:55.963 13PIL1:cam1:AutoIncrement devAsynInt32 process value 1
2014/08/15 22:21:55.963 13PIL1:cam1:AutoSave_RBV devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.964 13PIL1:cam1:AutoSave devAsynInt32 process value 0
2014/08/15 22:21:55.964 13PIL1:cam1:DeleteDriverFile_RBV devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.964 13PIL1:cam1:DeleteDriverFile devAsynInt32 process value 0
2014/08/15 22:21:55.964 13PIL1:cam1:ShutterControl_RBV devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.964 13PIL1:cam1:ShutterControl devAsynInt32 process value 0
2014/08/15 22:21:55.964 13PIL1:cam1:ThresholdAutoApply_RBV devAsynInt32::interruptCallbackInput new value=1
2014/08/15 22:21:55.964 13PIL1:cam1:ThresholdAutoApply devAsynInt32 process value 1
2014/08/15 22:21:55.964 13PIL1:cam1:PoolMaxBuffers devAsynInt32 process value=50
2014/08/15 22:21:55.964 13PIL1:cam1:FileNumber_RBV devAsynInt32::interruptCallbackInput new value=1
2014/08/15 22:21:55.964 13PIL1:cam1:FileNumber_write devAsynInt32 process value 1
2014/08/15 22:21:55.964 13PIL1:cam1:MinFlatField_RBV devAsynInt32::interruptCallbackInput new value=100
2014/08/15 22:21:55.964 13PIL1:cam1:MinFlatField devAsynInt32 process value 100
2014/08/15 22:21:55.964 13PIL1:cam1:NumCapture_RBV devAsynInt32::interruptCallbackInput new value=1
2014/08/15 22:21:55.964 13PIL1:cam1:NumCapture devAsynInt32 process value 1
2014/08/15 22:21:55.964 13PIL1:cam1:ArrayCallbacks_RBV devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.964 13PIL1:cam1:AutoIncrement_RBV devAsynInt32::getCallbackValue from ringBuffer value=1
2014/08/15 22:21:55.964 13PIL1:cam1:AutoSave_RBV devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.964 13PIL1:cam1:DeleteDriverFile_RBV devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.964 13PIL1:cam1:ShutterControl_RBV devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.964 13PIL1:cam1:ThresholdAutoApply_RBV devAsynInt32::getCallbackValue from ringBuffer value=1
2014/08/15 22:21:55.964 13PIL1:cam1:FileNumber_RBV devAsynInt32::getCallbackValue from ringBuffer value=1
2014/08/15 22:21:55.964 13PIL1:cam1:MinFlatField_RBV devAsynInt32::getCallbackValue from ringBuffer value=100
2014/08/15 22:21:55.964 13PIL1:cam1:NumCapture_RBV devAsynInt32::getCallbackValue from ringBuffer value=1
2014/08/15 22:21:55.970 pilatusDetector:readCamserver unexpected response from camserver, no OK, response=15 ERR Illegal exposure period
2014/08/15 22:21:55.973 13PIL1:cam1:NumExposures devAsynInt32 process value 1
2014/08/15 22:21:55.978 pilatusDetector:readCamserver unexpected response from camserver, no OK, response=15 ERR Illegal exposure period
2014/08/15 22:21:55.980 13PIL1:cam1:NumImages devAsynInt32 process value 1
2014/08/15 22:21:55.981 13PIL1:cam1:NumOscill devAsynInt32 process value 1
2014/08/15 22:21:55.981 13PIL1:cam1:FileFormat_RBV devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.981 13PIL1:cam1:FileFormat devAsynInt32 process value 0
2014/08/15 22:21:55.982 13PIL1:cam1:FileWriteMode_RBV devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.982 13PIL1:cam1:FileWriteMode devAsynInt32 process value 0
2014/08/15 22:21:55.982 13PIL1:cam1:FrameType_RBV devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.982 13PIL1:cam1:FrameType devAsynInt32 process value 0
2014/08/15 22:21:55.982 13PIL1:cam1:FileFormat_RBV devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.982 13PIL1:cam1:FileWriteMode_RBV devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.982 13PIL1:cam1:FrameType_RBV devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.986 pilatusDetector:readCamserver unexpected response from camserver, no OK, response=15 ERR Illegal exposure period
2014/08/15 22:21:55.989 13PIL1:cam1:GapFill_RBV devAsynInt32::interruptCallbackInput new value=-2
2014/08/15 22:21:55.989 13PIL1:cam1:GapFill devAsynInt32 process value 2
2014/08/15 22:21:55.989 13PIL1:cam1:ImageMode devAsynInt32 process value 2
2014/08/15 22:21:55.989 13PIL1:cam1:ShutterMode_RBV devAsynInt32::interruptCallbackInput new value=0
2014/08/15 22:21:55.989 13PIL1:cam1:ShutterMode devAsynInt32 process value 0
2014/08/15 22:21:55.989 13PIL1:cam1:GapFill_RBV devAsynInt32::getCallbackValue from ringBuffer value=-2
2014/08/15 22:21:55.989 13PIL1:cam1:ShutterMode_RBV devAsynInt32::getCallbackValue from ringBuffer value=0
2014/08/15 22:21:55.994 pilatusDetector:readCamserver unexpected response from camserver, no OK, response=15 ERR Illegal exposure period
2014/08/15 22:21:55.996 13PIL1:cam1:TriggerMode devAsynInt32 process value 0
auto_settings.sav: 1142 of 1142 PV's connected
epics> 2014/08/15 22:21:56.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:21:56.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:21:56.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:21:57.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:21:57.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:21:57.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:21:58.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:21:58.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:21:58.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:21:59.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:21:59.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:21:59.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:00.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:00.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:00.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:01.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:01.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:01.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:02.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:02.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:02.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:03.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:03.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:03.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:04.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:04.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:04.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:05.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:05.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:05.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:06.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:06.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:06.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:07.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:07.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:07.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:08.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:08.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:08.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:09.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:09.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:09.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:10.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:10.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:10.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:11.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:11.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:11.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:12.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:12.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:12.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:13.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:13.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:13.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:14.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:14.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:14.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:15.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:15.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:15.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:16.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:16.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:16.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:17.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:17.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:17.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:18.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:18.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:18.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:19.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:19.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:19.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:20.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:20.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:20.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:21.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:21.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:21.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:22.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:22.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:22.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:23.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:23.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:23.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:24.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:24.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:24.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:25.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:25.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:25.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:26.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:26.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:26.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:27.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:27.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:27.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:28.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:28.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:28.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:29.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:29.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:29.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:30.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:30.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:30.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:31.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:31.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:31.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:32.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:32.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:32.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:33.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:33.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:33.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:34.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:34.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:34.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:35.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:35.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:35.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:36.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:36.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:36.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:37.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:37.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:37.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:38.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:38.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:38.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:39.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:39.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:39.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:40.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:40.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:40.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:41.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:41.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:41.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:42.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:42.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:42.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:43.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:43.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:43.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:44.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:44.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:44.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:45.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:45.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:45.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
2014/08/15 22:22:46.411 13PIL1:cam1:PoolUsedMem devAsynFloat64 process value=0.756107
2014/08/15 22:22:46.411 13PIL1:cam1:PoolAllocBuffers devAsynInt32 process value=1
2014/08/15 22:22:46.411 13PIL1:cam1:PoolFreeBuffers devAsynInt32 process value=0
[root@pilatus iocPilatus]#
Areadetector IOC for Pilatus Specht, Eliot D.
Prev: Areadetector IOC for Pilatus Specht, Eliot D.
Next: Re: Windows installer error - bcastEnvChange.exe Ryan Pierce
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  <2014>  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026
Next: Future Minimum Requirements for IOCs Andrew Johnson
ANJ, 17 Dec 2015 · Home · News · About · Talk · Base · Modules · Extensions ·
