1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  <20172018  2019  2020  2021  2022  2023  2024  2025  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  <20172018  2019  2020  2021  2022  2023  2024  2025  2026
Subject: RE: Re: IOC Connecting Problem
To: Rod Nussbaumer <[email protected]>
Cc: [email protected]
Date: Thu, 23 Feb 2017 11:01:31 +0900
Rod:
Thanks a lot. Your comment is very helpful.
I have thought that It is different between Telnet and TCP communication, because there are two kind of communication in case of delay generator.
I attached delay generator's manual(http://www.thinksrs.com/downloads/PDFs/Manuals/DG645m.pdf).
As you can see, there are difference in page 44, operation 30, Table 14: NET Menu Options for TCP/IP Configuration.
Could you let me know what is different between Bare socket interface(port 5025) and Telnet interface(port 5024)?
In my case, I have used only port 5025 for connecting IOC.
I hope you have a nice day.
Best Regards,
Pilkyu
---------- Original Message ----------
 Date: 2017-02-23 01:56
 Subject: Re: IOC Connecting Problem
 To: "Pilkyu Jung" <[email protected]>
Telnet is simply doing the same kinds of TCP communication that the IOC
asynPort does. The connection seen by the oscilloscope is the same for a
telnet session as it is for an EPICS asynPort connection. In telnet, you
send commands interactively. The IOC sends commands as part of the
program algorithm. The data transferred is the same, and the scope
doesn't see any difference.
The port number is part of the IP address used to make the TCP
connection, so that is why it must be correct in order for the
connection to succeed.
I don't quite understand your last question. If you don't want to use a
telnet-like connection between the scope and the IOC, then you will need
to use some other kind of interface. Perhaps your scope is equipped with
GPIB or serial ports. These work equally well from the standpoint of
compatibility with EPICS asyn + streamDevice support.
I hope this helps.
--- rod.
On 02/21/2017 11:24 PM, Pilkyu Jung wrote:
> Rod:
> First of all, I appreciate your advice. It was very helpful.
> By the way, I have tried to use port 5024 and 5025.
> But, now IOC worked with port 5024, not port 5025.
> Can I ask you a question one more?
> When I use port 5024, the device worked with telnet automatically.
> Like a Image file(16:18:10) that I attached a image file.
> However, I guess it should more fast without connecting telnet.
> At least, when I connect other device, didn't connect telnet.
> and when I use port 5025, it doesn't work.
> So, Could you let me know how can I use this oscilloscope on IOC without
> telnet connecting like other devices?
> best regards,
> Pilkyu
> On 02/16/2017 11:15 PM, Rod Nussbaumer wrote:
>> Pilkyu:
>> In your telnet connection, you use port 5024, but in the asyn port
>> connection, you are attempting to use port 5025. This is quite likely
>> one cause of your problem.
>> Rod Nussbaumer
>> TRIUMF
>> Vancouver, Canada
>> On 02/15/2017 10:29 PM, Pilkyu Jung wrote:
>>> Hi all,
>>> First of all, I am trying to connect from a oscilloscope which name is
>>> Agilent Technologies DSO-X 3052A (500 MHz 4 GSa/s) to IOC.
>>> As you can see in the
>>> manual(http://literature.cdn.keysight.com/litweb/pdf/75019-97073.pdf ,
>>> http://www.keysight.com/upload/cmc_upload/All/3000_series_prog_guide.pdf),
>>> this oscilloscope support a few ways (e.g. the IO Library, Telnet
>>> Sockets, and using Browser Web Control) for sending SCPI.
>>> However, When I connect to IOC, I got a problem. Until now, I just
>>> connected only adjust IP address in st.cmd.
>>> Like this,
>>> drvAsynIPPortConfigure("P1",169.254.90.3:5025,0,0,0)
>>> When I connect by telnet, it is done successfully (please check the
>>> Screenshot...image file).
>>> I don't know why this device can't be connected to IOC in this case.
>>> Definitely I am done with this way before, but now I can't.
>>> Here is my question,
>>> 1. Normally, when you connect from device to IOC, what kind of method
>>> use?
>>> 2. In my case, How can I connect to IOC?
>>> Best Regards,
>>> Pilkyu Jung
RE: Re: IOC Connecting Problem Mark Rivers
Prev: New releases of areaDetector drivers Mark Rivers
Next: RE: Re: IOC Connecting Problem Mark Rivers
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  <20172018  2019  2020  2021  2022  2023  2024  2025  2026
Prev: Re: IOC Connecting Problem Pilkyu Jung
