1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  <2025>  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  <2025>  2026
Subject: Re: Protocol format issues in StreamDevice during COMBIVAC CM 52 device commissioning.
From: Marco Filho via Tech-talk <[email protected]>
To: "[email protected]" <[email protected]>, LONG FENG <[email protected]>
Date: Wed, 10 Dec 2025 13:45:09 +0000
out "RPV%(\$1:Channel)c?";
The manual says it should be in the format:
Reading command
S: Address Command <CR>
 E: [Parameter] <CR>
I think you are missing a comma?
I created an IOC with your db/test file combinations to test and my socket receives:
B'RPV3?\r'
If I correctly understand the manual it should be:
B'RPV,3?\r'
Is that right? If so,
out "RPV,%(\$1:Channel)c?";
Should fix the issue. Not sure though.
I really recommend connecting your IOC with some simple python socket server to see what it's actually sending:
import socket
HOST = '127.0.0.1'
PORT = 5001
with socket.socket(socket.AF_INET, socket.SOCK_STREAM) as s:
    s.bind((HOST, PORT))
    s.listen()
    conn, addr = s.accept()
    with conn:
        while True:
            data = "">
            print(data)
            if not data:
                break
            conn.sendall(data)
drvAsynIPPortConfigure("$(PORT)", "127.0.0.1:5001", 0, 0, 1)
That might make your life easier 🙂
Cheers,
Marco
From: Tech-talk <[email protected]> on behalf of LONG FENG via Tech-talk <[email protected]>
Sent: Wednesday, December 10, 2025 12:18
To: [email protected] <[email protected]>
Subject: Protocol format issues in StreamDevice during COMBIVAC CM 52 device commissioning.
Hi all,
In the communication protocol of the COMBIVAC CM 52, there is a specification as follows:
S: RPV[a]<CR>
E: b[,][TAB]x.xxxxE±xx
Here, a represents the channel number .1 = Channel 1 ,2 = Channel 2 ,3 = Channel 3 (The values 1, 2, and 3 are character digits).b represents the status.
I want to use StreamDevice to read data. The contents of my .db file are as follows:
record(mbbo, "$(P)$(R):Channel")
    field(DESC, "Select Channel")
    field(ZRVL, "49")
    field(ZRST, "Channel 1")
    field(ONVL, "50")
    field(ONST, "Channel 2")
    field(TWVL, "51")
    field(TWST, "Channel 3")
    field(VAL, "51")
    field(PINI, "YES")
    field(FLNK, "$(P)$(R):RPV")
record(ai, "$(P)$(R):RPV")
    field(DESC, "Reading pressure")
    field(DTYP, "stream")
    field(INP, "@test.proto readPressure($(P)$(R)) $(PORT)")
    field(SCAN, "1 second")
    field(FLNK, "$(P)$(R):Status")
record(mbbi, "$(P)$(R):Status")
    field(DESC, "status of channel")
    field(ZRVL, "0")
    field(ZRST, "OK")
    field(ONVL, "1")
    field(ONST, "BelowRange")
    field(TWVL, "2")
    field(TWST, "AboveRange")
    field(THVL, "3")
    field(THST, "ErrLo")
    field(FRVL, "4")
    field(FRST, "ErrHi")
    field(FVVL, "5")
    field(FVST, "SensorOff")
    field(SXVL, "6")
    field(SXST, "HVon")
    field(SVVL, "7")
    field(SVST, "SensorErr")
    field(EIVL, "9")
    field(EIST, "NoSensor")
    field(TEVL, "10")
    field(TEST, "NoTrig")
    field(TVVL, "12")
    field(TVST, "ErrPi")
    field(FFVL, "15")
    field(FFST, "OkDegas")
The contents of my .proto file are as follows:
OutTerminator = CR;
InTerminator = CR;
readPressure{
    in  "%(\$1:Status)d,\t%g";
readVersion{
    out "RVN";
    in  "%f";
 However, the following error occurs:
 2025/12/10 05:47:36.646640 Combivac1 Emission:fl:RPV: Input "?<09>P,<09>1" does not match format "%d"
2025/12/10 05:47:37.646725 Combivac1 Emission:fl:RPV: Input "?<09>P,<09>1" does not match format "%d"
2025/12/10 05:47:38.646532 Combivac1 Emission:fl:RPV: Input "?<09>P,<09>1" does not match format "%d"
2025/12/10 05:47:39.644943 Combivac1 Emission:fl:RPV: Input "?<09>P,<09>1" does not match format "%d"
What is the cause of this? Am I doing something wrong?Please help.
Feng
Re: Protocol format issues in StreamDevice during COMBIVAC CM 52 device commissioning. Marco Filho via Tech-talk
Protocol format issues in StreamDevice during COMBIVAC CM 52 device commissioning. LONG FENG via Tech-talk
Prev: Re: Protocol format issues in StreamDevice during COMBIVAC CM 52 device commissioning. Marco Filho via Tech-talk
Next: Re: Protocol format issues in StreamDevice during COMBIVAC CM 52 device commissioning. Marco Filho via Tech-talk
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  <2025>  2026
