1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  <2026>  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  <2026>
Subject: How to calculate checksum (three ASCII values) when using streamdevice
From: LiangChih Chiang via Tech-talk <tech-talk at aps.anl.gov>
To: tech-talk at aps.anl.gov
Date: Mon, 17 Aug 2026 13:53:21 +0800
Hi, EPICS mates,
I'm using streamdevice to communicate with the following device:
Vacuum gauge controller: Pfeiffer OmniControl 200
gauge model: Pfieffer PKR361
I don't know how to handle the checksums of commands/responses.
The command/response format is of all ASCII values.
The checksum is of three ASCII values.
The calculation method is: sum of all ASCII values, then modulo 256
For example: to get the vacuum pressure value:
Command: 1220074002=?110
122: address of the component gauge.
00: means this is a query command
740: means parameter pressure
02: means the length of the following data is 2
=?: means to query something
110: checksum
Response: 1221074006539212045
122: same as in command
10: means this is a response of query command
740: same as in command
06: means the length of the following data is 6
539212: -> 5.392x10^(20-12) hPa = 5.392x10^-8 hPa
045: checksum
Re: How to calculate checksum (three ASCII values) when using streamdevice Zimoch, Dirk via Tech-talk
Prev: Re: [EXTERNAL] Modbus and SScan acquisition synchronisation and timestamp query Ralph Lange via Tech-talk
Next: Re: How to calculate checksum (three ASCII values) when using streamdevice Zimoch, Dirk via Tech-talk
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  <2026>
Prev: Re: Oxford Cryostream 1000 IOC questions James P. Withrow via Tech-talk
ANJ, 17 Aug 2026 · Home · News · About · Talk · Base · Modules · Extensions ·