1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  <2011>  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  <2011>  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026
Subject: RE: problem with ip330-2-6 (A/D converter board) scanmode functionality to switch modes at runtime
To: "[email protected]" <[email protected]>, "[email protected]" <[email protected]>
Date: Tue, 5 Jul 2011 13:58:06 -0700
Hi Peter,
     Thanks a lot for the information. It sounds as if there is a problem in the ip330’s firmware.
     I am using 4 channels of the ip330. My normal gain setting for the ip330 is 0 (equivalent to gain = 1). For now, I have implemented the logic to switch to burst continuous mode by setting the scanmode to 3 (burst continuous), then setting the gain settings for the 4 channels to 1, then setting the gain settings back to 0. This issue only seems to come up intermittently, but we will watch out for it and see if your solution resolves this issue over the long term.
Cheers,
Jay Steele
Xradia Corporation
From: [email protected] [mailto:[email protected]] On Behalf Of [email protected]
Sent: Monday, July 04, 2011 1:48 AM
To: [email protected]
When I was using the ip330 in Continuous mode I found it necessary to set the gain a couple of times in my startup file in order to get it to run:
record(longout,"BL18I-EA-IONC-01:IGain")
        field(PINI,"YES")
        field(VAL,"0")
        field(DTYP,"asynInt32")
        field(OUT,"@asyn(ip330b 0)GAIN")
And in the startup file:
dbpf ("BL18I-EA-IONC-01:IGain","1")
dbpf ("BL18I-EA-IONC-01:IGain","0")
Note sure if this is relevant to you but it may be worth trying when you switch back to continuous mode.
Pete Leicester
Diamond Light Source
From: [email protected] [mailto:[email protected]] On Behalf Of Jay Steele
Sent: 29 June 2011 19:20
Subject: problem with ip330-2-6 (A/D converter board) scanmode functionality to switch modes at runtime
Hi EPICS folks,
     I’m seeing a problem with the EPICS module for the VME ADC board IP330, software version 2-6. I’m trying to switch between Burst Continuous and Burst Single modes for step scanning and hardware-synchronized fly scanning, respectively, at run time. When I process the provided ScanMode record, the board gets stuck in Burst Single mode no matter if I’m setting the Burst Continuous mode (value 3) or Burst Single mode (value 4).
    Has anybody else been successful at switching modes at run time for the IP330 board using this software?
Jay Steele
Xradia Corporation
The information in this email, including any attachments, is confidential and intended only for the recipient(s) listed. Any use of this email for any other purpose is prohibited. If you have received this email in error, please notify me immediately by reply email, delete this email, and do not disclose its contents to anyone.
problem with ip330-2-6 (A/D converter board) scanmode functionality to switch modes at runtime Jay Steele
RE: problem with ip330-2-6 (A/D converter board) scanmode functionality to switch modes at runtime peter.leicester
Prev: Build base 13-14-12-1 on Win32 Robert Bradford
Next: EPICS Collaboration Meeting Fall 2011, Reminder Dirk Zimoch
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  <2011>  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026
Prev: RE: problem with ip330-2-6 (A/D converter board) scanmode functionality to switch modes at runtime peter.leicester
Next: Turbo Pump controller Zheng, Susie