1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  <2013>  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  <2013>  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026
Subject: Re: Response Time
To: [email protected]
Date: Mon, 19 Aug 2013 17:27:44 -1000
If you (Yasemin) can log directly into the port the device is connected
to (you may have to turn off your IOC for that), if everything is
connected and set up correctly, you should be able to type a command
the device understands into it, and ^m and/or ^j. If you get the reply
you expect you're set up correct from the ethernet converter (or
terminal server) to the device. If you get no reply at all a breakout
box is your best way forward. If you get gibberish you can be pretty
sure that your cable is correct but one or more of the serial line
parameters isn't (as Emmanuel said).
Keep in mind that some devices are very configurable and some things
sometimes aren't quite as the manual says. E.g. we have 2 of the
same devices. You're supposed to be able to configure the parity,
yet one only speaks even parity and the other one only speaks odd.
Getting devices on serial lines to work sometimes is a bit like
banging your head against a wall.
Maren
On 08/19/13 14:47, Emmanuel Mayssat wrote:
Check the serial to ethernet converter. Is it RS232, RS422, or RS485?
(check physical switches on the converter)
Check the ethernet converter configuration to get the correct serial
parameters, baud rate, nb of stop bits, etc
Check the string terminator as well. The ethernet converter may append a
terminator which is not in the initial protocol.
If everything is configured correctly, changing interface from ethernet
to serial of a stream device should be seemless. (Of course, you changed
the physical connection, but also the asyn command from
drvAsynIPPortConfigure to drvAsynSerialPortConfigure, right?)
Emmanuel
Date: Mon, 19 Aug 2013 16:30:07 +0300
Subject: Response Time
Hello,
I've been trying to connect to my machine by RS232.
I use streamDevice and Asyn Module.
When we send command "*IDN?", we expect to get "THERMO OK" from the
machine.
I try the same connection on TCP/IP. I used serial to ethernet converter
and communicated with the machine. No problem occured. But when I
eliminated the converter and directly connected to the PC on RS232 port,
I got the following error when I send the command.
I guess the problem is about response time. How can I configure the
response time? Or if it is not response time problem, how can I fix it?
ERROR:
"2013/08/19 16:23:15.340949 thermoConn IDN: No reply from device within
1000 ms"
Response Time Yasemin Barutçu
RE: Response Time Emmanuel Mayssat
Prev: RE: Response Time Emmanuel Mayssat
Next: Channel Acess Ganesh Jangir
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  <2013>  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026
Next: asynPortDriver question Jörn Wüstenfeld