Subject: Re: problem connecting serial device with moxa 5150A
To: Sunny Badkul <[email protected]>
Date: Wed, 18 Jan 2017 12:51:22 +0000
Can you bring up the Moxa's web page in a browser?  If it pings but doesn't let you connect to port 4001, my guess is that it's not configured as "TCP Server Mode".
On 1/18/2017 7:37 AM, Sunny Badkul wrote:
Thank you Mark,
              I tried commands as you suggested and following is the result
~> ping 192.168.127.254
PING 192.168.127.254 (192.168.127.254) 56(84) bytes of data.
64 bytes from 192.168.
________________________________127.254: icmp_seq=54 ttl=255 time=0.742 ms
^C
--- 192.168.127.254 ping statistics ---
54 packets transmitted, 54 received, 0% packet loss, time 53000ms
rtt min/avg/max/mdev = 0.691/0.783/0.857/0.054 ms
Trying 192.168.127.254...
Also I attached screenshot of the configuration of Ethernet port which is connected to moxa. I am weak in Networking so thare are chances for mistake in configuration.
thanks
On Wed, Jan 18, 2017 at 5:28 PM, Mark Rivers <[email protected]> wrote:
Hi Sunny,
What happens if you try to ping the Moxa device from the Linux command line?
ping 192.168.127.254<MailScanner warning: numerical links are often malicious: http://192.168.127.254:4001>
What happens if you try to telnet to the port and type some commands for the Keithley?
telnet MailScanner warning: numerical links are often malicious: 192.168.127.254:4001<MailScanner warning: numerical links are often malicious: http://192.168.127.254:4001>
Are you sure that one of your Ethernet ports is correctly configured for the subnet with the 192.168.127.254<MailScanner warning: numerical links are often malicious: http://192.168.127.254:4001> subnet?
From: [email protected] [[email protected]] on behalf of Sunny Badkul [[email protected]]
Sent: Wednesday, January 18, 2017 4:56 AM
To: [email protected]
Subject: problem connecting serial device with moxa 5150A
hello all,
I am designing an IOC for my serial device which is an  keithley picoammeter. I am using Linux opensuse and EPICS base 3.14.12.5 and stream device. when I  am connecting my device with usb/serial cable it works file. st.cmd for above setup is
#!../../bin/linux-x86_64/K6485
## You may have to change K6485 to something else
## everywhere it appears in this file
epicsEnvSet("P","$(P=K6485:)")
< envPaths
epicsEnvSet("STREAM_PROTOCOL_PATH","${TOP}/db")
cd "${TOP}"
dbLoadDatabase "dbd/K6485.dbd"
K6485_registerRecordDeviceDriver pdbbase
dbLoadRecords "/home/pxbl21-1/programs/epics/modules/synApps_5_6/support/asyn-4-18/db/asynRecord.db" "P=$(P),R=asyn,PORT=L0,ADDR=-1,OMAX=0,IMAX=0"
# Set up ASYN ports
# drvAsynIPPortConfigure port ipInfo priority noAutoconnect noProcessEos
#drvAsynSerialPortConfigure("L0","/dev/ttyUSB0",0,0,0)
asynSetTraceMask("L0",-1,0x09)
asynSetOption("L0", -1, "baud", "9600")
asynSetOption("L0", -1, "bits", "8")
asynSetOption("L0", -1, "parity", "none")
asynSetOption("L0", -1, "stop", "1")
asynSetOption("L0", -1, "clocal", "Y")
asynSetOption("L0", -1, "crtscts", "Y")
asynOctetSetInputEos("L0", -1, "\r\n")
asynOctetSetOutputEos("L0", -1, "\r\n")
asynSetTraceIOMask("L0",-1,0x2)
dbLoadRecords "db/devK6485.db" "P=$(P),R=1:,PORT=L0,A=-1,NELM=1000"
#dbLoadRecords "/home/pxbl21-1/programs/epics/modules/synApps_5_6/support/asyn-4-18/db/asynRecord.db" "P=$(P),R=asyn,PORT=L0,ADDR=-1,OMAX=0,IMAX=0"
## Load record instances
#dbLoadRecords("db/xxx.db","user=pxbl21-1Host")
cd "${TOP}/iocBoot/${IOC}"
iocInit
## Start any sequence programs
#seq sncxxx,"user=pxbl21-1Host"
Now I want to connect my device through moxa 5150A board which is a Ethernet to serial converter having default IP address 192.168.127.254 I changed st.cmd and the error log is like --
epicsEnvSet("P","K6485:")
< envPaths
epicsEnvSet("ARCH","linux-x86_64")
epicsEnvSet("IOC","iocK6485")
epicsEnvSet("TOP","/home/pxbl21-1/programs/epics/bl21/K6485")
epicsEnvSet("ASYN","/home/pxbl21-1/programs/epics/modules/synApps_5_6/support/asyn-4-18")
epicsEnvSet("STREAM","/home/pxbl21-1/programs/epics/modules/synApps_5_6/support/stream-2-5-1")
epicsEnvSet("EPICS_BASE","/home/pxbl21-1/programs/epics/base-3.14.12.5")
epicsEnvSet("STREAM_PROTOCOL_PATH","/home/pxbl21-1/programs/epics/bl21/K6485/db")
cd "/home/pxbl21-1/programs/epics/bl21/K6485"
# Set up ASYN ports
drvAsynIPPortConfigure("L0","MailScanner warning: numerical links are often malicious: 192.168.127.254:4001<MailScanner warning: numerical links are often malicious: http://192.168.127.254:4001>",0,0,0)
asynSetTraceMask("L0",0,0x09)
#asynSetOption("L0", 0, "baud", "9600")
#asynSetOption("L0", 0, "bits", "8")
#asynSetOption("L0", 0, "parity", "none")
#asynSetOption("L0", 0, "stop", "1")
#asynSetOption("L0", 0, "clocal", "Y")
#asynSetOption("L0", 0, "crtscts", "Y")
#asynOctetSetInputEos("L0", 0, "\r\n")
#asynOctetSetOutputEos("L0", 0, "\r\n")
asynSetTraceIOMask("L0",0,0x2)
asynSetTraceMask("L0",0,255)
dbLoadRecords "db/devK6485.db" "P=K6485:,R=1:,PORT=L0,A=0,NELM=1000"
dbLoadRecords "/home/pxbl21-1/programs/epics/modules/synApps_5_6/support/asyn-4-18/db/asynRecord.db" "P=K6485:,R=asyn,PORT=L0,ADDR=0,OMAX=0,IMAX=0"
cd "/home/pxbl21-1/programs/epics/bl21/K6485/iocBoot/iocK6485"
iocInit
Starting iocInit
## EPICS R3.14.12.5 $Date: Tue 2015-03-24 09:57:35 -0500$
## EPICS Base built Sep 15 2016
2017/01/18 16:13:59.525 L0 addr -1 queueRequest priority 3 not lockHolder
2017/01/18 16:16:06.248 asynManager connect queueCallback port:L0
2017/01/18 16:16:06.248 Open connection to MailScanner warning: numerical links are often malicious: 192.168.127.254:4001<MailScanner warning: numerical links are often malicious: http://192.168.127.254:4001>  reason:0  fd:-1
2017/01/18 16:16:26.243 L0 addr -1 queueRequest priority 3 not lockHolder
2017/01/18 16:18:13.480 L0 K6485:1:Measure: pasynCommon->connect() failed: Can't connect to MailScanner warning: numerical links are often malicious: 192.168.127.254:4001<MailScanner warning: numerical links are often malicious: http://192.168.127.254:4001>: Connection timed out
2017/01/18 16:18:13.480 asynManager connect queueCallback port:L0
2017/01/18 16:18:13.480 Open connection to MailScanner warning: numerical links are often malicious: 192.168.127.254:4001<MailScanner warning: numerical links are often malicious: http://192.168.127.254:4001>  reason:0  fd:-1
2017/01/18 16:18:13.480 _main_ K6485:1:Measure: @init handler failed
2017/01/18 16:18:13.480 _main_ K6485:1:Measure: Record initialization failed
2017/01/18 16:18:13.481 K6485:asyn: queueRequest failed
2017/01/18 16:18:13.983 L0 addr -1 queueRequest priority 3 not lockHolder
2017/01/18 16:18:13.983 L0 schedule queueRequest timeout
epics> 2017/01/18 16:18:18.482 L0 addr -1 queueRequest priority 3 not lockHolder
2017/01/18 16:18:18.482 L0 schedule queueRequest timeout
2017/01/18 16:18:18.978 L0 asynManager:queueTimeoutCallback
2017/01/18 16:18:18.978 L0 addr -1 queueRequest priority 3 not lockHolder
2017/01/18 16:18:18.978 L0 schedule queueRequest timeout
2017/01/18 16:18:23.477 L0 asynManager:queueTimeoutCallback
2017/01/18 16:18:23.482 L0 addr -1 queueRequest priority 3 not lockHolder
2017/01/18 16:18:23.482 L0 schedule queueRequest timeout
2017/01/18 16:18:23.974 L0 asynManager:queueTimeoutCallback
2017/01/18 16:18:28.477 L0 asynManager:queueTimeoutCallback
2017/01/18 16:18:28.482 L0 addr -1 queueRequest priority 3 not lockHolder
2017/01/18 16:18:28.482 L0 schedule queueRequest timeout
2017/01/18 16:18:33.477 L0 asynManager:queueTimeoutCallback
2017/01/18 16:18:33.482 L0 addr -1 queueRequest priority 3 not lockHolder
2017/01/18 16:18:33.482 L0 schedule queueRequest timeout
2017/01/18 16:18:38.477 L0 asynManager:queueTimeoutCallback
2017/01/18 16:18:38.482 L0 addr -1 queueRequest priority 3 not lockHolder
2017/01/18 16:18:38.482 L0 schedule queueRequest timeout
2017/01/18 16:18:43.477 L0 asynManager:queueTimeoutCallback
2017/01/18 16:18:43.482 L0 addr -1 queueRequest priority 3 not lockHolder
2017/01/18 16:18:43.482 L0 schedule queueRequest timeout
2017/01/18 16:18:48.477 L0 asynManager:queueTimeoutCallback
2017/01/18 16:18:48.482 L0 addr -1 queueRequest priority 3 not lockHolder
2017/01/18 16:18:48.482 L0 schedule queueRequest timeout
2017/01/18 16:18:53.477 L0 asynManager:queueTimeoutCallback
2017/01/18 16:18:53.482 L0 addr -1 queueRequest priority 3 not lockHolder
2017/01/18 16:18:53.482 L0 schedule queueRequest timeout
2017/01/18 16:18:58.477 L0 asynManager:queueTimeoutCallback
2017/01/18 16:18:58.482 L0 addr -1 queueRequest priority 3 not lockHolder
2017/01/18 16:18:58.482 L0 schedule queueRequest timeout
2017/01/18 16:19:03.477 L0 asynManager:queueTimeoutCallback
2017/01/18 16:19:03.482 L0 addr -1 queueRequest priority 3 not lockHolder
2017/01/18 16:19:03.482 L0 schedule queueRequest timeout
2017/01/18 16:19:08.477 L0 asynManager:queueTimeoutCallback
2017/01/18 16:19:08.482 L0 addr -1 queueRequest priority 3 not lockHolder
2017/01/18 16:19:08.482 L0 schedule queueRequest timeout
2017/01/18 16:19:13.477 L0 asynManager:queueTimeoutCallback
2017/01/18 16:19:13.482 L0 addr -1 queueRequest priority 3 not lockHolder
2017/01/18 16:19:13.482 L0 schedule queueRequest timeout
I did configuration for moxa, setting baudrate 9600 and operating mode -- TCP server. My RS232 is straight through.
is there any problem in networking or I am missing something in IOC??
hoping for fast reply and solution.
thanks in advance
regards..
Sunny
John A. Priller             | Phone : (517) 908-7375
MSU Cyclotron Laboratory    | Email : [email protected]
640 S. Shaw Lane room 4303  | Web   : http://people.nscl.msu.edu/~priller/
East Lansing, MI 48824      |
problem connecting serial device with moxa 5150A Sunny Badkul
RE: problem connecting serial device with moxa 5150A Mark Rivers
Prev: Re: problem connecting serial device with moxa 5150A Sunny Badkul
Next: RE: problem connecting serial device with moxa 5150A Mark Rivers
Prev: RE: problem connecting serial device with moxa 5150A Mark Rivers
