Subject: Re: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway
From: Michael Davidsaver via Tech-talk <[email protected]>
To: "Hu, Yong" <[email protected]>
Date: Tue, 14 Jul 2020 07:42:47 -0700
Yong,
As a test, have you tried sending commands through the embedded
web server in the E5810?
When timing reply latency, it would be interesting to
compare with a command which should be quick, like "ID?".
If I recall correctly, it is also possible to
send vxi11 commands to the E5810 itself.  I believe
it responds to "*IDN?".  This removes the DMM
completely from the loop.  Unless this is also slow,
I wouldn't suspect the network.
Also, since I should have asked previously.  Is the
DMM the only GPIB device connected to this E5810?
On 7/14/20 4:51 AM, Hu, Yong via Tech-talk wrote:
> Just want to sum up what I have tried so far:
> 1) I git clone asyn & streamDevice from GitHub and rebuilt everything needed for HP3458A;
> 2) I tried Mark Engbretson’s suggestion by putting an old slower speed hub in between our upgraded network hardware and this older device. In my case, the system becomes even slower;
> 3) I tried Michael Davidsaver’s commands in his GitHub link as well as those commands (out "TARM HOLD;TRIG SYN;NRDGS 1,AUTO", etc.) in his previous email. I am sure I have used his commands properly. There is no significant performance improvement, it still takes about 3 seconds to get a floating number.
> The only thing I can think of is upgrading the firmware for both E5810 LAN/GPIB Gateway (I can see its version is A.01.06 from its web server) and HP3458A DMM, which I am reluctant to do at the moment.
> I recall the reason why we selected HP3458A Multimeter 10 years ago is that it has very high precision with up to 8.5 digits. Is anyone using any Ethernet-based modern Digital Multi-Meter (DMM) with 8.5 digits?
> Thanks,
> Yong
> *From: *Mark Rivers <[email protected]>
> *Date: *Friday, July 10, 2020 at 3:22 PM
> *To: *"Hu, Yong" <[email protected]>
> *Cc: *"[email protected]" <[email protected]>
> *Subject: *RE: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway
> Hi Yong,
>   *  2) Could you briefly explain what these mean: queueRequest timeout, vxiSetEos 1, vxiSetEos 0?
>   * queueRequest timeout
> The message is actually “schedule queueRequest timeout”
> 2020/07/08 15:16:31.567 L0 schedule queueRequest timeout
> That message means that asynManager is queueing an I/O request but scheduling a timeout in case the queued request does not run before the timeout.
>   * vxiSetEos 1
> This is the message:
> 2020/07/08 15:16:31.569 asynManager::portThread port=L0 callback
> 2020/07/08 15:16:31.569 L0 vxiSetEos 1
> \n
> 0a
> This is the code that writes the first part of that message.
>     asynPrintIO(pasynUser, ASYN_TRACE_FLOW, eos, eoslen,
>             "%s vxiSetEos %d\n",pvxiPort->portName,eoslen);
>     switch (eoslen) {
>     default:
>         asynPrint(pasynUser,ASYN_TRACE_ERROR,
>            "%s vxiSetEos illegal eoslen %d\n",pvxiPort->portName,eoslen);
>         return asynError;
>     case 1:
>         pdevLink->eos = (unsigned char)eos[0];
>         break;
>     case 0:
>         pdevLink->eos = -1;
>         break;
> So that part of the message means that eoslen=1.  It appears that the VXI-11 driver only allows single-character EOS strings, whereas in general asyn allows 2-character EOS strings.  What puzzles me is that in your protocol file you have this line:
> Terminator = "\r\n";
> So that is asking for a 2 character terminator, and the first character is \r.  How come drvVxi11.c is getting a 1 character terminator that is \n?  Is streamDevice smart enough to know that VXI11 only allows single-character terminators and changes it?
> The second part of the message above it shows the terminator both in escape form (\n) and in hex (0a).  I cannot find any code that is printing that. You said you are running asyn 4.23.0.  I wonder if there could be some local edit that is printing the EOS character itself in drvVXI11.c:: vxiSetEos()?
>   * The time difference between “vxiRead” and “vxiSetEos 1” is 3.4 seconds, which means it takes 3.4 seconds to read a floating number “-4.008121752E+00”.
> That may be because, as Michael Davidsaver said yesterday, you are actually using the wrong command to request the reading.  Perhaps the “DCV 10, 1E-6” command actually takes a long time for the device to execute, and if were sending the correct command it would be faster.
> Mark
> *From:*Hu, Yong <[email protected]>
> *Sent:* Wednesday, July 8, 2020 6:24 PM
> *To:* Mark Rivers <[email protected]>
> *Cc:* [email protected]
> *Subject:* Re: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway
> Hi Mark R.,
> Thanks for your suggestion. Yes, I found the version in /usr/lib/epics/include/asynDriver.h: 4.23.0.
> Yes, you are right. Both Asyn and StreamDevice are very old. I am going to upgrade them to the latest releases.
> I have set the trace mask to 0xFF. Attached please find the detailed output messages for a few cycles. Below is the simplified version with important timestamps highlighted.
> 1) The record is scanned at every 5-second so that we get 2020/07/08 15:16:31.567 à2020/07/08 15:16:36.567 à2020/07/08 15:16:41.567, which makes perfect sense. Then it jumps to 2020/07/08 15:16:51.567, which is 10-second gap meaning one scan is skipped. This kind of skipping is confirmed by Michael’s suggestion of setting TPRO. I guess “scan5: Active SR:C03-BI{DMM:1}V-I_” means the record is scanned at 5-second and its PACT is true so that the record processing is simply skipped and it waits for the next scan.
> 2) Could you briefly explain what these mean: queueRequest timeout, vxiSetEos 1, vxiSetEos 0? I am surprised to see “\n 0a” after vxiSetEos 1. The time difference between “vxiRead” and “vxiSetEos 1” is 3.4 seconds, which means it takes 3.4 seconds to read a floating number “-4.008121752E+00”.
> scan5: Process SR:C03-BI{DMM:1}V-I_
> *scan5: Active SR:C03-BI{DMM:1}V-I_*
> */2020/07/08 15:16:31.567 /*/L0 addr 6 queueRequest priority 0 not lockHolder/
> /2020/07/08 15:16:31.567 L0 schedule queueRequest timeout/
> /2020/07/08 15:16:31.567 asynManager::portThread port=L0 callback/
> /2020/07/08 15:16:31.567 L0 addr 6 queueRequest priority 0 not lockHolder/
> /2020/07/08 15:16:31.567 L0 6 vxiWrite numchars 14/
> /2020/07/08 15:16:31.569 L0 6 vxiWrite/
> /DCV 10, 1E-6/
> /DCV 10, 1E-6\r\n/
> /44 43 56 20 31 30 2c 20 31 45 2d 36 0d 0a /
> /2020/07/08 15:16:31.569 L0 addr 6 queueRequest priority 0 from lockHolder/
> /2020/07/08 15:16:31.569 L0 schedule queueRequest timeout/
> /2020/07/08 15:16:31.569 asynManager::portThread port=L0 callback/
> /2020/07/08 15:16:31.569 L0 vxiSetEos 1/
> /\n/
> /0a /
> /2020/07/08 15:16:34.961 L0 6 vxiRead/
> /-4.008121752E+00/
> /-4.008121752E+00\r\n/
> /2d 34 2e 30 30 38 31 32 31 37 35 32 45 2b 30 30 0d 0a /
> /2020/07/08 15:16:34.961 L0 vxiSetEos 0/
> */2020/07/08 15:16:36.567/*///L0 addr 6 queueRequest priority 0 not lockHolder/
> /2020/07/08 15:16:40.407 //L0 vxiSetEos 0/
> */2020/07/08 15:16:41.567/*///L0 addr 6 queueRequest priority 0 not lockHolder/
> /2020/07/08 15:16:46.957 //L0 vxiSetEos 0/
> */2020/07/08 15:16:51.567/*///L0 addr 6 queueRequest priority 0 not lockHolder *##note:* one scan is skipped/
> /2020/07/08 15:16:55.007 //L0 vxiSetEos 0/
> */2020/07/08 15:16:56.567/*///L0 addr 6 queueRequest priority 0 not lockHolder/
> *From: *Mark Rivers <[email protected] <mailto:[email protected]>>
> *Date: *Wednesday, July 8, 2020 at 10:12 AM
> *To: *"Hu, Yong" <[email protected] <mailto:[email protected]>>
> *Cc: *"[email protected] <mailto:[email protected]>" <[email protected] <mailto:[email protected]>>
> Hi Yong,
>   * PS: I had difficulty finding the exact version number of asyn/streamdevice since I built my IOC against Michael Davidsaver’s Debian packages (https://epics.nsls2.bnl.gov/debian/) and the packages provide everything under /usr/lib/epics/
> The Debian packages must also install the asyn header files, probably in /usr/include/epics.  Look at asynDriver.h, it should contain lines like the following:
> #define ASYN_VERSION       4
> #define ASYN_REVISION     39
> #define ASYN_MODIFICATION  0
> Let me know what version it is.  Since your streamDevice is more than 7 years old I suspect that asyn is very old as well.  If so, there have been quite a few fixes to the asyn VXI-11 support since 2013.  I suggest you build a test IOC with the latest version of asyn and streamDevice and see if he problem still occurs.
> You should probably enable all of the asyn trace flags on that driver, i.e. set the trace mask to 0xFF.  There may be some hints as to the timing issues if you do that.
> Mark
> *From:*Tech-talk <[email protected] <mailto:[email protected]>> *On Behalf Of *Hu, Yong via Tech-talk
> *Sent:* Wednesday, July 8, 2020 8:38 AM
> *To:* [email protected] <mailto:[email protected]>
> *Subject:* Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway
> Hello everyone,
> I have a simple StreamDevice protocol for an old GPIB-based device HP3458A which is connected to Agilent E5810 LAN/GPIB Gateway (100 Mbps) talking to a softIOC. Here is the record to read the voltage data at 0.2Hz (5 second).
> /#only showing the must-have fields/
> /record(ai, "SR:C03-BI{DMM:1}V-I_")/
> /    field(DTYP, "stream")/
> /    field(INP,  "@hp3458a.proto ReadData L0 6")/
> /    field(SCAN, "5 second")/
> /#StreamDevice protocol/
> /ReadData {/
> /         out "DCV  10, 1E-6";/
> /         in "%f";/
> And this is the EPICS shell output after I turn on the Asyn trace.
> /2020/07/07 17:55:46.280 L0 6 vxiWrite/
> /DCV  10, 1E-6/
> /2020/07/07 17:55:49.630 L0 6 vxiRead/
> /-4.016502336E+00/
> /2020/07/07 17:55:52.396 L0 6 vxiWrite/
> /DCV  10, 1E-6/
> /2020/07/07 17:55:55.756 L0 6 vxiRead/
> /-4.015675668E+00/
> /2020/07/07 17:56:00.680 L0 6 vxiWrite/
> /DCV  10, 1E-6/
> /2020/07/07 17:56:04.030 L0 6 vxiRead/
> /-4.014579704E+00/
> I get the expected response (i.e. -4.016502336E+00) after the command (DCV  10, 1E-6) is sent. However, as you see from the timestamps for each vxiWrite, the command is not sent at every 5-second. It takes more than 5 seconds and it has a pattern: 6.116, 8.284, 6.116, 8.284, …. ; the response time (vxiRead – vxiWrite) is consistent at 3.35 seconds although it seems ridiculously slow.
> So, the system is slower than expected. However, I am surprised that there is no any alarm for the record and no any warning message from the EPICS shell output when that record can not keep up its scanning rate. Is this kind of silence supposed to be? Just to be clear, the softIOC itself and the physical server are perfectly fine: I added the classical “dbExample1.db” and its calc record can be scanned at exactly 5-second.
> This system used to work as expected at 0.2Hz (5 second) despite it is ridiculously slow. Now it becomes even slower after our ITD upgraded the network switch. If anyone has experience on these old devices (HP3458A, E5810), please advise what I should do next.
> If the version of Asyn/StreamDevice matters for this case, here we go:
> /epics> dbior/
> /Driver: drvAsyn/
> /L0 multiDevice:Yes canBlock:Yes autoConnect:Yes/
> /    vxi11, host name: 10.0.132.67/
> /Driver: stream/
> /  StreamDevice 2.6.0 built May 29 2013 18:45:57/
> /  registered bus interfaces:/
> /    DebugInterface/
> /    DummyInterface/
> /AsynDriverInterface/
> PS: I had difficulty finding the exact version number of asyn/streamdevice since I built my IOC against Michael Davidsaver’s Debian packages (https://epics.nsls2.bnl.gov/debian/) and the packages provide everything under /usr/lib/epics/. I am sure there are many ways (ldd, strings, configure/RELEASE, etc.) to find out the version number of EPICS support module. I just found “dbior” seems a handy tool for reporting EPICS driver’s version. I am not sure if Mark Rivers has added this kind of report in his latest Asyn release.
> Best regards,
> Yong Hu
> NSLS-II Controls Group
Re: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway Hu, Yong via Tech-talk
RE: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway Mark Rivers via Tech-talk
Prev: Re: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway Mark Rivers via Tech-talk
Next: RE: Modbus alarms question Mark Rivers via Tech-talk
Prev: Re: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway Hu, Yong via Tech-talk
Next: Re: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway Hu, Yong via Tech-talk
