Hi Tony,
The error messages you posted ought to be self-explanatory if you
read them in order:
> error:
> IDEA ImpMik:IDEA:Status: Input "#31 hv error detect"
mismatch after 4 bytes
> IDEA ImpMik:IDEA:Status: got "hv error detect" where "ok"
was expected
> IDEA ImpMik:IDEA:SetFilterOn: Input "#2E impossible,
erro" mismatch after 1 byte
> IDEA ImpMik:IDEA:SetFilterOn: got "2E impossible, error"
where "0" was expected
Your protocol for reading the status from the device says:
> # Get_Status
> Get_Status {
> out "Get status";
> in "#%2i ok";
> out "ver";
> in "EFC_for_NanoESCA V01.03 B0115";
> out "Set mode remote";
> in "#04 remote mode";
Thus you're telling StreamDevice that after it sends a "Get status"
string, it will get back "#" followed by a 2-digit integer and the
word "ok". The first two error message lines explain that instead of
"ok" it's seeing the string "hv error detect". Have you looked at
the instrument that it's talking to? I would guess there's a problem
with some high voltage signal that it isn't seeing. You should
probably fix the protocol file so it can accept error strings like
that from the device and make them available in a stringin record
instead of always assuming that the instrument is working and will
never give you an error, but you might be able to resolve your
immediate issue by fixing whatever the instrument is complaining
about.
Always fix your first error first, then look at any error messages
you get in that state — later errors are often caused by earlier
ones and may go away once you've fixed the first problem.
HTH,
- Andrew
Complexity is free, it's Simplicity that takes work.
- Re: Unknown error with an equipment TONY YU via Tech-talk
- Unknown error with an equipment TONY YU via Tech-talk
- Re: Unknown error with an equipment Mark Rivers via Tech-talk
RE: Epics Archiver Appliance and Network Storage slowdown Manoussakis, Adamandios via Tech-talk
<2022>
2025
<2022>
2025
ANJ, 14 Sep 2022 |
Home
News
About
Base
Modules
Extensions
Distributions
Download
Search
IRMIS
Talk
Documents
Links
Licensing