1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  <2025>  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  <2025>  2026
Subject: Re: I implemented the Model 3 of EPICS motor driver support, but the VAL and RBV of motor record are zero after IOC reboot.
To: LiangChih Chiang <[email protected]>, Torsten Bögershausen <[email protected]>
Date: Fri, 3 Jan 2025 14:25:12 +0000
I don't think you have told us:
  • Are these stepper motors or DC servos
  • Do they have encoders?
Can you try temporarily removing the pseudo motor for energy, so there is only the real motor.  Does the problem exist?
Before resorting to wireshark, I recommend just turning on ASYN_TRACE_ERROR, ASYN_TRACE_FLOW, and ASYN_TRACEIO_DRIVER in your startup script right after creating the driver.   That may help to understand what is happening.  You may also want to add additional asynPrint statements in your driver to see what it is doing.
Sent: Friday, January 3, 2025 3:35 AM
To: Torsten Bögershausen <[email protected]>
Cc: Mark Rivers <[email protected]>; [email protected] <[email protected]>
> What is special with DCM:Z1.V ?
Not special.
ThetaM axis is the most important.
>The polling, if it works for motor 1,2,3...,
>shouldn't it work for motor 0 as well ?
Yeah. That bothers me a lot.
>Did you mention that you are using a "soft motor" here ?
There is pseudo motor representing the energy of x-ray beam,
whose value is converted to and from ThetaM(Bragg angle).
>If not, it may be an idea to use wireshark to see what is
>going on between the EPICS IOC and the motion controller.
Well, by using wireshark, I don't know what parts or messages I need to see.
On Fri, Jan 3, 2025 at 4:20 PM Torsten Bögershausen <[email protected]> wrote:
Thinking about it:
On 2025-01-03 09:09, Torsten Bögershausen wrote:
> Hej again,
> happy new year !
> Thanks for reporting back and sharing your experiences.
> I think that there are 2 problems/phenomenon :
> a) The read back position of  "DCM:ThetaM.RBV" is flickering.
>    You see this already when the IOC is running.
>    Why is that ?
>    Mechanics are suspected to be a contributer.
>    Mechanical load, high-precision encoder ?
>    Is the controller running in closed loop ?
> b) The initialisation is still flaky - sometimes RBV has a value of 0.0
>    I can see 3 ways to improve it:
>    b1) Increase the sleep() even more, 5.0 seconds ??
>    b2) Use the patched version of the motor from ESS,
>        which has been improved to handle this.
>        And comes with log-prints, which allow a better remote-debug
>        using this log-prints
>        https://github.com/EuropeanSpallationSource/motor
>    b3) I have seen the use of an EPICS (?) hook somewhere in one
>       of the motor-drivers. But after searching for hooks I still
>       can not find an example here on my disk or in old emails.
> Another question:
> How much is the motor moving ?
> /Torsten
> On 2025-01-03 02:54, LiangChih Chiang wrote:
>> I added epicsThreadSleep(1.0) after creating motor controller and axes.
>> I use camonitor to trace the VAL/RBV of motor records of three axes.
>> I rebooted ioc nine times.
>> Please see the following log.
>> I noticed that axes Z1 and Z2 successfully restored the VAL/RBV from
>> the controller every time.
>> But the behavior of axis ThetaM is strange.
>> Two out of nine reboots, the VAL field of ThetaM record is
>> successfully restored.
>> One out of nine reboots, the VAL field of ThetaM record is set to zero.
>> Six out of nine reboots, the VAL field of ThetaM record is set to
>> (previous value - 0.02).
>> Why -0.02?
>> DCM:Z1.VAL                 2025-01-02 16:28:46.483683 -0.04
>> DCM:Z1.RBV                 2025-01-02 16:28:46.483683 -0.0401
>> DCM:Z2.VAL                 2025-01-02 20:41:47.713687 0.0657
>> DCM:Z2.RBV                 2025-01-02 20:41:47.713687 0.0653
>> DCM:ThetaM.VAL             2025-01-02 16:28:50.353358 7.53376  <--
>> DCM:ThetaM.RBV             2025-01-02 16:28:50.353358 7.53376
>> ioc reboot
>> DCM:Z1.VAL                 2025-01-03 08:46:20.410653 -0.0401
>> DCM:Z1.RBV                 2025-01-03 08:46:20.410653 -0.04
>> DCM:Z2.VAL                 2025-01-03 08:46:20.350513 0.0653
>> DCM:Z2.RBV                 2025-01-03 08:46:20.350513 0.0657
>> DCM:ThetaM.VAL             2025-01-03 08:46:20.500699 7.53376   <--
>> same as previous
>> DCM:ThetaM.RBV             2025-01-03 08:46:20.500699 7.53376
>> ioc reboot
>> DCM:Z1.VAL                 2025-01-03 08:54:04.045355 -0.04
>> DCM:Z1.RBV                 2025-01-03 08:54:04.045355 -0.0401
>> DCM:Z2.VAL                 2025-01-03 08:54:03.985246 0.0657
>> DCM:Z2.RBV                 2025-01-03 08:54:03.985246 0.0653
>> DCM:ThetaM.VAL             2025-01-03 08:54:05.156115 7.51376   <--
>> different -0.02
>> DCM:ThetaM.RBV             2025-01-03 08:54:05.156115 7.52741   <--
>> motor move
>> DCM:ThetaM.RBV             2025-01-03 08:54:05.335235 7.52131
>> DCM:ThetaM.RBV             2025-01-03 08:54:05.505113 7.51621
>> DCM:ThetaM.RBV             2025-01-03 08:54:05.675255 7.51376
>> DCM:ThetaM.RBV             2025-01-03 08:54:05.845205 7.51361
>> DCM:ThetaM.VAL             2025-01-03 08:54:05.845205 7.51376
>> DCM:ThetaM.RBV             2025-01-03 08:54:05.995251 7.51356
>> DCM:ThetaM.RBV             2025-01-03 08:54:06.495140 7.51371
>> DCM:ThetaM.RBV             2025-01-03 08:54:06.795186 7.51376
>> DCM:ThetaM.RBV             2025-01-03 08:54:06.965241 7.51381
>> ioc reboot
>> DCM:Z1.VAL                 2025-01-03 09:00:47.829590 -0.0401
>> DCM:Z1.RBV                 2025-01-03 09:00:47.829590 -0.04
>> DCM:Z2.VAL                 2025-01-03 09:00:47.769508 0.0653
>> DCM:Z2.RBV                 2025-01-03 09:00:47.769508 0.0657
>> DCM:ThetaM.VAL             2025-01-03 09:00:49.279478 7.49377   <--
>> different -0.01999
>> DCM:ThetaM.RBV             2025-01-03 09:00:49.279478 7.49611   <--
>> motor move
>> DCM:ThetaM.RBV             2025-01-03 09:00:49.449660 7.49371
>> DCM:ThetaM.RBV             2025-01-03 09:00:49.619596 7.49351
>> DCM:ThetaM.VAL             2025-01-03 09:00:49.619596 7.49377
>> DCM:ThetaM.RBV             2025-01-03 09:00:49.939543 7.49356
>> DCM:ThetaM.RBV             2025-01-03 09:00:50.269661 7.49376
>> Manually move ThetaM to 7.553761
>> DCM:ThetaM.VAL             2025-01-03 09:04:05.044900 7.55378
>> DCM:ThetaM.VAL             2025-01-03 09:04:05.045646 7.55376
>> DCM:ThetaM.RBV             2025-01-03 09:04:05.371519 7.49566
>> DCM:ThetaM.RBV             2025-01-03 09:04:05.541574 7.50071
>> DCM:ThetaM.RBV             2025-01-03 09:04:05.711567 7.50711
>> DCM:ThetaM.RBV             2025-01-03 09:04:05.881592 7.51536
>> DCM:ThetaM.RBV             2025-01-03 09:04:06.061484 7.52551
>> DCM:ThetaM.RBV             2025-01-03 09:04:06.231575 7.53506
>> DCM:ThetaM.RBV             2025-01-03 09:04:06.402009 7.54296
>> DCM:ThetaM.RBV             2025-01-03 09:04:06.571571 7.54866
>> DCM:ThetaM.RBV             2025-01-03 09:04:06.741529 7.55221
>> DCM:ThetaM.RBV             2025-01-03 09:04:06.911574 7.55296
>> DCM:ThetaM.VAL             2025-01-03 09:04:06.911574 7.55376
>> DCM:ThetaM.RBV             2025-01-03 09:04:07.071671 7.55301
>> DCM:ThetaM.RBV             2025-01-03 09:04:07.241982 7.55326
>> DCM:ThetaM.RBV             2025-01-03 09:04:07.561637 7.55381
>> DCM:ThetaM.RBV             2025-01-03 09:04:07.881647 7.55346
>> DCM:ThetaM.RBV             2025-01-03 09:04:08.211724 7.55356
>> DCM:ThetaM.RBV             2025-01-03 09:04:08.531501 7.55376
>> ioc reboot
>> DCM:Z1.VAL                 2025-01-03 09:05:56.102804 -0.04
>> DCM:Z1.RBV                 2025-01-03 09:05:56.102804 -0.0401
>> DCM:Z2.VAL                 2025-01-03 09:05:56.042681 0.0657
>> DCM:Z2.RBV                 2025-01-03 09:05:56.042681 0.0653
>> DCM:ThetaM.VAL             2025-01-03 09:05:56.174482 7.55376   <--
>> same as previous
>> DCM:ThetaM.RBV             2025-01-03 09:05:56.174482 7.55376   <--
>> but motor move
>> DCM:ThetaM.RBV             2025-01-03 09:05:57.012778 7.55206
>> DCM:ThetaM.RBV             2025-01-03 09:05:57.502889 7.54196
>> DCM:ThetaM.RBV             2025-01-03 09:05:57.822823 7.54366
>> DCM:ThetaM.RBV             2025-01-03 09:05:57.992678 7.54791
>> DCM:ThetaM.RBV             2025-01-03 09:05:58.162720 7.55171
>> DCM:ThetaM.RBV             2025-01-03 09:05:58.342759 7.55266
>> DCM:ThetaM.RBV             2025-01-03 09:05:58.492806 7.55271
>> DCM:ThetaM.RBV             2025-01-03 09:05:58.662711 7.55371
>> DCM:ThetaM.RBV             2025-01-03 09:05:58.962828 7.55376
>> DCM:ThetaM.RBV             2025-01-03 09:05:59.132771 7.55381
>> DCM:ThetaM.RBV             2025-01-03 09:06:02.762853 7.55386
>> ioc reboot
>> DCM:Z1.VAL                 2025-01-03 09:09:37.385148 -0.0401
>> DCM:Z1.RBV                 2025-01-03 09:09:37.385148 -0.04
>> DCM:Z2.VAL                 2025-01-03 09:09:37.325204 0.0653
>> DCM:Z2.RBV                 2025-01-03 09:09:37.325204 0.0657
>> DCM:ThetaM.VAL             2025-01-03 09:09:38.354986 0         <--
>> different, zero!
>> DCM:ThetaM.RBV             2025-01-03 09:09:38.354986 7.55181   <--
>> motor move
>> DCM:ThetaM.RBV             2025-01-03 09:09:38.535008 7.54706
>> DCM:ThetaM.RBV             2025-01-03 09:09:38.704917 7.54026
>> DCM:ThetaM.RBV             2025-01-03 09:09:38.875048 7.53131
>> DCM:ThetaM.RBV             2025-01-03 09:09:39.045385 7.52046
>> DCM:ThetaM.RBV             2025-01-03 09:09:39.215134 7.50731
>> DCM:ThetaM.RBV             2025-01-03 09:09:39.385166 7.49361
>> DCM:ThetaM.RBV             2025-01-03 09:09:39.555112 7.48036
>> DCM:ThetaM.RBV             2025-01-03 09:09:39.725153 7.46636
>> DCM:ThetaM.RBV             2025-01-03 09:09:39.894942 7.45281
>> DCM:ThetaM.RBV             2025-01-03 09:09:40.064989 7.43961
>> DCM:ThetaM.RBV             2025-01-03 09:09:40.234971 7.42526
>> DCM:ThetaM.RBV             2025-01-03 09:09:40.405113 7.41221
>> DCM:ThetaM.RBV             2025-01-03 09:09:40.585072 7.39761
>> DCM:ThetaM.RBV             2025-01-03 09:09:40.755044 7.38401
>> DCM:ThetaM.RBV             2025-01-03 09:09:40.925113 7.37066
>> DCM:ThetaM.RBV             2025-01-03 09:09:42.095077 7.31426
>> DCM:ThetaM.VAL             2025-01-03 09:09:42.095077 7.31426  <-- I
>> stop the motor.
>> DCM:ThetaM.RBV             2025-01-03 09:09:46.705116 7.31421
>> ioc reboot
>> DCM:Z1.VAL                 2025-01-03 09:12:18.906995 -0.04
>> DCM:Z1.RBV                 2025-01-03 09:12:18.906995 -0.0401
>> DCM:Z2.VAL                 2025-01-03 09:12:18.846984 0.0657
>> DCM:Z2.RBV                 2025-01-03 09:12:18.846984 0.0653
>> DCM:ThetaM.VAL             2025-01-03 09:12:21.346979 7.29431   <--
>> different −0.01995
>> DCM:ThetaM.RBV             2025-01-03 09:12:21.346979 7.29431   <--
>> motor move
>> ioc reboot
>> DCM:Z1.VAL                 2025-01-03 09:18:20.230656 -0.0401
>> DCM:Z1.RBV                 2025-01-03 09:18:20.230656 -0.04
>> DCM:Z2.VAL                 2025-01-03 09:18:20.170608 0.0653
>> DCM:Z2.RBV                 2025-01-03 09:18:20.170608 0.0657
>> DCM:ThetaM.VAL             2025-01-03 09:18:20.311337 7.27437   <--
>> different −0.01994
>> DCM:ThetaM.RBV             2025-01-03 09:18:20.311337 7.29431   <--
>> motor move
>> DCM:ThetaM.RBV             2025-01-03 09:18:21.150538 7.29231
>> DCM:ThetaM.RBV             2025-01-03 09:18:21.320404 7.28771
>> DCM:ThetaM.RBV             2025-01-03 09:18:21.500407 7.28171
>> DCM:ThetaM.RBV             2025-01-03 09:18:21.670389 7.27681
>> DCM:ThetaM.RBV             2025-01-03 09:18:21.840686 7.27421
>> DCM:ThetaM.RBV             2025-01-03 09:18:22.010457 7.27406
>> DCM:ThetaM.VAL             2025-01-03 09:18:22.010457 7.27437
>> DCM:ThetaM.RBV             2025-01-03 09:18:22.330451 7.27421
>> DCM:ThetaM.RBV             2025-01-03 09:18:22.660350 7.27431
>> DCM:ThetaM.RBV             2025-01-03 09:18:22.960309 7.27436
>> DCM:ThetaM.RBV             2025-01-03 09:18:23.130896 7.27441
>> ioc reboot
>> DCM:Z1.VAL                 2025-01-03 09:22:00.722740 -0.04
>> DCM:Z1.RBV                 2025-01-03 09:22:00.722740 -0.0401
>> DCM:Z2.VAL                 2025-01-03 09:22:00.662695 0.0657
>> DCM:Z2.RBV                 2025-01-03 09:22:00.662695 0.0653
>> DCM:ThetaM.VAL             2025-01-03 09:22:02.912859 7.25442   <--
>> different −0.01995
>> DCM:ThetaM.RBV             2025-01-03 09:22:02.912859 7.25441   <--
>> motor move
>> ioc reboot
>> DCM:Z1.VAL                 2025-01-03 09:33:58.400417 -0.0401
>> DCM:Z1.RBV                 2025-01-03 09:33:58.400417 -0.04
>> DCM:Z2.VAL                 2025-01-03 09:33:58.340350 0.0653
>> DCM:Z2.RBV                 2025-01-03 09:33:58.340350 0.0657
>> DCM:ThetaM.VAL             2025-01-03 09:33:59.330344 7.23447   <--
>> different −0.01995
>> DCM:ThetaM.RBV             2025-01-03 09:33:59.330344 7.25241   <--
>> motor move
>> DCM:ThetaM.RBV             2025-01-03 09:33:59.500272 7.24781
>> DCM:ThetaM.RBV             2025-01-03 09:33:59.680584 7.24186
>> DCM:ThetaM.RBV             2025-01-03 09:33:59.850291 7.23701
>> DCM:ThetaM.RBV             2025-01-03 09:34:00.020355 7.23431
>> DCM:ThetaM.RBV             2025-01-03 09:34:00.190267 7.23416
>> DCM:ThetaM.VAL             2025-01-03 09:34:00.190267 7.23447
>> DCM:ThetaM.RBV             2025-01-03 09:34:00.510337 7.23431
>> DCM:ThetaM.RBV             2025-01-03 09:34:00.840415 7.23441
>> DCM:ThetaM.RBV             2025-01-03 09:34:01.140389 7.23446
>> DCM:ThetaM.RBV             2025-01-03 09:34:01.310255 7.23451
>> On Thu, Jan 2, 2025 at 11:26 AM LiangChih Chiang <[email protected]
>> <mailto:[email protected]>> wrote:
>>     Hi,
>>      >To explain a little more, it is important that the driver has
>>     completed the first poll cycle for the position and status of each
>>     axis before iocInit.
>>      >This is because the motor record looks at those positions during
>>     iocInit, and if there are non-zero it uses those values as the
>>     current position.
>>      >You can add the following to your startup script after you create
>>     the controller and axes to sleep for 1 second.
>>      >epicsThreadSleep(1.0)
>>     I did add epicsThreadSleep(1.0).
>>     There are 5 motors(axis 1 to 5) connected to the motor controller.
>>     After epicsThreadSleep(1.0) added, it seems that the VAL/RBV of
>>     motor records of axis 2 to 5 can be successfully loaded back from
>>     the controller after IOC reboot.
>>     But the behavior of axis 0 (this axis is the most important) is
>> strange.
>>     After IOC reboot, the motor record of axis 0 will move.
>>     It seems that the VAL field is set to a new value
>>     and I can't figure out the relationship between the new and old
>> values.
>>     The axis 0 is the main axis(Bragg angle) of a double crystal
>>     monochromator,
>>     and it has corresponding pseudo motor to represent the Energy of x-
>>     ray beam.
>>     On Wed, Nov 20, 2024 at 11:04 PM Mark Rivers
>>         each axis before iocInit.  This is because the motor record
>>         looks at those positions during iocInit, and if there are non-
>>         zero it uses those values as the current position.  You can add
>>         the following to your startup script after you create the
>>         Let us know if this does not fix the problem.
>>         Mark
>>         *From:* Tech-talk <[email protected] <mailto:tech-
>>         [email protected]>> on behalf of Torsten Bögershausen via
>>         Tech-talk <[email protected] <mailto:[email protected]>>
>>         *Sent:* Wednesday, November 20, 2024 5:15 AM
>>         *To:* LiangChih Chiang <[email protected]
>>         <mailto:[email protected]>>; [email protected]
>>         <mailto:[email protected]> <[email protected]
>>         *Subject:* Re: I implemented the Model 3 of EPICS motor driver
>>         reboot.
>>         Hej,
>>         without having seen ypur code, I would speculate that you run
>>         into a not-unknown issue using Model 3 motors, connected via
>>         a serial cable, USB or Ethernet?
>>         The value that the motorRecord needs is read from the driver,
>>         wich must have been able to read it from your hardware.
>>         Here are ESS we have a patched motor, which can handle this.
>>         If you use the upstream version, the easist workaround may be
>>         to add a sleep() after the createController() line in your st.cmd
>>         HTH
>>         /Torsten
>>         On 2024-11-20 09:47, LiangChih Chiang via Tech-talk wrote:
>>          > Hello, EPICS mates.
>>          > I implemented the Model 3 of EPICS motor driver support for
>>         some motor
>>          > controller,
>>          > but the VAL and RBV of the motor record are zero after IOC
>>         reboot.
>>          > I think the motor record doesn't restore the values of motor
>>         pulse and
>>          > encoder count from the motor controller after IOC reboot.
>>          > What member functions of asynMotorController and
>>         asynMotorAxis classes
>>          > should I implement?
>>          > Best regards.
Re: I implemented the Model 3 of EPICS motor driver support, but the VAL and RBV of motor record are zero after IOC reboot. LiangChih Chiang via Tech-talk
Re: I implemented the Model 3 of EPICS motor driver support, but the VAL and RBV of motor record are zero after IOC reboot. Torsten Bögershausen via Tech-talk
Prev: Re: I implemented the Model 3 of EPICS motor driver support, but the VAL and RBV of motor record are zero after IOC reboot. Torsten Bögershausen via Tech-talk
Next: Re: I implemented the Model 3 of EPICS motor driver support, but the VAL and RBV of motor record are zero after IOC reboot. LiangChih Chiang via Tech-talk
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  <2025>  2026
Prev: Re: I implemented the Model 3 of EPICS motor driver support, but the VAL and RBV of motor record are zero after IOC reboot. LiangChih Chiang via Tech-talk
