Hello Mark
Yes, we are using the serial port to communicate with the detector. I turned on asynTrace options and I noticed that the freezing and
unresponsive buttons issues are caused due to a timeout during acquisition, this is confirmed because “timeout handler” message gets printed almost every 1 second (not all the time), and after that the temperature PV gets processed. So I switched the scan
rate on the main GUI to passive just to test and the acquisition works fine for now.
Also, I noticed that the spectrum takes 1 or 2 seconds more than the real time, I looked into the MCA setup GUI and changed the “Read
spectrum every” setting to “Event”, seems a bit more responsive than before. I attached the current MCA settings screenshot. What do you think?
Best Regards,
Abdalla Al-Dalleh
Control Engineer
SESAME
Are you communicating with the XFlash over a serial port or over their QM100 high-speed interface?
If it is over a serial port then you should enable asynTrace to see the communication stream.
Hi
We have the MCA module running on a Windows 10 machine to control the XFlash detector from Bruker. We started to face the following issues:
When the IOC starts, the state is always acquiring without pressing the start button.
The buttons to erase/stop/start the acquisition have no effect, they do not stop or start the acquisition.
The real time and live time show weird values on startup, like in 10s or 100s. of seconds.
If I wait long enough, the acquisition stops but pressing the start button again, say for a real time of 1 second, it starts acquiring for fraction
of a second and then freezes, nothing is progressing.
Despite what is mentioned, the IOC log shows no error messages. The Bruker software seems to acquire just fine, but not tested thoroughly. We even
Installed the IOC on a different laptop and a PC and still the same behavior. Any idea on what might went wrong with the detector or IOC?
Best Regards,
Abdalla Al-Dalleh
Control Engineer
SESAME
Description: current-settings.png
- Re: Issues in MCA module with XFlash detector Mark Rivers via Tech-talk
- Issues in MCA module with XFlash detector Abdalla Ahmad via Tech-talk
RE: Issues with drvAsynIPServerPort for UDP Mark Rivers via Tech-talk
Fix buffer overflow issue in Epics 3.15 Soltani Farani, Bushra via Tech-talk
<2025>
2026
<2025>
2026
ANJ, 19 Mar 2026 |
Home
News
About
Talk
Base
Modules
Extensions
Distributions
Download
Documents
Links
Licensing