1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  <20062007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  <20062007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026
Subject: RE: Cost effective solution for monitoring a large number of temperatures
To: "Mark Rivers" <[email protected]>, "Josh Stein" <[email protected]>
Cc: "TechTalk EPICS" <[email protected]>
Date: Thu, 2 Feb 2006 11:02:14 -1000
Hi All
After many years of battling with serial devices in a noisy environment
I now tend towards devices which have an option for enabling a protocol
and/or a checksum or CRC. Doing that allows detecting and tossing bad
data, which is not feasible with pure ASCII devices. We have some
devices at the Keck observatory that occasionally get reprogrammed
simply from single bit changes in the data stream (distance is many
hundreds of feet and communications must pass through slip rings, which
are brushes dragging across copper strips)!
I do not know if the Keithley units have such features but...
AH
-----Original Message-----
Sent: Thursday, February 02, 2006 10:48 AM
To: Josh Stein
Cc: TechTalk EPICS
temperatures
Hi Josh,
I don't have hard numbers for the 2700.  On the older 2000 with a 10
channel card we can get a little faster than 1Hz if all 10 channels are
active I think (=10 readings per second). They are mechanical switches
for selecting the active input.
> From: Josh Stein [mailto:[email protected]]
> Sent: Thursday, February 02, 2006 2:16 PM
> To: Mark Rivers
> Cc: TechTalk EPICS
> Subject: Re: Cost effective solution for monitoring a large
> Hey Mark,
>   The Keithly 2700 looks like the ticket. The other solutions
> may be a
> bit more 'techy' but I'm aiming for COTS as much as possible. It also
> appears that Keithly is offering an 40 channel module with
> cold junction
> compensation now - I'll look into that one. What rates are
> you able to
> scan this beast at? I imagine it is hardware bound.
> Thanks,
> <josh>
> Mark Rivers wrote:
> >Hi Josh,
> >In synApps we support the Keithley 2700.  Each of those
> units takes 2 cards for a total of 40 thermocouples.  It
> costs about $2K, so $50 per thermocouple.  It can be extended
> with external multiplexors, but EPICS support would need to
> be written for those.
> >Mark
> >________________________________
> >Sent: Thu 1/19/2006 9:03 AM
> >To: TechTalk EPICS
> >Subject: Cost effective solution for monitoring a large
> >Controls experts,
> >  In the undulator hall of the LCLS we expect to monitor (what I
> >consider) a large number of temperature points (on the order of about
> >300). My past experience with thermocouples has been local
> installations
> >of between one and eight data points. It would be easy to just scale
> >that solution by a factor of 40, but I have to believe there are more
> >cost effective solutions.
> >  Can anyone with experience in this regard please point me
> in the right
> >direction? I'm looking for hardware that is, of course,
> EPICS compliant.
> >Thanks,
> ><josh>
Re: Cost effective solution for monitoring a large number of temperatures Josh Stein
Prev: RE: Cost effective solution for monitoring a large number of temperatures Mark Rivers
Next: RE: Cost effective solution for monitoring a large number of temperatures Mark Rivers
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  <20062007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026
Next: Re: Cost effective solution for monitoring a large number of temperatures Josh Stein
