1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  <2009>  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  <2009>  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026
Subject: Agilent E2050
To: <[email protected]>
Date: Tue, 14 Jul 2009 09:41:23 -0700
I am running a test soft IOC, which performs a few simple functions on a GPIB instrument.  The instrument is connected to Agilent E2050 gateway.  The IOC is periodically reading the Status Byte.  After a few hours the communication between the IOC and the device is broken.  This is what I get:
02009/07/14 08:50:24.312 L0 vxi11 clientIoCall errno Connection reset by peer clnt_stat 4
2009/07/14 08:50:24.312 L0 vxiDestroyDevLink RPC error : RPC: Unable to send; errno = Broken pipe
2009/07/14 08:50:24.312 L0 vxiDisconnectPort : RPC: Unable to send; errno = Broken pipe
2009/07/14 08:50:26.314 L0 vxiDestroyDevLink RPC error : RPC: Unable to send; errno = Broken pipe
L0 port not connected
2009/07/14 08:50:26.314 L0 asynGpib:pollAddr srqEnable
2009/07/14 08:50:26.314 MAGS:ASYN:IO: Write error, nout=0, L0 RPC failed
2009/07/14 08:50:26.314 L0 port not connected
2009/07/14 08:50:26.314 MAGS:ASYN:IO: Error L0 port is not connected
2009/07/14 08:50:26.314 MAGS:ASYN:IO: error  nread 162331380 L0 port is not connected
2009/07/14 08:50:26.314 MAGS:ASYN:IO: Overflow nread 162331380 L0 port is not connected
I have two E2050 devices and both exhibits the same behavior.  The length of time until the communication falls apart varies.  It could be as short as an hour or so, or as long as 10 hours.
Here is the E2050 configuration:
Configuration Parameters
  hostname:        E2050-esalangpib2 # Internet domain name
  hardware-addr:   0800091B1EEB    # Ethernet station address
  ip:              134.79.59.12    # Internet Protocol address
  default-gw:      134.79.59.1     # Default subnet gateway IP address
  subnet-mask:     255.255.252.0   # Network subnet mask
  syslog-svr:      0.0.0.0         # Syslog server IP address
  bootp:           OFF             # Obtain config via BOOTP/TFTP
  lan-timeout:     0               # LAN connect timeout in seconds
  io-timeout:      120             # Server I/O timeout in seconds
  allow:           *               # IP allow list
  hpib-address:    0               # HP-IB System Controller Address
  hpib-name:       gpib0           # HP-IB interface symbolic name
  hpib-unit:       7               # HP-IB logical unit number
Is this problem due to improper use of the gateway or are these devices malfunctioning?
Any ideas are most welcome.
Zen
Re: Agilent E2050 Benjamin Franksen
Prev: ANN: (no)fun-0-.4.1 Benjamin Franksen
Next: When an external device does not provide termination on TCP ASCII stream replies Allison, Stephanie
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  <2009>  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026
Next: Re: Agilent E2050 Benjamin Franksen
ANJ, 31 Jan 2014 · Home · News · About · Talk · Base · Modules · Extensions ·