As Ralph pointed out, you have set SCAN=I/O Intr. For this to work your
asyn port driver (called ENV in your case) must do callbacks to device
support when it receives a new value. Does this driver do so?
You can get information about this by typing
asynReport 10 ENV
That should show you what asyn clients (e.g. device support) have
registered for callbacks on which interfaces.
What happens if you change SCAN to "1 second"?
-----Original Message-----
[mailto:[email protected]] On Behalf Of Hinko Kocevar
Sent: Tuesday, May 11, 2010 9:07 AM
To: [email protected]
Subject: Re: No monitor refresh
Hello Ralph,
On 05/11/10 15:21, Ralph Lange wrote:
> Hello Hinko,
> what EPICS versions are you using for client and IOC?
IOC and client use 3.14.10 EPICS base version.
> Can you show a dump (dbpr <name>, 4) of the record that you are
> connecting to?
Sure, here it is:
epics> dbpr LIBERA01:ENV:ENV_AGC_MONITOR 4
ACKS: NO_ALARM ACKT: YES ASG: ASP:
0x00000000
BKPT: 0x00 COSV: NO_ALARM DESC: AGC DISA: 0
DISP: 0 DISS: NO_ALARM DISV: 1 DPVT:
0x50c72c00
DSET: 0x7c620840 DTYP: asynInt32 EIST: EISV:
NO_ALARM
EIVL: 0x0 ELST: ELSV: NO_ALARM ELVL: 0x0
EVNT: 0 FFST: FFSV: NO_ALARM FFVL: 0x0
FLNK:CONSTANT 0 FRST: FRSV: NO_ALARM FRVL: 0x0
FTST: FTSV: NO_ALARM FTVL: 0x0 FVST:
FVSV: NO_ALARM FVVL: 0x0 INP:INST_IO @asyn(ENV, 725, 1.0)
LALM: 1 LCNT: 0 LSET: 0xe8752e00 MASK:
4294967295
MLIS: 0x38d0320060d0320003000000 MLOK: 0x28700b00 MLST: 1
NAME: LIBERA01:ENV:ENV_AGC_MONITOR NIST: NISV:
NO_ALARM
NIVL: 0x0 NOBT: 0 NSEV: NO_ALARM NSTA:
NO_ALARM
ONST: AUTO ONSV: NO_ALARM ONVL: 0x1 ORAW: 1
PACT: 0 PHAS: 0 PINI: YES PPN:
0x00000000
PPNR: 0x00000000 PRIO: LOW PROC: 0 PUTF: 0
RDES: 0xa8a10300 RPRO: 0 RSET: 0x34b70a40 RVAL: 1
SCAN: I/O Intr SDEF: 1 SDIS:CONSTANT SEVR:
NO_ALARM
SHFT: 0 SIML:CONSTANT SIMM: NO SIMS:
NO_ALARM
SIOL:CONSTANT SPVT: 0x388a2f00 STAT: NO_ALARM SVAL: 0
SVST: SVSV: NO_ALARM SVVL: 0x0 SXST:
SXSV: NO_ALARM SXVL: 0x0 TEST: TESV:
NO_ALARM
TEVL: 0x0 THST: THSV: NO_ALARM THVL: 0x0
TIME: 2010-05-11 11:18:40.100440000 TPRO: 0 TSE: 0
TSEL:CONSTANT TTST: TTSV: NO_ALARM TTVL: 0x0
TVST: TVSV: NO_ALARM TVVL: 0x0 TWST:
TWSV: NO_ALARM TWVL: 0x0 UDF: 0 UNSV:
NO_ALARM
VAL: 1 ZRST: MANUAL ZRSV: NO_ALARM ZRVL: 0x0
> What field of that record are you connecting to?
I did no specify any field - so it is VAL, I guess by default.
Thank you!
Hinko
> Regards,
> Ralph
> On Tue 11 May 2010 6:26:48 Hinko Kocevar wrote:
>> Hi all,
>> I'm experiencing some strange behavior when using camonitor against
an
>> EPICS IOC.
>> For some reason a CA_PROTO_EVENT_ADD response from IOC to CA client
does
>> not arrive, or arrives several minutes later after the PV value has
>> changed. Note that first CA_PROTO_EVENT_ADD response is seen always,
>> only latter CA_PROTO_EVENT_ADD responses do not arrive (or arrive
late)?!
>> This behavior is not seen always - sometimes the IOC would promptly
>> respond with CA_PROTO_EVENT_ADD response when needed.
>> I can see the CA_PROTO_ECHO requests/responses fine on the same
>> camonitor channel link, only CA_PROTO_EVENT_ADD response is not seen.
>> Inspecting the IOC I can see that the value did change (on lower
levels)
>> and caput response is also valid - updated with new value.
>> I've ruled out the network problems, since the issue is raised even
when
>> using cross-over cable to communicate with the IOC. Actually even I
>> update the value in the IOC shell, the camonitor is not refreshed!!
>> Has someone experienced something similar?
>> Are there some conditions that need to fulfilled in order for IOC to
>> send a subsequent CA_PROTO_EVENT_ADD response(s)?
>> TIA & HAND,
>> Hinko
Hinko Kocevar
Technical support software engineer
Instrumentation Technologies
Velika pot 22, SI-5250 Solkan - Slovenia
T:+386 5 3352600, F:+386 5 3352601
mailto: [email protected]
http://www.i-tech.si - When your users demand stability
The information transmitted is intended solely for the addressee and may
contain confidential and/or privileged information. Any review,
retention,
disclosure or other use by persons other than the intended recipient is
prohibited. If you received this in error, please notify the sender and
delete all copies.
- Re: No monitor refresh Hinko Kocevar
- Re: No monitor refresh Ralph Lange
<2010>
2025
2026
RE: No monitor refresh Mark Rivers
<2010>
2025
2026