1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  <20182019  2020  2021  2022  2023  2024  2025  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  <20182019  2020  2021  2022  2023  2024  2025  2026
Subject: RE: Mythen M1K - Low fps
To: 'Nilson Pereira' <[email protected]>, "[email protected]" <[email protected]>, "[email protected]" <[email protected]>
Date: Thu, 15 Feb 2018 18:21:34 +0000
Hi Nilson,
I don't have any experience with this device, but I've just had a quick look at the driver.
Are you using Readmode=0 (raw) or Readmode=1 (corrected)?
This is the loop of the thread that reads the device.
    while (1) {
        /* Is acquisition active? */
        getIntegerParam(ADAcquire, &acquire);
        /* If we are not acquiring then wait for a semaphore that is given when acquisition is started */
        if (!acquire || !acquiring_)  {
            setIntegerParam(ADStatus, ADStatusIdle);
            callParamCallbacks();
            /* Release the lock while we wait for an event that says acquire has started, then lock again */
            asynPrint(this->pasynUserSelf, ASYN_TRACE_FLOW,
                "%s:%s: waiting for acquire to start\n", driverName, functionName);
            this->unlock();
            eventStatus = epicsEventWait(this->startEventId_);
            // setStringParam(ADStatusMessage, "Acquiring data");
            // setIntegerParam(ADNumImagesCounter, 0);
              // getIntegerParam(ADAcquire, &acquire);
              //printf("Read Mode: %d\tnModules: %d\t chanperline: %d\n", readmode_,this->nmodules,chanperline_);
            if (readmode_==0)       //Raw Mode
                nread_expect = sizeof(epicsInt32)*this->nmodules*(1280/chanperline_);
            else
                nread_expect = sizeof(epicsInt32)*this->nmodules*(1280);
            dataOK = 1;
            eventStatus = getStatus();
            setIntegerParam(ADStatus, eventStatus);
            if (eventStatus!=ADStatusError) {
              getDoubleParam(ADAcquireTime,&acquireTime);
              // printf("Acquisition start - expect %d\n",nread_expect);
              // Work on the cases of what are you getting from getstatus
              do {
                nread=0;
                if (readmode_==0)
                  strcpy(outString_, "-readoutraw");
                else
                  strcpy(outString_, "-readout");
                status = pasynOctetSyncIO->writeRead(pasynUserMeter_, outString_, strlen(outString_), (char *)detArray_,
                                        nread_expect, M1K_TIMEOUT+acquireTime, &nwrite, &nread, &eomReason);  //Timeout is M1K_TIMEOUT + AcquireTime
                //printf("nread_expected = %d\tnread = %d\n", nread_expect,nread);
                if(nread == nread_expect) {
                    this->lock();
                    dataOK = dataCallback(detArray_);
                    this->unlock();
                    if (!dataOK) {
                else {
                  //printf("Data not size expected ADStatus: %d\n",eventStatus);
                if(status != asynSuccess) {
                    asynPrint(pasynUserSelf, ASYN_TRACE_ERROR,
                          "%s:%s: error using readout command status=%d, nRead=%d, eomReason=%d\n",
                          driverName, functionName, status, (int)nread, eomReason);
              while (status == asynSuccess && (eventStatus==ADStatusAcquire||eventStatus==ADStatusReadout) && acquiring_);
           this->lock();
        if (eventStatus!=ADStatusError ) {
          printf("Acquisition finish\n");
          getIntegerParam(ADImageMode, &imageMode);
          if (imageMode == ADImageSingle || imageMode == ADImageMultiple) {
            printf("ADAcquire Callback\n");
            acquiring_ = 0;
            setIntegerParam(ADAcquire,  0);
        else {
          //Abort read
                "%s:%s: error timed out waiting for data\n",
          acquiring_ = 0;
            setAcquire(0);
I see a number of issues in this code.  But these will not impact performance.
- It has the lock() released much of the time.  It should only be released during the time-intensive operations of epicsEventWait and pasynOctetSyncIO->writeRead().
- It does not check the status return from pasynOctetSyncIO->writeRead()  before checking nread.  It should do it first.
- This line should be changed to asynPrint(pasynUserSelf, ASYN_TRACE_ERROR) because it is important to know if that happened.
-          This line should be changed to asynPrint(pasynUserSelf, ASYN_TRACEIO_DRIVER) so one can enable that message to get the information.
The inner do while() loop looks very fast and straightforward assuming pasynOctetSyncIO->writeRead() always succeeds, nread=nread_expect, and dataCallback() returns 1, which it should always do.
I think you need to turn on ASYN_TRACEIO_DRIVER to the underlying drvAsynIPPort and capture the communications with the device into a file.  You can do this with commands in st.cmd or with the asynRecord OPI screen.  That will let you look at the timing: is the device responding promptly, and is the IOC sending the next command promptly after receiving the previous data?  You may be correct that to get more than 100 frames/s it will be necessary to read multiple images per transfer, but you should get this diagnostic information before doing that.
> -----Original Message-----
> From: Nilson Pereira [mailto:[email protected]]
> Sent: Thursday, February 15, 2018 11:46 AM
> To: Mark Rivers <[email protected]>; [email protected]; [email protected]
> Subject: Re: Mythen M1K - Low fps
> Hello Mark and Mark,
> The image size is only around 3.1KB per frame (1028 bins with 24 bits
> each).
> I am only running 1 NDStardardArray plugin and it seems to be fine.
> There are no dropped frames that could be observed by the plugin and it
> is configured with blocking "Disable". The same behavior is observed
> with all plugins disabled.
> It Is the array count itself on the IOC that shows the discrepancy
> between the numbers of frames requested and received.
> On the detector side, I can use the same settings to acquire data within
> its own web interface application. The aplication is capable of
> acquiring up to 1000 fps with no loss. So I believe it should be between
> the detector and the IOC.
> Any suggestions?
> Nilson
> On 15.02.2018 18:31, Mark Rivers wrote:
> > Hi Nilson,
> > When you say it is dropping frames, can you be more explicit?  Can you tell where the
> frames are being dropped, i.e. between the detector and the IOC, or in an IOC plugin?
> > What plugins are you running in this IOC?
> > Are they configured with blocking=Enabled or Disabled?
> > What queue size are the plugins using?
> > Mark
> >> From: [email protected] [mailto:[email protected]] On
> Behalf
> >> Of Nilson Pereira
> >> Sent: Thursday, February 15, 2018 10:50 AM
> >> To: [email protected]
> >> Subject: Mythen M1K - Low fps
> >> Hello everyone,
> >> I started to work with a Dectris strip Detector Mythen2 X 1K and I am
> >> using the Mythen module from AreaDetector. This detector can reach
> >> 1000fps, however the IOC starts to drop frames with frequencies higher
> >> than 100fps.
> >> Has anyone experience this limitation with this type of detector?
> >> To avoid network issues, I have a straight network connection between
> >> the PC and detector.
> >> I also noticed that the IOC only requests 1 frame at a time when it
> >> could (according to the manual) requests multiples frames to avoid
> >> communication overheads. Has anyone tried implementing this?
> >> My versions:
> >> - Ubuntu 16.04
> >> - EPICS base 3.15.5
> >> - ADCore 2-6
> >> - ADMythen 2-0
> >> ________________________________
> >> Helmholtz-Zentrum Berlin für Materialien und Energie GmbH
> >> Mitglied der Hermann von Helmholtz-Gemeinschaft Deutscher Forschungszentren e.V.
> >> Aufsichtsrat: Vorsitzender Dr. Karl Eugen Huthmacher, stv. Vorsitzende Dr. Jutta
> Koch-
> >> Unterseher
> >> Geschäftsführung: Prof. Dr. Bernd Rech (kommissarisch), Thomas Frederking
> >> Sitz Berlin, AG Charlottenburg, 89 HRB 5583
> >> Postadresse:
> >> Hahn-Meitner-Platz 1
> >> D-14109 Berlin
> >> http://www.helmholtz-berlin.de
> Aufsichtsrat: Vorsitzender Dr. Karl Eugen Huthmacher, stv. Vorsitzende Dr. Jutta Koch-
> Unterseher
> Postadresse:
> D-14109 Berlin
Re: Mythen M1K - Low fps Nilson Pereira
RE: Mythen M1K - Low fps Mark Rivers
Prev: Re: Mythen M1K - Low fps Nilson Pereira
Next: Re: PMAC asyn autoconnect issue Bruno Martins
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  <20182019  2020  2021  2022  2023  2024  2025  2026
Next: Re: Mythen M1K - Low fps Nilson Pereira
ANJ, 16 Feb 2018 · Home · News · About · Talk · Base · Modules · Extensions ·
