1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  <2018>  2019  2020  2021  2022  2023  2024  2025  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  <2018>  2019  2020  2021  2022  2023  2024  2025  2026
Subject: RE: :tdir messages in ioc for Pmac
To: "'Gofron, Kazimierz'" <[email protected]>, "[email protected]" <[email protected]>
Date: Tue, 22 May 2018 19:08:20 +0000
We have a similar problem with the Newport XPS.  The problem occurs because the encoder is indicating that the motor is moving in the “wrong” direction because of overshoot, etc. and the motor record then stops the motion.  IMHO this is a design error in the motor record, it should not stop the motion when it thinks there is a problem, that is the job of the driver or controller.
We work around the problem by setting the NTM field to 0 by adding the following commands to the startup script after iocInit.
dbpf("13BMD:m89.NTM","0")
dbpf("13BMD:m90.NTM","0")
dbpf("13BMD:m91.NTM","0")
Note that I have not tested that this is still necessary with the latest release of the motor record, so perhaps the problem has been fixed and we no longer need this workaround.
I suggest you try this and see if it fixes the problem for you.
From: [email protected] [mailto:[email protected]] On Behalf Of Gofron, Kazimierz
Sent: Tuesday, May 22, 2018 1:37 PM
To: [email protected]
Subject: :tdir messages in ioc for Pmac
Perhaps someone resolved this problem.
My turbo pmac motors sometimes get 'stuck', mainly when tweaking them, and particularly when motion direction is reversed. Pressing tweak value several times increases requested position, but no motion occurs. Pressing 'Kill" results in motion to proceed to the requested position. In the same time, there are ':tdir' messages in the IOC itself:
epics> XF:11IDB-ES{Dif-Ax:Gam}Mtr:tdir = 1
XF:11IDB-ES{Dif-Ax:Gam}Mtr:tdir = 0
XF:11IDB-ES{Dif-Ax:Del}Mtr:tdir = 0
epics> XF:11IDB-ES{Dif-Ax:Del}Mtr:tdir = 0
EPICS:
  base-3.14.12.3-8
    streamdevice 2.6
    sncseq 2.1.17
    asyn 4.23
    vme 2.8.1
    vac 1.4.1
    std 3.2
    softGlue 2.4.1
    sscan 2.9
    mdautils 1.3.1
    optics 2.9.2
    motor 6.8.1
    modbus 2.7
    mca 7.5
    love 3.2.5
    ipUnidig 2.9
    i330 2.7
    ip 2.15
    ipac 2.12
    delaygen 1.1.0
    dc128V 2.7
    calc 3.4
    busy 1.6
System: Debian 7.11 (wheezy), 3.2.0-4-amd64 #1 SMP Debian 3.2.89-2 x86_64 GNU/Linux
epics> dbpr XF:11IDB-ES{Dif-Ax:Gam}Mtr 4
ACCL: 0.2           ACKS: NO_ALARM      ACKT: YES           ADEL: 0
ALST: 0             ASG:                ASP: (nil)          ATHM: 0
BACC: 0.5           BDST: 0             BKPT: 00            BVEL: 0
CARD: 0             CBAK: 0xe81cc0      CDIR: 0             CNEN: Disable
DCOF: 0             DESC: Gamma         DHLM: 0             DIFF: 0
DINP:CONSTANT       DIR: Pos            DISA: 0             DISP: 0
DISS: NO_ALARM      DISV: 1             DLLM: 0             DLY: 0
DMOV: 1             DOL:CONSTANT        DPVT: 0xe81db0      DRBV: 51.5416973976
DSET: 0x7f095de16280                    DTYP: asynMotor     DVAL: 51.5416973976
EGU: deg            ERES: 2.86479e-05   EVNT: 0             FLNK:CONSTANT 0
FOF: 0              FOFF: Variable      FRAC: 1             HHSV: NO_ALARM
HIGH: 0             HIHI: 0             HLM: 0              HLS: 0
HLSV: NO_ALARM      HOMF: 0             HOMR: 0             HOPR: 0
HSV: NO_ALARM       HVEL: 0             ICOF: 0             INIT:
JAR: 0              JOGF: 0             JOGR: 0             JVEL: 0
LCNT: 0             LDVL: 51.5416973976 LLM: 0              LLS: 0
LLSV: NO_ALARM      LOCK: NO            LOLO: 0             LOPR: 0
LOW: 0              LRLV: 0             LRVL: 1799144       LSET: 0xe88220
LSPG: Go            LSV: NO_ALARM       LVAL: 51.5416973976 LVIO: 0
MDEL: 0             MIP: 0              MISS: 0
MLIS: 90 c0 06 54 09 7f 00 00 60 b0 ff 00 00 00 00 00 ad 01 00 00
MLOK: c0 68 df 00 00 00 00 00           MLST: 0             MMAP: 0
MOVN: 0             MRES: 2.86479e-05   MSTA: 18691
NAME: XF:11IDB-ES{Dif-Ax:Gam}Mtr        NMAP: 0             NSEV: NO_ALARM
NSTA: NO_ALARM      NTM: YES            NTMF: 2             OFF: 0
OMSL: supervisory   OUT:INST_IO @asyn(motor,2)              PACT: 0
PCOF: 0             PHAS: 0             PINI: NO            POST:
PP: 0               PPN: (nil)          PPNR: (nil)         PREC: 4
PREM:               PRIO: LOW           PROC: 0             PUTF: 0
RBV: 51.5416973976  RCNT: 0             RDBD: 2.86479e-05   RDBL:CONSTANT
RDES: 0xc0d9b0      RDIF: 0             REP: 1799144        RHLS: 0
RINP:CONSTANT       RLLS: 0             RLNK:CONSTANT       RLV: 0
RMOD: Default       RMP: 1799144        RPRO: 0             RRBV: 1799144
RRES: 0             RSET: 0x7f095de160a0                    RTRY: 0
RVAL: 1799144       RVEL: 0             S: 0                SBAK: 0
SBAS: 0             SCAN: Passive       SDIS:CONSTANT       SET: Use
SEVR: NO_ALARM      SMAX: 0             SPMG: Go            SPVT: (nil)
SREV: 200           SSET: 0             STAT: NO_ALARM      STOO:CONSTANT
STOP: 0             STUP: OFF           SUSE: 0             SYNC: 0
TDIR: 1             TIME: <undefined>   TPRO: 0             TSE: 0
TSEL:CONSTANT       TWF: 0              TWR: 0              TWV: 0
UDF: 0              UEIP: No            UREV: 0.00572958    URIP: No
VAL: 51.5416973976  VBAS: 0             VELO: 0             VERS: 6.81
VMAX: 0             VOF: 0
epics>
epics> dbpr XF:11IDB-ES{Dif-Ax:Del}Mtr 4
CARD: 0             CBAK: 0xe81900      CDIR: 1             CNEN: Disable
DCOF: 0             DESC: Delta         DHLM: 0
DIFF: -6.5474440000024e-04              DINP:CONSTANT       DIR: Pos
DISA: 0             DISP: 0             DISS: NO_ALARM      DISV: 1
DLLM: 0             DLY: 0              DMOV: 1             DOL:CONSTANT
DPVT: 0xe819a0      DRBV: 2.0006547444  DSET: 0x7f095de16280
DTYP: asynMotor     DVAL: 2             EGU: deg            ERES: 2.86479e-05
EVNT: 0             FLNK:CONSTANT 0     FOF: 0              FOFF: Variable
FRAC: 1             HHSV: NO_ALARM      HIGH: 0             HIHI: 0
HLM: 0              HLS: 0              HLSV: NO_ALARM      HOMF: 0
HOMR: 0             HOPR: 0             HSV: NO_ALARM       HVEL: 0
ICOF: 0             INIT:               JAR: 0              JOGF: 0
JOGR: 0             JVEL: 0             LCNT: 0             LDVL: 2
LLM: 0              LLS: 0              LLSV: NO_ALARM      LOCK: NO
LOLO: 0             LOPR: 0             LOW: 0              LRLV: 0
LRVL: 69813         LSET: 0xe88200      LSPG: Go            LSV: NO_ALARM
LVAL: 2             LVIO: 0             MDEL: 0             MIP: 0
MISS: 0
MLIS: 20 cc 06 54 09 7f 00 00 90 b7 ff 00 00 00 00 00 a7 01 00 00
MLOK: 50 68 df 00 00 00 00 00           MLST: 0             MMAP: 0
MOVN: 0             MRES: 2.86479e-05   MSTA: 2307
NAME: XF:11IDB-ES{Dif-Ax:Del}Mtr        NMAP: 0             NSEV: NO_ALARM
OMSL: supervisory   OUT:INST_IO @asyn(motor,1)              PACT: 0
RBV: 2.0006547444   RCNT: 0             RDBD: 2.86479e-05   RDBL:CONSTANT
RDES: 0xc0d9b0      RDIF: -23           REP: 69836          RHLS: 0
RMOD: Default       RMP: 69836          RPRO: 0             RRBV: 69836
RVAL: 69813         RVEL: 0             S: 0                SBAK: 0
TDIR: 1             TIME: 2018-05-22 13:45:45.833378801     TPRO: 0
TSE: 0              TSEL:CONSTANT       TWF: 0              TWR: 0
TWV: 0.02           UDF: 0              UEIP: No            UREV: 0.00572958
URIP: No            VAL: 2              VBAS: 0             VELO: 0
VERS: 6.81          VMAX: 0             VOF: 0
epics>
Re: :tdir messages in ioc for Pmac Wayne Lewis
:tdir messages in ioc for Pmac Gofron, Kazimierz
Prev: Re: printing to console in Display Builder Kasemir, Kay
Next: Re: :tdir messages in ioc for Pmac Wayne Lewis
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  <2018>  2019  2020  2021  2022  2023  2024  2025  2026
Prev: :tdir messages in ioc for Pmac Gofron, Kazimierz
ANJ, 22 May 2018 · Home · News · About · Talk · Base · Modules · Extensions ·