1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  <20192020  2021  2022  2023  2024  2025  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  <20192020  2021  2022  2023  2024  2025  2026
Subject: Re: Streamdevice segmentation fault
To: "Brown, Garth" <[email protected]>
Date: Sun, 20 Oct 2019 16:16:50 +0000
Hi Garth,
Could you write a simple sever in Python or C that sends that string to the asyn/Stream client? If you do that does it crash the same way? If so, then we can reproduce the problem and track it down.
Sent from my iPhone
On Oct 15, 2019, at 6:06 PM, Brown, Garth via Tech-talk <[email protected]<mailto:[email protected]>> wrote:
After updating an IOC from:
streamdevice 2.5, asyn 4.21, base 3.14.12
to:
streamdevice 2.8.9 (also tested 2.7.7), asyn 4.32 (also tested 4.31 and 4.35), base 7.0.2
I’m seeing 100% repeatable segfaults after processing the first response from a Hybricon VME crate.
I’ve attached the db and protocol files I’m using.
The text it is processing when it crashes is:
S/N Z834ÿÿÿÿ   U/N      Q:SN=Z834    ,UN=     ,IP=172.027.005.094,V3=3308,V5=5005,V+12=12009,V-12=12415,V24=00000,T1=29,T2=27,T3=28,T4=0,F1=07230,F2=07380,F3=07410,F4=07290,F5=07290,F6=07290,F7=00000,F8=00000,F9=00000,OT=0,OV=0,UV=000000,FF=0,PS1=1,PS2=1,MSG=0,SW=1,PROT=00,BS=0,HV=1,SD=1,WP=1,OS=000,CODE=10156F8F,ENET=D6.10,PSFAIL=00,POH=34848.0,MAXTMP=61,MINTMP=06
;EV0001000000000000ET00000000EF000000000OT0OV0UV000000FF0PS11MSG0SW1BS0HV1SD1WP1OS000
You can get a better idea of the raw ASCII in the attached console log with ASYN debug output, or in the streamdevice debug log (too huge to attach) at: https://gist.github.com/SLACer-garth/6505fd173dd71d7cfa2620d09e325276
One quirk that stands out to me is that the line terminator is \r\000\n. It would not surprise me if that null character in the middle causes parsing problems. I have tried changing the protocol file to use just \r or just \n, and it still crashed.
The call stack  at the point where it segfaults is:
Thread #7 252221 [core: 0] (Suspended : Signal : SIGSEGV:Segmentation fault)
                AsynDriverInterface::startTimer() at AsynDriverInterface.cc:228<http://asyndriverinterface.cc:228> 0x49f9b9
                AsynDriverInterface::readRequest() at AsynDriverInterface.cc:845<http://asyndriverinterface.cc:845> 0x49c769
                StreamBusInterface::Client::busReadRequest() at StreamBusInterface.h:82 0x4c0180
                StreamCore::evalIn() at StreamCore.cc:921<http://streamcore.cc:921> 0x4ba493
                StreamCore::readCallback() at StreamCore.cc<http://StreamCore.cc>:1,092 0x4bb48b
                StreamBusInterface::readCallback() at StreamBusInterface.h:118 0x49f5a8
                AsynDriverInterface::readHandler() at AsynDriverInterface.cc<http://AsynDriverInterface.cc>:1,064 0x49d793
                AsynDriverInterface::handleRequest() at AsynDriverInterface.cc<http://AsynDriverInterface.cc>:1,510 0x49ee87
                AsynDriverInterface::handleRequest() at AsynDriverInterface.cc:243<http://asyndriverinterface.cc:243> 0x49fa4b
                portThread() at asynManager.c:879 0x492439
                start_routine() at osdThread.c:403 0x5e1f9c
                start_thread() at 0x7ffff79a4aa1
                clone() at 0x7ffff6902c4d
The pointer “timer” on line 226 of AsynDriverInterface.c (this is code in the streamdevice module) has a value which, when decoded as ASCII, contains part of the message from the crate, “SW1BS0HV1SD1WP1O”. It looks likely that inputBuffer.local was written beyond its limit. Right after inputBuffer.local in memory is inputBuffer.len. If you look at that memory location, decoded as ASCII, it’s that last line from the crate data, “EV00010000000000”. The “timer” pointer is a few memory addresses past there. But I haven’t been able to identify a root cause, the point where a buffer overflows or whatever it is. Has anyone seen this before?
Garth
<crat_hybricon_8slot.template>
<crat_hybricon_8slot.proto.db>
<HybriconStreamFaultConsole.txt>
RE: Streamdevice segmentation fault Brown, Garth via Tech-talk
Prev: Re: Get IP of IOC from Phoebus Heinz Junkes via Tech-talk
Next: EPICS on your mobile with React-Automation-Studio William Duckitt via Tech-talk
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  <20192020  2021  2022  2023  2024  2025  2026
Prev: Streamdevice segmentation fault Brown, Garth via Tech-talk
Next: RE: Streamdevice segmentation fault Brown, Garth via Tech-talk
ANJ, 21 Oct 2019 · Home · News · About · Talk · Base · Modules · Extensions ·
