1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  <20202021  2022  2023  2024  2025  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  <20202021  2022  2023  2024  2025  2026
Subject: RE: pco Camera USB and IEEE interface
To: "'Ivashkevych, Oksana'" <[email protected]>, "'Sintschuk, Michael'" <[email protected]>
Cc: 'tech-talk' <[email protected]>
Date: Mon, 11 May 2020 18:33:29 +0000
Ø  dbLoadRecords("C:/epics/inst-7.0.3.1/areaDetector-R3-9/ADCore/db/NDStdArrays.template", "P=PCO4000:,R=image1:,PORT=Image1, 1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PCO1,TYPE=Int16,FTVL=SHORT,NELEMENTS=11059200")
Ø  I looked at the datasheet and it says that the pixel depth is 14 bit for this camera, which requires 2 bytes.
Ø  4008x2672x2=21,418,752 and you allocated 11,059,200.
Ø  Try changing to NELEMENTS=21418752
Michael’s dbLoadRecords is actually fine.  Oksana has computed the size in bytes, but NELEMENTS is not bytes it is the number of elements in the waveform record, and in this case the FTVL is SHORT, so each element is 2 bytes.
Michael has actually allocated somewhat more elements than required, since  4008*2672 = 10709376, which is fewer than he allocated.
But more importantly Michael’s error is from the PCO driver, which knows nothing about the size of the waveform record being used by NDPluginStdArrays.  This problem is in the driver, not the plugins.
Sent: Monday, May 11, 2020 12:29 PM
To: Sintschuk, Michael <[email protected]>; Mark Rivers <[email protected]>
Cc: Bradnick, Ben (Tessella,RAL,TEC) <[email protected]>; tech-talk <[email protected]>
Hi Michael,
I did not see this error, but it might be related to the buffer you allocate for you image. It is possible the driver checks unsatisfactory size and reports the error.
You are using CA protocol Image.
dbLoadRecords("C:/epics/inst-7.0.3.1/areaDetector-R3-9/ADCore/db/NDStdArrays.template", "P=PCO4000:,R=image1:,PORT=Image1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PCO1,TYPE=Int16,FTVL=SHORT,NELEMENTS=11059200")
with X and Y :
epicsEnvSet("XSIZE",  "4008")
epicsEnvSet("YSIZE",  "2672")
And see it this fixes the problem.
With best regards,
Oksana
Sent: Monday, May 11, 2020 8:50 AM
Cc: Bradnick, Ben (Tessella,RAL,TEC) <[email protected]>; Ivashkevych, Oksana <[email protected]>; tech-talk <[email protected]>
Subject: AW: pco Camera USB and IEEE interface
Hello again,
I‘ve been working some weeks now with the pco1600-camera connected via GigE. It works fine with the ADPcoWin-IOC. Great software, thanks to the epics-community!
However, I also try to set up a pco4000-camera connected via Firewire. The IOC starts and finds the camera but everytime I try to acquire an image there is the error:
Failed to arm, PCO error, allocateBuffer returned a00a3006
I guess, there are some settings that I’ve missed…
The camera connects fine with the original camware from pco.
I already tried to set the “QSIZE” , “CBUFFS”, “XSIZE” or “YSIZE” values to different parameters, but with no look. Maybe someone already had this issue?
This is my IOC-startup:
C:\epics\inst-7.0.3.1\areaDetector-R3-9\ADPcoWin\iocs\pcowinIOC\iocBoot\iocPco4000Win> C:\epics\inst-7.0.3.1\areaDete
ctor-R3-9\ADPcoWin\iocs\pcowinIOC\bin\windows-x64-static\pcowinApp.exe st.cmd
< envPaths
epicsEnvSet("IOC","iocPcoWin")
epicsEnvSet("TOP","C:/epics/inst-7.0.3.1/areaDetector-R3-9/ADPcoWin/iocs/pcowinIOC")
epicsEnvSet("ADPCOWIN","C:/epics/inst-7.0.3.1/areaDetector-R3-9/ADPcoWin/iocs/pcowinIOC/../..")
epicsEnvSet("SUPPORT","C:/epics/inst-7.0.3.1")
epicsEnvSet("ASYN","C:/epics/inst-7.0.3.1/asyn-R4-39")
epicsEnvSet("AREA_DETECTOR","C:/epics/inst-7.0.3.1/areaDetector-R3-9")
epicsEnvSet("ADSUPPORT","C:/epics/inst-7.0.3.1/areaDetector-R3-9/ADSupport")
epicsEnvSet("ADCORE","C:/epics/inst-7.0.3.1/areaDetector-R3-9/ADCore")
epicsEnvSet("AUTOSAVE","C:/epics/inst-7.0.3.1/autosave-R5-10")
epicsEnvSet("BUSY","C:/epics/inst-7.0.3.1/busy-R1-7-2")
epicsEnvSet("CALC","C:/epics/inst-7.0.3.1/calc-R3-7-3")
epicsEnvSet("SSCAN","C:/epics/inst-7.0.3.1/sscan-R2-11-3")
epicsEnvSet("EPICS_BASE","C:/epics/base-7.0.3.1")
errlogInit(20000)
dbLoadDatabase("C:/epics/inst-7.0.3.1/areaDetector-R3-9/ADPcoWin/iocs/pcowinIOC/dbd/pcowinApp.dbd")
pcowinApp_registerRecordDeviceDriver(pdbbase)
# Prefix for all records
epicsEnvSet("PREFIX", "PCO4000:")
# The port name for the detector
epicsEnvSet("PORT",   "PCO1")
# Larger queue size may be need to stream to disk at full camera speed, up tp 2000
epicsEnvSet("QSIZE",  "20")
# The maximim image width; used for row profiles in the NDPluginStats plugin
# The maximim image height; used for column profiles in the NDPluginStats plugin
# The maximum number of time series points in the NDPluginStats plugin
epicsEnvSet("NCHANS", "2048")
# The maximum number of frames buffered in the NDPluginCircularBuff plugin
epicsEnvSet("CBUFFS", "500")
# The search path for database files
epicsEnvSet("EPICS_DB_INCLUDE_PATH", "C:/epics/inst-7.0.3.1/areaDetector-R3-9/ADCore/db")
# Define NELEMENTS to be enough for a 2048x2048x3 (color) image
epicsEnvSet("NELEMENTS", "11059200")
# pcoConfig(const char* portName, int maxBuffers, size_t maxMemory)
pcoConfig("PCO1", 0, 0, 8)
# pcoApiConfig(const char* portName)
pcoApiConfig("PCO1")
# Asyn tracing
asynSetTraceIOMask(PCO1, 0, 2)
#asynSetTraceMask($(PORT), 0, 0xFF)
#asynSetTraceFile($(PORT), 0, "asynTrace.out")
#asynSetTraceInfoMask($(PORT), 0, 0xf)
dbLoadRecords("C:/epics/inst-7.0.3.1/areaDetector-R3-9/ADPcoWin/iocs/pcowinIOC/../../db/pco.template", "P=PCO4000:,R=cam
1:,PORT=PCO1,ADDR=0,TIMEOUT=1")
dbLoadRecords("C:/epics/inst-7.0.3.1/areaDetector-R3-9/ADPcoWin/iocs/pcowinIOC/../../db/pco_device_firmware.template", "
P=PCO4000:,R=cam1:,PORT=PCO1,N=0")
P=PCO4000:,R=cam1:,PORT=PCO1,N=1")
P=PCO4000:,R=cam1:,PORT=PCO1,N=2")
P=PCO4000:,R=cam1:,PORT=PCO1,N=3")
P=PCO4000:,R=cam1:,PORT=PCO1,N=4")
P=PCO4000:,R=cam1:,PORT=PCO1,N=5")
P=PCO4000:,R=cam1:,PORT=PCO1,N=6")
P=PCO4000:,R=cam1:,PORT=PCO1,N=7")
# What other database files need to be loaded?
# Create a standard arrays plugin
NDStdArraysConfigure("Image1", 5, 0, "PCO1", 0, 0)
# Use this line for 8-bit data only
#dbLoadRecords("$(ADCORE)/db/NDStdArrays.template", "P=$(PREFIX),R=image1:,PORT=Image1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=$(P
ORT),TYPE=Int8,FTVL=CHAR,NELEMENTS=$(NELEMENTS)")
# Use this line for 8-bit or 16-bit data
# Use this line for 8-bit, 16-bit, or 32-bit data
ORT),TYPE=Int32,FTVL=LONG,NELEMENTS=$(NELEMENTS)")
# Load all other plugins using commonPlugins.cmd
< C:/epics/inst-7.0.3.1/areaDetector-R3-9/ADCore/iocBoot/commonPlugins.cmd
# This is an example file for creating plugins
# It uses the following environment variable macros
# Many of the parameters defined in this file are also in commonPlugins_settings.req so if autosave is being
# use the autosave value will replace the value passed to this file.
# $(PREFIX)      Prefix for all records
# $(PORT)        The port name for the detector.  In autosave.
# $(QSIZE)       The queue size for all plugins.  In autosave.
# $(XSIZE)       The maximum image width; used to set the maximum size for row profiles in the NDPluginStats plugin and
1-D FFT
#                   profiles in NDPluginFFT.
# $(YSIZE)       The maximum image height; used to set the maximum size for column profiles in the NDPluginStats plugin
# $(NCHANS)      The maximum number of time series points in the NDPluginStats, NDPluginROIStats, and NDPluginAttribute
plugins
# $(CBUFFS)      The maximum number of frames buffered in the NDPluginCircularBuff plugin
# $(MAX_THREADS) The maximum number of threads for plugins which can run in multiple threads. Defaults to 5.
# Create a netCDF file saving plugin
NDFileNetCDFConfigure("FileNetCDF1", 20, 0, "PCO1", 0)
dbLoadRecords("NDFileNetCDF.template","P=PCO4000:,R=netCDF1:,PORT=FileNetCDF1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PCO1")
# Create a TIFF file saving plugin
NDFileTIFFConfigure("FileTIFF1", 20, 0, "PCO1", 0)
dbLoadRecords("NDFileTIFF.template",  "P=PCO4000:,R=TIFF1:,PORT=FileTIFF1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PCO1")
# Create a JPEG file saving plugin
NDFileJPEGConfigure("FileJPEG1", 20, 0, "PCO1", 0)
dbLoadRecords("NDFileJPEG.template",  "P=PCO4000:,R=JPEG1:,PORT=FileJPEG1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PCO1")
# Create a NeXus file saving plugin
NDFileNexusConfigure("FileNexus1", 20, 0, "PCO1", 0)
dbLoadRecords("NDFileNexus.template", "P=PCO4000:,R=Nexus1:,PORT=FileNexus1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PCO1")
# Create an HDF5 file saving plugin
NDFileHDF5Configure("FileHDF1", 20, 0, "PCO1", 0)
dbLoadRecords("NDFileHDF5.template",  "P=PCO4000:,R=HDF1:,PORT=FileHDF1,ADDR=0,TIMEOUT=1,XMLSIZE=2048,NDARRAY_PORT=PCO1"
# Create a Magick file saving plugin
#NDFileMagickConfigure("FileMagick1", $(QSIZE), 0, "$(PORT)", 0)
#dbLoadRecords("NDFileMagick.template","P=$(PREFIX),R=Magick1:,PORT=FileMagick1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=$(PORT)")
# Create 4 ROI plugins
NDROIConfigure("ROI1", 20, 0, "PCO1", 0, 0, 0, 0, 0, 5)
dbLoadRecords("NDROI.template",       "P=PCO4000:,R=ROI1:,  PORT=ROI1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PCO1")
NDROIConfigure("ROI2", 20, 0, "PCO1", 0, 0, 0, 0, 0, 5)
dbLoadRecords("NDROI.template",       "P=PCO4000:,R=ROI2:,  PORT=ROI2,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PCO1")
NDROIConfigure("ROI3", 20, 0, "PCO1", 0, 0, 0, 0, 0, 5)
dbLoadRecords("NDROI.template",       "P=PCO4000:,R=ROI3:,  PORT=ROI3,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PCO1")
NDROIConfigure("ROI4", 20, 0, "PCO1", 0, 0, 0, 0, 0, 5)
dbLoadRecords("NDROI.template",       "P=PCO4000:,R=ROI4:,  PORT=ROI4,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PCO1")
# Create 8 ROIStat plugins
NDROIStatConfigure("ROISTAT1", 20, 0, "PCO1", 0, 8, 0, 0, 0, 0, 5)
dbLoadRecords("NDROIStat.template",   "P=PCO4000:,R=ROIStat1:  ,PORT=ROISTAT1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PCO1,NCHANS=
2048")
dbLoadRecords("NDROIStatN.template",  "P=PCO4000:,R=ROIStat1:1:,PORT=ROISTAT1,ADDR=0,TIMEOUT=1,NCHANS=2048")
dbLoadRecords("NDROIStatN.template",  "P=PCO4000:,R=ROIStat1:2:,PORT=ROISTAT1,ADDR=1,TIMEOUT=1,NCHANS=2048")
dbLoadRecords("NDROIStatN.template",  "P=PCO4000:,R=ROIStat1:3:,PORT=ROISTAT1,ADDR=2,TIMEOUT=1,NCHANS=2048")
dbLoadRecords("NDROIStatN.template",  "P=PCO4000:,R=ROIStat1:4:,PORT=ROISTAT1,ADDR=3,TIMEOUT=1,NCHANS=2048")
dbLoadRecords("NDROIStatN.template",  "P=PCO4000:,R=ROIStat1:5:,PORT=ROISTAT1,ADDR=4,TIMEOUT=1,NCHANS=2048")
dbLoadRecords("NDROIStatN.template",  "P=PCO4000:,R=ROIStat1:6:,PORT=ROISTAT1,ADDR=5,TIMEOUT=1,NCHANS=2048")
dbLoadRecords("NDROIStatN.template",  "P=PCO4000:,R=ROIStat1:7:,PORT=ROISTAT1,ADDR=6,TIMEOUT=1,NCHANS=2048")
dbLoadRecords("NDROIStatN.template",  "P=PCO4000:,R=ROIStat1:8:,PORT=ROISTAT1,ADDR=7,TIMEOUT=1,NCHANS=2048")
# Create a processing plugin
NDProcessConfigure("PROC1", 20, 0, "PCO1", 0, 0, 0)
dbLoadRecords("NDProcess.template",   "P=PCO4000:,R=Proc1:,  PORT=PROC1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PCO1")
# Create a TIFF file plugin to read dark and flatfield images into the processing plugin
NDFileTIFFConfigure("PROC1TIFF", 20, 0, "PCO1", 0)
dbLoadRecords("NDFileTIFF.template",  "P=PCO4000:,R=Proc1:TIFF:,PORT=PROC1TIFF,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PCO1")
# Create a scatter plugin
NDScatterConfigure("SCATTER1", 20, 0, "PCO1", 0, 0, 0)
dbLoadRecords("NDScatter.template",   "P=PCO4000:,R=Scatter1:,  PORT=SCATTER1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PCO1")
# Create a gather plugin with 8 ports
NDGatherConfigure("GATHER1", 20, 0, 8, 0, 0)
dbLoadRecords("NDGather.template",   "P=PCO4000:,R=Gather1:, PORT=GATHER1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PCO1")
dbLoadRecords("NDGatherN.template",   "P=PCO4000:,R=Gather1:, N=1, PORT=GATHER1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PCO1")
dbLoadRecords("NDGatherN.template",   "P=PCO4000:,R=Gather1:, N=2, PORT=GATHER1,ADDR=1,TIMEOUT=1,NDARRAY_PORT=PCO1")
dbLoadRecords("NDGatherN.template",   "P=PCO4000:,R=Gather1:, N=3, PORT=GATHER1,ADDR=2,TIMEOUT=1,NDARRAY_PORT=PCO1")
dbLoadRecords("NDGatherN.template",   "P=PCO4000:,R=Gather1:, N=4, PORT=GATHER1,ADDR=3,TIMEOUT=1,NDARRAY_PORT=PCO1")
dbLoadRecords("NDGatherN.template",   "P=PCO4000:,R=Gather1:, N=5, PORT=GATHER1,ADDR=4,TIMEOUT=1,NDARRAY_PORT=PCO1")
dbLoadRecords("NDGatherN.template",   "P=PCO4000:,R=Gather1:, N=6, PORT=GATHER1,ADDR=5,TIMEOUT=1,NDARRAY_PORT=PCO1")
dbLoadRecords("NDGatherN.template",   "P=PCO4000:,R=Gather1:, N=7, PORT=GATHER1,ADDR=6,TIMEOUT=1,NDARRAY_PORT=PCO1")
dbLoadRecords("NDGatherN.template",   "P=PCO4000:,R=Gather1:, N=8, PORT=GATHER1,ADDR=7,TIMEOUT=1,NDARRAY_PORT=PCO1")
# Create 5 statistics plugins
NDStatsConfigure("STATS1", 20, 0, "PCO1", 0, 0, 0, 0, 0, 5)
dbLoadRecords("NDStats.template",     "P=PCO4000:,R=Stats1:,  PORT=STATS1,ADDR=0,TIMEOUT=1,HIST_SIZE=256,XSIZE=4008,YSIZ
E=2672,NCHANS=2048,NDARRAY_PORT=PCO1")
NDTimeSeriesConfigure("STATS1_TS", 20, 0, "STATS1", 1, 23)
dbLoadRecords("C:/epics/inst-7.0.3.1/areaDetector-R3-9/ADCore/db/NDTimeSeries.template",  "P=PCO4000:,R=Stats1:TS:, PORT
=STATS1_TS,ADDR=0,TIMEOUT=1,NDARRAY_PORT=STATS1,NDARRAY_ADDR=1,NCHANS=2048,ENABLED=1")
NDStatsConfigure("STATS2", 20, 0, "ROI1",    0, 0, 0, 0, 0, 5)
dbLoadRecords("NDStats.template",     "P=PCO4000:,R=Stats2:,  PORT=STATS2,ADDR=0,TIMEOUT=1,HIST_SIZE=256,XSIZE=4008,YSIZ
NDTimeSeriesConfigure("STATS2_TS", 20, 0, "STATS2", 1, 23)
dbLoadRecords("C:/epics/inst-7.0.3.1/areaDetector-R3-9/ADCore/db/NDTimeSeries.template",  "P=PCO4000:,R=Stats2:TS:, PORT
=STATS2_TS,ADDR=0,TIMEOUT=1,NDARRAY_PORT=STATS2,NDARRAY_ADDR=1,NCHANS=2048,ENABLED=1")
NDStatsConfigure("STATS3", 20, 0, "ROI2",    0, 0, 0, 0, 0, 5)
dbLoadRecords("NDStats.template",     "P=PCO4000:,R=Stats3:,  PORT=STATS3,ADDR=0,TIMEOUT=1,HIST_SIZE=256,XSIZE=4008,YSIZ
NDTimeSeriesConfigure("STATS3_TS", 20, 0, "STATS3", 1, 23)
dbLoadRecords("C:/epics/inst-7.0.3.1/areaDetector-R3-9/ADCore/db/NDTimeSeries.template",  "P=PCO4000:,R=Stats3:TS:, PORT
=STATS3_TS,ADDR=0,TIMEOUT=1,NDARRAY_PORT=STATS3,NDARRAY_ADDR=1,NCHANS=2048,ENABLED=1")
NDStatsConfigure("STATS4", 20, 0, "ROI3",    0, 0, 0, 0, 0, 5)
dbLoadRecords("NDStats.template",     "P=PCO4000:,R=Stats4:,  PORT=STATS4,ADDR=0,TIMEOUT=1,HIST_SIZE=256,XSIZE=4008,YSIZ
NDTimeSeriesConfigure("STATS4_TS", 20, 0, "STATS4", 1, 23)
dbLoadRecords("C:/epics/inst-7.0.3.1/areaDetector-R3-9/ADCore/db/NDTimeSeries.template",  "P=PCO4000:,R=Stats4:TS:, PORT
=STATS4_TS,ADDR=0,TIMEOUT=1,NDARRAY_PORT=STATS4,NDARRAY_ADDR=1,NCHANS=2048,ENABLED=1")
NDStatsConfigure("STATS5", 20, 0, "ROI4",    0, 0, 0, 0, 0, 5)
dbLoadRecords("NDStats.template",     "P=PCO4000:,R=Stats5:,  PORT=STATS5,ADDR=0,TIMEOUT=1,HIST_SIZE=256,XSIZE=4008,YSIZ
NDTimeSeriesConfigure("STATS5_TS", 20, 0, "STATS5", 1, 23)
dbLoadRecords("C:/epics/inst-7.0.3.1/areaDetector-R3-9/ADCore/db/NDTimeSeries.template",  "P=PCO4000:,R=Stats5:TS:, PORT
=STATS5_TS,ADDR=0,TIMEOUT=1,NDARRAY_PORT=STATS5,NDARRAY_ADDR=1,NCHANS=2048,ENABLED=1")
# Create a transform plugin
NDTransformConfigure("TRANS1", 20, 0, "PCO1", 0, 0, 0, 0, 0, 5)
dbLoadRecords("NDTransform.template", "P=PCO4000:,R=Trans1:,  PORT=TRANS1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PCO1")
# Create an overlay plugin with 8 overlays
NDOverlayConfigure("OVER1", 20, 0, "PCO1", 0, 8, 0, 0, 0, 0, 5)
dbLoadRecords("NDOverlay.template", "P=PCO4000:,R=Over1:, PORT=OVER1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PCO1")
dbLoadRecords("NDOverlayN.template","P=PCO4000:,R=Over1:1:,NAME=ROI1,   SHAPE=1,O=Over1:,XPOS=PCO4000:ROI1:MinX_RBV,YPOS
=PCO4000:ROI1:MinY_RBV,XSIZE=PCO4000:ROI1:SizeX_RBV,YSIZE=PCO4000:ROI1:SizeY_RBV,PORT=OVER1,ADDR=0,TIMEOUT=1")
dbLoadRecords("NDOverlayN.template","P=PCO4000:,R=Over1:2:,NAME=ROI2,   SHAPE=1,O=Over1:,XPOS=PCO4000:ROI2:MinX_RBV,YPOS
=PCO4000:ROI2:MinY_RBV,XSIZE=PCO4000:ROI2:SizeX_RBV,YSIZE=PCO4000:ROI2:SizeY_RBV,PORT=OVER1,ADDR=1,TIMEOUT=1")
dbLoadRecords("NDOverlayN.template","P=PCO4000:,R=Over1:3:,NAME=ROI3,   SHAPE=1,O=Over1:,XPOS=PCO4000:ROI3:MinX_RBV,YPOS
=PCO4000:ROI3:MinY_RBV,XSIZE=PCO4000:ROI3:SizeX_RBV,YSIZE=PCO4000:ROI3:SizeY_RBV,PORT=OVER1,ADDR=2,TIMEOUT=1")
dbLoadRecords("NDOverlayN.template","P=PCO4000:,R=Over1:4:,NAME=ROI4,   SHAPE=1,O=Over1:,XPOS=PCO4000:ROI4:MinX_RBV,YPOS
=PCO4000:ROI4:MinY_RBV,XSIZE=PCO4000:ROI4:SizeX_RBV,YSIZE=PCO4000:ROI4:SizeY_RBV,PORT=OVER1,ADDR=3,TIMEOUT=1")
dbLoadRecords("NDOverlayN.template","P=PCO4000:,R=Over1:5:,NAME=Cursor1,SHAPE=1,O=Over1:,XPOS=junk,                  YPO
S=junk,                  XSIZE=junk,                   YSIZE=junk,                   PORT=OVER1,ADDR=4,TIMEOUT=1")
dbLoadRecords("NDOverlayN.template","P=PCO4000:,R=Over1:6:,NAME=Cursor2,SHAPE=1,O=Over1:,XPOS=junk,                  YPO
S=junk,                  XSIZE=junk,                   YSIZE=junk,                   PORT=OVER1,ADDR=5,TIMEOUT=1")
dbLoadRecords("NDOverlayN.template","P=PCO4000:,R=Over1:7:,NAME=Box1,   SHAPE=1,O=Over1:,XPOS=junk,                  YPO
S=junk,                  XSIZE=junk,                   YSIZE=junk,                   PORT=OVER1,ADDR=6,TIMEOUT=1")
dbLoadRecords("NDOverlayN.template","P=PCO4000:,R=Over1:8:,NAME=Box2,   SHAPE=1,O=Over1:,XPOS=junk,                  YPO
S=junk,                  XSIZE=junk,                   YSIZE=junk,                   PORT=OVER1,ADDR=7,TIMEOUT=1")
# Create 2 color conversion plugins
NDColorConvertConfigure("CC1", 20, 0, "PCO1", 0, 0, 0, 0, 0, 5)
dbLoadRecords("NDColorConvert.template", "P=PCO4000:,R=CC1:,  PORT=CC1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PCO1")
NDColorConvertConfigure("CC2", 20, 0, "PCO1", 0, 0, 0, 0, 0, 5)
dbLoadRecords("NDColorConvert.template", "P=PCO4000:,R=CC2:,  PORT=CC2,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PCO1")
# Create a circular buffer plugin
NDCircularBuffConfigure("CB1", 20, 0, "PCO1", 0, 500, 0)
dbLoadRecords("NDCircularBuff.template", "P=PCO4000:,R=CB1:,  PORT=CB1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=PCO1")
# Create an NDAttribute plugin with 8 attributes
NDAttrConfigure("ATTR1", 20, 0, "PCO1", 0, 8, 0, 0, 0)
dbLoadRecords("NDAttribute.template",  "P=PCO4000:,R=Attr1:,    PORT=ATTR1,ADDR=0,TIMEOUT=1,NCHANS=2048,NDARRAY_PORT=PCO
1")
dbLoadRecords("NDAttributeN.template", "P=PCO4000:,R=Attr1:1:,  PORT=ATTR1,ADDR=0,TIMEOUT=1,NCHANS=2048")
dbLoadRecords("NDAttributeN.template", "P=PCO4000:,R=Attr1:2:,  PORT=ATTR1,ADDR=1,TIMEOUT=1,NCHANS=2048")
dbLoadRecords("NDAttributeN.template", "P=PCO4000:,R=Attr1:3:,  PORT=ATTR1,ADDR=2,TIMEOUT=1,NCHANS=2048")
dbLoadRecords("NDAttributeN.template", "P=PCO4000:,R=Attr1:4:,  PORT=ATTR1,ADDR=3,TIMEOUT=1,NCHANS=2048")
dbLoadRecords("NDAttributeN.template", "P=PCO4000:,R=Attr1:5:,  PORT=ATTR1,ADDR=4,TIMEOUT=1,NCHANS=2048")
dbLoadRecords("NDAttributeN.template", "P=PCO4000:,R=Attr1:6:,  PORT=ATTR1,ADDR=5,TIMEOUT=1,NCHANS=2048")
dbLoadRecords("NDAttributeN.template", "P=PCO4000:,R=Attr1:7:,  PORT=ATTR1,ADDR=6,TIMEOUT=1,NCHANS=2048")
dbLoadRecords("NDAttributeN.template", "P=PCO4000:,R=Attr1:8:,  PORT=ATTR1,ADDR=7,TIMEOUT=1,NCHANS=2048")
NDTimeSeriesConfigure("ATTR1_TS", 20, 0, "ATTR1", 1, 8)
dbLoadRecords("C:/epics/inst-7.0.3.1/areaDetector-R3-9/ADCore/db/NDTimeSeries.template",  "P=PCO4000:,R=Attr1:TS:, PORT=
ATTR1_TS,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ATTR1,NDARRAY_ADDR=1,NCHANS=2048,ENABLED=1")
# Create an FFT plugin
NDFFTConfigure("FFT1", 20, 0, "PCO1", 0, 0, 0, 0, 0, 5)
dbLoadRecords("NDFFT.template", "P=PCO4000:, R=FFT1:, PORT=FFT1, ADDR=0, TIMEOUT=1, NDARRAY_PORT=PCO1, NAME=FFT1, NCHANS
=4008")
# Create 2 Codec plugins
NDCodecConfigure("CODEC1", 20, 0, "PCO1", 0, 0, 0, 0, 0, 5)
dbLoadRecords("NDCodec.template", "P=PCO4000:, R=Codec1:, PORT=CODEC1, ADDR=0, TIMEOUT=1, NDARRAY_PORT=PCO1")
NDCodecConfigure("CODEC2", 20, 0, "PCO1", 0, 0, 0, 0, 0, 5)
dbLoadRecords("NDCodec.template", "P=PCO4000:, R=Codec2:, PORT=CODEC2, ADDR=0, TIMEOUT=1, NDARRAY_PORT=PCO1")
set_requestfile_path("./")
set_requestfile_path("C:/epics/inst-7.0.3.1/areaDetector-R3-9/ADCore/ADApp/Db")
set_requestfile_path("C:/epics/inst-7.0.3.1/areaDetector-R3-9/ADCore/iocBoot")
set_savefile_path("./autosave")
set_pass0_restoreFile("auto_settings.sav")
set_pass1_restoreFile("auto_settings.sav")
save_restoreSet_status_prefix("PCO4000:")
dbLoadRecords("C:/epics/inst-7.0.3.1/autosave-R5-10/asApp/Db/save_restoreStatus.db", "P=PCO4000:")
# Optional: load NDPluginPva plugin
#NDPvaConfigure("PVA1", $(QSIZE), 0, "$(PORT)", 0, $(PREFIX)Pva1:Image, 0, 0, 0)
#dbLoadRecords("NDPva.template",  "P=$(PREFIX),R=Pva1:, PORT=PVA1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=$(PORT)")
# Must start PVA server if this is enabled
#startPVAServer
# Optional: load ffmpegServer plugin
#ffmpegServerConfigure(8081)
#ffmpegStreamConfigure("FfmStream1", 2, 0, "$(PORT)", 0, -1, 0)
#dbLoadRecords("$(FFMPEGSERVER)/db/ffmpegStream.template", "P=$(PREFIX),R=ffmstream1:,PORT=FfmStream1,NDARRAY_PORT=$(POR
T)")
#ffmpegFileConfigure("FfmFile1", 16, 0, "$(PORT)", 0, -1, 0)
#dbLoadRecords("$(FFMPEGSERVER)/db/ffmpegFile.template", "P=$(PREFIX),R=ffmfile1:,PORT=FfmFile1,NDARRAY_PORT=$(PORT)")
# Optional: load NDPluginEdge plugin
#NDEdgeConfigure("EDGE1", $(QSIZE), 0, "$(PORT)", 0, 0, 0, 0)
#dbLoadRecords("$(ADPLUGINEDGE)/db/NDEdge.template",  "P=$(PREFIX),R=Edge1:, PORT=EDGE1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=$(
PORT)")
#set_requestfile_path("$(ADPLUGINEDGE)/edgeApp/Db")
# Optional: load NDPluginCV plugin
#NDCVConfigure("CV1", $(QSIZE), 0, "$(PORT)", 0, 0, 0, 0, 0, $(MAX_THREADS=5))
#dbLoadRecords("$(ADCOMPVISION)/db/NDCV.template",  "P=$(PREFIX),R=CV1:, PORT=CV1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=$(PORT)"
#set_requestfile_path("$(ADCOMPVISION)/adcvApp/Db")
# Optional: load NDPluginBar plugin
#NDBarConfigure("BAR1", $(QSIZE), 0, "$(PORT)", 0, 0, 0, 0, 0, $(MAX_THREADS=5))
#dbLoadRecords("$(ADPLUGINBAR)/db/NDBar.template",  "P=$(PREFIX),R=Bar1:, PORT=BAR1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=$(PORT
#set_requestfile_path("$(ADPLUGINBAR)/barApp/Db")
# Optional: load scan records
dbLoadRecords("C:/epics/inst-7.0.3.1/sscan-R2-11-3/sscanApp/Db/scan.db", "P=PCO4000:,MAXPTS1=2000,MAXPTS2=200,MAXPTS3=20
,MAXPTS4=10,MAXPTSH=10")
set_requestfile_path("C:/epics/inst-7.0.3.1/sscan-R2-11-3/sscanApp/Db")
# Optional: load sseq record for acquisition sequence
dbLoadRecords("C:/epics/inst-7.0.3.1/calc-R3-7-3/calcApp/Db/sseqRecord.db", "P=PCO4000:, S=AcquireSequence")
set_requestfile_path("C:/epics/inst-7.0.3.1/calc-R3-7-3/calcApp/Db")
# Optional: load devIocStats records (requires DEVIOCSTATS module)
#dbLoadRecords("$(DEVIOCSTATS)/db/iocAdminSoft.db", "IOC=$(PREFIX)")
# Optional: load alive record (requires ALIVE module)
#dbLoadRecords("$(ALIVE)/aliveApp/Db/alive.db", "P=$(PREFIX),RHOST=192.168.1.254")
# Set the callback queue size to 5000, up from default of 2000 in base.
# This can be needed to avoid errors "callbackRequest: cbLow ring buffer full".
callbackSetQueueSize(5000)
set_requestfile_path("C:/epics/inst-7.0.3.1/areaDetector-R3-9/ADPcoWin/iocs/pcowinIOC/../../pcowinApp/Db")
iocInit()
Starting iocInit
## EPICS R7.0.3.1
## EPICS Base built Mar  9 2020
reboot_restore: entry for file 'auto_settings.sav'
reboot_restore: Found filename 'auto_settings.sav' in restoreFileList.
*** restoring from './autosave/auto_settings.sav' at initHookState 6 (before record/device init) ***
reboot_restore: done with file 'auto_settings.sav'
recDynLinkOut: waiting for CA context
recDynLinkOut: got CA context
*** restoring from './autosave/auto_settings.sav' at initHookState 7 (after record/device init) ***
# save things every thirty seconds
create_monitor_set("auto_settings.req", 30,"P=PCO4000:")
epics> auto_settings.sav: 2277 of 2277 PV's connected
pco.4000
                                        : Found 8 devices on camera
epics> 2020/05/11 13:53:06.730 Failed to arm, PCO error, allocateBuffer returned a00a3006
Again, thanks for the help!
AW: pco Camera USB and IEEE interface Sintschuk, Michael via Tech-talk
RE: pco Camera USB and IEEE interface Mark Rivers via Tech-talk
Re: pco Camera USB and IEEE interface Mark Rivers via Tech-talk
Re: pco Camera USB and IEEE interface Ivashkevych, Oksana via Tech-talk
Re: pco Camera USB and IEEE interface Bradnick, Ben (Tessella, RAL, TEC) via Tech-talk
Prev: Re: Using Jupyter notebooks for controlling experiments Jemian, Pete R. via Tech-talk
Next: AW: pco Camera USB and IEEE interface Sintschuk, Michael via Tech-talk
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  <20202021  2022  2023  2024  2025  2026
Prev: AW: pco Camera USB and IEEE interface Sintschuk, Michael via Tech-talk