Subject: RE: CAEN R1470ET HVPS - Asyn + Stream Device auto connect does not recover communication
To: "'Paduan Donadio, Marcio'" <[email protected]>
Date: Tue, 1 Sep 2020 22:34:16 +0000
Hi Paduan,
Ø  It looks like the equipment stops answering but Asyn still shows that the connection is opened and when it writes to the port the result is Asyn success.
If the device crashes or otherwise becomes unresponsive without properly closing the socket it will take a long time before the client (the drvAsynTCP driver) gets informed of this.  The messages it sends just queue up in the TCP stack on either the client or the server.  This just the way that TCP works.  That is why disconnectOnReadTimeout was added to asyn, and in general it is a good thing to use.  That way whenever the device reply times out asyn will close the socket and, if autoconnect is enabled, try to reopen the connection to the device.
Ø  If we close and open the port again, communication is reestablished.
What do you mean by that? How did you do it?
Ø  I’ve tried disconnectOnReadTimeout. Asyn disconnected after it, but then neither auto-connect or manual connection were able to open the connection again. The system is still in this condition as we were not authorized to reboot the IOC, yet.
I don’t understand why the disconnectOnReadTimeout disconnection would be any different from whatever you did when you said you “close and open the port again”.
That sounds like a truly dead device.  Can you ping the CAEN device? Can you open a manual telnet connection to it on port 1470, which seems to be the port you are using?  Those are both things you can try without restarting the IOC.
Do you believe that restarting the IOC will fix the problem, or do you think you will need to power-cycle the CAEN?
I don’t understand this sequence:
2020/08/28 14:15:55.193 hvm-ltuh-mc01:1470 write 29
$BD:0,CMD:MON,CH:5,PAR:VSET
2020/08/28 14:15:55.289 hvm-ltuh-mc01:1470 read 26
#BD:00,CMD:OK,VAL:3000.0
2020/08/28 14:16:00.266 HVM:LTUH:200:ASYN: exception 3, asynExceptionTraceMask
2020/08/28 14:16:00.283 HV0 set Eos 0
2020/08/28 14:16:00.323 asynManager::portThread port=HV0 callback
2020/08/28 14:16:00.393 HV0 addr -1 queueRequest priority 0 not lockHolder
2020/08/28 14:16:00.472 HV0 schedule queueRequest timeout
2020/08/28 14:16:00.534 asynManager::portThread port=HV0 callback
2020/08/28 14:16:00.604 hvm-ltuh-mc01:1470 read.
2020/08/28 14:16:00.657 HV0 get Eos 0
You said “Communication was ok, until it stopped and we set some masks to diagnose:”.  But communication does not appear to have stopped before you set those masks?  It had been less than 5 seconds since the last successful read, correct?  Was there some other reason the asynTrace was turned on then?
asynReport after using disconnectOnReadTimeout:
Cexp@ioc-ltuh-mc02>asynReport(4)
HV0 multiDevice:No canBlock:Yes autoConnect:Yes
    enabled:Yes connected:No numberConnects 146
That shows that it has connected 146 times.  Did you connect manually 146 times? That seems unlikely, so it seems that perhaps disconnectOnReadTimeout was doing what it is supposed to do, i.e. getting a read timeout, disconnecting and reconnecting?
From: Tech-talk <[email protected]> On Behalf Of Paduan Donadio, Marcio via Tech-talk
Sent: Tuesday, September 1, 2020 4:49 PM
To: Jeong Han Lee via Tech-talk <[email protected]>
Subject: CAEN R1470ET HVPS - Asyn + Stream Device auto connect does not recover communication
Hello, all.
I hope you can help me to decipher a scenario with an Ethernet device, Asyn, and Stream Device. It looks like the equipment stops answering but Asyn still shows that the connection is opened and when it writes to the port the result is Asyn success. Auto-connect does not act, I guess that is because Asyn is happy with the connection. If we close and open the port again, communication is reestablished. I’ve tried disconnectOnReadTimeout. Asyn disconnected after it, but then neither auto-connect or manual connection were able to open the connection again. The system is still in this condition as we were not authorized to reboot the IOC, yet.
Detailed data:
Asyn R4-32
Stream Device 2.8.9
2020/08/28 14:15:55.020 hvm-ltuh-mc01:1470 write 29
$BD:0,CMD:MON,CH:4,PAR:VSET
2020/08/28 14:15:55.116 hvm-ltuh-mc01:1470 read 26
#BD:00,CMD:OK,VAL:4500.0
2020/08/28 14:16:00.696 HV0 addr -1 queueRequest priority 0 not lockHolder
2020/08/28 14:16:00.775 HV0 set Eos 0
2020/08/28 14:16:00.816 HV0 schedule queueRequest timeout
2020/08/28 14:16:00.878 hvm-ltuh-mc01:1470 write.
2020/08/28 14:16:00.931 HV0 addr -1 queueRequest priority 0 not lockHolder
2020/08/28 14:16:01.010 wrote 29 to hvm-ltuh-mc01:1470, return asynSuccess.
2020/08/28 14:16:01.090 HV0 schedule queueRequest timeout
2020/08/28 14:16:01.152 HV0 set Eos 0
2020/08/28 14:16:01.192 HV0 addr -1 queueRequest priority 0 not lockHolder
2020/08/28 14:16:01.271 HV0 schedule queueRequest timeout
2020/08/28 14:16:01.333 HV0 addr -1 queueRequest priority 0 from lockHolder
2020/08/28 14:16:01.413 HV0 schedule queueRequest timeout
2020/08/28 14:16:01.474 HV0 addr -1 queueRequest priority 0 not lockHolder
2020/08/28 14:16:01.554 HV0 schedule queueRequest timeout
2020/08/28 14:16:01.615 HV0 addr -1 queueRequest priority 0 not lockHolder
2020/08/28 14:16:01.694 HV0 schedule queueRequest timeout
2020/08/28 14:16:01.756 asynManager::portThread port=HV0 callback
2020/08/28 14:16:01.825 HV0 addr -1 queueRequest priority 0 not lockHolder
2020/08/28 14:16:01.905 HV0 get Eos 0
2020/08/28 14:16:01.945 HV0 schedule queueRequest timeout
2020/08/28 14:16:02.008 HV0 set Eos 2
2020/08/28 14:16:02.053 HVM:LTUH:200:ASYN: exception 3, asynExceptionTraceMask
2020/08/28 14:16:02.136 HV0 addr -1 queueRequest priority 0 not lockHolder
2020/08/28 14:16:02.215 hvm-ltuh-mc01:1470 read.
2020/08/28 14:16:02.267 HV0 schedule queueRequest timeout
2020/08/28 14:16:02.330 HV0 addr -1 queueRequest priority 0 not lockHolder
2020/08/28 14:16:02.408 HV0 schedule queueRequest timeout
2020/08/28 14:16:02.470 HV0 addr -1 queueRequest priority 0 not lockHolder
2020/08/28 14:16:02.548 HV0 schedule queueRequest timeout
2020/08/28 14:16:05.582 HVM:LTUH:200:ASYN: exception 3, asynExceptionTraceMask
2020/08/28 14:16:05.612 HV0 addr -1 queueRequest priority 0 not lockHolder
2020/08/28 14:16:05.679 HV0 schedule queueRequest timeout
2020/08/28 14:16:05.740 HV0 addr -1 queueRequest priority 0 not lockHolder
2020/08/28 14:16:05.819 HV0 schedule queueRequest timeout
2020/08/28 14:16:05.881 HV0 addr -1 queueRequest priority 0 not lockHolder
2020/08/28 14:16:05.960 HV0 schedule queueRequest timeout
2020/08/28 14:16:06.021 HV0 addr -1 queueRequest priority 0 not lockHolder
2020/08/28 14:16:06.100 HV0 schedule queueRequest timeout
2020/08/28 14:16:06.162 HV0 set Eos 0
2020/08/28 14:16:06.202 HV0 addr -1 queueRequest priority 0 not lockHolder
2020/08/28 14:16:06.282 HV0 schedule queueRequest timeout
2020/08/28 14:16:06.343 HV0 addr -1 queueRequest priority 0 not lockHolder
2020/08/28 14:16:06.422 HV0 schedule queueRequest timeout
2020/08/28 14:16:06.484 asynManager::portThread port=HV0 callback
2020/08/28 14:16:06.554 HV0 addr -1 queueRequest priority 0 not lockHolder
2020/08/28 14:16:06.633 HV0 schedule queueRequest timeout
2020/08/28 14:16:06.695 HV0 addr -1 queueRequest priority 0 not lockHolder
2020/08/28 14:16:06.773 HV0 schedule queueRequest timeout
2020/08/28 14:16:06.835 HV0 addr -1 queueRequest priority 0 not lockHolder
2020/08/28 14:16:06.914 HV0 schedule queueRequest timeout
2020/08/28 14:16:06.975 HV0 addr -1 queueRequest priority 0 not lockHolder
2020/08/28 14:16:07.055 HV0 schedule queueRequest timeout
2020/08/28 14:16:31.102 HVM:LTUH:200:ASYN: exception 3, asynExceptionTraceMask
2020/08/28 14:16:31.125 HV0 set Eos 0
2020/08/28 14:16:31.160 asynManager::portThread port=HV0 callback
2020/08/28 14:16:31.229 HV0 addr -1 queueRequest priority 0 not lockHolder
2020/08/28 14:16:31.309 HV0 schedule queueRequest timeout
2020/08/28 14:16:31.370 asynManager::portThread port=HV0 callback
2020/08/28 14:16:31.440 hvm-ltuh-mc01:1470 read.
2020/08/28 14:16:31.493 HV0 get Eos 0
2020/08/28 14:16:31.533 HV0 set Eos 0
2020/08/28 14:16:31.573 hvm-ltuh-mc01:1470 write.
2020/08/28 14:16:31.626 wrote 29 to hvm-ltuh-mc01:1470, return asynSuccess.
2020/08/28 14:16:31.706 HV0 addr -1 queueRequest priority 0 not lockHolder
2020/08/28 14:16:31.786 HV0 set Eos 0
    enabled:Yes connected:Yes numberConnects 3
    nDevices 0 nQueued 49 blocked:No
    asynManagerLock:No synchronousLock:Yes
    exceptionActive:No exceptionUsers 170 exceptionNotifys 0
    traceMask:0xb traceIOMask:0x7 traceInfoMask:0x1
    interposeInterfaceList
        asynOctet pinterface 0x2e48048 drvPvt 0x2930310
    interfaceList
        asynCommon pinterface 0x2d5175c drvPvt 0x2926f60
        asynOption pinterface 0x2d51768 drvPvt 0x2926f60
        asynOctet pinterface 0x2927040 drvPvt 0x2926f60
    Port hvm-ltuh-mc01:1470: Connected
                    fd: 16
    Characters written: 31886349
       Characters read: 18628310
Asyn report a few minutes later:
interfaceList
               fd: 16
    Characters written: 31887627
After manually reopening the connection, the device restart sending characters (this was in another day):
2020/08/31 12:11:57.446 hvm-ltuh-mc01:1470 write 29
$BD:0,CMD:MON,CH:0,PAR:VSET
$BD:0,CMD:MON,CH:0,PAR:VSET\r\n
24 42 44 3a 30 2c 43 4d 44 3a 4d 4f 4e 2c 43 48 3a 30 2c 50
41 52 3a 56 53 45 54 0d 0a
2020/08/31 12:11:58.600 hvm-ltuh-mc01:1470 write 29
$BD:0,CMD:MON,CH:5,PAR:VMON
$BD:0,CMD:MON,CH:5,PAR:VMON\r\n
24 42 44 3a 30 2c 43 4d 44 3a 4d 4f 4e 2c 43 48 3a 35 2c 50
41 52 3a 56 4d 4f 4e 0d 0a
2020/08/31 12:11:59.765 hvm-ltuh-mc01:1470 write 29
$BD:0,CMD:MON,CH:1,PAR:VSET
$BD:0,CMD:MON,CH:1,PAR:VSET\r\n
24 42 44 3a 30 2c 43 4d 44 3a 4d 4f 4e 2c 43 48 3a 31 2c 50
2020/08/31 12:11:59.926 hvm-ltuh-mc01:1470 read 26
#BD:00,CMD:OK,VAL:0800.0
#BD:00,CMD:OK,VAL:0800.0\r\n
23 42 44 3a 30 30 2c 43 4d 44 3a 4f 4b 2c 56 41 4c 3a 30 38
30 30 2e 30 0d 0a
2020/08/31 12:12:00.122 hvm-ltuh-mc01:1470 write 29
$BD:0,CMD:MON,CH:0,PAR:STAT
$BD:0,CMD:MON,CH:0,PAR:STAT\r\n
41 52 3a 53 54 41 54 0d 0a
2020/08/31 12:12:00.345 hvm-ltuh-mc01:1470 read 25
#BD:00,CMD:OK,VAL:00001
#BD:00,CMD:OK,VAL:00001\r\n
This is how Asyn report looks like one things are going well:
    enabled:Yes connected:Yes numberConnects 4
    nDevices 0 nQueued 3 blocked:No
    traceMask:0x1 traceIOMask:0x1 traceInfoMask:0x1
    interfaceList
                    fd: 16
    Characters written: 32245553
       Characters read: 18922078
Dbpr when things are bad:
Cexp@ioc-ltuh-mc02>dbpr("HVM:LTUH:200:BoardName","4")
ACKS: INVALID       ACKT: YES           APST: On Change     ASG :
ASP : PTR 0x3651f20 BKLNK: ELL 1 [0x2ea1e00 .. 0x2ea1e00]   BKPT: 00
DESC: BoardName     DISA: 0             DISP: 0             DISS: NO_ALARM
DISV: 1             DPVT: PTR 0x35bbea0 DSET: PTR 0x2e4882c DTYP: stream
EVNT:               FLNK: CONSTANT
INP : INST_IO @caenN1470.proto readstring(0,BDNAME,HVM:LTUH:200) HV0
LCNT: 0             LSET: PTR 0x2fa7680 MLIS: ELL 2 [0x39950b8 .. 0x3994f00]
MLOK: 03 50 22 90   MPST: On Change     NAME: HVM:LTUH:200:BoardName
NSEV: NO_ALARM      NSTA: NO_ALARM      OLDSIMM: NO         OVAL: R1470ET
PACT: 1             PHAS: 0             PINI: NO            PPN : PTR 0x0
PPNR: PTR 0x0       PRIO: LOW           PROC: 0             PUTF: 0
RDES: PTR 0x2be1910 RPRO: 0             RSET: PTR 0x2e498e0 SCAN: Passive
SDIS: CONSTANT      SDLY: -1            SEVR: INVALID       SIML: CONSTANT
SIMM: NO            SIMPVT: PTR 0x0     SIMS: NO_ALARM      SIOL: CONSTANT
SPVT: PTR 0x0       SSCN: <nil>         STAT: TIMEOUT       SVAL:
TIME: 2020-08-31 13:09:47.237075064     TPRO: 0             TSE : 0
TSEL: CONSTANT      UDF : 0             UDFS: INVALID       VAL : R1470ET
Dbpr after using disconnectOnReadTimeout:
PACT: 0             PHAS: 0             PINI: NO            PPN : PTR 0x0
SPVT: PTR 0x0       SSCN: <nil>         STAT: COMM          SVAL:
TIME: 2020-08-31 13:14:51.686086457     TPRO: 0             TSE : 0
    nDevices 0 nQueued 28 blocked:No
    interfaceList
    Port hvm-ltuh-mc01:1470: Disconnected
                    fd: -1
    Characters written: 34072265
       Characters read: 20373454
Turning off disconnectOnReadTimeout and trying to manually disconnect didn’t work anymore:
2020/08/31 14:32:29.651 Can't connect to hvm-ltuh-mc01:1470: Connection timed out HV0 -1 autoConnect could not connect
2020/08/31 14:33:44.717 Can't connect to hvm-ltuh-mc01:1470: Connection timed out HV0 -1 autoConnect could not connect
2020/08/31 14:44:13.743 HVM:LTUH:200:ASYN: queueRequest failed
2020/08/31 14:44:13.745 HVM:LTUH:200:ASYN: queueRequest failed
2020/08/31 14:44:15.565 HVM:LTUH:200:ASYN: queueRequest failed
2020/08/31 14:44:15.567 HVM:LTUH:200:ASYN: queueRequest failed
2020/08/31 14:44:21.023 HVM:LTUH:200:ASYN: special queueRequest timeout
2020/08/31 14:44:27.077 HVM:LTUH:200:ASYN: special queueRequest timeout
2020/08/31 14:45:19.767 Can't connect to hvm-ltuh-mc01:1470: Connection timed out HV0 -1 autoConnect could not connect
2020/08/31 14:46:34.831 Can't connect to hvm-ltuh-mc01:1470: Connection timed out HV0 -1 autoConnect could not connect
RE: CAEN R1470ET HVPS - Asyn + Stream Device auto connect does not recover communication Mark Rivers via Tech-talk
CAEN R1470ET HVPS - Asyn + Stream Device auto connect does not recover communication Paduan Donadio, Marcio via Tech-talk
Prev: CAEN R1470ET HVPS - Asyn + Stream Device auto connect does not recover communication Paduan Donadio, Marcio via Tech-talk
Next: RE: CAEN R1470ET HVPS - Asyn + Stream Device auto connect does not recover communication Mark Rivers via Tech-talk
