Hi Arvinder,
Thank you for making an important point. I hadn't payed
attention because of using autosave; last values of PVs don't
change. I will change it in st.cmd as per your suggestion.
Hasan
On 19.10.2021 13:49, Palaha,
Arvinder (DLSLtd,RAL,LSCI) wrote:
Hi Hasan,
for what it's worth, when I changed a camera IOC to ADAravis,
the GevSCPSPacketSize always defaulted to 8228; the network port
it was connected to required 1500 or lower. So I put a dbpf
statement in the startup script setting it to 1500 upon IOC
restart.
Arvinder Palaha
Beamline Control Systems
DH G.63
T: +44-(0)1235-77-8137
E: [email protected]
Hi Mark and Giles,
Thank you for your responses. Network setup
was only on Bridged Network with MAC Address of ethernet
interface. When I set GevSCPSPacketSiz PV to 1500,
the problem is solved.
Hasan
On 17.10.2021 23:47, Knap, Giles
Hi Hasan,
I note that you are running on a virtual machine. What is
the network setup?
I have found that the GVCP control protocol will pass
through a NAT OK but the GVSP streaming protocol will not.
This is because the control protocol passes the IP and port
for the camera to connect to for streaming and the NAT does
not know about this.
If you VM has a virtual network, you would see control
working but streaming never connects and times out.
We are using Baumer CCD cameras with aravisGigE
driver and we have a AVT Manta G201-B camera. I'm
trying to add aravisGigE's st.cmd this camera but I
failed. The arv-tool-0.6 can recognize the camera and
template file can be built for the camera, but when I
run the start_epics file after changing st.cmd, I can
connect to camera but, cannot get raw data for images
and vision.
I have set to EPICS_CA_MAX_ARRAY_BYTES to .bashrc
and datasources of CSS as 1624*1234*32 and, I have set
new buffer memory values for different increasing
values and so, I couldn't get images. (I run
aravisGigE driver on virtual CentOS 7.9 on real Linux
Mint 21.04. Should I change buffer memory
values and EPICS_CA_MAX_ARRAY_BYTES
for Linux Mint? (This situation was not needed for
Baumer cameras.))
How can I solve this problem?
I'm sending related output of start_epics.
Thanks in advance.
Best regards.
Hasan.
output of start_epics:
[cam@localhost iocAravisGigE]$ ./start_epics
< envPaths
epicsEnvSet("IOC","iocAravisGigE")
epicsEnvSet("TOP","/opt/EPICS/support/Dedektor/aravisGigE/iocs/aravisGigEIOC")
epicsEnvSet("ARAVISGIGE","/opt/EPICS/support/Dedektor/aravisGigE/iocs/aravisGigEIOC/../..")
epicsEnvSet("SUPPORT","/opt/EPICS/support")
epicsEnvSet("ASYN","/opt/EPICS/support/asyn")
epicsEnvSet("AREA_DETECTOR","/opt/EPICS/support/Dedektor")
epicsEnvSet("ADSUPPORT","/opt/EPICS/support/Dedektor/ADSupport")
epicsEnvSet("ADCORE","/opt/EPICS/support/Dedektor/ADCore")
epicsEnvSet("AUTOSAVE","/opt/EPICS/support/autosave")
epicsEnvSet("BUSY","/opt/EPICS/support/busy")
epicsEnvSet("CALC","/opt/EPICS/support/calc")
epicsEnvSet("SNCSEQ","/opt/EPICS/support/sequencer")
epicsEnvSet("SSCAN","/opt/EPICS/support/sscan")
epicsEnvSet("DEVIOCSTATS","/opt/EPICS/support/iocStats")
epicsEnvSet("EPICS_BASE","/opt/EPICS/epics-base")
errlogInit(20000)
dbLoadDatabase("/opt/EPICS/support/Dedektor/aravisGigE/iocs/aravisGigEIOC/dbd/aravisGigEApp.dbd")
aravisGigEApp_registerRecordDeviceDriver(pdbbase)
# Prefix for all records
epicsEnvSet("PREFIX", "BL:DCC:")
# The port name for the detector
epicsEnvSet("PORT", "ARV1")
# The queue size for all plugins
epicsEnvSet("QSIZE", "20")
# The maximim image width; used for row
profiles in the NDPluginStats plugin
epicsEnvSet("XSIZE", "1624")
# The maximim image height; used for column
epicsEnvSet("YSIZE", "1234")
# The maximum number of time series points in
epicsEnvSet("NCHANS", "2048")
# The maximum number of frames buffered in the
NDPluginCircularBuff plugin
epicsEnvSet("CBUFFS", "500")
# The search path for database files
epicsEnvSet("EPICS_DB_INCLUDE_PATH",
"/opt/EPICS/support/Dedektor/ADCore/db")
dbLoadRecords("/opt/EPICS/support/Dedektor/aravisGigE/iocs/aravisGigEIOC/../../iocs/aravisGigEIOC/iocBoot/iocAravisGigE/Cam_Checker.db")
aravisCameraConfig("ARV1", "Allied Vision
Technologies-50-0503326564")
aravisCamera: Looking for camera 'Allied Vision
Technologies-50-0503326564'...
aravisCamera: Your tick frequency is 36000000
So your timestamp resolution is 27.777778 ns
aravisCamera: Done.
aravisCamera: Getting feature list...
aravisCamera: Done.
dbLoadRecords("/opt/EPICS/support/Dedektor/aravisGigE/iocs/aravisGigEIOC/../../db/AVT_Manta_G201B.template","P=BL:DCC:,R=cam1:,PORT=ARV1,ADDR=0,TIMEOUT=1")
#aravisCameraConfig("ARV2", "Baumer
Optronic-0402920514")
#dbLoadRecords("$(ARAVISGIGE)/db/Baumer_VLG20M.template","P=$(PREFIX),R=cam2:,PORT=ARV2,ADDR=0,TIMEOUT=1")
asynSetTraceMask("ARV1",0,0x21)
dbLoadRecords("/opt/EPICS/support/Dedektor/aravisGigE/iocs/aravisGigEIOC/../../db/aravisCamera.template",
#asynSetTraceMask("ARV2",0,0x21)
#dbLoadRecords("$(ARAVISGIGE)/db/aravisCamera.template",
# Create a standard arrays plugin
NDStdArraysConfigure("Image1", 5, 0, "ARV1", 0,
0)
# Allow for cameras up to 2048x2048x3 for RGB
dbLoadRecords("/opt/EPICS/support/Dedektor/ADCore/db/NDStdArrays.template",
"P=BL:DCC:,R=image1:,PORT=Image1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ARV1,TYPE=Int16,FTVL=SHORT,NELEMENTS=12582912")
# Load all other plugins using
commonPlugins.cmd
/opt/EPICS/support/Dedektor/ADCore/iocBoot/commonPlugins.cmd
# This is an example file for creating plugins
# It uses the following environment variable
macros
# Many of the parameters defined in this file
are also in commonPlugins_settings.req so if autosave
is being
# use the autosave value will replace the value
passed to this file.
# $(PREFIX) Prefix for all records
# $(PORT) The port name for the
detector. In autosave.
# $(QSIZE) The queue size for all
plugins. In autosave.
# $(XSIZE) The maximum image width; used
to set the maximum size for row profiles in the
NDPluginStats plugin and 1-D FFT
# profiles in NDPluginFFT.
# $(YSIZE) The maximum image height; used
to set the maximum size for column profiles in the
# $(NCHANS) The maximum number of time
series points in the NDPluginStats, NDPluginROIStats,
and NDPluginAttribute plugins
# $(CBUFFS) The maximum number of frames
buffered in the NDPluginCircularBuff plugin
# $(MAX_THREADS) The maximum number of threads
for plugins which can run in multiple threads.
Defaults to 5.
# Create a netCDF file saving plugin
NDFileNetCDFConfigure("FileNetCDF1", 20, 0,
"ARV1", 0)
dbLoadRecords("NDFileNetCDF.template","P=BL:DCC:,R=netCDF1:,PORT=FileNetCDF1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ARV1")
# Create a TIFF file saving plugin
NDFileTIFFConfigure("FileTIFF1", 20, 0, "ARV1",
0)
dbLoadRecords("NDFileTIFF.template",
"P=BL:DCC:,R=TIFF1:,PORT=FileTIFF1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ARV1")
# Create a JPEG file saving plugin
NDFileJPEGConfigure("FileJPEG1", 20, 0, "ARV1",
0)
dbLoadRecords("NDFileJPEG.template",
"P=BL:DCC:,R=JPEG1:,PORT=FileJPEG1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ARV1")
# Create a NeXus file saving plugin
NDFileNexusConfigure("FileNexus1", 20, 0,
"ARV1", 0)
dbLoadRecords("NDFileNexus.template",
"P=BL:DCC:,R=Nexus1:,PORT=FileNexus1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ARV1")
# Create an HDF5 file saving plugin
NDFileHDF5Configure("FileHDF1", 20, 0, "ARV1",
0)
dbLoadRecords("NDFileHDF5.template",
"P=BL:DCC:,R=HDF1:,PORT=FileHDF1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ARV1")
# Create a Magick file saving plugin
#NDFileMagickConfigure("FileMagick1", $(QSIZE),
0, "$(PORT)", 0)
#dbLoadRecords("NDFileMagick.template","P=$(PREFIX),R=Magick1:,PORT=FileMagick1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=$(PORT)")
# Create 4 ROI plugins
NDROIConfigure("ROI1", 20, 0, "ARV1", 0, 0, 0,
0, 0, 5)
dbLoadRecords("NDROI.template",
"P=BL:DCC:,R=ROI1:,
PORT=ROI1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ARV1")
NDROIConfigure("ROI2", 20, 0, "ARV1", 0, 0, 0,
0, 0, 5)
"P=BL:DCC:,R=ROI2:,
PORT=ROI2,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ARV1")
NDROIConfigure("ROI3", 20, 0, "ARV1", 0, 0, 0,
0, 0, 5)
"P=BL:DCC:,R=ROI3:,
PORT=ROI3,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ARV1")
NDROIConfigure("ROI4", 20, 0, "ARV1", 0, 0, 0,
0, 0, 5)
"P=BL:DCC:,R=ROI4:,
PORT=ROI4,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ARV1")
# Create 8 ROIStat plugins
NDROIStatConfigure("ROISTAT1", 20, 0, "ARV1",
0, 8, 0, 0, 0, 0, 5)
dbLoadRecords("NDROIStat.template",
"P=BL:DCC:,R=ROIStat1:
,PORT=ROISTAT1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ARV1,NCHANS=2048")
dbLoadRecords("NDROIStatN.template",
"P=BL:DCC:,R=ROIStat1:1:,PORT=ROISTAT1,ADDR=0,TIMEOUT=1,NCHANS=2048")
"P=BL:DCC:,R=ROIStat1:2:,PORT=ROISTAT1,ADDR=1,TIMEOUT=1,NCHANS=2048")
"P=BL:DCC:,R=ROIStat1:3:,PORT=ROISTAT1,ADDR=2,TIMEOUT=1,NCHANS=2048")
"P=BL:DCC:,R=ROIStat1:4:,PORT=ROISTAT1,ADDR=3,TIMEOUT=1,NCHANS=2048")
"P=BL:DCC:,R=ROIStat1:5:,PORT=ROISTAT1,ADDR=4,TIMEOUT=1,NCHANS=2048")
"P=BL:DCC:,R=ROIStat1:6:,PORT=ROISTAT1,ADDR=5,TIMEOUT=1,NCHANS=2048")
"P=BL:DCC:,R=ROIStat1:7:,PORT=ROISTAT1,ADDR=6,TIMEOUT=1,NCHANS=2048")
"P=BL:DCC:,R=ROIStat1:8:,PORT=ROISTAT1,ADDR=7,TIMEOUT=1,NCHANS=2048")
# Create a processing plugin
NDProcessConfigure("PROC1", 20, 0, "ARV1", 0,
0, 0)
dbLoadRecords("NDProcess.template",
"P=BL:DCC:,R=Proc1:,
PORT=PROC1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ARV1")
# Create a TIFF file plugin to read dark and
flatfield images into the processing plugin
NDFileTIFFConfigure("PROC1TIFF", 20, 0, "ARV1",
0)
"P=BL:DCC:,R=Proc1:TIFF:,PORT=PROC1TIFF,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ARV1")
# Create a scatter plugin
NDScatterConfigure("SCATTER1", 20, 0, "ARV1",
0, 0, 0)
dbLoadRecords("NDScatter.template",
"P=BL:DCC:,R=Scatter1:,
PORT=SCATTER1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ARV1")
# Create a gather plugin with 8 ports
NDGatherConfigure("GATHER1", 20, 0, 8, 0, 0)
dbLoadRecords("NDGather.template",
"P=BL:DCC:,R=Gather1:,
PORT=GATHER1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ARV1")
dbLoadRecords("NDGatherN.template",
"P=BL:DCC:,R=Gather1:, N=1,
"P=BL:DCC:,R=Gather1:, N=2,
PORT=GATHER1,ADDR=1,TIMEOUT=1,NDARRAY_PORT=ARV1")
"P=BL:DCC:,R=Gather1:, N=3,
PORT=GATHER1,ADDR=2,TIMEOUT=1,NDARRAY_PORT=ARV1")
"P=BL:DCC:,R=Gather1:, N=4,
PORT=GATHER1,ADDR=3,TIMEOUT=1,NDARRAY_PORT=ARV1")
"P=BL:DCC:,R=Gather1:, N=5,
PORT=GATHER1,ADDR=4,TIMEOUT=1,NDARRAY_PORT=ARV1")
"P=BL:DCC:,R=Gather1:, N=6,
PORT=GATHER1,ADDR=5,TIMEOUT=1,NDARRAY_PORT=ARV1")
"P=BL:DCC:,R=Gather1:, N=7,
PORT=GATHER1,ADDR=6,TIMEOUT=1,NDARRAY_PORT=ARV1")
"P=BL:DCC:,R=Gather1:, N=8,
PORT=GATHER1,ADDR=7,TIMEOUT=1,NDARRAY_PORT=ARV1")
# Create 5 statistics plugins
NDStatsConfigure("STATS1", 20, 0, "ARV1", 0, 0,
0, 0, 0, 5)
dbLoadRecords("NDStats.template",
"P=BL:DCC:,R=Stats1:,
PORT=STATS1,ADDR=0,TIMEOUT=1,HIST_SIZE=256,XSIZE=1624,YSIZE=1234,NCHANS=2048,NDARRAY_PORT=ARV1")
NDTimeSeriesConfigure("STATS1_TS", 20, 0,
"STATS1", 1, 23)
dbLoadRecords("/opt/EPICS/support/Dedektor/ADCore/db/NDTimeSeries.template",
"P=BL:DCC:,R=Stats1:TS:,
PORT=STATS1_TS,ADDR=0,TIMEOUT=1,NDARRAY_PORT=STATS1,NDARRAY_ADDR=1,NCHANS=2048,ENABLED=1")
NDStatsConfigure("STATS2", 20, 0, "ROI1", 0,
0, 0, 0, 0, 5)
"P=BL:DCC:,R=Stats2:,
PORT=STATS2,ADDR=0,TIMEOUT=1,HIST_SIZE=256,XSIZE=1624,YSIZE=1234,NCHANS=2048,NDARRAY_PORT=ARV1")
NDTimeSeriesConfigure("STATS2_TS", 20, 0,
"STATS2", 1, 23)
"P=BL:DCC:,R=Stats2:TS:,
PORT=STATS2_TS,ADDR=0,TIMEOUT=1,NDARRAY_PORT=STATS2,NDARRAY_ADDR=1,NCHANS=2048,ENABLED=1")
NDStatsConfigure("STATS3", 20, 0, "ROI2", 0,
0, 0, 0, 0, 5)
"P=BL:DCC:,R=Stats3:,
PORT=STATS3,ADDR=0,TIMEOUT=1,HIST_SIZE=256,XSIZE=1624,YSIZE=1234,NCHANS=2048,NDARRAY_PORT=ARV1")
NDTimeSeriesConfigure("STATS3_TS", 20, 0,
"STATS3", 1, 23)
"P=BL:DCC:,R=Stats3:TS:,
PORT=STATS3_TS,ADDR=0,TIMEOUT=1,NDARRAY_PORT=STATS3,NDARRAY_ADDR=1,NCHANS=2048,ENABLED=1")
NDStatsConfigure("STATS4", 20, 0, "ROI3", 0,
0, 0, 0, 0, 5)
"P=BL:DCC:,R=Stats4:,
PORT=STATS4,ADDR=0,TIMEOUT=1,HIST_SIZE=256,XSIZE=1624,YSIZE=1234,NCHANS=2048,NDARRAY_PORT=ARV1")
NDTimeSeriesConfigure("STATS4_TS", 20, 0,
"STATS4", 1, 23)
"P=BL:DCC:,R=Stats4:TS:,
PORT=STATS4_TS,ADDR=0,TIMEOUT=1,NDARRAY_PORT=STATS4,NDARRAY_ADDR=1,NCHANS=2048,ENABLED=1")
NDStatsConfigure("STATS5", 20, 0, "ROI4", 0,
0, 0, 0, 0, 5)
"P=BL:DCC:,R=Stats5:,
PORT=STATS5,ADDR=0,TIMEOUT=1,HIST_SIZE=256,XSIZE=1624,YSIZE=1234,NCHANS=2048,NDARRAY_PORT=ARV1")
NDTimeSeriesConfigure("STATS5_TS", 20, 0,
"STATS5", 1, 23)
"P=BL:DCC:,R=Stats5:TS:,
PORT=STATS5_TS,ADDR=0,TIMEOUT=1,NDARRAY_PORT=STATS5,NDARRAY_ADDR=1,NCHANS=2048,ENABLED=1")
# Create a transform plugin
NDTransformConfigure("TRANS1", 20, 0, "ARV1",
0, 0, 0, 0, 0, 5)
dbLoadRecords("NDTransform.template",
"P=BL:DCC:,R=Trans1:,
PORT=TRANS1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ARV1")
# Create an overlay plugin with 8 overlays
NDOverlayConfigure("OVER1", 20, 0, "ARV1", 0,
8, 0, 0, 0, 0, 5)
dbLoadRecords("NDOverlay.template",
"P=BL:DCC:,R=Over1:,
PORT=OVER1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ARV1")
dbLoadRecords("NDOverlayN.template","P=BL:DCC:,R=Over1:1:,NAME=ROI1,
SHAPE=1,O=Over1:,XPOS=BL:DCC:ROI1:MinX_RBV,YPOS=BL:DCC:ROI1:MinY_RBV,XSIZE=BL:DCC:ROI1:SizeX_RBV,YSIZE=BL:DCC:ROI1:SizeY_RBV,PORT=OVER1,ADDR=0,TIMEOUT=1")
dbLoadRecords("NDOverlayN.template","P=BL:DCC:,R=Over1:2:,NAME=ROI2,
SHAPE=1,O=Over1:,XPOS=BL:DCC:ROI2:MinX_RBV,YPOS=BL:DCC:ROI2:MinY_RBV,XSIZE=BL:DCC:ROI2:SizeX_RBV,YSIZE=BL:DCC:ROI2:SizeY_RBV,PORT=OVER1,ADDR=1,TIMEOUT=1")
dbLoadRecords("NDOverlayN.template","P=BL:DCC:,R=Over1:3:,NAME=ROI3,
SHAPE=1,O=Over1:,XPOS=BL:DCC:ROI3:MinX_RBV,YPOS=BL:DCC:ROI3:MinY_RBV,XSIZE=BL:DCC:ROI3:SizeX_RBV,YSIZE=BL:DCC:ROI3:SizeY_RBV,PORT=OVER1,ADDR=2,TIMEOUT=1")
dbLoadRecords("NDOverlayN.template","P=BL:DCC:,R=Over1:4:,NAME=ROI4,
SHAPE=1,O=Over1:,XPOS=BL:DCC:ROI4:MinX_RBV,YPOS=BL:DCC:ROI4:MinY_RBV,XSIZE=BL:DCC:ROI4:SizeX_RBV,YSIZE=BL:DCC:ROI4:SizeY_RBV,PORT=OVER1,ADDR=3,TIMEOUT=1")
dbLoadRecords("NDOverlayN.template","P=BL:DCC:,R=Over1:5:,NAME=Cursor1,SHAPE=1,O=Over1:,XPOS=junk,
YPOS=junk,
XSIZE=junk,
YSIZE=junk,
PORT=OVER1,ADDR=4,TIMEOUT=1")
dbLoadRecords("NDOverlayN.template","P=BL:DCC:,R=Over1:6:,NAME=Cursor2,SHAPE=1,O=Over1:,XPOS=junk,
YPOS=junk,
XSIZE=junk,
YSIZE=junk,
PORT=OVER1,ADDR=5,TIMEOUT=1")
dbLoadRecords("NDOverlayN.template","P=BL:DCC:,R=Over1:7:,NAME=Box1,
YPOS=junk,
XSIZE=junk,
YSIZE=junk,
PORT=OVER1,ADDR=6,TIMEOUT=1")
dbLoadRecords("NDOverlayN.template","P=BL:DCC:,R=Over1:8:,NAME=Box2,
YPOS=junk,
XSIZE=junk,
YSIZE=junk,
PORT=OVER1,ADDR=7,TIMEOUT=1")
# Create 2 color conversion plugins
NDColorConvertConfigure("CC1", 20, 0, "ARV1",
0, 0, 0, 0, 0, 5)
dbLoadRecords("NDColorConvert.template",
"P=BL:DCC:,R=CC1:,
PORT=CC1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ARV1")
NDColorConvertConfigure("CC2", 20, 0, "ARV1",
0, 0, 0, 0, 0, 5)
"P=BL:DCC:,R=CC2:,
PORT=CC2,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ARV1")
# Create a circular buffer plugin
NDCircularBuffConfigure("CB1", 20, 0, "ARV1",
0, 500, 0)
dbLoadRecords("NDCircularBuff.template",
"P=BL:DCC:,R=CB1:,
PORT=CB1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=ARV1")
# Create an NDAttribute plugin with 8
attributes
NDAttrConfigure("ATTR1", 20, 0, "ARV1", 0, 8,
0, 0, 0)
dbLoadRecords("NDAttribute.template",
"P=BL:DCC:,R=Attr1:,
PORT=ATTR1,ADDR=0,TIMEOUT=1,NCHANS=2048,NDARRAY_PORT=ARV1")
dbLoadRecords("NDAttributeN.template",
"P=BL:DCC:,R=Attr1:1:,
PORT=ATTR1,ADDR=0,TIMEOUT=1,NCHANS=2048")
"P=BL:DCC:,R=Attr1:2:,
PORT=ATTR1,ADDR=1,TIMEOUT=1,NCHANS=2048")
"P=BL:DCC:,R=Attr1:3:,
PORT=ATTR1,ADDR=2,TIMEOUT=1,NCHANS=2048")
"P=BL:DCC:,R=Attr1:4:,
PORT=ATTR1,ADDR=3,TIMEOUT=1,NCHANS=2048")
"P=BL:DCC:,R=Attr1:5:,
PORT=ATTR1,ADDR=4,TIMEOUT=1,NCHANS=2048")
"P=BL:DCC:,R=Attr1:6:,
PORT=ATTR1,ADDR=5,TIMEOUT=1,NCHANS=2048")
"P=BL:DCC:,R=Attr1:7:,
PORT=ATTR1,ADDR=6,TIMEOUT=1,NCHANS=2048")
"P=BL:DCC:,R=Attr1:8:,
PORT=ATTR1,ADDR=7,TIMEOUT=1,NCHANS=2048")
# Create an FFT plugin
NDFFTConfigure("FFT1", 3, 0, "ARV1", 0, 0, 0,
0, 0, 5)
dbLoadRecords("NDFFT.template", "P=BL:DCC:,
R=FFT1:, PORT=FFT1, ADDR=0, TIMEOUT=1,
NDARRAY_PORT=ARV1, NAME=FFT1, NCHANS=1624")
set_requestfile_path("./")
set_requestfile_path("/opt/EPICS/support/Dedektor/ADCore/ADApp/Db")
set_requestfile_path("/opt/EPICS/support/Dedektor/ADCore/iocBoot")
set_savefile_path("./autosave")
set_pass0_restoreFile("auto_settings.sav")
set_pass1_restoreFile("auto_settings.sav")
save_restoreSet_status_prefix("BL:DCC:")
dbLoadRecords("/opt/EPICS/support/autosave/asApp/Db/save_restoreStatus.db",
"P=BL:DCC:")
# Optional: load NDPluginPva plugin
#NDPvaConfigure("PVA1", $(QSIZE), 0, "$(PORT)",
0, $(PREFIX)Pva1:Image, 0, 0, 0)
#dbLoadRecords("NDPva.template",
"P=$(PREFIX),R=Pva1:,
PORT=PVA1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=$(PORT)")
# Must start PVA server if this is enabled
#startPVAServer
# Optional: load NDPluginEdge plugin
#NDEdgeConfigure("EDGE1", $(QSIZE), 0,
"$(PORT)", 0, 0, 0, 0)
#dbLoadRecords("$(ADPLUGINEDGE)/db/NDEdge.template",
"P=$(PREFIX),R=Edge1:,
PORT=EDGE1,ADDR=0,TIMEOUT=1,NDARRAY_PORT=$(PORT)")
#set_requestfile_path("$(ADPLUGINEDGE)/edgeApp/Db")
# Optional: load scan records
#dbLoadRecords("$(SSCAN)/sscanApp/Db/scan.db",
"P=$(PREFIX),MAXPTS1=2000,MAXPTS2=200,MAXPTS3=20,MAXPTS4=10,MAXPTSH=10")
#set_requestfile_path("$(SSCAN)/sscanApp/Db")
# Optional: load sseq record for acquisition
sequence
#dbLoadRecords("$(CALC)/calcApp/Db/sseqRecord.db",
"P=$(PREFIX), S=AcquireSequence")
#set_requestfile_path("$(CALC)/calcApp/Db")
# Optional: load devIocStats records (requires
DEVIOCSTATS module)
#dbLoadRecords("$(DEVIOCSTATS)/db/iocAdminSoft.db",
"IOC=$(PREFIX)")
# Optional: load alive record (requires ALIVE
module)
#dbLoadRecords("$(ALIVE)/aliveApp/Db/alive.db",
"P=$(PREFIX),RHOST=192.168.1.254")
macLib: macro ADPILATUS is undefined (expanding
string
set_requestfile_path("$(ADPILATUS)/prosilicaApp/Db"))
#asynSetTraceMask("$(PORT)",0,255)
#asynSetTraceMask("$(PORT)",0,3)
iocInit()
Starting iocInit
## EPICS R3.15.8
## EPICS Base built Aug 24 2021
reboot_restore: entry for file
'auto_settings.sav'
reboot_restore: Found filename
'auto_settings.sav' in restoreFileList.
*** restoring from
'./autosave/auto_settings.sav' at initHookState 6
(before record/device init) ***
reboot_restore: done with file
'auto_settings.sav'
2021/10/15 10:26:42.552
asynPortDriver:drvUserCreate: addr=0, cannot find
parameter CURSOR_VAL
BL:DCC:Stats1:CursorVal
devAsynFloat64::initCommon drvUserCreate
BL:DCC:Stats2:CursorVal
BL:DCC:Stats3:CursorVal
2021/10/15 10:26:42.553
BL:DCC:Stats4:CursorVal
BL:DCC:Stats5:CursorVal
aravisCamera: Adding feature DeviceVendorName
with value: Allied Vision Technologies
aravisCamera: Adding feature DeviceModelName
with value: Manta_G-201B
DeviceFirmwareVersion with value: 00.01.44.04
aravisCamera: Adding feature DeviceID with
value: 50-0503326564
aravisCamera: Adding feature DevicePartNumber
with value: E0020008
aravisCamera: Adding feature ModelName with
value: Manta_G-201B
'auto_settings.sav'
*** restoring from
'./autosave/auto_settings.sav' at initHookState 7
(after record/device init) ***
'auto_settings.sav'
2021/10/15 10:26:43.916
NDPluginDriver::writeInt32 cannot do ProcessPlugin, no
input array cached
NDPluginDriver::writeInt32 ERROR, status=3,
function=66, value=1, connectedToArrayPort_=1
input array cached
2021/10/15 10:26:43.918
input array cached
input array cached
2021/10/15 10:26:43.919
input array cached
input array cached
2021/10/15 10:26:43.920
input array cached
input array cached
2021/10/15 10:26:43.922
input array cached
input array cached
# save things every thirty seconds
create_monitor_set("auto_settings.req",
30,"P=BL:DCC:")
save_restore:readReqFile: unable to open file
aravisCamera_settings.req. Exiting.
sseq_settings.req. Exiting.
create_monitor_set("commonPlugin_settings.req",30,"P=BL:DCC:")
commonPlugin_settings.sav: 1070 of 1070 PV's
connected
auto_settings.sav: 1086 of 1086 PV's connected
epics>Bad Frame Status: Timeout
Bad Frame Status: Timeout, 10 msgs suppressed
Hasan Sansar
This e-mail and any attachments may contain
confidential, copyright and or privileged material, and are
for the use of the intended addressee only. If you are not
the intended addressee or an authorised recipient of the
addressee please notify us of receipt by returning the
e-mail and do not use, copy, retain, distribute or disclose
the information in or attached to the e-mail.
Any opinions expressed within this e-mail are those of the
individual and not necessarily of Diamond Light Source Ltd.
Diamond Light Source Ltd. cannot guarantee that this e-mail
or any attachments are free from viruses and we cannot
accept liability for any damage which you may sustain as a
result of software viruses which may be transmitted in or
with the message.
Diamond Light Source Limited (company no. 4375679).
Registered in England and Wales with its registered office
at Diamond House, Harwell Science and Innovation Campus,
Didcot, Oxfordshire, OX11 0DE, United Kingdom
Hasan Sansar
confidential, copyright and or privileged material, and are for
the use of the intended addressee only. If you are not the
intended addressee or an authorised recipient of the addressee
please notify us of receipt by returning the e-mail and do not
use, copy, retain, distribute or disclose the information in or
Diamond Light Source Ltd. cannot guarantee that this e-mail or
any attachments are free from viruses and we cannot accept
liability for any damage which you may sustain as a result of
software viruses which may be transmitted in or with the
message.
Diamond Light Source Limited (company no. 4375679). Registered
in England and Wales with its registered office at Diamond
House, Harwell Science and Innovation Campus, Didcot,
Hasan Sansar
- aravisGigE driver for AVT Manta camera Hasan SANSAR via Tech-talk
- Re: aravisGigE driver for AVT Manta camera Knap, Giles (DLSLtd,RAL,LSCI) via Tech-talk
- Re: aravisGigE driver for AVT Manta camera Hasan SANSAR via Tech-talk
- Re: aravisGigE driver for AVT Manta camera Palaha, Arvinder (DLSLtd, RAL, LSCI) via Tech-talk
RE: ADAravis multiple image mode fail for Basler raL12288-8gm Mark Rivers via Tech-talk
RE: ADAravis multiple image mode fail for Basler raL12288-8gm David Vine via Tech-talk
<2021>
2025
2026
Java & CA Pavel Cheblakov via Tech-talk
<2021>
2025
2026