Subject: RE: ADEiger Stream stop zmq?
To: Jesse Hopkins <[email protected]>
Date: Fri, 10 Dec 2021 14:28:20 +0000
Hi Jesse,
I have created a fix_stream branch on Github which I hope will fix the issue you are having with the zmq interface.  The zmq socket will now be created when StreamEnable is set to 1, and deleted when StreamEnable is set to 0.  It would be good if you can test the following:
-          If you do not enable the Stream interface then your zmq client should receive all of the messages
-          If you do enable the Stream interface then ADEiger should get the images.  You can check with NDPluginStats or other plugins.
-          If the Stream interface is disabled and you try to change DataSource to Stream then you should get an error message and it should not enable the Stream interface, which it previously would have done.
Sent: Thursday, December 9, 2021 5:05 PM
Cc: [email protected]
Hi Jesse,
I suspect you have already found an issue I was afraid of when I fixed the race condition in ADEiger R3-1.  Previously the zmq socket in ADEiger was only created when acquisition was started and the Stream interface was enabled.  That had a race condition, because the socket was only created after the Eiger might start sending data over the zmq interface.  The fix was to create the zmq socket in the thread that reads the zmq socket before acquisition starts, and whether or not the Stream interface is enabled.  We thought that would be OK, but I think it could be causing your problem with your client only getting every other zmq message.
Can you try rolling back to ADEiger R3-0 and see if the problem goes away?  If so then I will need to change the code so it only opens the zmq socket when the Stream interface is enabled, and closes it when it is disabled.
From: Tech-talk <[email protected]> On Behalf Of Jesse Hopkins via Tech-talk
Sent: Thursday, December 9, 2021 3:32 PM
To: EPICS Tech Talk <[email protected]>
Subject: ADEiger Stream stop zmq?
Hi folks,
We recently got an Eiger2 XE 9M and I've been using AreaDetector to control it. So far, if I'm just using AreaDetector by itself everything works great (thanks for that!). In the long run what I would like to do is use AreaDetector to control the exposures, but use my own program to read the Eiger Stream over zmq and process the data in real time. According to the documentation this should be possible.
However, when I run the Eiger from AreaDetector with DataSource set to None, as specified in the documentation, my zmq client only receives half of the messages, which is what I would expect with two PULL clients connected to a single PUSH client. For example, if I take two images, I see only frame 0 and the dseries-end messages but not the dheader or frame 1 messages. If I collect images under the same conditions without the IOC running (e.g. just stop the IOC and send an arm and trigger command directly to the api with all settings left the same) I get all of the expected messages.
I'm wondering if there's some setting I've missed (see attached screenshot of the caQtDM window) or if this is a bug.
Also, as a side note, you'll see in the caQtDM window that the DCU RAM Disk Free is reporting a very wrong number, I suspect it's an overflow error. If I query buffer_free (which I assume is that that is), I get the following:
{"access_mode":"r","unit":"B","value":85402976256,"value_type":"uint"}
I'm using:
ADCore: 3.11.0
ADEiger: 3.1.0
SIMPLON API: 1.8.0
Eiger firmware: release-2020.2.4
EPICS: 7.0.4.1
And the IOC is running on Debian 11
All the best.
- Jesse
Jesse Hopkins, PhD
Deputy Director
BioCAT, Sector 18
Advanced Photon Source
Re: [Ext] RE: ADEiger Stream stop zmq? Jesse Hopkins via Tech-talk
RE: ADEiger Stream stop zmq? Mark Rivers via Tech-talk
Prev: RE: ADEiger Stream stop zmq? Mark Rivers via Tech-talk
Next: ADPilatus writing cbf strangeness John Dobbins via Tech-talk
Next: Re: [Ext] RE: ADEiger Stream stop zmq? Jesse Hopkins via Tech-talk
