1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  <2022>  2023  2024  2025  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  <2022>  2023  2024  2025  2026
Subject: Re: Ways to know if instrument connected via TCP/IP through StreamDevice has lost connection
From: Kuldeep Joshi via Tech-talk <[email protected]>
To: [email protected]
Date: Fri, 9 Dec 2022 07:28:21 +0000
Hello Mark,
Can this be used to find out if the device connected on RS232 based serial port is powered down, or the communication is not responding.
Ie will Modbus like module do an auto reconnect on serial communication timeout. This problem may not occur on actual serial com ports.
However, for virtual comports, I have noticed at times the driver drops the backend TCP connection and to restore the connection I have to close the comport and then reconnect.
Here similar reconnect functionality for Modbus over TCP will be useful for serial port (in view of the virtual com port based devices)
Kuldeep
Date: Fri, 9 Dec 2022 02:51:07 +0000
To: EPICS tech-talk <[email protected]>, "Wang, Andrew"
        <[email protected]>
Message-ID:
        <DM5PR11MB20127608AD0E433E01A94181991C9@DM5PR11MB2012.namprd11.prod.outlook.com" target="_blank">DM5PR11MB20127608AD0E433E01A94181991C9@DM5PR11MB2012.namprd11.prod.outlook.com>
Content-Type: text/plain; charset="iso-8859-1"
Hi Andy,
> I feel like there is a better way than me trying to use a calcout record and check if any record has its STAT field set to COMM. Like, is it possible to monitor for connection status without needing to process a record whose device type is StreamDevice and check its STAT field?
Yes, there is a better way.
First, you probably want to use this line in your startup script:
asynSetOption("$(PORT)",0, "disconnectOnReadTimeout", "Y")
That will cause the port to disconnect if there is a timeout.  If you don't do that then the asyn won't know that the port is disconnected for a very long time on Linux, it will just keep giving "temporarily unavailable" messages.
In your startup script load an asyn record that is connected to that asyn port.  Here is an example:
dbLoadRecords("$(ASYN)/db/asynRecord.db", "P=IOC:, R=asyn1, PORT=ARIES1, ADDR=0, OMAX=256, IMAX=256")
You can monitor the  IOC:asyn1.CNCT field,  which will either be "Connect" or "Disconnect".
From: Tech-talk <[email protected]> on behalf of Wang, Andrew via Tech-talk <[email protected]>
Sent: Thursday, December 8, 2022 8:21 PM
Subject: Ways to know if instrument connected via TCP/IP through StreamDevice has lost connection
Hi all,
Basically, I was wondering if there is a better way to know if communication to an instrument has been lost. I understand that this can be done either by simplying looking at the IOC shell for disconnection errors or checking if the STAT field is set to COMM.
I was wondering if there is another method of determining if the instrument is still connected or not besides the two methods described above. Additionally, I would like to have a record called "is_instrument_connected", that holds a value like 1 to indicate the instrument is connected and a 0 if it is not. I feel like there is a better way than me trying to use a calcout record and check if any record has its STAT field set to COMM. Like, is it possible to monitor for connection status without needing to process a record whose device type is StreamDevice and check its STAT field?
Thank you,
Andy
URL: <http://mailman.aps.anl.gov/pipermail/tech-talk/attachments/20221209/0ce7a120/attachment.html>
Subject: Digest Footer
Tech-talk mailing list [email protected]
https://mailman.aps.anl.gov/mailman/listinfo/tech-talk
End of Tech-talk Digest, Vol 16, Issue 707
Prev: Re: Ways to know if instrument connected via TCP/IP through StreamDevice has lost connection Mark Rivers via Tech-talk
Next: StreamDevice redirection to ai record with ASLO question Kevin Peterson via Tech-talk
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  <2022>  2023  2024  2025  2026
Prev: Re: Ways to know if instrument connected via TCP/IP through StreamDevice has lost connection Wang, Andrew via Tech-talk
