1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  <20232024  2025  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  <20232024  2025  2026 
<== Date ==> <== Thread ==>

Subject: RE: Cannot do successful RS-232 communication with baud rates <2400 with VME and IP module
From: "Goetze, Kurt via Tech-talk" <[email protected]>
To: "Rivers, Mark L." <[email protected]>, Mrinal Bera <[email protected]>
Cc: Tech-talk <[email protected]>
Date: Thu, 2 Nov 2023 22:16:17 +0000

Thanks Mark.  It’s being configured and loaded into ioc3idb, which is running synApps 6.2.1.  Whether it actually works I don’t know.  We’ll need to ask sec.3 staff to verify.

 

serial.cmd:tyGSAsynInit("serial21","UART_2",4, 1200,'O',1,7,'N',"\r","\n")   /* LS330 */

serial.cmd:# Lakeshore 330 Temperature Controller

serial.cmd:dbLoadRecords("$(IP)/ipApp/Db/LakeShore330.db","P=3idb:,Q=TC1,PORT=serial21")

 

Kurt

 

 

From: Mark Rivers <[email protected]>
Sent: Thursday, November 2, 2023 4:43 PM
To: Goetze, Kurt <[email protected]>; Mrinal Bera <[email protected]>
Cc: Tech-talk <[email protected]>
Subject: RE: Cannot do successful RS-232 communication with baud rates <2400 with VME and IP module

 

I tested with my VxWorks system in the sector 13 electronics lab today.  I reproduced the behavior Mrinal is seeing.  This is all different hardware from Mrinal’s.

 

I put a loopback from pin 2 to 3 and used the asyn record to send a string and read the response.  It works fine with the following baud rates:

2400

4800

9600

19200

38400

 

300, 600, and above 38400 report that the baud rate is not supported on the VxWorks console.

 

1200 does not report that it is unsupported, but the readback is incorrect, it does not match the output.

 

 

If 1200 was previously working at APS beamlines then we need to figure out if there was a change to the driver that broke it.

 

Mark

 

 

From: Tech-talk <[email protected]> On Behalf Of Goetze, Kurt via Tech-talk
Sent: Monday, October 30, 2023 5:21 PM
To: Mrinal Bera <[email protected]>
Cc: Tech-talk <[email protected]>
Subject: RE: Cannot do successful RS-232 communication with baud rates <2400 with VME and IP module

 

Hi Mrinal,

 

Mark Rivers made the suggestion to change the settings.  Which is a good suggestion of course.  To that I would add, for me, the next thing I would do is get on the port and sniff the traffic. 

 

Let’s go off-line (off tech-talk) for now, you can bring me your boards and I can test them if you want.

 

Kurt

 

 

From: Mrinal Bera <[email protected]>
Sent: Monday, October 30, 2023 5:15 PM
To: Goetze, Kurt <[email protected]>
Cc: Tech-talk <[email protected]>
Subject: Re: Cannot do successful RS-232 communication with baud rates <2400 with VME and IP module

 

Hi Kurt,

Although the RS-232 settings for Lakeshore 330 are (Baud rate:1200bps, Parity: Odd, Stop bits:1, Data bits: 7), I tried to test the loopback with your suggestions (data bits: 8, parity: none), and got different input/outputs for 1200 baud rate as shown in the attached image. With 2400 and above baud rates the input/output values are the same as expected.

 

The IP-Octal modules I am trying to set are old modules from SBS. We have ordered some new ones (IP-520) from Acromag, but I haven’t yet received them. We will look into the Moxa option as well.

 

Thanks and regards,

Mrinal

 

 

 

On Oct 30, 2023, at 4:47 PM, Goetze, Kurt <[email protected]> wrote:

 

Hi Mrinal,

 

ioc3idb in sec.3 is running a LakeShore330 at 1200 baud, using the VME/IP-Octal.

 

Mostly these days we use Moxa “device servers” to run RS232 ports, I see some examples of the LakeShore330 being run this way here.

 

Just as a longshot, what company built your IP-Octal-232 modules?  Did these come from the old supplier (SBS?), or did they come from an “Artisan” type company that does re-pops of legacy hardware?  If the latter, I would be suspicious.

 

IIRC we actually moved away from the IP-Octal-232 in favor of an Acromag (IP520?) board, towards the end of our VME era.  These days much prefer a Moxa.

 

Kurt

 

 

 

From: Mrinal Bera <[email protected]> 
Sent: Monday, October 30, 2023 4:15 PM
To: Goetze, Kurt <[email protected]>
Cc: Tech-talk <[email protected]>
Subject: Re: Cannot do successful RS-232 communication with baud rates <2400 with VME and IP module

 

Hi Kurt, 

I checked the setup with two IP-Octal-232 modules, and they both showed similar behavior. I have been using Lakeshore so far but with soft IOC using on a Raspberry Pi computer. This is the first time I am trying to interface it through VME and the IP-232 module. Can you tell me which beamline at APS is using Lakeshore through VME?

 

Thanks,

Mrinal 

 

On Oct 30, 2023, at 3:08 PM, Goetze, Kurt <[email protected]> wrote:

 

Hi Mrinal,

 

Your loopback test results suggest a hardware problem to me.  Have you tried the test using a different (or new) IP-Octal-232 card?  Also, I would send more and different characters, to see if any patterns or other failure modes emerge.

 

LakeShore 330’s have been used here at the APS, at 1200 kbps, from what I can tell searching through ioc setup files.

 

Kurt

 

From: Tech-talk <[email protected]> On Behalf Of Mrinal Bera via Tech-talk
Sent: Monday, October 30, 2023 1:32 PM
To: [email protected]
Subject: Cannot do successful RS-232 communication with baud rates <2400 with VME and IP module

 

Hello,
While configuring RS-232 communication through XM-octal breakout boards with TVME-200 (with IP-Octal serial module) and MVME3100 using EPICS 7.0 and vxWorks-6.9.4, I did some quick tests with asynRecord. I found the following:
1) With loopback communication(when I sorted the 2 and 3 terminals of the RS-232 ports) with the baud rates of 2400 and above, I received the same input and output values (as shown in the 1st attached image). That shows that RS-232 communication is OK.
2) With the baud rate of 1200, the Output and Input values of asyRecord are different (as shown in the 2nd attached image). I suspect either the tyGSModule or TVME-200 with IP-Octal Serial board has some issue with communication baud rates below 2400.

When I checked the IP-Octal Serial module documentation I found the baud rate information as ‘Programmable in 18 steps to 34.8kbps’. I dug further into the information regarding the baud rate of the chip used in the module (i.e SCC2698B). I found that the chip supports 26 fixed baud rates from 50 to 38.4kbps. Hence, I think 1200 kbps should be supported by the chip and IP-Octal Serial module.


I need the baud rate of 1200kbps to interface the Lakeshore 330 temperature controller with EPICS on VME hardware. Did anybody try to interface Lakeshore 330 with VME and IP modules?


Best regards,
Mrinal

<image001.png>

 

<image002.png>

 


References:
Cannot do successful RS-232 communication with baud rates <2400 with VME and IP module Mrinal Bera via Tech-talk
RE: Cannot do successful RS-232 communication with baud rates <2400 with VME and IP module Goetze, Kurt via Tech-talk
Re: Cannot do successful RS-232 communication with baud rates <2400 with VME and IP module Mrinal Bera via Tech-talk
RE: Cannot do successful RS-232 communication with baud rates <2400 with VME and IP module Goetze, Kurt via Tech-talk
Re: Cannot do successful RS-232 communication with baud rates <2400 with VME and IP module Mrinal Bera via Tech-talk
RE: Cannot do successful RS-232 communication with baud rates <2400 with VME and IP module Goetze, Kurt via Tech-talk
RE: Cannot do successful RS-232 communication with baud rates <2400 with VME and IP module Mark Rivers via Tech-talk

Navigate by Date:
Prev: RE: Cannot do successful RS-232 communication with baud rates <2400 with VME and IP module Mark Rivers via Tech-talk
Next: AlliedVision Manta G-235B Miroslaw Dach via Tech-talk
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  <20232024  2025  2026 
Navigate by Thread:
Prev: RE: Cannot do successful RS-232 communication with baud rates <2400 with VME and IP module Mark Rivers via Tech-talk