1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  <2026>  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  <2026>
Subject: Re: RTEMS/MVME2100 OMS58 card does not exist! error
From: Miroslaw Dach via Tech-talk <tech-talk at aps.anl.gov>
To: Michael Davidsaver <mdavidsaver at gmail.com>
Cc: EPICS Tech-Talk <tech-talk at aps.anl.gov>
Date: Mon, 13 Apr 2026 23:54:30 -0700
Hi Michael and Andrew,
Please find the summary below:
On MVME2100 (MPC8240 + Tundra Universe II CA91C142), ANY software write to
the Universe II's outbound (PCI-to-VME) window registers (LSI CTL, BS, BD,
or TO) causes all subsequent VME coupled reads to fail with S_TA=1 (PCI
Signaled Target Abort) and VERR=0 (no VME cycle generated).  Only the
PPCBug firmware's factory power-on configuration produces working VME
cycles.  The RTEMS BSP's normal boot sequence — vmeUniverseReset() followed
by BSP_VMEOutboundPortCfg() — triggers this failure.
In more details:
The Universe II claims PCI transactions (asserts DEVSEL#) but signals
Target Abort without ever generating a VME bus cycle.  The chip appears to
have hidden internal state in its outbound window state machine that is not
captured in the register values and cannot be restored by software.
The workaround:
We override BSP_vme_config() (weak symbol) and wrap vmeUniverseReset()
and BSP_vme2local_adrs() via the --wrap linker option:
  - __wrap_vmeUniverseReset(): no-op, preserves PPCBug config
  - BSP_vme_config(): calls BSP_VMEInit() (sets theOps), configures DBAT1
    for PPCBug's PCI range (0xA0000000), installs IRQ manager.  Does NOT
    call BSP_VMEOutboundPortCfg().
  - __wrap_BSP_vme2local_adrs(): hardcoded VME-to-local address translation
    for PPCBug's window layout (the Universe II driver's xlateAddr cannot
    reverse PPCBug's translation offsets due to 32-bit wrapping, and the
    AM code doesn't match because PPCBug uses non-privileged mode)
Makefile:
  USR_LDFLAGS_RTEMS += -Wl,--wrap=vmeUniverseReset
  USR_LDFLAGS_RTEMS += -Wl,--wrap=BSP_vme2local_adrs
Board details: MVME2101, MPC8240 @200MHz, 32MB SDRAM, Bus Clock 67MHz,
PPCBug ENV at factory defaults, Universe II at PCI BAR0=0xFCDFF000,
RTEMS 6.0.0, EPICS Base 7.0.10.
We don't know if this is specific to our board/revision or affects all
MVME2100 systems.  If anyone has successfully used vmeUniverseReset() +
BSP_VMEOutboundPortCfg() on MVME2100 with RTEMS 6, we'd be interested
to hear about it.
for our OMS based project we have added the following files in the omsApp/src directory:
file: Makefile
oms_DBD += motorRecord.dbd
oms_DBD += devOms.dbd
oms_DBD += omsSetupRegister.dbd
oms_DBD += epicsvme.dbd
oms_DBD += epicspci.dbd
oms_SRCS += omsSetupRegister.cpp
ifeq ($(T_A),RTEMS-mvme2100)
USR_LDFLAGS += -Wl,--wrap=vmeUniverseReset
USR_LDFLAGS += -Wl,--wrap=BSP_vme2local_adrs
endif
file: omsSetupRegister.dbd
registrar(omsSetupRegistrar)
variable(drvOMSdebug)
file: omsSetupRegister.cpp
 * omsSetupRegister.cpp
 * 1) IOCsh registration for omsSetup() (missing from upstream drvOms.cc).
 * 2) MVME2100 VME bridge fix for RTEMS 6:
 *    - Override BSP_vme_config() (weak symbol) to prevent BSP from
 *      reprogramming the Universe II outbound windows.  ANY write to
 *      the outbound LSI registers (CTL, BS, BD, or TO) destroys hidden
 *      internal state and causes S_TA=1 (Signaled Target Abort) on all
 *      subsequent VME coupled reads.  Only PPCBug's factory power-on
 *      configuration produces working VME cycles.
 *    - Wrap vmeUniverseReset() to a no-op (it disables all windows).
 *    - Wrap BSP_vme2local_adrs() to provide address translation for
 *      PPCBug's window layout (the Universe II driver's xlateAddr
 *      cannot reverse PPCBug's translation offsets due to 32-bit wrap).
 *    - Set DBAT1 to cover PPCBug's PCI address range (0xA0000000).
 * 3) Diagnostic IOCsh commands: vmeProbeTest, vmeScanA16
#include <stdio.h>
#include <string.h>
#include <iocsh.h>
#include <epicsExport.h>
#include <devLib.h>
#include <epicsMMIO.h>
/* RTN_STATUS is defined in motor.h (typedef enum RTN_VALUES) */
#include "motor.h"
/* Forward-declare omsSetup from drvOms.cc */
extern RTN_STATUS omsSetup(int num_cards, unsigned addrs, unsigned vector,
                           int int_level, int scan_rate);
/*  MVME2100 VME bridge fix (RTEMS 6 / Universe II)                    */
#if defined(__rtems__) && defined(__PPC__)
extern "C" {
#include <bsp.h>          /* defines mot_ppc_mvme2100 via bspopts.h — required
                             before VMEConfig.h for correct BSP_VME_BAT_IDX */
#include <libcpu/bat.h>
#include <libcpu/io.h>
#include <bsp/VMEConfig.h>
#include <bsp/VME.h>
extern "C" {
 * __wrap_vmeUniverseReset — linked via -Wl,--wrap=vmeUniverseReset
 * Prevents the BSP from resetting the Universe II during BSP_VMEInit().
 * vmeUniverseReset() disables all outbound/inbound windows, clears
 * interrupt routing, and sets CRT/U2SPEC — destroying PPCBug's config.
void __wrap_vmeUniverseReset(void)
    printf("*** vmeUniverseReset: WRAPPED — preserving PPCBug defaults ***\n");
 * Override BSP_vme_config (weak symbol in librtemsbsp.a).
 * The default BSP_vme_config calls vmeUniverseReset (our no-op wrapper),
 * then BSP_VMEOutboundPortCfg to reprogram windows at 0x9xxx PCI addresses.
 * Those reprogrammed windows cause S_TA=1 on MVME2100.
 * Our override calls BSP_VMEInit (to set up theOps function table), sets
 * DBAT1 for PPCBug's PCI range, and installs the IRQ manager — but does
 * NOT reprogram any outbound windows.
 * PPCBug factory-default outbound windows (MVME2100 manual Ch.5):
 *   LSI1: PCI 0x81000000-0x9FFFFFFF → VME A32 0x01-0x1F
 *   LSI2: PCI 0xA0000000-0xA1FFFFFF → VME A24 0xF0-0xF1
 *   LSI3: PCI 0xAFFF0000-0xAFFFFFFF → VME A16 0xFFFF
void BSP_vme_config(void)
    /* Fix stdout buffering: Newlib on RTEMS defaults to fully buffered,
     * which causes output from iocsh commands (vmeread, dbl, etc.) to
     * get stuck in the buffer and never appear on the serial console.
     * Switch to line-buffered so output appears after each newline. */
    setvbuf(stdout, NULL, _IOLBF, 0);
    printf("=== Custom BSP_vme_config for MVME2100 ===\n");
    /* BSP_VMEInit finds Universe II on PCI bus (via pci_find_device),
     * stores register base in vmeUniverse0BaseAddr, enables PCI BM+MS,
     * sets theOps = uniOpsRec.
     * Internally calls vmeUniverseReset — our no-op wrapper. */
    if (BSP_VMEInit()) {
        printf("  BSP_VMEInit FAILED!\n");
        return;
    printf("  BSP_VMEInit OK\n");
    /* DBAT1: cover PPCBug A24 (0xA0) + A16 (0xAFFF) PCI range */
    setdbat(BSP_VME_BAT_IDX, 0xA0000000, 0xA0000000, 0x10000000, IO_PAGE);
    printf("  DBAT%d: 0xA0000000/256MB -I-G\n", BSP_VME_BAT_IDX);
    /* DO NOT call BSP_VMEOutboundPortCfg — PPCBug defaults preserved.
     * ANY write to Universe II outbound LSI registers causes S_TA=1. */
    BSP_VMEOutboundPortsShow(0);
    BSP_VMEInboundPortsShow(0);
    /* Install VME interrupt manager */
    printf("  Installing VME IRQ manager...\n");
    if (BSP_VMEIrqMgrInstall() < 0)
        printf("  WARNING: BSP_VMEIrqMgrInstall error\n");
    else
        printf("  VME IRQ manager installed\n");
    printf("=== Custom BSP_vme_config complete ===\n");
 * __wrap_BSP_vme2local_adrs — linked via -Wl,--wrap=BSP_vme2local_adrs
 * The Universe II driver's vmeUniverseXlateAddr cannot reverse PPCBug's
 * translation offsets (32-bit wrapping) and doesn't match the AM code
 * (PPCBug uses SUPER=0, EPICS requests supervisory AM).  We provide
 * hardcoded translations for PPCBug's known window layout.
extern int __real_BSP_vme2local_adrs(
    unsigned long am, unsigned long vmeaddr, unsigned long *plocaladdr);
int __wrap_BSP_vme2local_adrs(
    switch (am) {
    case 0x2D:  /* A16 supervisory */
    case 0x29:  /* A16 non-privileged */
        *plocaladdr = 0xAFFF0000UL + (vmeaddr & 0xFFFF);
        return 0;
    case 0x3D:  /* A24 supervisory */
    case 0x39:  /* A24 non-privileged */
        if (vmeaddr < 0x02000000UL) {
            *plocaladdr = 0xA0000000UL + vmeaddr;
            return 0;
        break;
    case 0x0D:  /* A32 supervisory */
    case 0x09:  /* A32 non-privileged */
        if (vmeaddr >= 0x01000000UL && vmeaddr < 0x20000000UL) {
            *plocaladdr = 0x81000000UL + (vmeaddr - 0x01000000UL);
            return 0;
        break;
    return __real_BSP_vme2local_adrs(am, vmeaddr, plocaladdr);
} /* extern "C" */
/*  Diagnostic helpers (read-only Universe II register access)         */
static volatile unsigned int *universeRegs = NULL;
static inline unsigned int univ_bswap32(unsigned int v)
    return ((v & 0xFF) << 24) | ((v & 0xFF00) << 8) |
           ((v >> 8) & 0xFF00) | ((v >> 24) & 0xFF);
static inline unsigned int univRead(int offset)
    return univ_bswap32(universeRegs[offset / 4]);
static int findUniverseRegs(void)
    volatile unsigned int *candidate = (volatile unsigned int *)0xFCDFF000UL;
    unsigned int id = bswap32(candidate[0]);
    if ((id & 0x0000FFFF) == 0x10E3 || (id & 0xFFFF0000) == 0x10E30000) {
        universeRegs = candidate;
        return 0;
    return 0;
/*  VME diagnostic at startup (registrar, runs before iocInit)         */
static void installFixedVmeProbe(void)
    printf("VME diag: registrar running — PPCBug defaults should be active\n");
    if (findUniverseRegs() != 0) return;
    /* Read-only diagnostics */
        unsigned int pci_csr = univRead(0x004);
        printf("VME diag: PCI_CSR=0x%08X (BM=%d MS=%d S_TA=%d)\n",
               pci_csr, (pci_csr>>2)&1, (pci_csr>>1)&1, (pci_csr>>27)&1);
    /* Show PPCBug's outbound windows */
        int port;
        printf("VME diag: Outbound windows:\n");
        for (port = 0; port < 4; port++) {
            int off = 0x100 + port * 0x14;
            unsigned int ctl = univRead(off);
            if (!((ctl >> 31) & 1)) continue;
            printf("  LSI%d: CTL=0x%08X BS=0x%08X BD=0x%08X TO=0x%08X\n",
                   port, ctl, univRead(off+4), univRead(off+8), univRead(off+0xC));
    /* Quick VME test */
        volatile void *pLocal = NULL;
        unsigned char probeVal = 0;
        long status;
        status = devBusToLocalAddr(atVMEA16, 0xFC00, &pLocal);
        if (status == 0) {
            status = devReadProbe(1, pLocal, &probeVal);
            printf("VME diag: devReadProbe A16 0xFC00 @ %p = %s (0x%02X)\n",
                   pLocal, status == 0 ? "OK" : "FAIL", probeVal);
        } else {
            printf("VME diag: devBusToLocalAddr A16 0xFC00 FAILED\n");
#endif /* __rtems__ && __PPC__ */
/*  omsSetup IOCsh registration                                        */
static const iocshArg omsArg0 = {"num_cards",  iocshArgInt};
static const iocshArg omsArg1 = {"addrs",      iocshArgInt};
static const iocshArg omsArg2 = {"vector",     iocshArgInt};
static const iocshArg omsArg3 = {"int_level",  iocshArgInt};
static const iocshArg omsArg4 = {"scan_rate",  iocshArgInt};
static const iocshArg * const omsArgs[5] = {
    &omsArg0, &omsArg1, &omsArg2, &omsArg3, &omsArg4
static const iocshFuncDef omsSetupFuncDef = {"omsSetup", 5, omsArgs};
static void omsSetupCallFunc(const iocshArgBuf *args)
    omsSetup(args[0].ival, (unsigned)args[1].ival, (unsigned)args[2].ival,
             args[3].ival, args[4].ival);
/*  vmeProbeTest - diagnostic IOCsh command                            */
static void vmeProbeTestFunc(int vmeAddr, int wordSize)
    volatile void *pPhysical = NULL;
    long status;
    unsigned char  val8  = 0;
    unsigned short val16 = 0;
    unsigned int   val32 = 0;
    int ws;
    printf("vmeProbeTest: VME A16 address 0x%04X, wordSize=%d\n",
           (unsigned)vmeAddr, wordSize);
    status = devBusToLocalAddr(atVMEA16, (size_t)vmeAddr, &pPhysical);
    if (status != 0) {
        printf("  devBusToLocalAddr FAILED, status=%ld\n", status);
        return;
    printf("  local addr = %p\n", pPhysical);
    for (ws = 1; ws <= 4; ws <<= 1) {
        void *valPtr;
        switch (ws) {
            case 1: valPtr = &val8;  break;
            case 2: valPtr = &val16; break;
            default: valPtr = &val32; break;
        status = devReadProbe(ws, pPhysical, valPtr);
        if (status == 0) {
            switch (ws) {
                case 1: printf("  %d-byte: 0x%02X  OK\n", ws, val8); break;
                case 2: printf("  %d-byte: 0x%04X  OK\n", ws, val16); break;
                default: printf("  %d-byte: 0x%08X  OK\n", ws, val32); break;
        } else {
            printf("  %d-byte: FAIL\n", ws);
static const iocshArg probeArg0 = {"vmeAddr",  iocshArgInt};
static const iocshArg probeArg1 = {"wordSize", iocshArgInt};
static const iocshArg * const probeArgs[2] = { &probeArg0, &probeArg1 };
static const iocshFuncDef vmeProbeTestDef = {"vmeProbeTest", 2, probeArgs};
static void vmeProbeTestCallFunc(const iocshArgBuf *args) {
    int ws = args[1].ival;
    if (ws < 1 || ws > 4) ws = 1;
    vmeProbeTestFunc(args[0].ival, ws);
/*  vmeScanA16 - scan entire A16 space                                 */
static void vmeScanA16Func(void)
    long status;
    int addr, found = 0;
    printf("vmeScanA16: Scanning A16 0x0000..0xFF00 (step 0x100)...\n");
    for (addr = 0; addr < 0x10000; addr += 0x100) {
        status = devBusToLocalAddr(atVMEA16, (size_t)addr, &pPhysical);
        if (status != 0) continue;
        status = devReadProbe(1, pPhysical, &val8);
        if (status == 0) {
            printf("  A16 0x%04X -> %p -> 0x%02X  FOUND\n",
                   addr, pPhysical, val8);
            found++;
    printf("  %d device(s) found.\n", found);
static const iocshFuncDef vmeScanA16Def = {"vmeScanA16", 0, NULL};
static void vmeScanA16CallFunc(const iocshArgBuf *args) { vmeScanA16Func(); }
/*  Registrar                                                          */
static void omsSetupRegistrar(void)
    installFixedVmeProbe();
#endif
    iocshRegister(&omsSetupFuncDef, &omsSetupCallFunc);
    iocshRegister(&vmeProbeTestDef, &vmeProbeTestCallFunc);
    iocshRegister(&vmeScanA16Def, &vmeScanA16CallFunc);
extern "C" {
    epicsExportRegistrar(omsSetupRegistrar);
Best Regards
Mirek
On Mon, Apr 13, 2026 at 2:03 PM Michael Davidsaver <mdavidsaver at gmail.com> wrote:
On 4/13/26 9:34 AM, Johnson, Andrew N. via Tech-talk wrote:
Hi Mirek,
This is good news, could you explain a bit more what you were doing before you found this out?
I assume you restored the PPCBug settings to factory defaults to get it to work?
Is it possible to use any PPCBug settings to configure the VMEbus windows on the Universe-2 chip for RTEMS, or do they have to stay exactly as factory-programmed?
fyi. It turns out that I have a copy of a PPCBug manual (Feb. 2001).  Apparently the "ENV" command can be used to read-back VME bridge settings without updating.
...  If the ENV command is invoked without the D
option, you are prompted to configure all operational parameters. You may
change the displayed value by typing a new value, followed by the Return
key. To leave the field unaltered, press the Return key without typing a
new value.
I’m trying to judge how necessary it might be to change the driver and/or RTEMS BSPs for the VME interface chips to be configurable by IOC applications. Some of that work might need BSP changes or even deeper OS modifications as I don’t know if RTEMS on the MVME boards configures the MMU or PCI host bridge at all, and changing the memory mappings to open windows to the VME could require altering those settings.
- Andrew
Complexity comes for free, Simplicity you have to work for.
On 4/11/26, 5:14 PM, "Tech-talk" <tech-talk-bounces at aps.anl.gov> wrote:
Hi All,
I found the issue with the OMS card. Actually, the issue wasn't with the card itself but with way how the MVME2100 accesses the VME bus.
The  Tundra Universe II initialization in the original RTEMS bsp for mvme2100 does not work for me. At least the  mvme2100 version which I use:
Only PPCBug's factory power-on configuration - written by firmware during hardware initialization, never touched by software - produces working VME cycles.
If anybody is facing a similar problem with mvme2100 I have the wrapper functions that fix the issue.
Best Regards
Mirek
On Fri, Apr 3, 2026 at 9:33 PM Miroslaw Dach <mdach at lbl.gov> wrote:
Hello,
I am struggling with converting the mvme2100 with OMS VS4 card system from vxWorks 5.1 to RTEMS 6 EPICS 7.0.10.
For RTEMS, I use the same omsSetup() call in the startup script as in vxWorks:
Welcome to rtems-6.0.0 (PowerPC/Generic (classic FPU)/mvme2100) on MVME 2100
pci : Configuring interrupt routing for 'MVME 2100'
pci : Device 0:0x00:0 routed to interrupt_line 16
pci : Device 0:0x0d:0 routed to interrupt_line 23
pci : Device 0:0x0e:0 routed to interrupt_line 17
Cleared PCI errors: pci_stat was 0x20a0
WARNING: unable to allocate page table, keeping DBAT0
WARNING: unable to setup page tables VME bridge must share PCI space
OpenPIC Version 1.2 (1 CPUs and 24 IRQ sources) at 0x4242800640
OpenPIC Vendor 0 (Unknown), Device 0 (Unknown), Stepping 1
Overriding NumSources (24) from configuration with 16
OpenPIC timer frequency is 8333344 Hz
BATs must not overlap; area 0x002e0000..0x00300000 hits DBAT 0
Universe II PCI-VME bridge detected at 0xfcdff000, IRQ 23
Universe Master Ports:
Port  VME-Addr   Size       PCI-Adrs   Mode:
0:    0x20000000 0x0e000000 0x90000000 A32, D64 [MBLT], Dat, Sup
1:    0x00000000 0x00ff0000 0x9f000000 A24, D64 [MBLT], Dat, Sup
2:    0x00000000 0x00010000 0x9fff0000 A16, D64, Dat, Sup
Universe Slave Ports:
0:    0xc0000000 0x02000000 0x00000000 A32, Pgm, Dat, Sup, Usr, PWEN, PREN
Registering /dev/console as minor 0 (==/dev/ttyS0)
ntpd: RTEMS service init
# omsSetup(nCards, baseAddress, intVectBase, intLevel, pollRate)
var drvOMSdebug 9
omsSetup(1, 0xFC00, 180, 5, 10)
iocInit()
Starting iocInit
## EPICS R7.0.10.1
## Rev. R7.0.10
## Rev. Date Git: 2026-03-14 20:49:46 -0700
recGblRecordError: motor_init_record_com(): card does not exist! Illegal field value PV: LS1:MOT
vmeread 0xfc00 16 8 16
Reading from 0x0000fc00 A16 D8
Mapped to 0x9ffffc00
0x00000000 00000000 00000000 00000000 00000000
*** Bus errors occurred ***
The vxWorks based system with the same hardware works fine and shows:
localAddr = 0
new symbol "localAddr" added to symbol table.
sysBusToLocalAdrs(0x2d, 0xFC00, &localAddr)
value = 0 = 0x0
d localAddr, 0x10, 1
f1fffc00:  ff 0d ff 00 ff b0 ff 48 ff b4 ff ff ff ff ff ff   *.......H........*
value = 21 = 0x15
It looks like VxWorks handles VME memory mapping differently than RTEMS, but devLib2 should manage all of that properly.
I use the following modules:
MOTOR        = $(SUPPORT)/motor-R7-3-1
DEVLIB2      = /usr/local/epics/R7.0.10/modules/soft/devlib2
EPICS_BASE   = /usr/local/epics/R7.0.10/base
I found a similar post on the tech talk:
I am wondering what the solution was?
Best Regards
Mirek
RTEMS/MVME2100 OMS58 card does not exist! error Miroslaw Dach via Tech-talk
Re: RTEMS/MVME2100 OMS58 card does not exist! error Miroslaw Dach via Tech-talk
Re: RTEMS/MVME2100 OMS58 card does not exist! error Johnson, Andrew N. via Tech-talk
Re: RTEMS/MVME2100 OMS58 card does not exist! error Michael Davidsaver via Tech-talk
Prev: Re: RTEMS/MVME2100 OMS58 card does not exist! error Michael Davidsaver via Tech-talk
Next: A new E-Traveler platform with modern tech stack Wang, Lin via Tech-talk
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  <2026>
Next: Turning off despeckle filter in ADKinetix Prabudhya Bhattacharyya via Tech-talk
ANJ, 16 Apr 2026 · Home · News · About · Talk · Base · Modules · Extensions ·