1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  <2023>  2024  2025  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  <2023>  2024  2025  2026
Subject: Re: timestampField not retrieved using CA provider for pvac::ClientChannel
From: Michael Davidsaver via Tech-talk <[email protected]>
To: "Bisegni, Claudio" <[email protected]>
Cc: "[email protected]" <[email protected]>, "Veseli, Sinisa" <[email protected]>
Date: Sun, 7 May 2023 22:20:57 -0700
On 5/6/23 11:11, Bisegni, Claudio wrote:
Hi Michael,
Unfortunally my application could talk with PVs that are hosted only on CA protocol. So, I have to manage also this case.
I hope that you can encourage those responsible for these PVs to upgrade at some point soon.
I could execute two monitor on CA PV for listen over two different set of fields and recombine the full structure on client side.
Could be a clean solution?
"Clean" is relative.  The usual way this is handled (assuming a fairly recent epics-base)
is to create two subscriptions.  In the language of libca, one with a DBR_TIME_* type and
event mask DBE_VALUE|DBE_ALARM, and a second with a DBR_CTRL_* type and DBE_PROPERTY.
The dbf_type_to_DBR_*() type promotion macros can help here.
Once both subscriptions have delivered at least the initial update, then you will have a
complete view.  There are some edge cases since these two CA subscriptions may not be as
well synchronized as the one PVA subscription would be.  However, in practice this probably
won't be an issue since DBE_PROPERTY updates are relatively rare.
Of course, if you have an older epics-base where libca doesn't (fully) support DBE_PROPERTY,
the conventional thing to do is to issue a one time GET for DBR_CTRL_* instead of a
second subscription.
Claudio
*Date: *Saturday, 6 May 2023 at 08:28
*To: *Bisegni, Claudio <[email protected]>
*Cc: *Andrew Johnson <[email protected]>, Veseli, Sinisa <[email protected]>, [email protected] <[email protected]>
*Subject: *Re: timestampField not retrieved using CA provider for pvac::ClientChannel
Hello Claudio,
On 5/5/23 12:50, Andrew Johnson via Tech-talk wrote:
The Channel Access network layer can only retrieve specific combinations of metadata corresponding to the DBR types which the native CA code implements, so it's not possible to fetch the timeStamp and alarm data at the same time as other kinds of data such  as the display, control or alarm limits. Unfortunately the CA provider doesn't give any errors if you ask for a combination that it can't provide, it just gives you a subset of the metadata — it might be better if it threw an exception, but it doesn't. I think that fully explains your symptoms.
I think it is worth emphasizing that this CA protocol limitation is _not_ shared by the PVA protocol.
What you tried should work if you were able to use the PVA tools with their native PVA protocol.
Have you tested to see if the PVs you are interested in are also served over PVA?
- Andrew
On 5/5/23 2:44 PM, Bisegni, Claudio wrote:
Should be a my mistake because if I use the pvget with the specific parameter it work:
developer@e058f4bc2090:/workspace$ pvget -p ca -r "field(field,timeStamp,alarm)" ca:variable:a
ca:variable:a structure
    alarm_t alarm
        int severity 0
int status 0
  string message
    time_t timeStamp 2023-05-05 19:22:59.521
        long secondsPastEpoch 1683314579
int nanoseconds 521323913
int userTag 0
I am digging into it.
Anyway, thank you for point me in the right direction
Claudio
*Date: *Friday, 5 May 2023 at 12:24
*To: *Andrew Johnson <[email protected]>, Veseli, Sinisa <[email protected]>, [email protected] <[email protected]>
I fixed the error reported by Sinisa (the reference to other channel for the sum)
I changed the pv name and not I’m trying to get data from ca:variable:a(below) could be some problem on c++ layer of pvData?
I done:
caput ca:variable:a.PROC 1
and the I got:
structure
    double value 1
alarm_t alarm
int severity 0
int status 0
string message
    display_t display
double limitLow 0
double limitHigh 0
string description
        string format F6.0
string units
    control_t control
double limitLow 0
double limitHigh 100
double minStep 0
valueAlarm_t valueAlarm
boolean active false
double lowAlarmLimit nan
double lowWarningLimit nan
double highWarningLimit nan
double highAlarmLimit nan
 int lowAlarmSeverity 0
int lowWarningSeverity 0
int highWarningSeverity 0
int highAlarmSeverity 0
byte hysteresis 0
regards
Claudio
*Date: *Friday, 5 May 2023 at 11:46
*To: *Bisegni, Claudio <[email protected]>, Veseli, Sinisa <[email protected]>, [email protected] <[email protected]>
Hi Claudio,
Please try running the command
     caput ca:variable:sum.PROC 1
and then run your client again. I suspect that the timestamp of your CALC record is still undefined i.e. 0,0 and that's why it isn't getting displayed.
- Andrew
On 5/5/23 1:23 PM, Bisegni, Claudio via Tech-talk wrote:
     Hi yes sure:
     This is my code:
     auto ca_provider = std::make_unique<pvac::ClientProvider>("ca", epics::pvAccess::ConfigurationBuilder().push_env().build());
     auto channel = std::make_shared<pvac::ClientChannel>(provider.connect(pv_name, opt));
     auto op = channel->get(this, pvd::createRequest("field()"));
     I need to specify that I connect to a IOCs launched with “sofIOC -d xyz.db” this si the content of the DB:
     record(ao, "ca:variable:a")
         field(DRVH, "100")
         field(DOL, "0")
         field(PINI, "YES")
     record(ao, "ca:variable:b")
         field(DRVH, "100")
         field(DOL, "0")
         field(PINI, "YES")
     record(calc, "ca:variable:sum")
         field(SCAN, "1 second")
         field(INPA, "variable:a")
         field(INPB, "variable:b")
         field(CALC, "A+B")
     I got this as answer on get operation:
     structure
         double value 3
         alarm_t alarm
             int severity 0
             int status 0
             string message
         display_t display
             double limitLow 0
             double limitHigh 0
             string description
             string format F6.0
             string units
         control_t control
             double limitLow 0
             double limitHigh 0
             double minStep 0
             byte hysteresis 0
     *From: *Veseli, Sinisa <[email protected]> <mailto:[email protected] <mailto:[email protected]>>
     *Date: *Friday, 5 May 2023 at 11:05
     *To: *Bisegni, Claudio <[email protected]> <mailto:[email protected] <mailto:[email protected]>>, [email protected] <[email protected]> <mailto:[email protected] <mailto:[email protected]>>
     I am not sure I understand what is going on. Perhaps you could include your code so we can see what exactly you are trying to do?
     Also, if you'd like to get all fields, the string 'field()' or just empty string '' should work:
     >>> print(c.get('field()'))
     epics:nt/NTEnum:1.0
     enum_t value
       int index 0
       string[] choices [Off, On]
     alarm_t alarm
       int severity 0
       int status 0
       string message
     time_t timeStamp
       long secondsPastEpoch 1683129086
       int nanoseconds 834120910
       int userTag 0
     Siniša Veseli
     Scientific Software Engineering & Data Management
     Advanced Photon Source
     Argonne National Laboratory
     [email protected]
     (630)252-9182
     *From:*Bisegni, Claudio <[email protected]> <mailto:[email protected] <mailto:[email protected]>>
     *Sent:* Friday, May 5, 2023 12:56 PM
     *To:* Veseli, Sinisa <[email protected]> <mailto:[email protected] <mailto:[email protected]>>; [email protected] <[email protected]> <mailto:[email protected] <mailto:[email protected]>>
     *Subject:* Re: timestampField not retrieved using CA provider for pvac::ClientChannel
     I tried to set request field as:
     pvd::createRequest("field(value, timeStamp, alarm, display, control, valueAlarm)
     but I didn’t get the timeStamp Field in PVStructure. The request field work well for all other (for CA provider) for the PVA provider, instead, I always got every field also if I put only “filed(value)” in the request.
     I’m wrong somenting….
     Claudio
     *From: *Bisegni, Claudio <[email protected]> <mailto:[email protected] <mailto:[email protected]>>
     *Date: *Friday, 5 May 2023 at 10:44
     *To: *Veseli, Sinisa <[email protected]> <mailto:[email protected] <mailto:[email protected]>>, [email protected] <[email protected]> <mailto:[email protected] <mailto:[email protected]>>
     Thank you for yor help, is effect on get operation I use the default request, so Iat this point I need to use pvd::createRequest("field(value, timeStamp, alarm, display, control, valueAlarm)"));to retrieve all field
     I’ll give a try!
     Thank you very much.
     Claudio
     *Date: *Friday, 5 May 2023 at 09:55
     *To: *[email protected] <[email protected]> <mailto:[email protected] <mailto:[email protected]>>, Bisegni, Claudio <[email protected]> <mailto:[email protected] <mailto:[email protected]>>
     I am not sure exactly why is there a difference in behavior between caget and pvget, but you should be able to get timestamps if you request them explicitly:
     $ caget -a S01-DAQTBT:Strm:DataCaptureC
     S01-DAQTBT:Strm:DataCaptureC   2023-05-03 10:51:26.834121 Off
     PV request string like field(value,timeStamp) should work from C++:
     >>> from pvapy import *
     >>> c = Channel('S01-DAQTBT:Strm:DataCaptureC', CA)
     >>> print(c.get('field(value,timeStamp)'))
     epics:nt/NTEnum:1.0
       enum_t value
           int index 0
       time_t timeStamp
           int userTag 0
     Hope this helps,
     Sinisa
     Siniša Veseli
     [email protected]
     (630)252-9182
     *From:*Tech-talk <[email protected]> <mailto:[email protected] <mailto:[email protected]>> on behalf of Bisegni, Claudio via Tech-talk <[email protected]> <mailto:[email protected] <mailto:[email protected]>>
     *Sent:* Friday, May 5, 2023 11:39 AM
     *To:* [email protected] <[email protected]> <mailto:[email protected] <mailto:[email protected]>>
     *Subject:* timestampField not retrieved using CA provider for pvac::ClientChannel
     Hello EPICS tech,
     I'm working with last epics release and I have notice that making a get operation for an IOC using a CA provider for crete a pvac::ClientChannel I don't receive the timestamp field:
     I noticed that also the official caget has the problem:
     developer@e058f4bc2090:/workspace$ pvget variable:sum
     variable:sum 2023-05-05 16:34:54.607 3
     developer@e058f4bc2090:/workspace$ caget variable:sum
     variable:sum 3
     so is a client library problem or a server or something else?
     Thank
     you in advance.
     Claudio
Complexity is free, it's Simplicity that takes work.
Re: timestampField not retrieved using CA provider for pvac::ClientChannel Bisegni, Claudio via Tech-talk
Re: timestampField not retrieved using CA provider for pvac::ClientChannel Veseli, Sinisa via Tech-talk
Re: timestampField not retrieved using CA provider for pvac::ClientChannel Andrew Johnson via Tech-talk
Re: timestampField not retrieved using CA provider for pvac::ClientChannel Michael Davidsaver via Tech-talk
Prev: Job Opportunity at the Canadian Light Source Gillian Black via Tech-talk
Next: Re: Why does libca.so depend on libreadline.so? Michael Davidsaver via Tech-talk
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  <2023>  2024  2025  2026
Prev: Re: timestampField not retrieved using CA provider for pvac::ClientChannel Bisegni, Claudio via Tech-talk
Next: Re: timestampField not retrieved using CA provider for pvac::ClientChannel Bisegni, Claudio via Tech-talk