Dear Florian,
the isegHAL-remote library provides a blocking access to the isegHAL
over TCP/IP (secured with SSL sockets). Therefore all data has to be
serialized.
It *should* be possible to open multiple TCP connections by using
multiple iseg_connects() with unique session names and then concurrently
operate on these unique sessions names with iseg_getItem() and
iseg_setItem().
But if the same session is reused, all isegHAL-remote calls must be
serialized.
When the IOC is running on the crate controller, you use the
isegHAL-client library, and there the access is already protected with
mutexes.
But of course the access is much faster here, because no network is
involved.
Best regards,
Andre
Am 30.05.2023 um 16:06 schrieb Florian Feldbauer:
Hey Andre,
is there any difference between the isegHAL-remote library and the
"local" isegHAL library in regards to the iseg_getItem method?
I'm a bit puzzled that there is no issue with threading in the IOC
running on the crate controller itself because both IOCs use the same
device support module. Just a different version of the isegHal library
beneath.
Cheers,
Florian
On 5/30/23 15:59, André Hartmann wrote:
Dear Yann,
I've never used the isegHAL-Remote library with multiple threads, and
I assume that threads can cause problems.
What you could try is to put a mutex around writeReadSynchronous(),
but note that this will block all other threads during the network
operation (which can fail, so in worst case it blocks for the timeout).
I hope that helps.
Best regards,
Andre
Am 29.05.2023 um 10:10 schrieb Florian Feldbauer:
Hey Yann,
as for the AO record: The initialization failes, because the
"Quality" of the iseg items returns 000 (=ISEG_ITEM_QUALITY_INVALID)
Maybe the module you are using does not support setting the RampSpeed
and thus the isegHal return INVALID qualtiy for it.
That would explain, why it is the only record producing this error.
As for the output recrods being monitored via the polling thread:
The isegHAL offers multiple interfaces to the outside world, like the
EPICS interface (devIsegHal), the browser-based frontend from ISEG,
and some others. So Voltages/Currents and all other configurations
can be changed outside of EPICS and we wanted to have a way,
that our EPICS records stay "synced" with the settings of the module.
It can be quite confusing, when the measured voltage is 0 although
the set voltage is 1.2kV and the setOn bit is set in EPICS (but not
in the module).
Or why the set voltage shows 1.2kV but the HV only goes up to 1.0kV
(because someone set Vset to 1.0kV via another interface).
Unfortunately I cannot answer your question about Qt.
Personally I do not use Qt in any of my own projects and we usually
use the IOCs running on the crate controllers for our isegHV modules.
With them we never had this issue.
Best regards,
Florian
Am 28.05.2023 um 23:13 schrieb Yann Mandza via Tech-talk:
Hello all,
I have a ICSMINI 2 box from iseg am trying to interface to via the
HAL services utilities using HAL service libraries from iseg, I was
able to recompiled for my system.
So far I tried to connect to the box using the provided example and
it worked.
However I tried the epics module from florian
https://github.com/ffeldbauer/devIsegHal
<https://github.com/ffeldbauer/devIsegHal>, so i could disable the
embedded IOC and run mine outside.
I was able to compiled the module and get the IOC connected to the
box. However I’m having two main issues.
1 - error at record initialization for AO record
ISEG:7200081:0:32:VoltageRampSpeed: Error while reading item
property '0.32.VoltageRampSpeed' (Q: 000)
recGblRecordError: ao: init_record Error (514,11) PV:
ISEG:7200081:0:32:VoltageRampSpeed ( is this a result from the
above? Then sorting out the above should clear this error)
it’s strange because only AO return this error.
I read on the deviseghal GitHub page that AO record were added to
the polling thread that also handle I/O interrupt.
I wonder why that would be?
This excerpt is from the page
“It is possible that control parameters change during operation. For
example, if a trip occurs the corresponding setON bit in the channel
control register will be set to 0.
These changes are monitored by devIsegHal through a polling thread.
Each output record (except for stringout records) is automatically
registered to this thread and their values are checked for updates
on the isegHAL.
If a value has changed the VAL field and timestamp of the record
will be set to the new values”
I’m a bit confused why output record would need that.
2- Qt Socket Library return this warning “QSocketNotifier: Socket
notifiers cannot be enabled or disabled from another thread", it
seems the IOC starts several threads to read/write information to
the iseg system.
One can trace this issues till the writeReadSynchronous method
within the isegHal-service library. I used QT5 to compile the library.
Can the device support protect against this or should iseg update
their code to handle multi thread access to the socket?
Yann
André Hartmann, Dipl.-Ing. (FH)
Softwareentwicklung / Software Development
E-Mail: [email protected]
Tel: +49.351.26996-43 | Fax: +49.351.26996-21
iseg Leistungselektronik Sachsen GmbH
www.iseg-hp.de
Bautzner Landstr. 23, 01454 Radeberg / Rossendorf, Germany
Geschäftsführer / Managing directors: Dr. Tanju Gleisberg, Dr. Martin
Rentzsch
Amtsgericht / Lower district court: Dresden HRB 33563
Diese E-Mail enthält vertrauliche und/oder rechtlich geschützte
Informationen. Wenn Sie nicht der richtige Adressat sind oder
diese E-Mail irrtümlich erhalten haben, informieren Sie bitte
sofort den Absender und vernichten Sie diese Mail. Das unerlaubte
Kopieren sowie die unbefugte Weitergabe dieser Mail ist nicht
gestattet.
This e-mail may contain confidential and/or privileged information.
If you are not the intended recipient (or have received this e-mail
in error) please notify the sender immediately and destroy this e-mail.
Any unauthorized copying, disclosure or distribution of the material
in this e-mail is strictly forbidden.
- Issues with epics module and ISEG HAL service libraries Yann Mandza via Tech-talk
- Re: Issues with epics module and ISEG HAL service libraries Florian Feldbauer via Tech-talk
- Re: Issues with epics module and ISEG HAL service libraries André Hartmann via Tech-talk
Re: More precise timing for record processing Sobhani, Alex via Tech-talk
<2023>
2025
2026
<2023>
2025
2026