1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  <2025>  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  <2025>  2026
Subject: RE: Handling Polling and CPU Bottleneck for Large Image Capture in areaDetector
To: "Kim, Kuktae" <[email protected]>, tech-talk <[email protected]>
Date: Mon, 28 Apr 2025 16:37:09 +0000
Hi Kuktae,
  • We set up Polling features (ReadStatus.SCAN) so users can monitor sensor and device temperatures for interlocks. However, polling seems to block other tasks and lowers the image capture rate from ~12 Hz to ~7 Hz.
The asyn port is locked during the poll cycle, so that will block the read thread and slow down acquisition.  ReadStatus.SCAN reads all the GenICam features which you probably don’t need to do.
  • We added the SCAN field of GC_SensorTemperature and GC_DeviceTemperature records as SCAN = "1 second", but now get warnings like "GC_SensorTemperature is not writable" in the IOC shell.
    Could you advise us on how to properly handle this situation?
You should be setting the SCAN field of GC_SensorTemperature_RBV and GC_DeviceTemperature_RBV.  Those are the input records.
  • According to the manual, the camera can capture ~11 Hz (standard mode) and ~26 Hz (high-speed mode). We currently reach up to ~13 Hz, but plugin threads are hitting 99.9% CPU usage.
    Is there a way to use multithreading for plugins to reduce CPU load and increase the capture rate?
You should first disable all the plugins and see how fast you can acquire with no plugins running.  Or you can simply disable ArrayCallbacks in the driver.
Then you should determine which plugins really need to run at the full frame rate, and which can run at reduced rate.  For the ones that are OK to run at 5 Hz, for example, you can set the MinCallbackTime to 0.2 seconds.
For plugins that need to run at the full rate you can run most of them with multiple threads.  In the startup script you set MaxThreads and at runtime you can set NumThreads.  If NumThreads > 1 then you need to think about the fact the the NDArrays can be processed out of order.  You can then enable sorting if that is an issue.  This is all documented here:
https://areadetector.github.io/areaDetector/ADCore/NDPluginDriver.html
The release notes for ADCore R3-0 lists the plugins that support multiple threads and those that do not.
https://github.com/areaDetector/ADCore/blob/master/RELEASE.md
What is the camera model?
Can you send screenshots of the OPI screen for the detector and for the ROI1 and CC1 plugins?
From: Tech-talk <[email protected]> On Behalf Of Kim, Kuktae via Tech-talk
Sent: Monday, April 28, 2025 10:00 AM
To: tech-talk <[email protected]>
Subject: Handling Polling and CPU Bottleneck for Large Image Capture in areaDetector
Hello,
We are developing an areaDetector IOC for a large sensor camera (6144 x 6144, 16-bit depth, ~75 MB per image) and would like some advice on a few issues:
  1. Polling Impact:
  1. CPU Bottleneck:
Photos are attached for reference:
< CPU Bottleneck: Some plugin threads reach up to 99.9% CPU usage, limiting the capture rate >
< Polling Impact: Polling causes overall CPU usage to decrease and lowers image capture performance >
Thank you for your help!
Best regards,
Kuktae
  1.
Re: Handling Polling and CPU Bottleneck for Large Image Capture in areaDetector Kim, Kuktae via Tech-talk
Prev: Re: Waveform to single ai (again) Érico Nogueira Rolim via Tech-talk
Next: FYI, the msys2 compilation approach is not working Wang, Andrew via Tech-talk
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  <2025>  2026
Prev: Handling Polling and CPU Bottleneck for Large Image Capture in areaDetector Kim, Kuktae via Tech-talk
Next: Re: Handling Polling and CPU Bottleneck for Large Image Capture in areaDetector Kim, Kuktae via Tech-talk