1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  <20092010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  <20092010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026
Subject: RE: How to get array and float at the same time by sscan
To: "??" <[email protected]>, <[email protected]>
Date: Fri, 17 Apr 2009 06:18:27 -0500
> I haven't set preset real time and the value is 0. Normally I use 0.5s to counting at each point.
> But in order to confirm whether scan1.DnnPV later than scanH.DnnPV, I use 5s to counting at each point.
This is the problem.  The sscan record will not go on to the next point until each detector is "done".  That means that the putCallback to the detector trigger has completed its callback, which happens when the record calls recGblFwdLink.  For the MCA record that does not happen until acquisition stops.  If you did not set a preset time then I don't understand how the sscan record was ever going on to the next point.  Had you set some other completion mechanism, such as preset counts in one or more ROIs?
> You are right. The status and data read update rates affect the time of reading mca. Now, I set all of them as 0.1s.
The status read update controls the rate at which the mca record polls the hardware to see if acquisition is complete (as well as other values such as elapsed times, etc.).  The read data rate is used to control reading the spectra.  However, it is only needed to see the intermediate spectra before acquisition is done.  The mca record always reads the data when acquisition completes, to read data can be Passive, and the spectra will still be updated when acquisition completes.  It really slows things down to set read data to 0.1s, you should not do this.  That is not your problem, you will still have data that is not in sync.  You need to figure out why the scan is going to the next point when the preset times are 0, and set if it works correctly when read data is passive, read status is 0.1s and you set a preset real time of 10 s.  Then you can increase read data to 1 second or something so operators get an update at a reasonable rate.
Sent: Fri 4/17/2009 3:00 AM
To: Mark Rivers; [email protected]
Subject: Re: How to get array and float at the same time by sscan
Hi Mark,
>What are you setting as the preset real time (or live time) for the MCA?  How long should the MCA be counting at each point?
I haven't set preset real time and the value is 0. Normally I use 0.5s to counting at each point. But in order to confirm whether scan1.DnnPV later than scanH.DnnPV, I use 5s to counting at each point.
>What have you set for the status and data read update rates in the MCA record?
Thank you very much!
Best wishes,
Yan Fen
----- Original Message -----
To: "??" <[email protected]>; <[email protected]>
Sent: Friday, April 17, 2009 12:00 PM
Sent: Thu 4/16/2009 9:51 PM
Hi Mark,
In order to confirm whether scan1.DnnPV later than scanH.DnnPV, I set scan1.DDLY as 0 and scanH.DDLY as 5.
The order of operations: (I just watch the interface)
move motor
mda.VAL turn to 0 and mda.ERTM turn to 0
after about 5 seconds, scanH.CPT turn to 2048
after about 1 second, all the scan1.DnnCV, scan1.P1DV and scan1.R1CV change values
then mda.VAL turn to 0 and mda.ERTM from about 6 to 0
The process above mentioned just by watching. May be the truth is others. Would you mind give me some suggestions?
Thank you!
Best wishes,
Yan Fen
To: [email protected] ; [email protected]
Sent: Thursday, April 16, 2009 9:34 PM
> Now, I want to save the ADC.VAL, DAC.VAL and mca.VAL at the same time. But I can't get it.
> I find the data of ADC is later than the data of mca evenif  I set scan1.DDLY as 0. So please tell me how to gain the float data and the array data at the same time.
How much later is the ADC than the MCA?  What times are you looking at to determine this?
> I find the value of RnROI field are different from the value of specture in the saved file.
> That means scan1.DnnPV and scanH.DnnPV be saved in the different time even if I set scan1.DDLY as 0 second.
> So would you mind tell me how to set the value of scanH.DDLY, scan1.PDLY,  scan1.DDLY and scan2.PDLY?
>  I don't know wheater they should gain in the same time or not.
I don't understand how the RnROI field could be different from that of the spectra.  The RnROI field is computed when the MCA stops acquiring, which is also when the sscan record reads both the spectra and the ROIs.
What is the evidence that the ROIs are incorrect?
From: [email protected] [mailto:[email protected]] On Behalf Of ??
Sent: Wednesday, April 15, 2009 9:12 PM
Subject: How to get array and float at the same time by sscan
Hi everyone,
I set up a 2-D scan as follows:
scan1 record: Drive is X motor VAL field, trigger is scanH.EXSC field. I set its detectors to be ADC.VAL, DAC.VAL and so on.
scanH record: No positions. Detector trigger is the field for starting  MCA. The detector is mca.VAL which is the array .
Now, I want to save the ADC.VAL, DAC.VAL and mca.VAL at the same time. But I can't get it. I find the data of ADC is later than the data of mca evenif  I set scan1.DDLY as 0. So please tell me how to gain the float data and the array data at the same time.
Thank you!
Best wishes.
Yan Fen
Yan Fen
Shanghai Synchrotron Radiation Facility
Shanghai Institute of Applied Physics
Chinese Academy of Sciences
Zhangheng Rd 239,Pudong Dist.,
Shanghai 201204
China
Tel:+86-21-33933191
How to get array and float at the same time by sscan 闫芬
RE: How to get array and float at the same time by sscan Mark Rivers
Re: How to get array and float at the same time by sscan 闫芬
Re: How to get array and float at the same time by sscan ??
Prev: Re: How to get array and float at the same time by sscan ??
Next: record processing delay Patrick Thomas
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  <20092010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  2026
Next: How to get array and float at the same time by sscan 闫芬
ANJ, 31 Jan 2014 · Home · News · About · Talk · Base · Modules · Extensions ·
