Hi Peter,
You are loading a local .db file, sum.db. I don't know what that files contains. But it clearly is setting the .OUT field of the motor record to the string "@asyn(test-se1-1, 0)".
That is not correct, again because you are not using asyn device support for the motor record. You need to set the .OUT field of the motor record to a string like the following:
field(OUT,"#C0 S0 @")
C0 means "card 0", i.e. the first PM304 card in the system. S0 means signal 0, i.e. the first axis on this controller. That is the syntax used by all Model 1 motor drivers.
I think you are basing your database on asyn_motor.db. That database can only be used for Model 2 and Model 3 drivers. For Model 1 drivers like the one you are using you must use motor.db, not asyn_motor.db.
Sent: Wednesday, December 11, 2013 10:24 PM
To: Mark Rivers; [email protected]
Subject: RE: Mclennan PM600 motor controller
Hi Mark
Thank you very much for the assistance. As you may have gathered, EPICS is not a full time task for me, so assistance is much appreciated.
Indeed I am using the Mclennan driver. Your direction allowed me to progress greatly, but I am at another stumbling block. I have managed to communicate and move the motor through an asynRecord device by simply connecting to port "test-se-1" and address "0". The .OEOS field must be set to "\r\n" for the commands to actually be acted upon (otherwise it just reads back the input command). However, when I try connecting through a motor record, I get the trace below on IOC start up:
# Test SE (test-se1)
drvAsynIPPortConfigure("test-se1-1", "172.16.0.108:5300")
# Add these lines for asynTrace debugging
asynSetTraceIOMask("test-se1-1",0,2)
asynSetTraceMask("test-se1-1",0,9)
# Test for Mclennan PM600 stepper motor controller
# PM304Setup(controller count, poll rate (Hz))
# PM304Config(card being configured, asyn port name, number of axes)
PM304Setup(1,5)
PM304Config(1, "test-se1-1", 1)
## Load record instances
dbLoadRecords("../../db/sum.db","INST=pitest")
Can't set "test:buncher:motor.OUT" to "@asyn(test-se1-1, 0)"
Error at or before ")" in file "../../db/sum.db" line 34
iocInit()
Starting iocInit
## EPICS R3.14.12.3 $Date: Mon 2012-12-17 14:11:47 -0600$
## EPICS Base built Aug 13 2013
Illegal field value PV: test:buncher:motor motor_init_record_com(): card does not exist!
I'm aware the last line means there is a communication issue, but I have no idea about the "Can't set to @asyn" line. I assume from asyn_motor.db that that is correct format for that field. In any case, I have tried a few different ways, but no dice.
The full db record is:
# Test record for Mclennan PM600 Stepper Motor driver
record(motor, "test:buncher:motor") {
field(DTYP, "Mclennan PM304")
field(OUT, "@asyn(test-se1-1, 0)")
field(RRES, "1.0")
field(PREC, "0")
field(TWV, "10")
field(VELO, "50")
Regards
Peter
-----Original Message-----
Sent: Wednesday, 11 December 2013 4:04 PM
To: Peter Linardakis; [email protected]
Hi Peter,
I assume you are using the McLennan driver that is included in synApps/motor/motorApp/MclennanSrc/?
If so, then the problem is the DTYP field in the motor record in your database. You specified
field(DTYP, "asynMotor")
DTYP "asynMotor" is only for Model 2 and 3 motor drivers, where the communication between the motor record device support and the driver uses the standard asyn interfaces.
The Mclennan driver is an old Model 1 driver. This is the correct definition of DTYP for the McLennan motors.
You can find that required string ("Mclennan PM304") by looking in the DBD file your application is including:
pitest_DBD += devMclennanMotor.dbd
That file, motorApp/MclennanSrc/devMclennanMotor.dbd, contains the following line:
device(motor,VME_IO,devPM304,"Mclennan PM304")
That tells you what the device support string name is.