Michael:
We've been using the DTM-151's with stream device support successfully.
I am glad to hear that we are not the only ones to encounter problems
with the proprietary network scheme, which we implemented initially
using fibre for three devices. After fighting with the system for many
months, I concluded that the protocol implementation in the DTM-151 is
faulty.
Since connecting each to a dedicated serial port on our IOC, they have
been problem free. Our implementation is pretty much identical to
Stephanie's. We included support for range selection and RESET, which is
also quite simple. Ask me if you need example code.
Rod Nussbaumer
TRIUMF,
Vancouver, Canada.
On 03/30/2015 01:25 PM, Allison, Stephanie wrote:
Hi Michael,
We have integrated the DTM-151 teslameter ... somewhat. We found it best to connect each one directly to the terminal server. These teslameters are not so tolerant of radiation. We only got 2 out of 5 working properly (the other 3 were in higher radiation places).
The proto file:
# Stream device protocol file for Group3 Hall Probe
# Harvey Rarback
# July 31, 2009
WriteTimeout = 10000;
ReadTimeout = 10000;
ReplyTimeout = 10000;
ExtraInput = Ignore;
OutTerminator = "";
InTerminator = CR;
field {
out "F\r"; # send request
# in "A%*i"; # fiber interface echo
in "%f"; # device returns field
temp {
out "T\r"; # send request
in "%f"; # device returns temperature
The database:
record(ai, "$(DEV):MagField") {
field(DESC, "Hall Probe field")
field(DTYP, "stream")
field(INP, "@hallprobe.proto field $(PORT)")
field(SCAN, "$(FIELD_SCAN_RATE)")
field(EGU, "T")
field(PREC, "5")
field(MDEL, "$(MDEL)")
field(PINI, "YES")
record(ai, "$(DEV):MagTemp") {
field(DESC, "Hall Probe temperature")
field(INP, "@hallprobe.proto temp $(PORT)")
field(SCAN, "$(TEMP_SCAN_RATE)")
field(EGU, "C")
field(PREC, "1")
field(PINI, "YES")
I have some notes from Harvey Rarback who has since retired:
Teslameter Jumper, Dip Switch, and Rotary Switch Settings:
Refer to the user manual, and use defaults for option and handshake jumper pins. If the teslameter is connected directly to the terminal server or PC serial port, set T/M pins to terminal mode. If teslameter is connected to the DPM display, set them to modem mode (default). Set rotary switch to "E" (9600 baud). Set S1-1 to S1-5 to the device address. To choose 8 data bits, no parity, and 1 stop bit, set S1-6 to On, S1-7 to Off, and S1-8 to On. Set S2-1 to Off (on-demand transmission), S2-2 to S2-4 to On (select terminator = carriage return, double terminator = enabled, echo = On), and S2-5 to S2-8 to Off (units = tesla, no units symbol, no filtering, defaults = no action).
DPM Teslameter Display Instructions
Change the address of the DPM with the two cryptically labeled front panel switches, ROLL (R) and SELECT (S). When the address is changed, the display shows the corresponding teslameter output and ignores all other addresses.
To change the address:
Hold R down and press and release S and then release R
Do this again
Press S to increment the address, maximum 31
Press R to decrement the address, minimum 0
Do step 1 again to return to field display
Stephanie Allison
-----Original Message-----
From: [email protected] [mailto:[email protected]] On Behalf
Of Michael Davidsaver
Sent: Monday, March 30, 2015 12:30 PM
To: [email protected]
Subject: support for DTM-151?
Has anyone integrated a Digital Teslameter 151 from Group3?
http://www.group3technology.com/index.php/products/product/digital-teslameter-151
- Re: support for DTM-151? Jesus Vasquez
- support for DTM-151? Michael Davidsaver
- RE: support for DTM-151? Allison, Stephanie
unable to obtain context information about the PV Thomas Schösser
Re: building base-3.15.1: perl -typemap Andrew Johnson
<2015>
2025
2026
<2015>
2025
2026