Hi all,
A (relatively) more concise and complete summary of my current issue with StreamDevice:
EPICS 3.14.12.2 IOC using StreamDevice 2.5 and asyn 4-32.
Creates connection to an NPort 6450 4-port serial/Ethernet server using this call:
drvAsynIPPortConfigure("SCR_BTS35_-TMP_D1612", "192.168.54.218:4006")
(the port # was changed from the default so my development IOC rather than the live IOC can connect)
A 9600 baud RS-232 connection from the Moxa to the Edwards STP-iX3006 Turbomolecular pump.
The basics of the pump serial protocol: (### = serial transmission block # starting at 001)
IOC: <STX>###?<cmd-char><ETX><8-bit-checksum>
Pump: <ACK>
if < 255 bytes of data:
Pump: <STX>###<space><cmd-char><data><ETX><8-bit-checksum>
else:
Pump: <STX>###<space><cmd-char><data><ETB><8-bit-checksum>
IOC: <ACK>
While > 255 data bytes left:
Pump: <STX>###<another-255-bytes-data><ETB><8-bit-checksum>
IOC: <ACK>
When <= 255 bytes left:
Pump: <STX>###<remaining-data><ETX><8-bit-checksum>
IOC: <ACK>
When the total bytes in a transmission block > 112, the Moxa splits it in to multiple packets.
NOTE: There is no evidence of any delay between the characters within a serial transmission block from the pump, yet even when I change settings in the Moxa that should result in larger packets (the max packet size is set to 1024, the time to wait for more
bytes before sending a packet is set to 1000 ms, the delimiter character is set to <ETX>, ...), it always breaks up any data stream > 112 bytes in to separate packets. If it were not for the fact that it (supposedly) supports a packet size up to 1024 bytes,
I would assume that (as odd a # as that is) it was the max # of data bytes it could send in one packet.
The StreamDevice issue related to this:
For a protocol linked to a periodically processed record, StreamDevice will correctly match the pattern of the "in" commands to the contents of the transmission blocks from the pump, even when individual blocks have been split up in to multiple TCP packets.
However, this does not appear to be true for protocols that are linked to records using "I/O Intr". For such protocols, StreamDevice appears to require a 1-to-1 relationship between the packets and the patterns in the "in" commands. If you know the # and
size of each packet, this can be used to define a working protocol (see example below). But if those values change for whatever reason, then the protocols that depend on them will stop working.
Example from my protocol file:
## times are in milliseconds
LockTimeout = 20000; # max wait for exclusive access to the device
WriteTimeout = 500; # max wait to be able to write to the device
ReplyTimeout = 1000; # max wait until the start of a reply
ReadTimeout = 1000; # max idle time after start of reply
PollPeriod = 200; # How often to poll for new input for "I/O Intr" mode records
ExtraInput = Ignore;
MaxInput = 0;
##-----[ReadEventsWithTime]----------------------------------------------------
#===== Record for this protocol must process periodically =====
##--- StreamDevice properly matches the "in" patterns to the separate
##--- transmission blocks even when the contents of each block has
##--- been split across multiple TCP packets.
ReadEventsWT {
out "\x02" "001?}" "\x03%<xor8ff>";
in "\x06\x02" "001 }%253c" "\x17%*1c";
out "\x06";
in "\x02" "002%*151c" "\x03%*1c";
out "\x06";
#=====Records for these protocols must use "I/O Intr" SCAN =====
#--- The data for these are in the 1st packet of the 1st transmission block
curNumErrs { in "\x02" "001 }%2x"; } # current # of errors
maxNumErrs { in "\x02" "001 }%*2c%2x"; } # max # of errors
err1Type { in "\x02" "001 }%*4c%2x"; } # Error 1 #
err1Date { in "\x02" "001 }%*8c%6c"; } # Error 1 date (YYMMDD)
err1Time { in "\x02" "001 }%*14c%4c"; } # Error 1 time (HHMM)
err2Type { in "\x02" "001 }%*24c%2x"; } # Error 2 #
err2Date { in "\x02" "001 }%*28c%6c"; } # Error 2 date (YYMMDD)
err2Time { in "\x02" "001 }%*34c%4c"; } # Error 2 time (HHMM)
err3Type { in "\x02" "001 }%*44c%2x"; } # Error 3 #
err3Date { in "\x02" "001 }%*48c%6c"; } # Error 3 date (YYMMDD)
err3Time { in "\x02" "001 }%*54c%4c"; } # Error 3 time (HHMM)
err6Type { in "\x02" "001 }%*104c%2x"; } # Error 6 #
##--- The following fail to match the remaining values in the 1st transmission block
##--- that are in the 2nd TCP packet
#err6Date { in "\x02" "001 }%*108c%6c"; } # Error 6 date (YYMMDD)
#err6Time { in "\x02" "001 }%*114c%4c"; } # Error 6 time (HHMM)
##--- The following work to get the desired values from the 1st transmission block
err6Date { in "\x02" "001 }%*106c"; in "%*2c%6c"; } # Error 6 date (YYMMDD)
err6Time { in "\x02" "001 }%*106c"; in "%*8c%4c"; } # Error 6 time (HHMM)
Sample of the related records:
record(stringin, "_$(SYS)_$(SUB):$(DEV)_$(INST):ReadEventsWT") {
field(DESC, "List of events/errors with times")
field(SCAN, "5 second")
field(DTYP, "stream")
field(INP, "@Edwards_STP_TMP.proto ReadEventsWT $(PORT)")
record(longin, "$(SYS)_$(SUB):$(DEV)_$(INST):HWER_RD_CNT") {
field(DESC, "Number of HW err events")
field(EGU, "NumErrs")
field(SCAN, "I/O Intr")
field(INP, "@Edwards_STP_TMP.proto curNumErrs $(PORT)")
record(longin, "$(SYS)_$(SUB):$(DEV)_$(INST):HWER_RD_TYP1") {
field(DESC, "hdw error 1 event type")
field(INP, "@Edwards_STP_TMP.proto err1Type $(PORT)")
record(stringin, "$(SYS)_$(SUB):$(DEV)_$(INST):HWER_RD_DT1") {
field(DESC, "hdw error 1 event date")
field(INP, "@Edwards_STP_TMP.proto err1Date $(PORT)")
record(stringin, "$(SYS)_$(SUB):$(DEV)_$(INST):HWER_RD_TM1") {
field(DESC, "hdw error 1 event time")
field(INP, "@Edwards_STP_TMP.proto err1Time $(PORT)")
record(longin, "$(SYS)_$(SUB):$(DEV)_$(INST):HWER_RD_TYP6") {
field(DESC, "hdw error 6 event type")
field(INP, "@Edwards_STP_TMP.proto err6Type $(PORT)")
record(stringin, "$(SYS)_$(SUB):$(DEV)_$(INST):HWER_RD_DT6") {
field(DESC, "hdw error 6 event date")
field(INP, "@Edwards_STP_TMP.proto err6Date $(PORT)")
record(stringin, "$(SYS)_$(SUB):$(DEV)_$(INST):HWER_RD_TM6") {
field(DESC, "hdw error 6 event time")
field(INP, "@Edwards_STP_TMP.proto err6Time $(PORT)")