1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  <20202021  2022  2023  2024  2025  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  <20202021  2022  2023  2024  2025  2026
Subject: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway
Date: Wed, 8 Jul 2020 13:38:02 +0000
Hello everyone,
I have a simple StreamDevice protocol for an old GPIB-based device HP3458A which is connected to Agilent E5810 LAN/GPIB Gateway (100 Mbps) talking to a softIOC. Here is the record to read the voltage data at 0.2Hz (5 second).
#only showing the must-have fields
record(ai, "SR:C03-BI{DMM:1}V-I_")
    field(DTYP, "stream")
    field(INP,  "@hp3458a.proto ReadData L0 6")
    field(SCAN, "5 second")
#StreamDevice protocol
ReadData {
         out "DCV  10, 1E-6";
         in "%f";
And this is the EPICS shell output after I turn on the Asyn trace.
2020/07/07 17:55:46.280 L0 6 vxiWrite
DCV  10, 1E-6
2020/07/07 17:55:49.630 L0 6 vxiRead
-4.016502336E+00
2020/07/07 17:55:52.396 L0 6 vxiWrite
DCV  10, 1E-6
2020/07/07 17:55:55.756 L0 6 vxiRead
-4.015675668E+00
2020/07/07 17:56:00.680 L0 6 vxiWrite
DCV  10, 1E-6
2020/07/07 17:56:04.030 L0 6 vxiRead
-4.014579704E+00
I get the expected response (i.e. -4.016502336E+00) after the command (DCV  10, 1E-6) is sent. However, as you see from the timestamps for each vxiWrite, the command is not sent at every 5-second. It takes more than 5 seconds and it has a pattern: 6.116, 8.284, 6.116, 8.284, …. ; the response time (vxiRead – vxiWrite) is consistent at 3.35 seconds although it seems ridiculously slow.
So, the system is slower than expected. However, I am surprised that there is no any alarm for the record and no any warning message from the EPICS shell output when that record can not keep up its scanning rate. Is this kind of silence supposed to be? Just to be clear, the softIOC itself and the physical server are perfectly fine: I added the classical “dbExample1.db” and its calc record can be scanned at exactly 5-second.
This system used to work as expected at 0.2Hz (5 second) despite it is ridiculously slow. Now it becomes even slower after our ITD upgraded the network switch. If anyone has experience on these old devices (HP3458A, E5810), please advise what I should do next.
If the version of Asyn/StreamDevice matters for this case, here we go:
epics> dbior
Driver: drvAsyn
L0 multiDevice:Yes canBlock:Yes autoConnect:Yes
    vxi11, host name: 10.0.132.67
Driver: stream
  StreamDevice 2.6.0 built May 29 2013 18:45:57
  registered bus interfaces:
    DebugInterface
    DummyInterface
AsynDriverInterface
PS: I had difficulty finding the exact version number of asyn/streamdevice since I built my IOC against Michael Davidsaver’s Debian packages (https://epics.nsls2.bnl.gov/debian/ ) and the packages provide everything under /usr/lib/epics/. I am sure there are many ways (ldd, strings, configure/RELEASE, etc.) to find out the version number of EPICS support module. I just found “dbior” seems a handy tool for reporting EPICS driver’s version. I am not sure if Mark Rivers has added this kind of report in his latest Asyn release.
Best regards,
Yong Hu
NSLS-II Controls Group
RE: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway Mark Rivers via Tech-talk
Re: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway Ralph Lange via Tech-talk
Re: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway Michael Davidsaver via Tech-talk
Prev: Process at Init of mbbo-record Nonn, Patrick via Tech-talk
Next: RE: Process at Init of mbbo-record Mark Rivers via Tech-talk
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  <20202021  2022  2023  2024  2025  2026
Prev: Re: Process at Init of mbbo-record Nonn, Patrick via Tech-talk
Next: RE: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway Mark Rivers via Tech-talk