Hi Mark,
This is the complete output (a few comments are removed). The hp3458a DMM is using GPIB address 6.
#!../../bin/linux-x86/hp3458a
epicsEnvSet("EPICS_CA_AUTO_ADDR_LIST", "NO")
epicsEnvSet("EPICS_CA_ADDR_LIST", "10.0.153.255")
< envPaths
epicsEnvSet("ARCH","linux-x86")
epicsEnvSet("IOC","iochp3458a")
epicsEnvSet("TOP","/direct/iocs_diagnostics/hp3458a")
epicsEnvSet("EPICS_BASE","/usr/lib/epics")
cd /direct/iocs_diagnostics/hp3458a
dbLoadDatabase "dbd/hp3458a.dbd"
hp3458a_registerRecordDeviceDriver pdbbase
epicsEnvSet "STREAM_PROTOCOL_PATH", "protocol"
vxi11Configure("L0","10.0.132.67",0,10,"gpib0")
dbLoadRecords("/usr/lib/epics/db/asynRecord.db", "P=SR:C03-BI,R={DMM}Asyn,PORT=L0,ADDR=6,OMAX=0,IMAX=0")
dbLoadRecords "db/hp3458a.db"
iocInit
Starting iocInit
## EPICS R3.14.12.3-6 $Date: Mon 2012-12-17 14:11:47 -0600$
## EPICS Base built May 22 2013
cas warning: Configured TCP port was unavailable.
cas warning: Using dynamically assigned TCP port 54813,
cas warning: but now two or more servers share the same UDP port.
cas warning: Depending on your IP kernel this server may not be
cas warning: reachable with UDP unicast (a host's IP in EPICS_CA_ADDR_LIST)
epics>
There are a few other functions, such as reading identification number, setting input buffer, etc. in the stream protocol file, but the records associated with those functions are processed on-command (by
caput), not periodically scanned. So, what really matters is the ai record and the function ReadData() in my original message. I do have the configurations of terminator and timeout in my protocol file as below:
locktimeout = 10000;
replytimeout = 10000;
readtimeout = 10000;
writetimeout = 10000;
MaxInput = 0;
Terminator = "\r\n";
Hi Yong,
Please send the complete output when your IOC starts. In your original message was that the complete stream protocol file?
Hi Mark R.,
Thanks for your suggestion. Yes, I found the version in /usr/lib/epics/include/asynDriver.h: 4.23.0.
Yes, you are right. Both Asyn and StreamDevice are very old. I am going to upgrade them to the latest releases.
I have set the trace mask to 0xFF. Attached please find the detailed output messages for a few cycles. Below is the simplified version with important timestamps highlighted.
1) The record is scanned at every 5-second so that we get 2020/07/08 15:16:31.567
à 2020/07/08 15:16:36.567
à 2020/07/08 15:16:41.567, which makes perfect sense. Then it jumps to 2020/07/08 15:16:51.567,
which is 10-second gap meaning one scan is skipped. This kind of skipping is confirmed by Michael’s suggestion of setting TPRO. I guess “scan5: Active SR:C03-BI{DMM:1}V-I_” means the record is scanned at 5-second and its PACT is true so that the record processing
is simply skipped and it waits for the next scan.
2) Could you briefly explain what these mean:
queueRequest timeout, vxiSetEos 1, vxiSetEos 0? I am surprised to see “\n 0a” after vxiSetEos 1. The time difference between “vxiRead” and “vxiSetEos 1” is 3.4 seconds, which means it takes 3.4 seconds to read a floating number
“-4.008121752E+00”.
scan5: Process SR:C03-BI{DMM:1}V-I_
L0 addr 6 queueRequest priority 0 not lockHolder
2020/07/08 15:16:31.567 L0 schedule queueRequest timeout
2020/07/08 15:16:31.567 asynManager::portThread port=L0 callback
2020/07/08 15:16:31.567 L0 addr 6 queueRequest priority 0 not lockHolder
2020/07/08 15:16:31.567 L0 6 vxiWrite numchars 14
2020/07/08 15:16:31.569 L0 6 vxiWrite
DCV 10, 1E-6
DCV 10, 1E-6\r\n
44 43 56 20 31 30 2c 20 31 45 2d 36 0d 0a
2020/07/08 15:16:31.569 L0 addr 6 queueRequest priority 0 from lockHolder
2020/07/08 15:16:31.569 L0 schedule queueRequest timeout
2020/07/08 15:16:31.569 asynManager::portThread port=L0 callback
2020/07/08 15:16:31.569 L0 vxiSetEos 1
\n
0a
2020/07/08 15:16:34.961 L0 6 vxiRead
-4.008121752E+00
-4.008121752E+00\r\n
2d 34 2e 30 30 38 31 32 31 37 35 32 45 2b 30 30 0d 0a
2020/07/08 15:16:34.961 L0 vxiSetEos 0
2020/07/08 15:16:40.407
L0 vxiSetEos 0
2020/07/08 15:16:46.957
L0 vxiSetEos 0
##note: one scan is skipped
2020/07/08 15:16:55.007
L0 vxiSetEos 0
2020/07/08 15:16:56.567
Hi Yong,
- PS: I had difficulty finding the exact version number of asyn/streamdevice since I built my IOC against Michael Davidsaver’s Debian packages (https://epics.nsls2.bnl.gov/debian/
) and the packages provide everything under /usr/lib/epics/
The Debian packages must also install the asyn header files, probably in /usr/include/epics. Look at asynDriver.h, it should contain lines like the following:
#define ASYN_VERSION 4
#define ASYN_REVISION 39
#define ASYN_MODIFICATION 0
Let me know what version it is. Since your streamDevice is more than 7 years old I suspect that asyn is very old as well. If so, there have been quite a few fixes to the asyn VXI-11 support
since 2013. I suggest you build a test IOC with the latest version of asyn and streamDevice and see if he problem still occurs.
You should probably enable all of the asyn trace flags on that driver, i.e. set the trace mask to 0xFF. There may be some hints as to the timing issues if you do that.
On Behalf Of Hu, Yong via Tech-talk
Sent: Wednesday, July 8, 2020 8:38 AM
To: [email protected]
Subject: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway
Hello everyone,
I have a simple StreamDevice protocol for an old GPIB-based device HP3458A which is connected to Agilent E5810 LAN/GPIB Gateway (100 Mbps) talking to a softIOC. Here is the record to read the voltage data
at 0.2Hz (5 second).
#only showing the must-have fields
record(ai, "SR:C03-BI{DMM:1}V-I_")
field(DTYP, "stream")
field(INP, "@hp3458a.proto ReadData L0 6")
field(SCAN, "5 second")
#StreamDevice protocol
ReadData {
out "DCV 10, 1E-6";
in "%f";
And this is the EPICS shell output after I turn on the Asyn trace.
2020/07/07 17:55:46.280 L0 6 vxiWrite
DCV 10, 1E-6
2020/07/07 17:55:49.630 L0 6 vxiRead
-4.016502336E+00
2020/07/07 17:55:52.396 L0 6 vxiWrite
DCV 10, 1E-6
2020/07/07 17:55:55.756 L0 6 vxiRead
-4.015675668E+00
2020/07/07 17:56:00.680 L0 6 vxiWrite
DCV 10, 1E-6
2020/07/07 17:56:04.030 L0 6 vxiRead
-4.014579704E+00
I get the expected response (i.e. -4.016502336E+00) after the command (DCV 10, 1E-6) is sent. However, as you see from the timestamps for each vxiWrite, the command is not sent at every 5-second. It takes
more than 5 seconds and it has a pattern: 6.116, 8.284, 6.116, 8.284, …. ; the response time (vxiRead – vxiWrite) is consistent at 3.35 seconds although it seems ridiculously slow.
So, the system is slower than expected. However, I am surprised that there is no any alarm for the record and no any warning message from the EPICS shell output when that record can not keep up its scanning
rate. Is this kind of silence supposed to be? Just to be clear, the softIOC itself and the physical server are perfectly fine: I added the classical “dbExample1.db” and its calc record can be scanned at exactly 5-second.
This system used to work as expected at 0.2Hz (5 second) despite it is ridiculously slow. Now it becomes even slower after our ITD upgraded the network switch. If anyone has experience on these old devices
(HP3458A, E5810), please advise what I should do next.
If the version of Asyn/StreamDevice matters for this case, here we go:
epics> dbior
Driver: drvAsyn
L0 multiDevice:Yes canBlock:Yes autoConnect:Yes
vxi11, host name: 10.0.132.67
Driver: stream
StreamDevice 2.6.0 built May 29 2013 18:45:57
registered bus interfaces:
DebugInterface
DummyInterface
AsynDriverInterface
) and the packages provide everything under /usr/lib/epics/. I am sure there are many ways (ldd, strings, configure/RELEASE, etc.) to find out the version number of EPICS support module. I just found “dbior” seems a handy tool for reporting EPICS driver’s
version. I am not sure if Mark Rivers has added this kind of report in his latest Asyn release.
Best regards,
Yong Hu
NSLS-II Controls Group
