1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  <2022>  2023  2024  2025  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  <2022>  2023  2024  2025  2026
Subject: Re: Unknown error with an equipment
To: Mark Rivers <[email protected]>, Andrew Johnson <[email protected]>
Date: Fri, 18 Mar 2022 11:15:13 +0100
Hi Mark and Andrew,
I added asynSetTraceMask("IDEA",-1,0x9) and asynSetTraceIOMask("IDEA",-1,0x2) in st.cmd to monitor the outputs from IDEA. Here is the first few lines from the output:
2022/03/18 10:45:52.753 terminal.chamber6:10002 write 10
Get ekin\r\n
2022/03/18 10:45:52.763 terminal.chamber6:10002 read 7
0.000\r\n
2022/03/18 10:45:52.765 terminal.chamber6:10002 write 9
Get hsa\r\n
2022/03/18 10:45:52.774 terminal.chamber6:10002 read 8
1638.400
2022/03/18 10:45:52.775 terminal.chamber6:10002 read 2
\r\n
2022/03/18 10:45:53.206 terminal.chamber6:10002 write 12
Get filter\r\n
2022/03/18 10:45:53.215 terminal.chamber6:10002 read 8
#07 filt
2022/03/18 10:45:53.216 terminal.chamber6:10002 read 7
er on\r\n
2022/03/18 10:45:53.297 terminal.chamber6:10002 write 11
Get range\r\n
2022/03/18 10:45:53.306 terminal.chamber6:10002 read 8
#06 rang
2022/03/18 10:45:53.307 terminal.chamber6:10002 read 7
e x10\r\n
2022/03/18 10:45:53.998 terminal.chamber6:10002 read 8
#01 ok\r\n
2022/03/18 10:45:53.999 terminal.chamber6:10002 write 5
ver\r\n
2022/03/18 10:45:54.010 terminal.chamber6:10002 read 24
EFC_for_NanoESCA V01.03
2022/03/18 10:45:54.012 terminal.chamber6:10002 read 7
B0115\r\n
2022/03/18 10:45:54.013 terminal.chamber6:10002 write 17
Set mode remote\r\n
2022/03/18 10:45:54.022 terminal.chamber6:10002 read 8
#04 remo
I realize that the default get and set voltage is exceed the range of the operation (terminal.chamber6:10002 read 8 1638.400), which can't be handled by the power supply. The maximum value in the DB file should be 200. The problem is that I cannot set the output voltage properly. When I try to set the voltage in CSS, there is no response from the changing. Below shows the logging.
2022/03/18 11:04:40.283 terminal.chamber6:10002 write 18
Set hsa 0.000000\r\n
2022/03/18 11:04:40.292 terminal.chamber6:10002 read 12
#0E HSA ok\r\n
2022/03/18 11:04:40.734 terminal.chamber6:10002 write 9
Get hsa\r\n
2022/03/18 11:04:40.744 terminal.chamber6:10002 read 8
1638.400
2022/03/18 11:04:41.228 terminal.chamber6:10002 write 9
Get hsa\r\n
2022/03/18 11:04:41.237 terminal.chamber6:10002 read 10
1638.400\r\n
I would like to know how to initialize the set values to the devices in the db file. Below the the corresponding db.
record(ai, "$(P):$(NAME):Epass") {
  field(SCAN, ".5 second")
  field(DTYP, "stream")
  field(INP, "@IDEA.proto Get_HSA $(interface)")
  field(PREC, "3")
  field(EGU, "V")
  field(HOPR, "200")
  field(LOPR, "0")
  field(AOFF, "0")
  field(ASLO, "0.5239717")
record(ao, "$(P):$(NAME):SetEpass") {
  field(OUT, "@IDEA.proto Set_HSA $(interface)")
  field(PREC, "3")
  field(EGU, "V")
  field(HOPR, "200")
  field(LOPR, "0")
  field(AOFF, "0")
  field(HIHI, "200")
  field(LOLO, "30")
  field(HIGH, "200")
  field(LOW, "30")
Best,
Tony
Andrew Johnson <[email protected]> 於 2022年3月17日週四 下午9:12寫道:
Hi Tony,
The error messages you posted ought to be self-explanatory if you read them in order:
> error:
> IDEA ImpMik:IDEA:Status: Input "#31 hv error detect" mismatch after 4 bytes
> IDEA ImpMik:IDEA:Status: got "hv error detect" where "ok" was expected
> IDEA ImpMik:IDEA:SetFilterOn: Input "#2E impossible, erro" mismatch after 1 byte
> IDEA ImpMik:IDEA:SetFilterOn: got "2E impossible, error" where "0" was expected
Your protocol for reading the status from the device says:
> # Get_Status
> Get_Status {
>     out "Get status";
>     in  "#%2i ok";
>     out "ver";
>     in "EFC_for_NanoESCA V01.03 B0115";
>     out "Set mode remote";
>     in "#04 remote mode";
Thus you're telling StreamDevice that after it sends a "Get status" string, it will get back "#" followed by a 2-digit integer and the word "ok". The first two error message lines explain that instead of "ok" it's seeing the string "hv error detect". Have you looked at the instrument that it's talking to? I would guess there's a problem with some high voltage signal that it isn't seeing. You should probably fix the protocol file so it can accept error strings like that from the device and make them available in a stringin record instead of always assuming that the instrument is working and will never give you an error, but you might be able to resolve your immediate issue by fixing whatever the instrument is complaining about.
Always fix your first error first, then look at any error messages you get in that state — later errors are often caused by earlier ones and may go away once you've fixed the first problem.
HTH,
- Andrew
Complexity is free, it's Simplicity that takes work.
Re: Unknown error with an equipment Mark Rivers via Tech-talk
Unknown error with an equipment TONY YU via Tech-talk
Re: Unknown error with an equipment TONY YU via Tech-talk
Re: Unknown error with an equipment Andrew Johnson via Tech-talk
Prev: Re: Timestamp test questions on Linux gjzhai--- via Tech-talk
Next: Re: StripTool license file Andrew Johnson via Tech-talk
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  <2022>  2023  2024  2025  2026
Prev: Re: Unknown error with an equipment Andrew Johnson via Tech-talk
Next: Re: Unknown error with an equipment Mark Rivers via Tech-talk