1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  <20232024  2025  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  <20232024  2025  2026
Subject: Re: USBCTR08 ch0 scaler counts missing
From: Keith Surrena via Tech-talk <[email protected]>
To: Mark Rivers <[email protected]>, "[email protected]" <[email protected]>
Date: Thu, 24 Aug 2023 13:45:16 +0000
Thanks, Mark. I overlooked that my patch panel has all of the gates connected at a singular BNC node. Removing CG0 makes it function as expected now. I suspect you’re right that SPEC configures the USBCTR08 differently and allows for a different schematic. I noticed that the counter output waveform differs in that with the measComp driver the counter output waveform is a square pulse of a defined count time whereas with spec, there is usually a 150us initialization pulse that precedes the counter output timing pulse.
Could you confirm another behavior for me? If I’m using the preset counters on ch 1-7 with the idle state low, the counter output waveform goes high and then stays high even when the count completes. If I initiate another count sequence, it will dip low at the start, go high for the duration of the count, and then stay high until I once again doing another count. Is this typical or perhaps another incompatibility of my patch panel?
Keith Surrena
CHESS
Sent: Wednesday, August 23, 2023 4:25 PM
To: Keith Surrena <[email protected]>; [email protected] <[email protected]>
Subject: RE: USBCTR08 ch0 scaler counts missing
Hi Keith,
  • I have TMR0 set at 1 MHz connected to my counter 0 input and my counter 0 output connected to the gate.
When you say “connected to the gate” what do you mean?  The EPICS driver assumes that the counter 0 output is connected to the gate of all channels except channel 0, i.e. channels 1-7.  That is documented here: https://epics-modules.github.io/measComp/measCompUSBCTRDoc.html#scaler-record-support
  • I can complete a count sequence by using the preset counters on counters 1-7 connected to TMR0 but not c0.
When you do preset counts on counters 1-7 that is done in software, not in hardware.  The driver reads a stream of  count values from the scalers and stops when any counter exceeds its present.  So it will always count a little bit more than the present counts.
Because counter 0 is used in preset mode, and is connected to the gate of all other counters it will count for exactly the requested number of counts and all other channels will stop at exactly the same time.
I just configured my USBCTR-08 as follows:
  • TMR0 set at 1e6 Hz and connected to CTR0.
  • TMR1 set at 1e5 Hz and connected to CTR1.
  • TMR2 set at 1e4 Hz and connected to CTR2.
  • TMR3 set at 1e3 Hz and connected to CTR3.
I set the scaler.TP field to 1.0 second and the scaler.FREQ field to 1e6 Hz.  The scaler.G1 field is 1 so it is a preset.  .G2-.G8 are 0 so they are not presets.
When I start the scaler record and let it finish counting I see the screen in the attached file.
Note that channels 1-4 have exactly the expected counts: 1000000, 100000, 10000, 1000.
I am not sure if SPEC and the measComp driver program the UCBCTR-08 in exactly the same way, and if SPEC expects the module to be wired the same way.
From: Tech-talk <[email protected]> On Behalf Of Keith Surrena via Tech-talk
Sent: Wednesday, August 23, 2023 2:27 PM
To: [email protected]
Subject: USBCTR08 ch0 scaler counts missing
I’m having an issue where counter 0 of my USBCTR08 is not registering counts when I attempt to do a basic scaler count sequence with the MEDM (count time 1s, count preset 1000000). I have TMR0 set at 1 MHz connected to my counter 0 input and my counter 0 output connected to the gate. I can complete a count sequence by using the preset counters on counters 1-7 connected to TMR0 but not c0. I’ve verified my TMR0 signal with an oscope, the numbers and timing on the other preset counters makes sense, and this same hardware setup works at our beamlines that use the USBCTR with SPEC, so I’m fairly confident in the hardware and signal patching. Version info: scaler 4.0, measComp R4.1, and epics 7.0.7.
Keith Surrena
CHESS
RE: USBCTR08 ch0 scaler counts missing Mark Rivers via Tech-talk
USBCTR08 ch0 scaler counts missing Keith Surrena via Tech-talk
Prev: Phoebus configuration and PV connection 刘晓敏 via Tech-talk
Next: RE: USBCTR08 ch0 scaler counts missing Mark Rivers via Tech-talk
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  <20232024  2025  2026
Prev: RE: USBCTR08 ch0 scaler counts missing Mark Rivers via Tech-talk
