1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  <20252026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  <20252026
Subject: Re: Issues in MCA module with XFlash detector
To: Abdalla Ahmad <[email protected]>, "[email protected]" <[email protected]>
Date: Mon, 14 Jul 2025 13:24:41 +0000
Hi Abdalla,
Sorry for the delay in replying.
Setting “Read spectrum every” setting to “Event” does not actually do anything, since you have probably not triggered any Events in your IOC.  You should just set it to Passive.  The spectrum will always be read when acquisition completes as long as Check Done is not Passive.
If you are collecting for more than a few seconds and want to see periodic spectrum updates then you can set “Read spectrum every” to 5 seconds and it won't slow things down too much.
What baud rate are you using?  Is it possible to increase the baud rate to improve communication speed?
Sent: Monday, June 30, 2025 2:03 AM
To: Mark Rivers <[email protected]>; [email protected] <[email protected]>
Subject: RE: Issues in MCA module with XFlash detector
Hello Mark
Yes, we are using the serial port to communicate with the detector. I turned on asynTrace options and I noticed that the freezing and unresponsive buttons issues are caused due to a timeout during acquisition, this is confirmed because “timeout handler” message gets printed almost every 1 second (not all the time), and after that the temperature PV gets processed. So I switched the scan rate on the main GUI to passive just to test and the acquisition works fine for now.
Also, I noticed that the spectrum takes 1 or 2 seconds more than the real time, I looked into the MCA setup GUI and changed the “Read spectrum every” setting to “Event”, seems a bit more responsive than before. I attached the current MCA settings screenshot. What do you think?
Best Regards,
Abdalla Al-Dalleh
Control Engineer
SESAME
Sent: Tuesday, June 24, 2025 4:19 PM
To: Abdalla Ahmad <[email protected]>; [email protected]
Hi Abdallah,
Are you communicating with the XFlash over a serial port or over their QM100 high-speed interface?
If it is over a serial port then you should enable asynTrace to see the communication stream.
Sent: Tuesday, June 24, 2025 4:55 AM
To:
Cc: Mark Rivers <
[email protected]>
Subject: Issues in MCA module with XFlash detector
Hi
We have the MCA module running on a Windows 10 machine to control the XFlash detector from Bruker. We started to face the following issues:
  • When the IOC starts, the state is always acquiring without pressing the start button.
  • The buttons to erase/stop/start the acquisition have no effect, they do not stop or start the acquisition.
  • The real time and live time show weird values on startup, like in 10s or 100s. of seconds.
  • If I wait long enough, the acquisition stops but pressing the start button again, say for a real time of 1 second, it starts acquiring for fraction
    of a second and then freezes, nothing is progressing.
Despite what is mentioned, the IOC log shows no error messages. The Bruker software seems to acquire just fine, but not tested thoroughly. We even
Installed the IOC on a different laptop and a PC and still the same behavior. Any idea on what might went wrong with the detector or IOC?
Best Regards,
Abdalla Al-Dalleh
Control Engineer
SESAME
RE: Issues in MCA module with XFlash detector Abdalla Ahmad via Tech-talk
Re: Issues in MCA module with XFlash detector Mark Rivers via Tech-talk
Prev: Re: Help operating ADHamamatsuDCAM IOC Marco Filho via Tech-talk
Next: Re: help requested about motorPI module Dariush Hampai via Tech-talk
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  <20252026
Prev: RE: Issues in MCA module with XFlash detector Abdalla Ahmad via Tech-talk
Next: RE: Issues in MCA module with XFlash detector Abdalla Ahmad via Tech-talk
