1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  <2026>  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  <2026>
Subject: Re: [EXTERNAL] Support for Gamma Vacuum MPCq ion pump controller over Ethernet
From: "Webb, Brad via Tech-talk" <tech-talk at aps.anl.gov>
To: "rivers at cars.uchicago.edu" <rivers at cars.uchicago.edu>, "tech-talk at aps.anl.gov" <tech-talk at aps.anl.gov>
Date: Mon, 26 Jan 2026 12:38:43 +0000
Hi Mark
See attached. It is what we use at SNS for Gamma ION QPC Digitel:
-Brad
On Fri, 2026-01-23 at 22:21 +0000, Mark Rivers via Tech-talk wrote:
Folks, I have been using this old C device support to talk to Gamma Vacuum MPC ion pump controllers. https: //github. com/epics-modules/ip/blob/master/ipApp/src/devMPC. c https: //github. com/epics-modules/ip/blob/master/ipApp/Db/MPC. db This only
Folks,
This only talks via RS-232, but many of their newer modules support Ethernet.  The Ethernet message format is somewhat different.  Has anyone written StreamDevice support for the Ethernet protocol?
Thanks
Terminator = CR;
#InTerminator = LF;
ReplyTimeout = 10000;
ReadTimeout = 10000;
#in "^[0-9]{2} OK 00 ([0-9]+.[a-zA-Z0-9-]+) (TORR|MBAR|PASCAL) [a-zA-Z0-9]{2}"
#in "%*s%*s%*s%s%*s%*s";
#in "\$1 OK 00 %s QPC %0<sum8>";
getModel
    out "~ \$1 01 %01.0<sum8>";
    in "\$1 OK 00 %/DIGITEL QPC/ %0<sum8>";
getFWV
    out "~ \$1 02 %01.0<sum8>";
    in "\$1 OK 00 FIRMWARE VERSION: %s %0<sum8>";
getCurrent
    out "~ \$1 0A \$2 %01.0<sum8>";
    in "\$1 OK 00 %e %*s %0<sum8>";
getPressure
    out "~ \$1 0B \$2 %01.0<sum8>";
getVoltage
    out "~ \$1 0C \$2 %01.0<sum8>";
    in "\$1 OK 00 %d %0<sum8>";
getPumpStatus
    out "~ \$1 0D \$2,00 %01.0<sum8>";
# ~ 08 11 1 00
# 08 OK 00 300 L/S 63
getPumpSize
    out "~ \$1 11 \$2 %01.0<sum8>";
    in "\$1 OK 00 %d %*s %0<sum8>";
getPumpPressureFactor
    out "~ \$1 1D \$2 %01.0<sum8>";
    in "\$1 OK 00 %f %0<sum8>";
getAutoRestartStr
    out "~ \$1 34 \$2 %01.0<sum8>";
    in "\$1 OK 00 %s %0<sum8>";
getPumpCurrentLimit
    out "~ \$1 43 \$2 %01.0<sum8>";
getSetpoint
    out "~ \$1 3b \$2 %01.0<sum8>";
    in "\$1 OK 00 \$3,1,\$3,%(\$4:IP\$5:Relay\$2OnPressure)e,%(\$4:IP\$5:Relay\$2OffPressure)e,%(\$4:IP\$5:Relay\$2State)d %0<sum8>";
SetSetpoint
    out "~ \$1 3b \$3,1,\$3,%(A)e,%(B)e %01.0<sum8>";
    in "\$1 OK 00 %0<sum8>";
HvTurnOn
    out "~ \$1 37 \$2 %01.0<sum8>";
HvTurnOff
    out "~ \$1 38 \$2 %01.0<sum8>";
file gammaSerial.template
 pattern
  {P, N, S1, S2, ADDR, PORT}
  {LDmp_Vac, 01, 01, 1, 06, PORT06}
  {LDmp_Vac, 04, 02, 2, 06, PORT06}
  {MDmp_Vac, 01, 01, 1, 07, PORT07}
  {HEBT_Vac, 04, 02, 2, 07, PORT07}
  {HEBT_Vac, 06, 01, 1, 08, PORT08}
  {HEBT_Vac, 08, 02, 2, 08, PORT08}
  {HEBT_Vac, 10, 01, 1, 09, PORT09}
  {HEBT_Vac, 12, 02, 2, 09, PORT09}
  {HEBT_Vac, 14, 01, 1, 0A, PORT10}
  {HEBT_Vac, 16, 02, 2, 0A, PORT10}
  {HEBT_Vac, 17, 01, 1, 0B, PORT11}
  {HEBT_Vac, 18, 02, 2, 0B, PORT11}
  {HEBT_Vac, 20, 01, 1, 0C, PORT12}
  {HEBT_Vac, 22, 02, 2, 0C, PORT12}
  {HEBT_Vac, 2B, 01, 1, 0D, PORT13}
  {HEBT_Vac, 4B, 02, 2, 0D, PORT13}
  {HEBT_Vac, 8B, 01, 1, 0E, PORT14}
  {HEBT_Vac, 14B, 02, 2, 0E, PORT14}
  {HEBT_Vac, 16B, 01, 1, 0F, PORT15}
  {HEBT_Vac, 17B, 02, 2, 0F, PORT15}
  {HEBT_Vac, 20B, 01, 1, 10, PORT16}
  {HEBT_Vac, 09, 02, 2, 10, PORT16}
record(stringin, "${P}:IP${N}:Model")
 field(DTYP, "stream")
 field(INP, "@gamm1.proto getModel(${ADDR}) ${PORT}")
 field(SCAN, "1 second")
record(stringin, "${P}:IP${N}:FirmwareVersion")
 field(INP, "@gamm1.proto getFWV(${ADDR}) ${PORT}")
record(ai, "${P}:IP${N}:Current")
 field(INP, "@gamm1.proto getCurrent(${ADDR},${S1}) ${PORT}")
 field(PREC, "0")
 field(EGU, "AMPS")
record(ai, "${P}:IP${N}:Pressure")
 field(INP, "@gamm1.proto getPressure(${ADDR},${S1}) ${PORT}")
 field(PREC, "0")
 field(EGU, "Torr")
record(ai, "${P}:IP${N}:Voltage")
 field(INP, "@gamm1.proto getVoltage(${ADDR},${S1}) ${PORT}")
 field(PREC, "0")
 field(EGU, "VOLTS")
record(mbbi, "${P}:IP${N}:Status")
 field(INP, "@gamm1.proto getPumpStatus(${ADDR},${S1}) ${PORT}")
 field(ZRVL, "0")
 field(ONVL, "1")
 field(TWVL, "2")
 field(THVL, "3")
 field(FRVL, "4")
#String Vals
 field(ZRST, "STANDBY")
 field(ONST, "STARTING")
 field(TWST, "RUNNING")
 field(THST, "COOLDOWN")
 field(FRST, "ERROR")
 #String Severities
 field(ZRSV, "NO_ALARM")
 field(ONSV, "MINOR")
 field(TWSV, "NO_ALARM")
 field(THSV, "MINOR")
 field(FRSV, "MAJOR")
record(ai, "${P}:IP${N}:PumpSize")
 field(INP, "@gamm1.proto getPumpSize(${ADDR},${S1}) ${PORT}")
 field(PREC, "0")
 field(EGU, "L/S")
record(ai, "${P}:IP${N}:PressureFactor")
 field(INP, "@gamm1.proto getPumpPressureFactor(${ADDR},${S1}) ${PORT}")
 field(PREC, "2")
record(stringin, "${P}:IP${N}:AutoRestart")
 field(INP, "@gamm1.proto getAutoRestartStr(${ADDR},${S1}) ${PORT}")
record(ai, "${P}:IP${N}:CurrentLimit")
 field(INP, "@gamm1.proto getPumpCurrentLimit(${ADDR},${S1}) ${PORT}")
 field(PREC, "0")
 field(EGU, "mA")
record(ai, "${P}:IP${N}:GetSetpoint")
 field(INP, "@gamm1.proto getSetpoint(${ADDR},${S1},${S2},${P},${N}) ${PORT}")
 field(PREC, "0")
record(ai, "${P}:IP${N}:Relay${S1}OnPressure")
 field(EGU, "Torr")
record(ai, "${P}:IP${N}:Relay${S1}OffPressure")
 field(EGU, "Torr")
record(ai, "${P}:IP${N}:Relay${S1}State")
record(calcout, "${P}:IP${N}:Relay${S1}OnCalc")
 field(SCAN, "Passive")
 field(INPA, "${P}:IP${N}:Relay${S1}OnPressureSetpoint")
 field(INPB, "${P}:IP${N}:Relay${S1}OffPressure")
 field(CALC, "0")
 field(OUT, "@gamm1.proto SetSetpoint(${ADDR},${S1},${S2}) ${PORT}")
record(ao, "${P}:IP${N}:Relay${S1}OnPressureSetpoint")
 field(FLNK, "${P}:IP${N}:Relay${S1}OnCalc")
 field(PREC, "2")
record(calcout, "${P}:IP${N}:Relay${S1}OffCalc")
 field(INPA, "${P}:IP${N}:Relay${S1}OnPressure")
 field(INPB, "${P}:IP${N}:Relay${S1}OffPressureSetpoint")
 field(CALC, "0")
record(ao, "${P}:IP${N}:Relay${S1}OffPressureSetpoint")
 field(FLNK, "${P}:IP${N}:Relay${S1}OffCalc")
 field(PREC, "2")
record(bo, "${P}:IP${N}:HighVoltageOn")
 field(OUT, "@gamm1.proto HvTurnOn(${ADDR},${S1}) ${PORT}")
 field(HIGH, "1")
record(bo, "${P}:IP${N}:HighVoltageOff")
 field(OUT, "@gamm1.proto HvTurnOff(${ADDR},${S1}) ${PORT}")
 field(HIGH, "1")
Support for Gamma Vacuum MPCq ion pump controller over Ethernet Mark Rivers via Tech-talk
Prev: RE: Investigating Archiver Appliance lost events Abdalla Ahmad via Tech-talk
Next: Re: Issues with PVXS Mark Rivers via Tech-talk
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  2022  2023  2024  2025  <2026>
Prev: Re: Support for Gamma Vacuum MPCq ion pump controller over Ethernet LiangChih Chiang via Tech-talk
Next: Deployment of the Archiver Appliance LONG FENG via Tech-talk