On 04/17/2018 06:18 AM, Jörn Dreyer wrote:
> ~ $caget -d 2 HIBEF:XFEL:image1:ArrayData | more
> Native data type: DBF_FLOAT
> Request type: DBR_FLOAT
> Element count: 1000
> Value: 2.02038e-39 0 ...
Don't ask me why, but I was annoyed at seeing (again) this garbage element
in what should be a zero'd array. So I traced it back. The issue is in
camessage.c in read_reply(). Specifically, that
> dbr_size_n(X,0) == dbr_size_n(X,1)
as the underlying dbr_size[] includes the size of meta-data and one value element.
So dbChannel_get_count() correctly sets item_count to zero. Then read_reply()
zeros part, but all, of the buffer. In this case data_size includes the first
value element, which isn't set by dbChannel_get_count(), and not isn't zeroed
here.
> ca_uint32_t data_size = dbr_size_n(pevext->msg.m_dataType, item_count);
> memset((char *) pPayload + data_size, 0, payload_size - data_size);
This is only in the !autosize case (get or non-dynamic monitor).
I'm not sure where to fix this, as there would seem to be several possibilities,
depending on whether or not dbr_size_n() should change.