1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  <2018>  2019  2020  2021  2022  2023  2024  2025  2026  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  <2018>  2019  2020  2021  2022  2023  2024  2025  2026
Subject: Re: Stream Device: Parse response that includes checksum of command string
To: David Vine <[email protected]>
Cc: EPICS <[email protected]>
Date: Fri, 3 Aug 2018 09:38:05 +0200
Hi David,
At the moment, StreamDevice cannot calculate the checksum of something that is not in the current I/O line.
In your case I see three possibilities. But before I show them, one correction seems to be necessary. According to the "MeCom Protocol Specification" document (https://www.meerstetter.ch/de/category/19-mecom-communication-protocol-specifications), "The Frame Checksum is represented by four HEX digits." Your example uses binary format (2 bytes). You need the 0 flag: "%0<xmodem>". Now that all bytes are ASCII, I would put the "\r" in the Terminator. This saves you from relying on timeout to detect the end of the input line.
Now the three possibilities:
a. Pre-calculate your checksum. As the output line is constant (does not contain any formats), the checksum is constant as well.
    IN "#0115AA?IF257D";
b. Ignore the input checksum. The sequence "\?" stands for "any byte".
    IN "#0115AA%\?\?\?\?";
c. Post-process the string. First use the string to calculate the checksum, then after the checksum modify the string with regsub:
    IN "#0115AA?IF%0<xmodem>%#/?IF//\r"
However, according to the "MeCom Protocol Specification", this is not what you need. In the example on page 13 it says that the reply checksum is 342D, not 257D. Thus I guess the checksum is indeed calculated from the current input (which is unfortunately incomplete in the example, so that I cannot confirm this). Also the reply in the example starts with "!", not with "#".
I guess what you really need is this:
Terminator = "\r";
OUT "#0115AA?IF%0<xmodem>";
IN  "!0115AA%20c%0<xmodem>";
If you see something starting with "#", can it be that the device echos the request before sending the reply? In that case you would need:
Terminator = "\r";
IN  "#0115AA?IF%0<xmodem>";
Also the document recommends to use a changing "sequence number". You can use a calc record (e.g. with the name "SEQ") to count the sequence (FLNK to it from any I/O record) and reference it using redirection:
Terminator = "\r";
OUT "#01%(SEQ)04X?IF%0<xmodem>";
IN  "!01%(SEQ)=04X%20c%0<xmodem>";
The "=" in "%(SEQ)=04X" means that the value of record "SEQ" should be formatted and compared to the input rather than parsing the input and storing the value in the record.
Here is the counting calc record:
record(calc, "SEQ")
    field(VAL,  "0x15AA")
    field(INPA, "SEQ")
    field(CALC, "(A+1)&0xffff")
Or with recent EPICS versions:
record(calc, "SEQ")
    field(CALC, "(VAL+1)&0xffff")
Best regards,
Dirk
On 03.08.2018 02:44, David Vine wrote:
Hi all,
i'm writing device support for the MeCom protocol for a TEC driver from Meerstetter Engineering.
In order to parse the response correctly I need to calculate a checksum thats not included in the response. This is because the response includes the checksum of the command and not the response.
For example:
Command format:
# address sequence_number command CRC16 terminator
# 01 15AA ?IF 257D \r
Response is:
# address sequence_number CRC16 terminator
# 01 15AA 257D \r
So to parse the response I need to calculate a CRC that's not included in the response. For example:
OUT "#0115AA?IF%<xmodem>\r"
IN "#0115AA%<xmodem>\r" but the CRC is for the OUT not the IN.
Any ideas would be much appreciated.
David
Best regards,
David Vine
(925) 326 7503
[email protected] <mailto:[email protected]>
5750 Imhoff Drive, Suite I  |  Concord, CA 94520  |  USA <https://maps.google.com/?q=5750+Imhoff+Drive,+Suite+I%C2%A0+%7C%C2%A0+Concord,+CA+94520%C2%A0+%7C%C2%A0+USA&entry=gmail&source=g>
http://www.Sigray.com <http://www.sigray.com/>
This e-mail and any attachments are provided for the sole use of the intended recipient(s), and may contain information that is confidential, privileged, proprietary or otherwise protected by law. If you are not the intended recipient of this message, you are not authorized to read, print, retain, copy, forward or otherwise disseminate this message or any part of it. If you have received this e-mail in error, please notify the sender immediately by reply e-mail and delete the original message from your mail system. Thank you.
Stream Device: Parse response that includes checksum of command string David Vine
Prev: Re: dbGetPdbAddrFromLink dropped from 3.15 again Michael Davidsaver
Next: RE: Stream Device: Parse response that includes checksum of command string [email protected]
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  <2018>  2019  2020  2021  2022  2023  2024  2025  2026
Prev: RE: Stream Device: Parse response that includes checksum of command string [email protected]
Next: DBE_PROPERTY and Gateway Dominic Oram - UKRI STFC
ANJ, 03 Aug 2018 · Home · News · About · Talk · Base · Modules · Extensions ·