1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  <20222023  2024  2025  Index 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  <20222023  2024  2025
Subject: Re: Ways to know if instrument connected via TCP/IP through StreamDevice has lost connection
From: "Wang, Andrew via Tech-talk" <tech-talk at aps.anl.gov>
To: Mark Rivers <rivers at cars.uchicago.edu>, EPICS tech-talk <tech-talk at aps.anl.gov>
Date: Mon, 12 Dec 2022 22:11:28 +0000
Mark,
This is extremely helpful. I will let you know if I have follow up questions.
Andy
Sent: Thursday, December 8, 2022 6:51 PM
To: EPICS tech-talk <tech-talk at aps.anl.gov>; Wang, Andrew <wang126 at llnl.gov>
Hi Andy,
> I feel like there is a better way than me trying to use a calcout record and check if any record has its STAT field set to COMM. Like, is it possible to monitor for connection status without needing to process a record whose device type is StreamDevice and check its STAT field?
Yes, there is a better way.
First, you probably want to use this line in your startup script:
asynSetOption("$(PORT)",0, "disconnectOnReadTimeout", "Y")
That will cause the port to disconnect if there is a timeout.  If you don't do that then the asyn won't know that the port is disconnected for a very long time on Linux, it will just keep giving "temporarily unavailable" messages.
In your startup script load an asyn record that is connected to that asyn port.  Here is an example:
dbLoadRecords("$(ASYN)/db/asynRecord.db", "P=IOC:, R=asyn1, PORT=ARIES1, ADDR=0, OMAX=256, IMAX=256")
You can monitor the  IOC:asyn1.CNCT field,  which will either be "Connect" or "Disconnect".
From: Tech-talk <tech-talk-bounces at aps.anl.gov> on behalf of Wang, Andrew via Tech-talk <tech-talk at aps.anl.gov>
Sent: Thursday, December 8, 2022 8:21 PM
Subject: Ways to know if instrument connected via TCP/IP through StreamDevice has lost connection
Hi all,
Basically, I was wondering if there is a better way to know if communication to an instrument has been lost. I understand that this can be done either by simplying looking at the IOC shell for disconnection errors or checking if the STAT field is set to COMM.
I was wondering if there is another method of determining if the instrument is still connected or not besides the two methods described above. Additionally, I would like to have a record called "is_instrument_connected", that holds a value like 1 to indicate the instrument is connected and a 0 if it is not. I feel like there is a better way than me trying to use a calcout record and check if any record has its STAT field set to COMM. Like, is it possible to monitor for connection status without needing to process a record whose device type is StreamDevice and check its STAT field?
Thank you,
Andy
Ways to know if instrument connected via TCP/IP through StreamDevice has lost connection Wang, Andrew via Tech-talk
Re: Ways to know if instrument connected via TCP/IP through StreamDevice has lost connection Mark Rivers via Tech-talk
Prev: Phoebus Alarm Server on CentOS 9? William F Badgett Jr via Tech-talk
Next: Checking CPU and memory usage for an IOC Wang, Andrew via Tech-talk
Index: 1994  1995  1996  1997  1998  1999  2000  2001  2002  2003  2004  2005  2006  2007  2008  2009  2010  2011  2012  2013  2014  2015  2016  2017  2018  2019  2020  2021  <20222023  2024  2025
Prev: Re: Ways to know if instrument connected via TCP/IP through StreamDevice has lost connection Mark Rivers via Tech-talk
Next: Re: Ways to know if instrument connected via TCP/IP through StreamDevice has lost connection Kuldeep Joshi via Tech-talk
ANJ, 12 Dec 2022 Valid HTML 4.01! · Home · News · About · Base · Modules · Extensions · Distributions ·
