Subject: Re: [Scopes] BMP image record??
To: Pavel Masloff <[email protected]>
Date: Fri, 04 May 2012 07:23:01 -0700
Pavel:
We must be talking about different things. If your waveform is 2500 points, and takes 73.5 KB, that works out to 29.4 bytes per point. Sounds like very high resolution! Even if that waveform was transferred in some ASCII format, that is still very verbose.
I've never tried to measure the GPIB throughput, but I am fairly certain that the overall throughput in any case that I've encountered is limited by the scope itself, and the operation of the scope in terms of its triggering setup and the signal it is measuring. On some scopes (eg TDS 380), I can swamp it's CPU with GPIB communications, and effectively cripple its functionality. That is an old scope, and I doubt that modern scopes would be similarly afflicted. Perhaps the hardcopy function would be a good case for a throughput test.
   ---   rod.
Pavel Masloff wrote:
Hi Rod,
We have 2500 points per waveform. Our PCs are equipped with COM ports
only, no GPIB cards. One waveform is 73,5 KB to be exact. The RS232
connection speed is maximum 19200 Kb/s at maximum. For the time being I
just want to check whether the experiment went fine, no false
triggering, etc. In the future, I would like to analyze the waveforms in
matlab via CA. So this will be handy.
What is the speed of the GPIB interface on Tek scopes, anyway?
On Fri, May 4, 2012 at 1:42 AM, Rod Nussbaumer <[email protected]
<mailto:[email protected]>> wrote:
    Pavel:
    Every Tektronix scope I've set up communications with had a 500
    point waveform (as well as some substantially larger ones).
    Presently, I'm acquiring those at 10 waveforms per second (over
    GPIB). Are you sure that would not suffice for you? Is GPIB an
    option? Much much faster.
       ---   rod.
        Mark,
        A waveform weighs 70kB. 4 waveforms = 280 kB. 280 vs 40. I would
        have to
        wait a couple of minutes till the data is processed. 10-15
        seconds for
        the screen dmp.
        On Thu, May 3, 2012 at 11:13 PM, Mark Rivers
        <[email protected] <mailto:[email protected]>
        <mailto:[email protected].__edu
        <mailto:[email protected]>>> wrote:
            It may be time to back up a step.  Why not transfer the
        waveforms
            themselves, rather than a screen dump?  There are others in the
            EPICS community who have already done that for the scopes
        you are using.
            __ __
Best regards,
Pavel Maslov, MS
Controls Engineer at Pulsed power Lab
Efremov Institute for Electro-Physical Apparatus
St. Petersburg, Russia
Mobile: +7 (951) 672 22 19
Landline: +7 (812) 461 01 01
RE: [Scopes] BMP image record?? Mark Rivers
[Scopes] BMP image record?? Pavel Masloff
Re: [Scopes] BMP image record?? Dirk Zimoch
Re: [Scopes] BMP image record?? Pavel Masloff
Re: [Scopes] BMP image record?? Rod Nussbaumer
Prev: Problem installing CA for QT manta santosh aditya santosh_aditya
Next: Re: ioc with genSub record: compiling error Eric Norum
Prev: Re: [Scopes] BMP image record?? Pavel Masloff
Next: RE: [Scopes] BMP image record?? Mark Rivers